Hello,
Any results so far?
My preliminary results:
I have ordered 2 times (8 pcs each) from GigaJoe s seller.
https://www.eevblog.com/forum/metrology/ad587lq-from-aliexpress/msg1677863/#msg1677863He needed both times 1-2 weeks until I could see in the tracking system that he had sent them out.
When I asked him for delivery he answered something like he wanted to find the best source for me.
So obviously the seller does not have them on stock: so quality may vary.
Each reference has a different date code.
Devices are desoldered with partly bent pins.
Output voltage on all 16 devices is within +/-3 mV from 10V. So even with ageing within 5mV datasheet spec.
Pin 3 shows 0 V (so no temperature output signal like on some band-gaps)
Pin 8 and pin 5 are around 7V so obviously it is a buried zener reference.
1/f noise is around 4uVpp.
up to now 2 devices have frequently (every few seconds) popcorn noise. (but have not evaluated all data up to now).
It seems that the popcorn candidates had been marked with a black dot.
On the first batch I thought that it was some flux residue from desoldering so I had removed it.
On the 2nd batch I can correlate the marked device with the popcorn noise.
T.C. is still under evaluation.
Measured 12 devices up to now with < 2.5 ppm/K over +/-15 deg C span around room temperature.
2 devices showed up to 2-4 ppm (ageing) drift during T.C. measurement.
with best regards
Andreas