the LT2400 which you and branadic used for TC measurement works, at this moment, approximately what average resolution could those produce now (with your improvements)? 5.5D ?
Hello,
that depends.
Mainly on the effort which you put into each single device.
And the comparison to any x.x digit device is difficult.
You should compare it more to a old Solartron or similar
with long integration times on high resolution
than with a modern multi-slope instrument.
If you just build it and do full scale adjustment it is actually something like a 5.5 digit instrument.
Noise for a single measurement is relatively high:
- 10uVpp or 1.8uV RMS in 0..5V range corresponding to 20uVpp 3.6 uV RMS in 0..10V range with a LTC1043 divider.
With averaging over one minute you go down to around 0.5uV RMS for a 7V stable reference
in 10V range which is about factor 5 above a HP3458A.
What I am doing is the following:
Buy plenty of good voltage references (AD586LQ).
Select them for T.C. , hysteresis and popcorn noise.
So from 10 references you get (in average) around 2 which are suitable:
(T.C. < 1 ppm/K (better 0.3 ppm/K), hyst < 1ppm, popcorn < 2uVpp)
Put the best of them onto ADCs which have a NTC for temperature correction.
Adjust them for T.C. (3rd order correction), INL, Gain.
The newer devices I also adjust for offset and offset drift (then paired with a LTC1043 divider cirquit).
Adjustment takes around 4-6 week-ends for each device.
Then you have simply to wait the run-in phase (switched on) until the AD586 has stabilized.
(typically 5000 hours)
You can shorten this phase partly by loading the AD586 with a 15 mA current cyclically.
(30 minutes on during 2 hours cycle).
After that you typicaly have to adjust the gain. (T.C. and INL normally only need to be checked).
What you get (typically):
- Noise 0.5uV eff @10V (1 minute integration time)
(about factor 2 above a 1 minute average of a 6.5 digit DMM)
- INL < 1ppm (that what is specced on many 6.5 to 7.5 digit DMMs)
- Long term stability about 1-2 ppm/year (also comparable to well aged LM399 based instruments).
- Excellent temperature stability which is essential in my lab.
(from 18-33 deg C max 1 ppm drift)
That is why I cannot use a 6.5 DMM in 10V range for my T.C. measurements.
The drift over one day is usual several PPMs on a K2000 or 34401A with my lab conditions.
- even better stability you get for ratiometric measurements.
But nothing comes for free.
A 6.5 DMM might be cheaper when counting all the gear you need to adjust a ADC.
And even then for exact absolute measurements there is no Auto-Zero implemented up to now.
So the 2-6uV offset you have to measure before/after each measurement.
But all in all with long integration times you have something like a 6.5 digit instrument with
some features (T.C.) being even much better.
With best regards
Andreas