Anyone come across an issue where the frequency drifts when burst is enabled?

I have my SDG2042X connected to an external 10MHz reference. I set a 10 MHz sine wave on CH1 and enabled the output. Feeding both the reference and SDG CH1 output to my oscilloscope I see the signal phases are stable relative to each other. Next, I switch to CH2 and enable burst (output not yet enabled on CH2). On the oscilloscope I see CH1 now drifting from the input reference. The CH1 keeps drifting until I disable burst on CH2 and toggle the output of CH1 on/off (or change freq on CH1).

I'm running the latest firmware I see v 2.01.01.35R3B2.

I looks like that the burst freq. alters the coherent behavior against the reference. May check whether using ARB it gets any better, while IMHO to control the coherent behavior.

I am waiting on Siglent now about one year to fix the square wave & low deviation FM/PM modulation bug....

so far as FPGA gets on the limit:

a) offered to send back for refund
b) trial FW so only one channel possible

from the face looser, nothing happens so far as  :-- :-- :--  :palm:
 
 
The following users thanked this post: TurboTom

Online TurboTom

  • Super Contributor
  • ***
  • Posts: 1416
  • Country: de
Re: The Siglent SDG2042X Thread
« Reply #1476 on: February 09, 2022, 12:39:09 pm »
Due to the similarities of Siglent's AWG series, I checked for the reported problems on my SDG6000X. And I guess nobody's surprised I found them... I posted the details in the '6000 thread.
 

Offline jeraymond

  • Contributor
  • Posts: 39
  • Country: ca
Re: The Siglent SDG2042X Thread
« Reply #1477 on: February 09, 2022, 08:27:01 pm »
Due to the similarities of Siglent's AWG series, I checked for the reported problems on my SDG6000X. And I guess nobody's surprised I found them... I posted the details in the '6000 thread.

Looking more closely, the signal actually drifts from the external reference even without enabling Burst on the second channel. The drift is much, much slower but still there. Do you see that too on your device?
 

Online TurboTom

  • Super Contributor
  • ***
  • Posts: 1416
  • Country: de
Re: The Siglent SDG2042X Thread
« Reply #1478 on: February 09, 2022, 08:57:39 pm »
No, in case of the SDG6000X, the drift only seems to depend on the burst period, it doesn't change at all if burst mode is enabled or disabled. And as reported in the '6000 thread, I only found this effect if the AWG is configured in "Phase Locked Mode". So, regarding this issue, there appear to be some differences in the way these two instruments behave after all...
 

Offline dorkshoei

  • Frequent Contributor
  • **
  • Posts: 510
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1479 on: February 10, 2022, 01:24:28 am »
Is there a cheap soft storage case that the SDG2042X will fit into.   Cheap == $40 or less.
 

Offline nike75

  • Contributor
  • Posts: 16
  • Country: bg
Re: The Siglent SDG2042X Thread
« Reply #1480 on: February 10, 2022, 07:16:57 am »
I use this one

https://www.aliexpress.com/item/32835074432.html?spm=a2g0o.9042311.0.0.27424c4dvjXYyp

produced for photo gears, but very usable for SDG2042X

I ordered 3 years ago and price was $25, now is much more, but may be found from other store.
 

Offline jeraymond

  • Contributor
  • Posts: 39
  • Country: ca
Re: The Siglent SDG2042X Thread
« Reply #1481 on: February 10, 2022, 01:52:58 pm »
No, in case of the SDG6000X, the drift only seems to depend on the burst period, it doesn't change at all if burst mode is enabled or disabled. And as reported in the '6000 thread, I only found this effect if the AWG is configured in "Phase Locked Mode". So, regarding this issue, there appear to be some differences in the way these two instruments behave after all...

Change the Phase Mode to Independent also fixed the drift issue with Burst enabled on the second channel. Siglent has a App Note about it here: https://siglentna.com/USA_website_2014/FAQ/SDG/SmoothSDGOutputs_01172016.pdf

 
The following users thanked this post: egonotto, tautech

Offline SignalArray

  • Contributor
  • Posts: 12
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1482 on: February 10, 2022, 03:36:24 pm »

try this one from HD:

https://www.homedepot.com/p/RIDGID-22-in-Pro-Organizer-Black-221735/205440492

remove those bins and if needed modify inside of lid with dremel a little.  It should fit.

if not try next size:
https://www.homedepot.com/p/RIDGID-22-in-Pro-Tool-Box-Black-221733/205440481#overlay

this one fits 2 instruments of similar size.(with foam)
 

Offline dorkshoei

  • Frequent Contributor
  • **
  • Posts: 510
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1483 on: February 10, 2022, 11:44:07 pm »

try this one from HD:
"Is there a cheap soft storage case that the SDG2042X will fit into."

I use this one

https://www.aliexpress.com/item/32835074432.html?spm=a2g0o.9042311.0.0.27424c4dvjXYyp
Thanks.  Right kind of material.  I'd prefer something without straps.   Yes I realize I can cut them off.

« Last Edit: February 10, 2022, 11:46:00 pm by dorkshoei »
 

Offline nike75

  • Contributor
  • Posts: 16
  • Country: bg
Re: The Siglent SDG2042X Thread
« Reply #1484 on: February 11, 2022, 05:54:18 am »
When I search for bag, my important point was fit and drop protection. Not so easy to find bag with right sizes if not designed exactly for this gear.
Actually materials is not so bad, this case are padded of all sizes with shockproof pads (it is still designed for photographic equipment)
I have photographic equipment, have branded bags, also bought Chinese - this is completely satisfactory for storing and carrying sensitive equipment.

Its disadvantages are:
- the back straps as well as the hand handle feel a little weaker and unreliable.
- the position of the hand handle is placed at one end and when carried in the hand slightly tilts the case.

As for the back straps - I sewed mine for the back so they wouldn't get in the way, but you can cut them.
« Last Edit: February 11, 2022, 08:40:11 pm by nike75 »
 

Offline n3mmr

  • Regular Contributor
  • *
  • Posts: 121
  • Country: se
Re: The Siglent SDG2042X Thread
« Reply #1485 on: February 15, 2022, 12:50:43 pm »
Is there a repository of downloadable arbitrary waveforms somewhere for this generator?

All I want for X-mas, or possibly my Easter Egg, is a set of waveforms to emulate/simulate UART rs232 or ttl  or so style character strings, at different baud rates.
« Last Edit: February 15, 2022, 12:53:29 pm by n3mmr »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29028
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1486 on: February 15, 2022, 01:07:19 pm »
Is there a repository of downloadable arbitrary waveforms somewhere for this generator?
You might find something useful in this thread:
https://www.eevblog.com/forum/testgear/siglent-sdg1025-matlab-programming-arb-waveform/

Dunno if they work with the later X model arbs but they all I got.  :-[
Avid Rabid Hobbyist.
Out of office and unavailable for a few days.
 

Offline n3mmr

  • Regular Contributor
  • *
  • Posts: 121
  • Country: se
Re: The Siglent SDG2042X Thread
« Reply #1487 on: February 16, 2022, 06:25:52 am »
Is there a repository of downloadable arbitrary waveforms somewhere for this generator?
You might find something useful in this thread:
https://www.eevblog.com/forum/testgear/siglent-sdg1025-matlab-programming-arb-waveform/

Dunno if they work with the later X model arbs but they all I got.  :-[

I couldn't find any ready-made waveform descriptions, just files stating the uart properties for something to be done in Matlab. It might even just be serial interface descriptions for connecting the instrument control toolbox to a siglent generator.
« Last Edit: February 16, 2022, 06:38:52 am by n3mmr »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29028
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1488 on: February 16, 2022, 07:03:13 am »
Is there a repository of downloadable arbitrary waveforms somewhere for this generator?
You might find something useful in this thread:
https://www.eevblog.com/forum/testgear/siglent-sdg1025-matlab-programming-arb-waveform/

Dunno if they work with the later X model arbs but they all I got.  :-[

I couldn't find any ready-made waveform descriptions, just files stating the uart properties for something to be done in Matlab. It might even just be serial interface descriptions for connecting the instrument control toolbox to a siglent generator.
Nothing in the inbuilt waveforms that are of use ?
Have you experimented using EasyWaveX ?

I'll ask Siglent if they have anything novel they can share.
Avid Rabid Hobbyist.
Out of office and unavailable for a few days.
 

Offline SignalArray

  • Contributor
  • Posts: 12
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1489 on: February 16, 2022, 03:46:32 pm »
will re-flashing to older version and then "converting " to SDG2122 (as per procedure outlined on this thread)  work for recently purchased SDG2042?

i'm planning to buy SDG2042.
 

Offline Martin.M

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
  • in Tek we trust
    • vintage Tek collection
Re: The Siglent SDG2042X Thread
« Reply #1490 on: February 16, 2022, 06:01:57 pm »
I am the next one,
SDG2082X is on the way  :)
 

Offline dorkshoei

  • Frequent Contributor
  • **
  • Posts: 510
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1491 on: February 16, 2022, 07:00:04 pm »
will re-flashing to older version and then "converting " to SDG2122 (as per procedure outlined on this thread)  work for recently purchased SDG2042?

i'm planning to buy SDG2042.

Read the hack thread.   This is from Feb 9.   Seven days ago.
https://www.eevblog.com/forum/testgear/siglent-sdg2042x-hack-door-closed/msg3996982/#msg3996982
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29028
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1492 on: March 18, 2022, 05:20:45 pm »
Is there a repository of downloadable arbitrary waveforms somewhere for this generator?
You might find something useful in this thread:
https://www.eevblog.com/forum/testgear/siglent-sdg1025-matlab-programming-arb-waveform/

Dunno if they work with the later X model arbs but they all I got.  :-[

I couldn't find any ready-made waveform descriptions, just files stating the uart properties for something to be done in Matlab. It might even just be serial interface descriptions for connecting the instrument control toolbox to a siglent generator.
Nothing in the inbuilt waveforms that are of use ?
Have you experimented using EasyWaveX ?

I'll ask Siglent if they have anything novel they can share.
Had these for a bit but sorry been too busy to fully test them.
CAN and I2C BIN files that you can import directly into 2ch SDG models.
NOTE, remove txt file extension required to upload here before uploading to SDG.

Another project for the adventurous; Hearts
https://www.eevblog.com/forum/chat/a-valentine_s-day-activity-for-your-scope-and-function-generator/
Reply #4 has a doc file with the required AWG settings.

Further project; enter these settings into you SDG and save them then download the BIN files and post them here.
TIA
Avid Rabid Hobbyist.
Out of office and unavailable for a few days.
 
The following users thanked this post: TurboTom, 2N3055, Mortymore

Offline MorpTester

  • Newbie
  • Posts: 2
  • Country: de
Re: The Siglent SDG2042X Thread
« Reply #1493 on: April 25, 2022, 07:10:00 am »
Sorry in case this was discussed and I missed it - how to do a screenshot with the SDG2042X onto the front USB drive? There is nothing mentioned about it in the manual.
 

Online ebastler

  • Super Contributor
  • ***
  • Posts: 6786
  • Country: de
Re: The Siglent SDG2042X Thread
« Reply #1494 on: April 25, 2022, 11:34:12 am »
Sorry in case this was discussed and I missed it - how to do a screenshot with the SDG2042X onto the front USB drive? There is nothing mentioned about it in the manual.

I am not aware of a direct "screenshot to USB" function; but a solution for "screenshot via SCPI" was mentioned in this thread a while ago: https://www.eevblog.com/forum/testgear/the-siglent-sdg2042x-thread/msg3577567/#msg3577567.

Disclaimer: I have not tried this myself.
« Last Edit: April 25, 2022, 11:36:05 am by ebastler »
 

Offline MorpTester

  • Newbie
  • Posts: 2
  • Country: de
Re: The Siglent SDG2042X Thread
« Reply #1495 on: April 25, 2022, 01:51:18 pm »
Thanks.

Edit: The command in the next post does indeed work, but it's quite complicated to finally receive and store a valid graphics file. This screenshot has been taken measuring a GPS diciplined source, after performing the Timebase Calibration of the Siglent with that source.


« Last Edit: May 08, 2022, 11:43:15 am by MorpTester »
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29028
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1496 on: April 25, 2022, 08:07:42 pm »
Yes it's a bit annoying the lack of consistency across the Siglent websites  ::)
 https://siglentna.com/operating-tip/sdg-remote-scpi-command-to-request-a-screen-image-2/

SCPI command: SCDP
Avid Rabid Hobbyist.
Out of office and unavailable for a few days.
 
The following users thanked this post: MorpTester

Offline Kanbus

  • Contributor
  • Posts: 26
  • Country: dk
Re: The Siglent SDG2042X Thread
« Reply #1497 on: May 01, 2022, 02:23:17 pm »
Hi
I have for hours tried upgrading the bandwith on my sdg2042x to no avail since the info seems scattered about and there is no guide for the sdg2042x

I take from one post that the telnet route dont work anymore and you could use the pyton script also used for the scopes.

I used this to unlock my sds1104x-e but i'm really confused now.

what do you use for scope id and how to get it?
should you type the model number since its not one of the listed ones?
which one of the keys in the script is for the 120Mhz bandwith?


With great power comes great smell
 

Offline blurpy

  • Regular Contributor
  • *
  • Posts: 236
  • Country: no
Re: The Siglent SDG2042X Thread
« Reply #1498 on: May 01, 2022, 03:50:13 pm »
Hi
I have for hours tried upgrading the bandwith on my sdg2042x to no avail since the info seems scattered about and there is no guide for the sdg2042x
There is a link to the last page of the hack thread just a few posts up, which describes the process step by step.
 

Offline Kanbus

  • Contributor
  • Posts: 26
  • Country: dk
Re: The Siglent SDG2042X Thread
« Reply #1499 on: May 01, 2022, 08:19:02 pm »
Thanks. next problem is it will not show the usb stick, i beeps and says usb stick removed when removing it thoug - i've tried several. It worked fine earlier.
With great power comes great smell
 