Ever wondered if your Temperature measurements are accurate?
I have been casually looking for one of these for a long while to test Thermocouples, RTD's and other Temperature measuring devices. Below Freezing and from 0-100 can be done with reasonable accuracy using more conventional cludges such as Ice/Water or a combination of both. For commercial Ovens, Kilns or one of the areas I play with Coffee Roasters being able to accurately test measurement devices well above that are needed.
This example turned up on the other side of Australia out of Perth and was what I consider a bargain at $150USD (only bidder
![Confused :-//](https://www.eevblog.com/forum/Smileys/default/confused0024.gif.pagespeed.ce.5xOqKkq0Co.gif)
) to my door with a range of different diameter inserts compared to $700-1k generally without inserts/blocks. Its only downside is for higher temperature kilns as it has a limit of 600C but for what I do not really an issue and above that most of the common thermocouples are 'reasonably' linear so if you track them below fairly safe above will be close. If I develop a need for higher temperature calibration then back to the bay of evil again.
I will get around to putting a more accurate 4 wire RTD on it over the next few days but overall it while it is well out of current NATA (NIST) Cal it seems to be close to its +-1C accuracy spec with +-0.1C stability.
Closest currently available model (0-200C)
https://www.sika.net/en/products/test-and-calibration-instruments/temperature-calibrators/series-tp-basic/item/965-temperature-calibrator-type-tp-18200e.html#technical-features Programming is simple and the operation manual is fairly much ok to use on this model.
Basic construction is made up of a hot and an electronic side kept apart by a small airgap but the outside case remains cool even at fairly high temperatures on both so the electronics is really only at ambient. The Spagetti will be getting some attention or some cable ties at a minimum
![Flogging A Dead Horse :horse:](https://www.eevblog.com/forum/Smileys/default/DeadHorse.gif.pagespeed.ce.kBWvHKqCd1.gif)
The main Block takes an age to cooldown so you really want to take your calibration points on the way up. More accurate testing to come.