zhtoor:
I think you need to listen to everyone and maybe think about another way besides wet cells.
I would suggest don't even worry about spending limited resources on getting a wet cell calibrated. You don't really have any equipment that's going to see any benefit - and it's very very unlikely you're going to successfully get a wet cell to the lab and back without destroying the calibrated measure anyway. In other words: The cost to benefit ratio is very, very low for getting a calibrated absolute measure on a wet cell.
The other problem is with wet cells - the more you try to take a measure, the more you're changing the discharge curve.
For your situation getting something like a working 3456a (or similar), get that calibrated to within a few ppm (usually very inexpensive), and you're miles ahead for having a known stable way to measure your wet cells to a reasonably accurate absolute value. That's probably the most bang for your buck to get some ppm measuring ability. Those old meters are big and slow - but they can offer really good yearly drift specs (often better than 3458a). Used, functional 731b's work very very well also and they don't break the bank - those can work well down to very low ppm per year drift also. Which is still going to outperform and outlast your wet cells if you are using them at all.
We just measured a 731b recently that really wasn't drifting more than a ppm per year, and they don't get too picky if you keep them powered on or not. A bit noisier than a 732a or battery, but they can be rock-solid stable as a value Vref.
You can still use the wet cells as a fairly quiet voltage source as required - as long as you're willing to discharge. They will -always- discharge, not matter what you think you're doing to draw no energy...
But I wouldn't use the wet cells to transfer an absolute voltage measure to your location. There's a ton of better ways to do it that are more realistic for your modest uncertainty of measure capability.