Author Topic: Test Equipment Anonymous (TEA) group therapy thread  (Read 17753881 times)

vk6zgo, Henrik_V, xrunner and 94 Guests are viewing this topic.

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29149
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97300 on: August 11, 2021, 08:41:29 pm »
Working the intermittent sweep mag issue on the horizontal board. See schematic. I went ahead and changed Q554 betting that transistor was going leaky or weird when hot. Seems like a good candidate anyway because the sweep mag switches are in that vicinity.

The Siglent is monitoring ramp trace from A Sweep at Q543. That waveform should not vary unless the Time/Div deck is exercised. The Fluke is monitoring the +100V current.

So far.....after about 20 minutes or so.....no mag variance. Will let it run and see what happens. 




Is the little flat top on the waveform acceptable ?
IME they were always a perfect sawtooth/triangle waveform.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97301 on: August 11, 2021, 08:43:02 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 
The following users thanked this post: med6753, mnementh, Mortymore, bd139, Kosmic, cyclin_al

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97302 on: August 11, 2021, 08:45:57 pm »
@bd send you one ? Have a spare 5d Mark 2 and a spare 500d ...

Will pass if it’s canon  :-DD

Got my eye on a Nikon mirrorless rabbit hole at the moment  :palm:
Yes, both are Canon bodies.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 
The following users thanked this post: bd139

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97303 on: August 11, 2021, 08:46:44 pm »
Upcoming TEA(cquisition) alert - teaser pic:      Got the photo from the seller, hopefully the unit will arrive in a few days

Bonus points for recognizing the instrument or the make. (that's too easy). Won't disclose before it is on my bench.
Japan's finest spec an like Tony's?

mnem
 >:D
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97304 on: August 11, 2021, 08:49:59 pm »


https://www.dx.com/p/creative-aircraft-gyro-support-car-air-outlet-mobile-phone-support-mouse-navigation-universal-bracket-2746652.html

OMG... this has to be the height of something... and the nadir of something else. It is the "connected age" equivalent of a propeller beanie; a dash-vent smartPwn holder with decorative propellers.  :o

There is just a universe of fail here; starting with the egregious pointlessness of it all, leading to the inevitable incapability to perform its basic function and finishing with the fact it is designed such that it will place spinning propellers right over the touchscreen.  |O

And somebody got paid to design this fucking thing. :palm:

mnem

« Last Edit: August 11, 2021, 08:52:41 pm by mnementh »
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 
The following users thanked this post: Specmaster, bd139

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97305 on: August 11, 2021, 08:57:14 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

   The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Yeah, it looks like he tried to patch it with heat-shrink like I did the first time I broke one. ;)

It actually does affect usability; when the contacts inevitably spread, it will get intermittent and eventually one half will break off. Then it will fail to detect wrong lead placement as it is supposed to warn you aboot. The latter is why the contact ring is split into two halves this way.

You want to get a new one on the way ASAP; this will quickly become a favorite tool and it deserves to be properly maintained. ;)

mnem
 :-DMM
« Last Edit: August 11, 2021, 09:01:15 pm by mnementh »
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline med6753

  • Super Contributor
  • ***
  • Posts: 11320
  • Country: us
  • Tek nut
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97306 on: August 11, 2021, 09:01:03 pm »

Is the little flat top on the waveform acceptable ?
IME they were always a perfect sawtooth/triangle waveform.

It is a perfect sawtooth. At the point it was measured it's inverted and agrees with the pulse shown on the schematic.
An old gray beard with an attitude.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97307 on: August 11, 2021, 09:07:37 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Ah yes common failure mode on these. The last 87 I had suffered the same problems and I replaced them.

Unfortunately fairly expensive to replace  :(
 

Offline AaronLee

  • Regular Contributor
  • *
  • Posts: 229
  • Country: kr
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97308 on: August 11, 2021, 09:08:30 pm »
I may have finally had a break thru.  :-+ The Type 547 ran as shown in the prior entry for over 2.5 hours without a hitch. It appears I was right about Q554 causing the intermittent sweep mag issue.

Back on it's feet as shown here. I can't proceed with re-building the vertical until the parts arrive. And it appears that the critical 105V zener won't show up until next week. I'll probably keep powering it up for periods to be absolutely certain all other issues are finally fixed. :-/O



Beautiful scope. Probably a stupid newbie question, but what are the yellow caps on the trigger input/output/etc. for?
 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4536
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97309 on: August 11, 2021, 09:11:19 pm »
I may have finally had a break thru.  :-+ The Type 547 ran as shown in the prior entry for over 2.5 hours without a hitch. It appears I was right about Q554 causing the intermittent sweep mag issue.

Back on it's feet as shown here. I can't proceed with re-building the vertical until the parts arrive. And it appears that the critical 105V zener won't show up until next week. I'll probably keep powering it up for periods to be absolutely certain all other issues are finally fixed. :-/O



Beautiful scope. Probably a stupid newbie question, but what are the yellow caps on the trigger input/output/etc. for?

They are protective caps, protecting the BNC connectors.

Edit:
Can be found here for example: https://www.ebay.com/itm/121136673516

« Last Edit: August 11, 2021, 09:14:54 pm by BU508A »
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 
The following users thanked this post: AaronLee

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97310 on: August 11, 2021, 09:19:38 pm »
Protects them from splatter when...

mnem
... never mind.
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline med6753

  • Super Contributor
  • ***
  • Posts: 11320
  • Country: us
  • Tek nut
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97311 on: August 11, 2021, 09:20:05 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Ah yes common failure mode on these. The last 87 I had suffered the same problems and I replaced them.

Unfortunately fairly expensive to replace  :(

Geez, I'm surrounded by Gorillas who break Fluke input terminals. Never broke a one and mine is 24 years old.

Do you drag it around by the test leads caveman style?  :P :-DD
An old gray beard with an attitude.
 
The following users thanked this post: Specmaster, Neomys Sapiens, bd139, ch_scr, AVGresponding

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97312 on: August 11, 2021, 09:20:56 pm »
Nah, that's not the real thing.

Also, as I commented on the video, a clock that does not talk NTP or PTP (and no %&€&%(#"%€ SNTP, if I may, please. ), or has a GNSS rx, should at least have a 10MHz/1PPS ref input so it'll actually show the time instead of an opinion. Seriously underwhelmed by most such designer efforts.

Fully agree. My (diy) clocks usually understand DCF77 (built-in receicer or input for demodulated pulses), my proprietary CAN "homebus" protocol, GPS PPS + NMEA, or NTP (broadcast by a local server).

Right on brothers! I mean, if it's more than ±100ns from UTC can you really call it a clock? Amiright, eh?  :)

It just means the display is more than 30m from the clock element :) (c~=1ns/ft)
And before my inner vision a virtual Grace Hopper appears holding 30cm of wire and exclaiming 'THIS is one nanosecond!'
In vacuum only.
Of course. And in Austria it is a mere 5cm.  (Huuusch, huuuusch...)

Hey... I came here from Texas, where everything is bigger, all the womens are beeyootiful, and every man has at least one nanosecond. ;)

mnem
"Here; hold muh beer and watch this!"
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97313 on: August 11, 2021, 09:21:59 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Ah yes common failure mode on these. The last 87 I had suffered the same problems and I replaced them.

Unfortunately fairly expensive to replace  :(

Geez, I'm surrounded by Gorillas who break Fluke input terminals. Never broke a one and mine is 24 years old.

Do you drag it around by the test leads caveman style?  :P :-DD

Just like yer woman does with you, big guy.  :-DD

mnem
we all know you like it rough... >:D
« Last Edit: August 11, 2021, 09:24:20 pm by mnementh »
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 2999
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97314 on: August 11, 2021, 09:24:56 pm »
Upcoming TEA(cquisition) alert - teaser pic:      Got the photo from the seller, hopefully the unit will arrive in a few days

Bonus points for recognizing the instrument or the make. (that's too easy). Won't disclose before it is on my bench.
Japan's finest spec an like Tony's?

mnem
 >:D

Looks like a Yokogawa scope to me, has a floppy drive to the right of the LCD screen but not the DL15xx series or earlier as they are CRT based, must be a later one from DL16xx or DL17xx series but I've not seen inside one of them.

David
« Last Edit: August 11, 2021, 09:35:00 pm by factory »
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29149
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97315 on: August 11, 2021, 09:26:33 pm »

Is the little flat top on the waveform acceptable ?
IME they were always a perfect sawtooth/triangle waveform.

It is a perfect sawtooth. At the point it was measured it's inverted and agrees with the pulse shown on the schematic.
OK so you are happy with it then.
Yes I saw some schematic waveforms had flat tops of varying degrees where they sometimes are post s/div settings whereas at the root of the sweep generator they should be a perfect triangle which is where I normally checked for the few I fixed that were without a sweep at all.
Carry on.......  :)
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97316 on: August 11, 2021, 09:34:41 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

   The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Yeah, it looks like he tried to patch it with heat-shrink like I did the first time I broke one. ;)

It actually does affect usability; when the contacts inevitably spread, it will get intermittent and eventually one half will break off. Then it will fail to detect wrong lead placement as it is supposed to warn you aboot. The latter is why the contact ring is split into two halves this way.

You want to get a new one on the way ASAP; this will quickly become a favorite tool and it deserves to be properly maintained. ;)

mnem
 :-DMM
I already have taken steps towards fixing that aspect, more on that later as it unfolds  ;), but I think you are mistaken on the warning system for the incorrect lead / switch position, the manual only makes mention of that for current jacks and is only operated by a lead being inserted in either of these jacks and the switch being in the incorrect position for the lead position. That means that this feature will still operate even if current is being measured, but the input jack selected could mean the meter could be overloaded, i.e., the lead is in the Amps jack, but the switch is in the uA position, the alert is still operational
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97317 on: August 11, 2021, 09:39:59 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Ah yes common failure mode on these. The last 87 I had suffered the same problems and I replaced them.

Unfortunately fairly expensive to replace  :(
      https://www.ebay.com/itm/262483877746

This one looks like the genuine article; single-piece overmolded red plastic, not multipart grey plastic like the Chinese knockoffs, and I'm pretty sure I recognize the username. Funny, the cheap knockoffs cost more than this guy now. Saw a couple at $50 plus shipping.  :palm:

mnem
 :-/O
« Last Edit: August 11, 2021, 09:42:41 pm by mnementh »
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97318 on: August 11, 2021, 09:49:57 pm »
By the time it gets here that’s £45 spent  :scared:
 
The following users thanked this post: mnementh

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97319 on: August 11, 2021, 09:52:45 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

   The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Yeah, it looks like he tried to patch it with heat-shrink like I did the first time I broke one. ;)

It actually does affect usability; when the contacts inevitably spread, it will get intermittent and eventually one half will break off. Then it will fail to detect wrong lead placement as it is supposed to warn you aboot. The latter is why the contact ring is split into two halves this way.

You want to get a new one on the way ASAP; this will quickly become a favorite tool and it deserves to be properly maintained. ;)

mnem
 :-DMM
I already have taken steps towards fixing that aspect, more on that later as it unfolds  ;), but I think you are mistaken on the warning system for the incorrect lead / switch position, the manual only makes mention of that for current jacks and is only operated by a lead being inserted in either of these jacks and the switch being in the incorrect position for the lead position. That means that this feature will still operate even if current is being measured, but the input jack selected could mean the meter could be overloaded, i.e., the lead is in the Amps jack, but the switch is in the uA position, the alert is still operational

The split contact is how it knows there's a lead plugged into the connector. When you open her up you'll see what I'm talking aboot; the two halves have separate traces. If you're plugged into Ω/V and turn it to A/mA/uA it warns you too.

Like I said; once they spread, contact becomes intermittent and eventually one breaks off completely. Obvi, the ones that get used all the time die first.

This is part of the cost of ownership on these beasts; despite all med's arguments to the contrary, it is a common failure mode and most of the [FLUKE] pirates know to look for it. That is probably the reason you got it at the price you got it. ;)

Don't let it stop you from fixing it properly and loving it tho. It's a great meter and I still miss my 87 from time to time. 8)

mnem
 :-DMM
« Last Edit: August 11, 2021, 09:56:07 pm by mnementh »
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97320 on: August 11, 2021, 09:54:24 pm »
By the time it gets here that’s £45 spent  :scared:

Yup. That's why the [FLUKE] flipper-pirates weren't interested. :o

mnem
 :-/O
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 
The following users thanked this post: bd139

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97321 on: August 11, 2021, 09:56:03 pm »
As promised, here are a few pics of the 85 doing voltage checks, everything works just as it should. The supplied leads were complete crap and even had a resistance of 4 ohms :palm: They are now in the bin, but I don't have a shortage of leads at the moment, so I'm OK.

The seller did say that the input jacks had been damaged, but these are available as spares, but they do not actually detract from it being used.

Ah yes common failure mode on these. The last 87 I had suffered the same problems and I replaced them.

Unfortunately fairly expensive to replace  :(
      https://www.ebay.com/itm/262483877746

This one looks like the genuine article; single-piece overmolded red plastic, not multipart grey plastic like the Chinese knockoffs, and I'm pretty sure I recognize the username. Funny, the cheap knockoffs cost more than this guy now. Saw a couple at $50 plus shipping.  :palm:

mnem
 :-/O
That is the correct one OK, but the price equates to £34.65 with shipping then there is the risk of customs duty to be paid on top, that is more than 60% of the price I paid for the complete 85  |O As I said, it does work perfectly OK as it is and should continue to do so forever and a day. If my plan comes to fruition, I'll explain how I intend to put that small defect right when it happens.  :-DD
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 

Offline mansaxel

  • Super Contributor
  • ***
  • Posts: 3558
  • Country: se
  • SA0XLR
    • My very static home page
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97322 on: August 11, 2021, 10:00:49 pm »
Right on brothers! I mean, if it's more than ±100ns from UTC can you really call it a clock? Amiright, eh?  :)

I'm friends with a bunch of the people who make UTC(SP) and distribute it to the country as "Swedish normal time" (although it's still UTC on the wire, of course).  They scratch the smiley from above. 100ns is a lot of deviation to them.

Offline mnementh

  • Super Contributor
  • ***
  • Posts: 17541
  • Country: us
  • *Hiding in the Dwagon-Cave*
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97323 on: August 11, 2021, 10:06:36 pm »
      https://www.ebay.com/itm/262483877746

This one looks like the genuine article; single-piece overmolded red plastic, not multipart grey plastic like the Chinese knockoffs, and I'm pretty sure I recognize the username. Funny, the cheap knockoffs cost more than this guy now. Saw a couple at $50 plus shipping.  :palm:

mnem
 :-/O
That is the correct one OK, but the price equates to £34.65 with shipping then there is the risk of customs duty to be paid on top, that is more than 60% of the price I paid for the complete 85  |O As I said, it does work perfectly OK as it is and should continue to do so forever and a day. If my plan comes to fruition, I'll explain how I intend to put that small defect right when it happens.  :-DD
Yeah, I tried heat-shrink and CA as well as a bit of semi-rigid spray bottle tubing I found that was exactly the right diameter to fit over the contacts. Nothing stays put. The problem is that the overmolded plastic actually supports the contacts, not the other way around.

I don't know if that guy will ever be able to get genuine [FLUKE] connector bridges again once those sell out. :-// Myself, I'd just suck it up and do the fix right. That meter is worth it.

If you pot the sockets with epoxy such that sleeved leads can no longer be used, the [FLUKE] gods will smite you, and I will cheer. ;)

mnem
 :horse:
alt-codes work here:  alt-0128 = €  alt-156 = £  alt-0216 = Ø  alt-225 = ß  alt-230 = µ  alt-234 = Ω  alt-236 = ∞  alt-248 = °
 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4536
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #97324 on: August 11, 2021, 10:11:29 pm »
Just sniped ten LM723H/883Q NOS from National for ca. 17.50 Euro (plus shipping).  :D  :-+

https://www.ebay.de/itm/402994733209

“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 
The following users thanked this post: Wolfgang, mnementh, Neomys Sapiens, wolfy007, ch_scr, syau