Hello,
update of ADR1399LS8 on PCB#1 ageing 8kHr after putting into Hammond die cast enclosure and with temperature compensation (sensor on bottom of the hammond case)
see also
https://www.eevblog.com/forum/metrology/adr1399-reference/msg5226999/#msg5226999The ADR1399LS8 have now 11.3 kHrs in total operating time. Tempco is calculated out for the ageing curves.
These LS8-devices are drifting upwards.
I had again a 5-6 ppm voltage shift on the 2 LM399 (only for comparison) on the 2nd Multiplexer (MAX4052A).
This time I had no obvious reason for the shift. So it was very unreasonable that the Multiplexer should have died again.
I reflected what I did last time:
- Switching power off and on (on another desk) to do some measurements.
- Measurements did not show any obvious error.
- Exchange the MAX4052A. (with power off).
- reassembling and putting all back to the measurement desk.
So I decided to do power cycling only and actually this "cured" the 5-6 ppm voltage shift.
Obviously there is some "latchup effect" which affects the upper channels.
But what can the reason for this latchup effect be?
I had introduced some level shifters to reduce the self heating of the multiplexers when having 5V logic levels.
But the level shifters are relatively high ohmic (BSS138 with 330K pull-up resistors on the 15V side) to reduce self heating.
Simulation with a similar FET (BSS123) showed around 20 us rise time with 20 pF for wiring and input capacitances of the multiplexers.
Should this be the reason for the latchup-effects?
with best regards
Andreas