Author Topic: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator  (Read 136318 times)

0 Members and 5 Guests are viewing this topic.

Offline rifkum

  • Contributor
  • Posts: 22
  • Country: us
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #375 on: February 08, 2024, 02:16:43 pm »
Quote
The problem with the UTG962E is that the output is reduced for all output levels, whether they are near to maximum or not. OK, I can live with that, but the amplitude reading on the front panel should reflect this reduced level. You shouldn't have to remember to take 6dB off the reading when switching on AM.

Agree!  I guess I will make a label to stick on the top saying "With AM modulation enabled, level out is 6 dB below setpoint" or something like that.

I think Uni-T firmware could/should be updated to simply boost the output by 6dB and flash warning saying "with AM Mod the max output is reduced"

This seems like what Dave Jones would call a "trap for young players".

regards
dm
« Last Edit: February 08, 2024, 02:23:25 pm by rifkum »
 

Offline 807

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: gb
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #376 on: February 08, 2024, 02:30:19 pm »
Idealy this generator should be programmed to do what the "big boys" do, & only reduce the carrier when it's set to within 6dB of it's maximum output level.

At the very least the firmware should have been written to show this reduced level on the Amplitude display. Surely that wouln't have taken much coding? :-//
 

Offline rifkum

  • Contributor
  • Posts: 22
  • Country: us
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #377 on: February 08, 2024, 03:22:53 pm »
@807,
Yes!! and I believe your second option is best since the normal workflow is to turn on Wave output level first and choose your desired level then second is to go to Mod and choose your modulation type.  And the assumption is that you will get your chosen modulated level out.

Still going to try to "hound" Uni-T tech support to correct the firmware but not holding my breath on that possibility.

Cheers
dm

Note: see my "fix" in below image
« Last Edit: February 08, 2024, 03:27:30 pm by rifkum »
 
The following users thanked this post: 807, Markus2801A

Offline rifkum

  • Contributor
  • Posts: 22
  • Country: us
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #378 on: February 09, 2024, 03:32:09 am »
Well, I got an initial response from Uni-T on my tech support question about the signal dropping 6 dB when Mod is enabled...

Quote
Message I sent:

There is a problem with the UTG962 signal generator.
When AM modulation is turned on the signal level drops by 6 dB
This happens even if the "depth" of modulation is 0%
Please confirm that this is a problem

Test conditions... 
Wave: Sine, Freq 14 MHz
Mod Sine, Freq 1 KHz

Turning Mod off/on causes the output signal to drop by 6 dB

Their Reply...

Hi
Thanks for asking the question!

AM modulation is different from FM modulation , the signal was transmitted as side band ,and the side band level will be suppressed. With 100% point of modulation , the primary side band signal will be compressed 6dB .

At the point of 100% modulation, the spectrum appears identical to a normal double-sideband AM transmission, with the center carrier at 0 dB reference level, and both side bands on either side of the primary carrier at -6 dB.

Please  refer to the link to get more theoretical explanations :
https://en.wikipedia.org/wiki/Single-sideband_modulation

 

And here is my reply back to them...

Quote
Thank you for the reply.
I agree that the side bands may be 6 dB down but in the case of the UTG962 it is the MAIN Carrier that is dropping by 6 dB when MOD is enabled no matter what the % modulation.
The center carrier SHOULD NOT DROP!!  And the level out should match what the display shows.

Please do a check on your end with a sample unit and you will see that the Center carrier drops by 6 dB when Mod is enabled.
This should not happen!!

The display on the UTG962 still shows the initial set value even though the actual level out has dropped.
This is a firmware bug!  The MAIN carrier should still show the correct value of the WAVE as displayed on the SINE output screen.

Please let me know how your test performs.

Thank you for looking into this.

Hopefully they will actually do some testing.
I'll report if I hear anything else.

Someone please speak up if I am in error here about this being a firmware issue!! 
« Last Edit: February 09, 2024, 10:42:37 pm by rifkum »
 
The following users thanked this post: 807

Offline 807

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: gb
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #379 on: February 09, 2024, 02:24:50 pm »
Be interesting to see how they reply to your follow up.

Out of interest, I looked at a similar AWG, the Owon DGE2070. It appears that range of generators also behave the same as the UTG962. Here's what their manual says about AM modulation...

"In 0% modulation, the output amplitude is half of the specified value. In 100% modulation, the output amplitude is equal to the specified value."

Owon have got it wrong too IMO.
 

Offline TheoB

  • Regular Contributor
  • *
  • Posts: 62
  • Country: nl
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #380 on: February 11, 2024, 10:13:06 am »
I received my UTG962 and noticed that the DC offset voltage is out of spec by a lot.
I wanted to use the device as a 25 Ohm power source. A power source that can be programmed to any DC, ramp or noise voltage.
The specification says +/-3% of offset value +/-6% of amplitude value +/-2mV
I interpret this as +/-3% of the DC value +/-2mV if I set the amplitude to a small value as is the case with a DC "wave" output setting. Anyhow I get up to 12% error at +/-600mV.
I can send the device back as being not properly calibrated, but wanted to know first what others see. Simply put it in HiZ mode and measure the DC output voltage at some DC values. See my error plot.
« Last Edit: February 11, 2024, 10:22:46 am by TheoB »
 

Offline wasedadoc

  • Super Contributor
  • ***
  • Posts: 1504
  • Country: gb
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #381 on: February 11, 2024, 11:03:13 am »
Notwithstanding the offset observations, the output impedance is always 50 Ohms regardless of the setting. Changing that setting only makes the display show what will (or should) be found across such an external load.

If you truly want 25 Ohms, connect a 50 Ohm resistor from output to ground and set the dial to 50 Ohm.
 

Offline TheoB

  • Regular Contributor
  • *
  • Posts: 62
  • Country: nl
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #382 on: February 11, 2024, 12:22:52 pm »
I know, I have a 50 Ohm load connected to lower the impedance to 25 Ohm. My dut is a lower power tranceiver capable of running on a CR2032 battery. So 25 Ohm is ok.
The offset error however, is not the same for HiZ and when terminated with 50 Ohm. For DC the generator source is clearly not a voltage source with 50 Ohm output impedance only.
 

Offline Skyguard

  • Newbie
  • Posts: 3
  • Country: ar
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #383 on: February 28, 2024, 08:35:46 pm »
Hello!
I want to make a query.
I have a UTG962e that is slightly frequency shifted.
Approximately 30 hz when it is emitting 10,000,000 hz.-
Does anyone know of a hidden menu command or other way to correct this small difference?
Thank you so much
 

Offline MechatrommerTopic starter

  • Super Contributor
  • ***
  • Posts: 11701
  • Country: my
  • reassessing directives...
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #384 on: February 28, 2024, 09:09:27 pm »
how about typing 9999970 Hz? ;D or check the clock?
Nature: Evolution and the Illusion of Randomness (Stephen L. Talbott): Its now indisputable that... organisms “expertise” contextualizes its genome, and its nonsense to say that these powers are under the control of the genome being contextualized - Barbara McClintock
 

Offline electr_peter

  • Supporter
  • ****
  • Posts: 1344
  • Country: lt
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #385 on: February 28, 2024, 09:47:30 pm »
IIRC, there was a post (and video) how to add 10MHz reference input to UTG962.
 

Offline Skyguard

  • Newbie
  • Posts: 3
  • Country: ar
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #386 on: March 02, 2024, 12:39:17 am »
how about typing 9999970 Hz? ;D or check the clock?
:-DD :-DD
 

Offline onx

  • Newbie
  • Posts: 6
  • Country: ua
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #387 on: June 28, 2024, 12:34:52 pm »
Hi. Anybody tested the new firmware(V3.06.004 2024.05.22) for UTG900E? https://gduts.sharepoint.com/:u:/s/Instrument/EVdQIMZHnSlLn7esjvCAI_UBOF3qx3EdnlhCS4G_SXEg4A?e=rK1ZYv
It can be flashed with new software https://gduts.sharepoint.com/:u:/s/Instrument/EYV72CjGaS9Np-0jMMUsJ1oB1448POvP4xGFZuOymlAcnQ?e=JzdY36
Source: https://instruments.uni-trend.com/download
------------------
Release Notes:
UTG900E Firmware Upgrade

Firmware version: V3.06.004
Logic version:V3.01.003
Compatibility:Hardware version V1.03

Update:
1.1. Modification: The calibration data of 1kHz is used for 100kHz and the following frequency points. This change is to optimize the equipment for the flatness index "≤ 100kHz: 0.1dB", that is, relative to the 1kHz amplitude tolerance ± 1.157945%.
2、Optimize the flying line problem of synchronous waveform
------------------

I flashed my unit and firmware was 1.08 and changed to 3.06, but I can't update FPGA firmware (Logic) all time have error. :-BROKE
I purchased my unit this Year and in EEPROM date is:
UTG9Oct 18 2022 12:06:04
Main board: V1.03
« Last Edit: June 28, 2024, 06:45:54 pm by onx »
 
The following users thanked this post: 807, Markus2801A, Yuriy V.

Offline Yuriy V.

  • Contributor
  • Posts: 23
  • Country: ru
    • trengtor custom
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #388 on: June 29, 2024, 01:49:49 pm »
And they ignored older hardware versions. Not surprised.
« Last Edit: June 30, 2024, 05:15:37 pm by Yuriy V. »
 
The following users thanked this post: 807

Offline 807

  • Frequent Contributor
  • **
  • Posts: 274
  • Country: gb
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #389 on: June 30, 2024, 09:19:33 pm »
And they ignored older hardware versions. Not surprised.

Same here. My 962E is Hardware Version 1.01  :--
 

Offline onx

  • Newbie
  • Posts: 6
  • Country: ua
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #390 on: July 01, 2024, 11:06:48 am »
And they ignored older hardware versions. Not surprised.

Same here. My 962E is Hardware Version 1.01  :--

The Hardware Version is stored in EPROM (what you can see in about menu), and it has nothing to do with reality. Because the version of the board is the same for everyone and is 1.03, which is the real version of Hardware Version.
« Last Edit: July 01, 2024, 11:12:02 am by onx »
 

Offline RogerG

  • Contributor
  • Posts: 30
  • Country: de
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #391 on: July 01, 2024, 11:24:13 am »
You mean, we can try the update even when HW1.01 is displayed?

Did anyone do that with success?
 

Offline onx

  • Newbie
  • Posts: 6
  • Country: ua
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #392 on: July 01, 2024, 11:44:46 am »
I did :)
And it works perfect. But FPGA (Logic) firmware I can't update. All time error. But the main firmware updated good.
 

Offline RogerG

  • Contributor
  • Posts: 30
  • Country: de
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #393 on: July 01, 2024, 11:48:12 am »
Thx for the reply.

What does that mean for the normal operation of the UTG?
Working or not?
Every boot with error message, or just an error message while flashing the update?
 

Offline onx

  • Newbie
  • Posts: 6
  • Country: ua
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #394 on: July 01, 2024, 11:52:19 am »
Only if you try to update FPGA (Logic) you will have error. Maybe in your case, it can update.
Boot is ok, work is ok.
« Last Edit: July 01, 2024, 12:10:12 pm by onx »
 

Offline RogerG

  • Contributor
  • Posts: 30
  • Country: de
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #395 on: July 01, 2024, 12:01:13 pm »
Thx !
 

Offline TomKatt

  • Frequent Contributor
  • **
  • Posts: 508
  • Country: us
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #396 on: July 01, 2024, 12:12:05 pm »
The Hardware Version is stored in EPROM (what you can see in about menu), and it has nothing to do with reality. Because the version of the board is the same for everyone and is 1.03, which is the real version of Hardware Version.
That seems like an odd thing to do - why bother recording an older hardware version in EPROM if the version is actually the same for all devices?  If that were the case, why bother changing the reported hw version to 1.03 - because if they are all the same, that should really mean everyone has v1.01 ?

The boards may appear identical, but there may be some hw differences that are not easily seen by eye.
Several Species of Small Furry Animals Gathered Together in a Cave and Grooving with a PICt
 

Offline onx

  • Newbie
  • Posts: 6
  • Country: ua
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #397 on: July 01, 2024, 12:49:51 pm »
Yes, it is true. But the version has never changed. Every new and old device is sold with version 1.01 in the menu.
My device was purchased in 2024. The only thing I noticed that was different from the old version was the newer date in EPROM (old version date EPROM 2021 new version date is 2022).
Example old EPROM and new EPROM
« Last Edit: July 01, 2024, 12:58:44 pm by onx »
 

Offline RogerG

  • Contributor
  • Posts: 30
  • Country: de
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #398 on: July 01, 2024, 02:58:36 pm »
Both updates worked at my UTG purchased in 2022.

It was initially on
3.02
1.01
3.01

and now it has
3.06
1.01
3.01

FPGA update was running without error too, but it's still remaining the same version.
After reset it was on chinese language.

Sadly the jitter on higher frequencies is still present. That was the reason (and hope) for the update.
« Last Edit: July 01, 2024, 03:00:52 pm by RogerG »
 

Offline Yuriy V.

  • Contributor
  • Posts: 23
  • Country: ru
    • trengtor custom
Re: UNI-T UTG932/UTG962 200MSa/s Function Arbitrary Waveform Generator
« Reply #399 on: July 02, 2024, 07:00:35 am »
Fail to FPGA program update.
Any ideas?
« Last Edit: July 02, 2024, 07:56:10 am by Yuriy V. »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf