I only found options under utility to calibrate the touch screen. Where is the internal self calibration?
On page 54 of service manual.
Channel Self Adjustment
This adjustment is for internal control channel and signal channel. Through
feedback circuits of the feedback channel and signal channel, this adjustment
helps to obtain the actual working performance for components from signal
channels.
1. After the feedback channel adjustment completed, disconnect the two
BNC cables on CH1 and CH2.
2. Input the password “123654” under the system info interface to show the
SelfCal menu and perform the self adjustment.
3. Once the “Self Adjust” begins, the generator turns to adjustment interface
and the progress bar displays on the screen. In about 20 seconds it will
reach 100%, which indicates completing of the self-adjust. The progress
bar will disappear automatically after 2 seconds.