Hello,
Today teardown of a unlucky 7A29 plugin.
I was troubleshooting it and started to take pics as
reference for the reassembly process, so here are them.
The plugin is a single channel vertical amplifier for the 7000
series oscilloscopes, if used with the 7104 or R7103
the bandwidth is a nice round 1GHz.
This unit has a broken position knob, and some
ranges that dont work.
The internal construction is nice, with the signal path
built more like water piping than electrical circuit
On the right side there is the switching assemblies ,
the readout switch top rear, and amplifer chain.
The amplifers are the square ceramic hybrid ics.
On the left side there is the delay line, and bottom
side of the main board.
The signal path is simple: the signal enters from
the front bnc, goes trough the delay line and
into the bottom switching assembly that contains
the AC/DC selection and input protection relay.
From the first assembly it goes to the second
switching assembly on top, where it is attenuated
and then to the amplifiers.
Here is the problem, in all it's glory, it is the attenuator.
Contains three group of switches and resistors.
The stages provide attenuation factors of 2x 5x and 10x,
while they keep the input impedance 50ohm.
Each stage is be combined with the others to
give a total of 7 ranges in 1-2-5 sequence,
from 10mV to 1V per division.
This is the camshaft that actuates the switches
in sequence, through metallic rollers.
This is the other side of the actuators.
And finally the actuators removed from their
sliding vane, here can be seen that the
actuators push on 4 pins each: two on
top and two on bottom, and each
pin has it's trimming screw.
Time to remove the input and output connectors.
And remove one of the covers. On the bottom
plate there are the gold plated switches, and
on the top there are the resistive attenuators,
printed on ceramic boards.
The metallic strings, are pushed down to short,
or up to touch the attenuators.
Here the hybrid attenuators, on the left the 5x,
center 10x and right 2x.
The problem is the attenuators are
fried, in particular the 2x has the resistors
to ground opened. The left and right
pads go to ground, while top and bottom
are inserted in the signal path when
the 2x is selected:
The 10x has part of the resistors overheated
and off-spec, bottom of the picture:
the 5x is fine, the black lines are
the cuts used to trim resistors to
the correct value (laser?):
I hope I will find some spares...
That's all for now.
Fabio.