Author Topic: Test Equipment Anonymous (TEA) group therapy thread  (Read 17872515 times)

0 Members and 32 Guests are viewing this topic.

Offline worsthorse

  • Super Contributor
  • ***
  • Posts: 1237
  • Country: us
  • aina varma, usein väärin
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38700 on: September 12, 2019, 04:32:28 am »
Well my hopes of "touch it up and sell it" for the 2831D DMM were dashed tonight. I did a quick check and found it out of spec by a little for DC volts, low for resistance readings, and a little high on AC Volts. But I found the service manual and there are adjustments for each, so no problem!

Unfortunately, the meter doesn't zero on the AC ranges and I can't adjust the 10V DC range within spec. In both cases the adjustment pot is hard against one limit. Worse yet, When I removed the metal shield on the main circuit board, I discovered that the plastic trimmer screw on one of the pots had been stripped by some gorilla. I guess he/she/it was trying to force the adjustment past the limit. I can still turn it but I should probably replace it.

Time to cozy up with the schematics. Tomorrow I will start checking the power supply and reference voltages.
specialization is for insects.
 

Offline worsthorse

  • Super Contributor
  • ***
  • Posts: 1237
  • Country: us
  • aina varma, usein väärin
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38701 on: September 12, 2019, 04:34:19 am »
I left mine on over supper just now and I found it pegged downscale.  :-//

Does yours drift more when you touch the thermistor body? This is metal part just behind the N connector. Mine seems to move a lot when I apply body heat to this area.

Oh, wow. When I hold onto the thermistor, it goes almost full scale in a matter of seconds. So much for, "Design of the mount minimizes adverse effects from environmental temperature changes during measurement," and "Each mount contains two series pairs of thermistors, which are matched to cancel the effects of drift with ambient temperature change."

Tapping or moving the mount and cable around makes the needle wiggle, too.

I don't know what to expect since I haven't owned these things before. Maybe such sensitivity is normal when set to measure tens of µW?

I am watching this discussion with some interest. I have three HP power sensors, from microwatts to watts, and three meters of various sorts. I want to start working on them once I finish the last of the DMMs (two down, one on the bench, two to go).
specialization is for insects.
 

Offline bitseekerTopic starter

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38702 on: September 12, 2019, 07:00:32 am »
Cool. Feel free to join the party, wch!
TEA is the way. | TEA Time channel
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38703 on: September 12, 2019, 07:02:27 am »
HP power sensors is one graveyard I do not set foot in  :-DD

Edit: cheap, new 34460A here: https://cpc.farnell.com/keysight-technologies/34460a/multimeter-digital-bench-6-1-2/dp/IN0671269 (£200 off usual price). I honestly did consider it and bouncing the 34401A for a few moments  :-DD. This doesn’t have Ethernet or rear terminals though like the 34461A though.
« Last Edit: September 12, 2019, 07:44:22 am by bd139 »
 

Offline Mr. Scram

  • Super Contributor
  • ***
  • Posts: 9810
  • Country: 00
  • Display aficionado
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38704 on: September 12, 2019, 08:21:07 am »
HP power sensors is one graveyard I do not set foot in  :-DD

Edit: cheap, new 34460A here: https://cpc.farnell.com/keysight-technologies/34460a/multimeter-digital-bench-6-1-2/dp/IN0671269 (£200 off usual price). I honestly did consider it and bouncing the 34401A for a few moments  :-DD. This doesn’t have Ethernet or rear terminals though like the 34461A though.
I'd rather take the DMM6500.
 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4537
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38705 on: September 12, 2019, 08:34:08 am »
HP power sensors is one graveyard I do not set foot in  :-DD

Edit: cheap, new 34460A here: https://cpc.farnell.com/keysight-technologies/34460a/multimeter-digital-bench-6-1-2/dp/IN0671269 (£200 off usual price). I honestly did consider it and bouncing the 34401A for a few moments  :-DD. This doesn’t have Ethernet or rear terminals though like the 34461A though.
I'd rather take the DMM6500.

Nice review from Shariar of the very similar DAQ6510:

http://thesignalpath.com/blogs/2019/08/04/keithley-daq6510-6-5-digit-data-acquisition-multimeter-review-teardown-experiments/
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 
The following users thanked this post: Mr. Scram

Offline Mr. Scram

  • Super Contributor
  • ***
  • Posts: 9810
  • Country: 00
  • Display aficionado
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38706 on: September 12, 2019, 08:35:55 am »
Nice review from Shariar of the very similar DAQ6510:

http://thesignalpath.com/blogs/2019/08/04/keithley-daq6510-6-5-digit-data-acquisition-multimeter-review-teardown-experiments/
I've watched that one a handful of times. TSP is obviously always a joy too. The DAQ is a bit too pricey for my specific needs. I'd probably rather add it as a second unit when and if I outgrow the first one.
« Last Edit: September 12, 2019, 08:37:35 am by Mr. Scram »
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38707 on: September 12, 2019, 08:39:35 am »
300 more galactic credits though here and you have to spend on acquisition modules if you really want to make use of it as it is intended.

Edit: also as I think I mentioned elsewhere, the thing is pretty slow in switched config when you include settling time and switching time. Better to buy a couple of basic DMMs and cheaper switches. That’s if you care about test system throughput.
« Last Edit: September 12, 2019, 08:42:05 am by bd139 »
 

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38708 on: September 12, 2019, 09:10:55 am »
300 more galactic credits though here and you have to spend on acquisition modules if you really want to make use of it as it is intended.

Edit: also as I think I mentioned elsewhere, the thing is pretty slow in switched config when you include settling time and switching time. Better to buy a couple of basic DMMs and cheaper switches. That’s if you care about test system throughput.
It is also rather spartan on the specs, 750VDC, come on it cannot be that hard to take it upto at least 1,000V and only a 3A current rating, I know that is about normal for a bench meter, but again, its hardly difficult to stick a larger shunt inside is now?

I know that these ranges are not always used very often, but it is nice to have them there for the odd time that you do need to go higher.  >:D
 
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 
The following users thanked this post: bd139

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4537
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38709 on: September 12, 2019, 09:11:29 am »
300 more galactic credits though here and you have to spend on acquisition modules if you really want to make use of it as it is intended.

Edit: also as I think I mentioned elsewhere, the thing is pretty slow in switched config when you include settling time and switching time. Better to buy a couple of basic DMMs and cheaper switches. That’s if you care about test system throughput.

You mean something like the Aneng AN8009?

https://www.aliexpress.com/item/32839539223.html

“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 
The following users thanked this post: bd139

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38710 on: September 12, 2019, 09:14:39 am »
 :palm: :-DD

Was thinking more a couple of 34460A's or something :)

 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4537
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38711 on: September 12, 2019, 09:25:55 am »
Ah! When you are saying "basic DMM" and thinking of "I'm cheap" the Aneng was the first one which was coming to my mind.
I was wondering where the GPIB connection will be, though.  :-//   :-DD
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline Mr. Scram

  • Super Contributor
  • ***
  • Posts: 9810
  • Country: 00
  • Display aficionado
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38712 on: September 12, 2019, 09:26:28 am »
300 more galactic credits though here and you have to spend on acquisition modules if you really want to make use of it as it is intended.

Edit: also as I think I mentioned elsewhere, the thing is pretty slow in switched config when you include settling time and switching time. Better to buy a couple of basic DMMs and cheaper switches. That’s if you care about test system throughput.
Just buy four DMM6500s. That TSP system link is interesting.
 

Offline Zucca

  • Supporter
  • ****
  • Posts: 4558
  • Country: it
  • EE meid in Itali
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38713 on: September 12, 2019, 10:34:28 am »
cheap, new 34460A here: https://cpc.farnell.com/keysight-technologies/34460a/multimeter-digital-bench-6-1-2/dp/IN0671269 (£200 off usual price). I honestly did consider it and bouncing the 34401A

I sold all my four 34401A for kept my 34461A, now everytime I press the on button I have to wait until it boots; with the 34401A I was ready to go immediately.
Nothing critical, but it could be annoying for someone.
Can't know what you don't love. St. Augustine
Can't love what you don't know. Zucca
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38714 on: September 12, 2019, 10:45:11 am »
Broken Telequipment D75 in Hammersmith £20. Do I need another analogue scope?  :-DD

Worst case it’s got £20 worth of dual FETs, pots and knobs in it.

Edit: too late; I did it  :-DD. Still not got over the loss of the D83’s and this is basically a D83 with a 465 tube in it. I have the benefit now of a huge amount of spares.
« Last Edit: September 12, 2019, 10:58:24 am by bd139 »
 

Offline nixiefreqq

  • Super Contributor
  • ***
  • Posts: 1028
  • Country: us
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38715 on: September 12, 2019, 11:05:32 am »
Broken Telequipment D75 in Hammersmith £20. Do I need another analogue scope?  :-DD

Worst case it’s got £20 worth of dual FETs, pots and knobs in it.

Edit: too late; I did it  :-DD. Still not got over the loss of the D83’s and this is basically a D83 with a 465 tube in it. I have the benefit now of a huge amount of spares.

we are agog in south central PA.

what happened with the 1740a?
free range primate
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29210
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38716 on: September 12, 2019, 11:05:45 am »
Still not got over the loss of the D83’s and this is basically a D83 with a 465 tube in it.
Me neither, it's the only CRO I now wish I'd hung onto despite that I always had my head under its hood !  :horse:
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38717 on: September 12, 2019, 11:08:34 am »
Broken Telequipment D75 in Hammersmith £20. Do I need another analogue scope?  :-DD

Worst case it’s got £20 worth of dual FETs, pots and knobs in it.

Edit: too late; I did it  :-DD. Still not got over the loss of the D83’s and this is basically a D83 with a 465 tube in it. I have the benefit now of a huge amount of spares.

we are agog in south central PA.

what happened with the 1740a?

Still here. It was waiting for an LM723 which just came through the door five minute ago so will get into that now.

Still not got over the loss of the D83’s and this is basically a D83 with a 465 tube in it.
Me neither, it's the only CRO I now wish I'd hung onto despite that I always had my head under its hood !  :horse:

Yeah a working one is a glorious thing. But rare  :-DD

This one appear to have an intact timebase control so I want to carefully dismantle it and measure so I can get a replica 3d printed. Lots of demand for that. The CT curve tracer and all the plugins use them and they are rare as anything!
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38718 on: September 12, 2019, 11:44:10 am »
Still on holiday. Woohoo. That means 1740A time. So put the new LM723 in. Fortunately this one is socketed so it didn't mean removing the power supply again and was really easy.

So firstly the new LM723 regulators installed. Fortunately one of them was still good:



The repaired and re-connectored power supply. 6-way connectors were cheaper so I just got them. Also made for easier debugging which was handy:



Fired it up and voltages were slightly too high so I trimmed them down again to 15.000V and all the rails are now spot on.

Anyway first time this is running under its own power supply:



Seems to work ok. The timebase needs a bit of a squirt of contact cleaner however. Sort of expected with the ganky switches in there.

Next steps are:

1. I'm going to add a small heatsink made out of a bit of bent alu sheet to the 15V rectifier as it runs very hot. This should stop it taking the rest of the damn thing out.
2. Clean timebase out and calibrate it.
3. Feeeet. beanflying's 1740A feet will be printed by a friend of mine at some point in the next couple of weeks. He's got a relatively decent printer and is doing it for free if I set up his WiFi extender :)
 
The following users thanked this post: xrunner, med6753, Kosmic, grbk

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38719 on: September 12, 2019, 11:52:22 am »
I left mine on over supper just now and I found it pegged downscale.  :-//

Does yours drift more when you touch the thermistor body? This is metal part just behind the N connector. Mine seems to move a lot when I apply body heat to this area.

Oh, wow. When I hold onto the thermistor, it goes almost full scale in a matter of seconds. So much for, "Design of the mount minimizes adverse effects from environmental temperature changes during measurement," and "Each mount contains two series pairs of thermistors, which are matched to cancel the effects of drift with ambient temperature change."

Tapping or moving the mount and cable around makes the needle wiggle, too.

I don't know what to expect since I haven't owned these things before. Maybe such sensitivity is normal when set to measure tens of µW?

Maybe. It’s good to know that mine isn’t the only one. Mine seems to measure power just fine, agreeing almost exactly with the 436A meter I have in short term. It just drifts a lot more. 
 

Offline BU508A

  • Super Contributor
  • ***
  • Posts: 4537
  • Country: de
  • Per aspera ad astra
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38720 on: September 12, 2019, 12:01:16 pm »
Nice job.  :)  :-+

Fired it up and voltages were slightly too high so I trimmed them down again to 15.000V and all the rails are now spot on.


Why haven't you activated the 6 1/2 digit mode of your gorgeous 34401A?  :-/O  :-DD
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline xrunner

  • Super Contributor
  • ***
  • Posts: 7612
  • Country: us
  • hp>Agilent>Keysight>???
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38721 on: September 12, 2019, 12:04:11 pm »
Nice job.  :)  :-+

Fired it up and voltages were slightly too high so I trimmed them down again to 15.000V and all the rails are now spot on.


Why haven't you activated the 6 1/2 digit mode of your gorgeous 34401A?  :-/O  :-DD

Waiting on the new display.  :popcorn:
I told my friends I could teach them to be funny, but they all just laughed at me.
 

Offline med6753

  • Super Contributor
  • ***
  • Posts: 11320
  • Country: us
  • Tek nut
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38722 on: September 12, 2019, 12:05:31 pm »
Still on holiday. Woohoo. That means 1740A time. So put the new LM723 in. Fortunately this one is socketed so it didn't mean removing the power supply again and was really easy.

So firstly the new LM723 regulators installed. Fortunately one of them was still good:



The repaired and re-connectored power supply. 6-way connectors were cheaper so I just got them. Also made for easier debugging which was handy:



Fired it up and voltages were slightly too high so I trimmed them down again to 15.000V and all the rails are now spot on.

Anyway first time this is running under its own power supply:



Seems to work ok. The timebase needs a bit of a squirt of contact cleaner however. Sort of expected with the ganky switches in there.

Next steps are:

1. I'm going to add a small heatsink made out of a bit of bent alu sheet to the 15V rectifier as it runs very hot. This should stop it taking the rest of the damn thing out.
2. Clean timebase out and calibrate it.
3. Feeeet. beanflying's 1740A feet will be printed by a friend of mine at some point in the next couple of weeks. He's got a relatively decent printer and is doing it for free if I set up his WiFi extender :)

Looks good!  :-+ I thought you mentioned that one of the pass transistors had taken a dump? No? It was just the 723?
An old gray beard with an attitude.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23059
  • Country: gb
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38723 on: September 12, 2019, 12:30:42 pm »
Thought it was a pass transistor but it was the drive current from the LM723. I stuffed a DMM in series with base and it was only driving 4mA through whereas for the beta it should have been about 30mA. That was causing rails to collapse.

Anyway made a heatsink quick. Took longer to get the bloody screw in than make the heatsink. Disclaimer: I have zero metalworking tools here so this was literally aviation shears, file and hand drill :-DD

 

Offline xrunner

  • Super Contributor
  • ***
  • Posts: 7612
  • Country: us
  • hp>Agilent>Keysight>???
Re: Test Equipment Anonymous (TEA) group therapy thread
« Reply #38724 on: September 12, 2019, 12:41:16 pm »
Not bad at all!
I told my friends I could teach them to be funny, but they all just laughed at me.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo