In general, the results of setting up the multimeter according to the service manual are encouraging. Testing and calculation of DCV, OHM and INL correction factors takes about 30-40 minutes in automatic mode. However, I want to draw the attention of everyone who is going to use this technique to the fact that the implementation of paragraph "14.5. Resistance transfer adjustment" will require a new calibration of the multimeter by an external resistance standard!
In addition, I categorically do not recommend measuring with a multimeter in the ranges of 10, 100 and 1000 MΩ, using a 4-wire connection scheme, as well as 2-wire, but with the AZERO function enabled. If, nevertheless, zero correction is required in the range of 10 MΩ (at 100 and 1000 MΩ it is always disabled), then it is better to carry it out once, turning it off before the main measurement(s).