Another one from the HP Harrison 6824A bipolar power supply / amplifier.
Now as an amplifier.
Situation: Load Test.
Input: 2.5V RMS sine wave 400Hz
Output: variable depending on amplification factor (adjustable)
First screenshot: oscillogram of the output waveform at 118,4V peak-to-peak. No distortion/clipping visible:
Second screenshot: output waveform at 121,6V peak-to-peak. Beginning of clipping at the positive peak visible:
Third screenshot: output waveform at 125,6V peak-to-peak. Clipping at the positive peak clearly visible. No clipping at the negative peak. Hmm.
Fourth screenshot: output waveform at a slightly higher amplification factor:
This looks strange to me; asymetrical and with a jump back like a limiter kicking in delayed -inspite of the negative half of the amplifier being able to deliver the peak voltage (?).
Fifth screenshot: output waveform with an even higher amplification factor:
The limit seems to be stable, the sine wave gets cut off "earlier" with higher voltage levels as it is to be expected.
I'm curious why the amplifier of that 6824A doesn't clip symmetrically.
Have to explore the service manual parts more deeply..
Didn't notice that my siglent gave hints to me while capturing screenshots..