Author Topic: Fluke Tir32 continuous (re)calibration ?  (Read 542 times)

0 Members and 1 Guest are viewing this topic.

Offline darkstar49Topic starter

  • Frequent Contributor
  • **
  • Posts: 309
Fluke Tir32 continuous (re)calibration ?
« on: April 05, 2023, 01:09:06 pm »
Hi,

my TIR32 works more or less OK, temps seem OK (things from the freezer, boiling water, etc...), but every 30 secs or so, it freezes and says 'calibration' for a few seconds, and then works fine again...
I think to recall having seen similar behaviour on Fluke thermal cams of similar generation (Ti25,...), but not at this frequency.

It doesn't make the cam unusable, but it's somehow annoying.

I asked Fluke about their calibration: 710€ or 1065€ (excl. VAT), for a CALNET vs. CALINTL calibration respectively (flagged as 'ISO9001 traceable' for the first, and 'Accredited' for the 2nd ??).

Not being even sure that a calibration is gonna fix the issue, that's sounds quite a lot to me...

Any thoughts/suggestions ?
 

Offline railrun

  • Regular Contributor
  • *
  • Posts: 113
Re: Fluke Tir32 continuous (re)calibration ?
« Reply #1 on: April 05, 2023, 03:00:21 pm »
Does the intervalls get longer when the unit is on for a while?
I don‘t have one, but this is the internal calibration with a shutter. At the beginning the device warms up and it compensate the temperature drift. After a few minutes the temperature should be stable and it need less calibrations.
 

Offline DaJMasta

  • Super Contributor
  • ***
  • Posts: 2318
  • Country: us
    • medpants.com
Re: Fluke Tir32 continuous (re)calibration ?
« Reply #2 on: April 05, 2023, 03:52:55 pm »
Depending on how much access you have to the configuration, usually there are two methods they use to trigger NUC calibration - either a certain amount of change in temperature of the core, a certain number of frames taken, or both.  As previously mentioned, if the interval gets longer over time when the camera is operating, it's probably the delta temperature method - as the core warms up to its equilibrium point, the change slows and the interval gets longer.  If it's a fixed timeframe, it's likely just counting a thousand or two frames and then automatically rerunning the calibration.  A hybrid of the two is probably most useful (recalibrate after a certain temperature change or after a large number of frames), and they are likely just values in configuration firmware, but I don't know if every system will actually give you access as the user.

This behavior doesn't have to do with the radiometric calibration units get to read out temperatures correctly.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf