Thanks for the suggestion.  I'll take a look at that thread.

Shariar's video, however, is the one I linked above, where he does exactly what I am doing...

 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #251 on: January 14, 2019, 04:50:29 am »
Thanks for the suggestion.  I'll take a look at that thread.

Shariar's video, however, is the one I linked above, where he does exactly what I am doing...
Yep, well watch it again from 37.20 where Shahriar explains that changing settings will impact the normalization and he recommends once the settings are as you want then to re normalize for accurate results.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline rmel

  • Regular Contributor
  • *
  • Posts: 50
  • Country: us
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #252 on: January 14, 2019, 05:20:50 am »
Ah yes, the dots are starting to connect.

I appreciate the patience you and everyone else has shown in helping me to understand what was happening.
 
The following users thanked this post: tautech

Offline rmel

  • Regular Contributor
  • *
  • Posts: 50
  • Country: us
Reference Offset and Trace View
« Reply #253 on: January 20, 2019, 08:26:53 am »
Ok, here's another behavior that seems weird to me.

Can someone tell me what I did wrong here?

Here's what I did:

I am feeding a square wave at 30 MHz through a 20 dB attenuator to the SA.

Preset -> Amplitude -> Reference Offset -> 20dB
Peak -> Peak table on
Trace -> View

Then the peak values in the Peak Table seem to go nuts and the waveform disappears.

Here's a video I captured:

https://gfycat.com/SmallGivingAmericanwarmblood

 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #254 on: March 25, 2019, 07:29:57 pm »
SVA1015X Defpom overview video:

https://youtu.be/VXXlJkPwxSw
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1424
  • Country: de
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #255 on: March 25, 2019, 08:40:59 pm »
Sorry but using two *10 scope probes to "demo measure" with the SVA1015X's VNA function in an unpowered CB radio corpse is, well, a little "unresultful"   :palm:
This RF stuff definitely isn't Defpom's business, clicking through the menus isn't what really helps someone seriously interested in a purchase. Regarding things one doesn't feel "at home" with, better do yourself a favor and let others who are more into it do this job.
Watching this clip is a waste of time, and probably producing it was even more so for Defpom. He's got much better reviews in his channel than this one...

Cheers and no offense intended  :phew:
 
The following users thanked this post: tinhead, egonotto, Wolfgang, 1Ghz, 2N3055, TK, bb37

Offline TheDefpom

  • Frequent Contributor
  • **
  • Posts: 748
  • Country: nz
  • YouTuber Nerd - I Fix Stuff
    • The Defpom's Channel
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #256 on: April 22, 2019, 10:49:16 am »
Sorry but using two *10 scope probes to "demo measure" with the SVA1015X's VNA function in an unpowered CB radio corpse is, well, a little "unresultful"   :palm:
This RF stuff definitely isn't Defpom's business, clicking through the menus isn't what really helps someone seriously interested in a purchase. Regarding things one doesn't feel "at home" with, better do yourself a favor and let others who are more into it do this job.
Watching this clip is a waste of time, and probably producing it was even more so for Defpom. He's got much better reviews in his channel than this one...

Cheers and no offense intended  :phew:

What I was trying to demonstrate was the filtering effects of the tuned circuits in the RF section of the radio, and the ability to see that on the Siglent, no I wasn't trying to obtain accurate and true measurements, and at no time did I say that was what I was doing.

My reviews are done in a way to try and replicate how a new user would find the gear, something that is easy enough to capture as sometimes (such as in this case) I am not an expert in its usage having only limited experience with the particular item.
Cheers Scott

Check out my Electronics Repair, Mailbag, or Review Videos at https://www.youtube.com/TheDefpom
 
The following users thanked this post: TurboTom

Offline jemangedeslolos

  • Frequent Contributor
  • **
  • Posts: 386
  • Country: fr
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #257 on: April 26, 2019, 04:56:50 pm »
Hello,

I have some trouble with UHF RFID developpement and it is maybe the time for me to buy a this kind of tool.

Almost All Siglent instruments have been hacked.
Is this the case also for this one ?

The start price is pretty decent but It will be nice to have EMI precompliance, Advanced Measurement Kit and VNA too  >:D
 

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 438
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #258 on: April 27, 2019, 08:07:46 pm »
I just posted a video showing the power consumption, power factor, and line current waveforms for the SVA1015X spectrum/vector network analyzer.

I also demonstrate a consistent hard crash mode when the VNA mode is turned on while the preamp is on in the spectrum analyzer mode.  This crash requires a power cycling to reset.



If you missed it I also posted a review of the vector network analyzer mode a few weeks ago



I hope you find it interesting.

Chris
« Last Edit: April 27, 2019, 08:16:20 pm by graybeard »
 
The following users thanked this post: tinhead, Performa01, edigi

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #259 on: April 29, 2019, 01:10:08 am »
I also demonstrate a consistent hard crash mode when the VNA mode is turned on while the preamp is on in the spectrum analyzer mode.  This crash requires a power cycling to reset.


I hope you find it interesting.

Chris
More so amusing.  :)
Yes the bug needs fixing to protect the operator from their own usage errors.  ;)

Thanks for finding these silly UI errors for Siglent to fix. 
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline hendorog

  • Super Contributor
  • ***
  • Posts: 1626
  • Country: nz
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #260 on: April 29, 2019, 04:45:57 am »
@graybeard

IIRC it is possible to use custom cal kit in the VNA mode of the SVA - for a competent unix user.

The cal kit definition is stored in a file on the file system, and could be backed up and modified to suit another cal kit.

Also IIRC the definitions are stored in a file as mag/phase data, so any standard which one has measurement data for can be used as a cal standard.
IMHO this is a more comprehensive method than using the traditional Cx/Lx model which is what I think you are proposing. I do agree Siglent should implement custom cal kits in the UI.
 

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 438
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #261 on: May 02, 2019, 04:06:29 pm »
@graybeard

IIRC it is possible to use custom cal kit in the VNA mode of the SVA - for a competent unix user.

The cal kit definition is stored in a file on the file system, and could be backed up and modified to suit another cal kit.

Also IIRC the definitions are stored in a file as mag/phase data, so any standard which one has measurement data for can be used as a cal standard.
IMHO this is a more comprehensive method than using the traditional Cx/Lx model which is what I think you are proposing. I do agree Siglent should implement custom cal kits in the UI.

The gain and phase information is the best method, especially at much higher frequencies where a lumped model is inadequate.  However in the frequency range of the SVA1015x lumped models work well.

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 438
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Review of the SVA1015X Network Analyzer Option Firmware Update
« Reply #262 on: June 03, 2019, 07:53:39 am »
Siglent is releasing new firmware for the SVA1015X spectrum/vector network analyzer today.
In the following video I review firmware changes to the Network Analyzer Option.  This firmware fixes flaws and adds excellent new capability to the SVA including port extensions, multiple simultaneous plot types, additional calibration types including user cals, on-screen annotations, and a read only FTP server.



You can get a copy of my presentation here if you want to follow along.

I hope you enjoy it.

Chris

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #263 on: June 03, 2019, 08:07:45 pm »
And here's the link to the new SVA1015X firmware.

Version 2.2.1.2.1
38.8 MB
https://www.siglentamerica.com/download/9029/

Release notes
New Functionality:
 Add CNR, Harmonic measurement in AMK options for SA Mode
 Add Port Extensions for VNA Mode
 Add QAM, MSK, PSK modulation analysis for MA Mode
 Add basic SCPI commands for all Modes
 Add Screen Annotation in Display menu
 Add FTP server for File operation
Improvements:
 Optimize UI, refresh and response for all Modes
 Optimize SCPI commands, such as INIT, *OPC?, *ESR, Abort, etc.
 Optimize Log scale axes, Peak search, Auto tune, TG normalization for SA Mode
 Optimize S11 and S21 simultaneous measurement, calibration control, and user define cal kit for VNA Mode
 Optimize multi traces, and multi-format overlay display for VNA and MA Mode
 Optimize Power Saving, Web Server,
Solved Issues:
 Fixed OBW Transmit Freq Error
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline hendorog

  • Super Contributor
  • ***
  • Posts: 1626
  • Country: nz
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #264 on: June 03, 2019, 08:37:07 pm »
Looks like a very big jump in the UI and features!

Is there s-file export yet? Looking at the video it appears to be a csv data dump, but not in a standard s1p format. Is that correct?
Is there a plan to implement Enhanced Response calibration type?

With Enhanced Response calibration and a good attenuator, to improve load match, I believe you can get results almost as good as a full 2 port calibration.




 

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 438
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #265 on: June 03, 2019, 09:16:31 pm »
Looks like a very big jump in the UI and features!

Is there s-file export yet? Looking at the video it appears to be a csv data dump, but not in a standard s1p format. Is that correct?
Is there a plan to implement Enhanced Response calibration type?

With Enhanced Response calibration and a good attenuator, to improve load match, I believe you can get results almost as good as a full 2 port calibration.

The data export is only CSV, there is no S1P file generated.
   
If I wanted a S1P file I would just set up an EXCEL or Libre Office template to do the translation.

Chris
« Last Edit: June 03, 2019, 09:22:00 pm by graybeard »
 

Offline hendorog

  • Super Contributor
  • ***
  • Posts: 1626
  • Country: nz
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #266 on: June 03, 2019, 09:51:18 pm »
Looks like a very big jump in the UI and features!

Is there s-file export yet? Looking at the video it appears to be a csv data dump, but not in a standard s1p format. Is that correct?
Is there a plan to implement Enhanced Response calibration type?

With Enhanced Response calibration and a good attenuator, to improve load match, I believe you can get results almost as good as a full 2 port calibration.

The data export is only CSV, there is no S1P file generated.
   
If I wanted a S1P file I would just set up an EXCEL or Libre Office template to do the translation.

Chris

Yes of course. But why not? It is the standard format if you want to integrate the data into any other software.

But it certainly is nice they have now made some progress on the basic stuff thanks to your prodding.

Edit: Pro's and cons of Enhanced Response cal for a T/R VNA is covered in here on page 53:
http://anlage.umd.edu/Microwave%20Measurements%20for%20Personal%20Web%20Site/Agilent%20NWA%20Basics%205965-7917E.pdf
« Last Edit: June 03, 2019, 09:59:28 pm by hendorog »
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11754
  • Country: us
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #267 on: June 04, 2019, 12:11:27 am »
Any chance we will see a 2-port 3GHz version any time soon? 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #268 on: June 04, 2019, 01:04:54 am »
Any chance we will see a 2-port 3GHz version any time soon?
No Chinese whispers about the SVA3000X models yet other than the one spotted at Electronica last year.
Even the beta testers that I chat to haven't seen one yet.
Some months away yet I guess.  :(
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #269 on: June 14, 2019, 01:55:12 am »
Any chance we will see a 2-port 3GHz version any time soon?
No Chinese whispers about the SVA3000X models yet other than the one spotted at Electronica last year.
Even the beta testers that I chat to haven't seen one yet.
Some months away yet I guess.  :(
Update.

Based on this post:
Here's a snapshot of the Siglent showcase at Electronica 2018.

The SVA3032X prototype analyzes a 16QAM signal produced by the new SSG3032X-IQE, which in turn is using an SDG6000X with I/Q option as the modulation source.

The display of the SVA3032X has been brought to the big monitor using the internal webserver and a notebook computer hidden under the counter ;)

We thought a SVA3032X was coming however recently the Chinese website has been updated to indicate a SVA1032X is actually coming.

Noticeable improvements:
Vector network analysis mode with frequencies ranging from a minimum of 100 kHz to a maximum of 3.2 GHz.
DANL -161 dBm/Hz
« Last Edit: June 14, 2019, 03:28:24 am by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #270 on: June 18, 2019, 07:29:47 pm »
And here's the link to the new SVA1015X firmware.

Version 2.2.1.2.1

Some families asked me for this patch. It should enable telnet as before.

Test and report.
« Last Edit: June 19, 2019, 03:28:32 pm by tv84 »
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #271 on: June 18, 2019, 08:49:45 pm »
No Chinese whispers about the SVA3000X models yet other than the one spotted at Electronica last year.
Even the beta testers that I chat to haven't seen one yet.
Some months away yet I guess.  :(

BTW, no need for whispers.

This latest FW (which is quite different from previous ones) has the explanations inside:   :popcorn:

SVA1015X
SVA3032X
SVA7075X
   :scared:

And 5 other options besides what is currently used:

RTA, RCV, SVA, NA and MA

 
The following users thanked this post: tautech

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #272 on: June 18, 2019, 09:05:06 pm »
No Chinese whispers about the SVA3000X models yet other than the one spotted at Electronica last year.
Even the beta testers that I chat to haven't seen one yet.
Some months away yet I guess.  :(

BTW, no need for whispers.

This latest FW (which is quite different from previous ones) has the explanations inside:   :popcorn:

SVA1015X
SVA3032X
SVA7075X
   :scared:

And 5 other options besides what is currently used:

RTA, RCV, SVA, NA and MA
Well yeah that's real interesting as we know from recent info on the Chinese site SVA1032X is not too far away and it seems a 3k series is indeed coming too plus a 7k series.  :o
Drooling..........
« Last Edit: June 18, 2019, 09:10:42 pm by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #273 on: June 18, 2019, 09:11:45 pm »
from recent info on the Chinese site SVA1032X is not too far away

From what I see in the latest FW, I'll bet that "SVA1032X" is a typo. However it would be great if I'm wrong...
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29193
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SVA1015X 1.5GHz Spectrum & Vector Network Analyzer
« Reply #274 on: June 18, 2019, 09:21:38 pm »
from recent info on the Chinese site SVA1032X is not too far away

From what I see in the latest FW, I'll bet that "SVA1032X" is a typo. However it would be great if I'm wrong...
Check the Cn datasheet.  ;)
I listed some of the SVA1032X improvements a few posts back.