« Reply #2125 on: January 05, 2023, 06:15:49 pm »
What Python program is suggested? I see IDOL and several other available?
 

Offline tubularnut

  • Regular Contributor
  • *
  • Posts: 225
  • Country: gb
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2126 on: January 05, 2023, 06:19:35 pm »
Search for Python in the box at the top of this page. Search within  ‘this topic’

As you scroll down the results, you’ll find the script.
« Last Edit: January 05, 2023, 06:22:21 pm by tubularnut »
 
The following users thanked this post: Crambone, DL4JY

Offline DeDe

  • Contributor
  • Posts: 18
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2127 on: January 05, 2023, 06:55:54 pm »
Python-Script: https://www.online-python.com/PX3v0J8nbs
Select the System|System Info menu on your device and read the "Host ID" (not serial) and modify the script on that web-site.
Uncomment your model, e.g. Model = 'SSA3000X_Plus'
Execute script and then you should be able with menu option System|Load Option to enter the keys you want (e.g. ALL + 3032).

Regarding not working backup SSA3000X+_backup_11411.ADS - of course script will only work under following conditions;:
  • Before crossflash on original SS3000X *Plus* devices with V3.2.x.y.z firmware
  • Plugged in USB-Stick formatted with a single FAT32 partition
  • USB-Stick should allow write access: try with File|Create Folder if you can create a new folder on your stick

Here some useful links:
Python Script Discussion a few pages back in this thread: https://www.eevblog.com/forum/testgear/siglent-ssa3000x-spectrum-analyzers/msg4355635/#msg4355635
Another list of steps in this thread: https://www.eevblog.com/forum/testgear/siglent-ssa3000x-spectrum-analyzers/msg4355635/#msg4355635
Very detailed description form electr_peter: https://www.eevblog.com/forum/testgear/siglent-ssa3000x-spectrum-analyzers/msg3563577/#msg3563577
Newer checklist from user uski in the "Hack it.." thread: https://www.eevblog.com/forum/testgear/hack-of-sigllent-spectrum-analyzer-ssa3021x/msg4596574/#msg4596574
 
The following users thanked this post: Crambone

Offline Major1206

  • Contributor
  • Posts: 13
  • Country: pl
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2128 on: January 05, 2023, 07:06:14 pm »
Hi everyone,
I wasn't sure about crossflash, so first did PYTHON to have ALL + 3032.
2: Full Backup
3: Crossflash from SSA 5.1R20 to SVA 5.1R1 using TV84 hex mod
4: No need to python again - ALL permanent, 3.2GHz working, VNA working
Issue: no Telnet for me either SSA telnet.ads notr SVA telnet.ads -> can't copy na_1p, na_2p ,
Question un-crossflash -> copy cal -> crosflash again? Or sth else?

Edit:
found solution -> SCPI 'DEBTTT'
telnet IP port 23
username:root paswd:ding1234
Currently I am struggling with:
Quote
To fix this, it was necessary to copy those 2 files na_1p + na_2p to /usr/bin/siglent/firmdata0/cali
Now I have permission issue to firmdata0/cali " Read-only file system"
« Last Edit: January 05, 2023, 08:19:57 pm by Major1206 »
 

Offline DeDe

  • Contributor
  • Posts: 18
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2129 on: January 05, 2023, 08:15:18 pm »
You need to make filesystem writable, login via telnet and:
mount -o remount,rw /usr/bin/siglent/firmdata0

BTW: this was also required for me as a first step in order to perform a successful backup with SSA3000X+_backup_11411.ADS on the USB stick - make U-disk0 fs writable:
mount -o remount,rw /usr/bin/siglent/usr/mass_storage/U-disk0

After you are done just restart the device once, then all filesystems are read only again.
 
The following users thanked this post: Major1206

Offline Thomas

  • Regular Contributor
  • *
  • Posts: 118
  • Country: no
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2130 on: January 05, 2023, 08:17:02 pm »
No more time to hand in homework...

To close things off:

"SSA3000X", "SSA3000X_Plus", "SVA1000X", "SSA3000R"

Summarizing the collected information :)

Script to run:
https://www.online-python.com/PX3v0J8nbs

And added to the attachment (as if the links stopped working).

I used this (excellent!) script to enhance my aging SSA 3021X (non-Plus).

Analyzer data:
SW1: 1.3.9.8
SW2: 20201230-1
SW3: 000000D1
HW:   07.03.00
Bought with TG included permanently and 48 hours of trial for the other options (expired now).

I uncommented Model = 'SSA3000X' and filled in HOSTID as instructed, ran the script and got a list of license keys for all options. Plain and simple :)
But I am only able to activate options 3032 and EMI, even though that is great in itself.
I don't understand why AMK and Refl won't activate. I get an "Invalid License!" message on the screen when entering the license key for these.

The order of activating licenses is irrelevant, right? I think we're supposed to be able to buy these and activate at any time?
In hindsight I realize I should have activated 3032 and then ALL, but I find it hard to believe this is essential.
Any pointers or suggestions are welcome!
 

Offline Major1206

  • Contributor
  • Posts: 13
  • Country: pl
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2131 on: January 05, 2023, 08:23:41 pm »
Jeez,
i was trying
"mount -o remount,rw /usr/bin/siglent"
and mount -o remount,rw /usr/bin/siglent/firmdata0/cali"
 |O

Edit
restored classic telnet: "/sbin/telnetd -l /bin/sh -p 10101" & and "/sbin/telnetd -1 /bin/sh -p 10101 &" are not the same
« Last Edit: January 05, 2023, 08:30:54 pm by Major1206 »
 

Offline el_man

  • Contributor
  • Posts: 19
  • Country: bg
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2132 on: January 05, 2023, 10:22:39 pm »
Thank so much for your help DeDe!

I succeeded in everything and the upgrade is finished successfully now i have SVA1032X
But unfortunately It seem like something is missing in VNA part maybe This VNA cable :-//
The Smith chart was total Mess and after adding the two configuration files na_1p na_2p it becomes more or less normal at least now it is in the contour of the chart :D
But the port Doesn't react to any Load 50ohm Close - Nothing!

Where is this "VNA cable" and is there a fix to this?
Any other guy with same version of this firmware 3.2.2.5.1R20 Does Experience the same problem?
My unit is quite New - manufacture date is 13.12.2022 Maybe because of the unlucky date they stop putting the cables ;)
or at lest at This date :'(

Any help will be Highly appreciated!
 

Offline DeDe

  • Contributor
  • Posts: 18
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2133 on: January 06, 2023, 11:45:44 am »
To check the internal TG cable you would have to open the device, see here:
https://www.eevblog.com/forum/testgear/siglent-ssa3000x-spectrum-analyzers/msg3412882/#msg3412882

But first check what RoV has written:

S11 Port1 is the left TG port
You must first perform a full 1-port calibration and then activate it (turn correction on) as well
You need a fairly decent CAL kit for that

I don't own one of this Siglent-specific CAL.-Kits like F503ME (ca. 400€/$) but I purchased an SMA-Cal-Kit for 40€ from Amazon which even was fully characterized.
I applied it as "User 1" CAL-Kit and set all the parameter.
For a simple test you simply try with some low-cost Short/Open/Load Cal-Kit.

Instrument should display a blue "Cor"  in the upper left corner of the screen and not something like "---" or "Off".


 

Offline DL4JY

  • Regular Contributor
  • *
  • Posts: 68
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2134 on: January 06, 2023, 04:16:17 pm »
No more time to hand in homework...

To close things off:

"SSA3000X", "SSA3000X_Plus", "SVA1000X", "SSA3000R"

Summarizing the collected information :)

Script to run:
https://www.online-python.com/PX3v0J8nbs

And added to the attachment (as if the links stopped working).

I used this (excellent!) script to enhance my aging SSA 3021X (non-Plus).

Analyzer data:
SW1: 1.3.9.8
SW2: 20201230-1
SW3: 000000D1
HW:   07.03.00
Bought with TG included permanently and 48 hours of trial for the other options (expired now).

I uncommented Model = 'SSA3000X' and filled in HOSTID as instructed, ran the script and got a list of license keys for all options. Plain and simple :)
But I am only able to activate options 3032 and EMI, even though that is great in itself.
I don't understand why AMK and Refl won't activate. I get an "Invalid License!" message on the screen when entering the license key for these.

The order of activating licenses is irrelevant, right? I think we're supposed to be able to buy these and activate at any time?
In hindsight I realize I should have activated 3032 and then ALL, but I find it hard to believe this is essential.
Any pointers or suggestions are welcome!

Reply #2007
 

Offline Major1206

  • Contributor
  • Posts: 13
  • Country: pl
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2135 on: January 06, 2023, 07:25:52 pm »
It looks like they calibrated VNA oly up to 1.5 Ghz.
On a full span, the response look funny, but one calibrated you should see decent response for 50ohm load and through measurement.
Best
 

Offline Thomas

  • Regular Contributor
  • *
  • Posts: 118
  • Country: no
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2136 on: January 06, 2023, 08:22:35 pm »
No more time to hand in homework...

To close things off:

"SSA3000X", "SSA3000X_Plus", "SVA1000X", "SSA3000R"

Summarizing the collected information :)

Script to run:
https://www.online-python.com/PX3v0J8nbs

And added to the attachment (as if the links stopped working).

I used this (excellent!) script to enhance my aging SSA 3021X (non-Plus).

Analyzer data:
SW1: 1.3.9.8
SW2: 20201230-1
SW3: 000000D1
HW:   07.03.00
Bought with TG included permanently and 48 hours of trial for the other options (expired now).

I uncommented Model = 'SSA3000X' and filled in HOSTID as instructed, ran the script and got a list of license keys for all options. Plain and simple :)
But I am only able to activate options 3032 and EMI, even though that is great in itself.
I don't understand why AMK and Refl won't activate. I get an "Invalid License!" message on the screen when entering the license key for these.

The order of activating licenses is irrelevant, right? I think we're supposed to be able to buy these and activate at any time?
In hindsight I realize I should have activated 3032 and then ALL, but I find it hard to believe this is essential.
Any pointers or suggestions are welcome!

Why don't I just answer myself, since I found the solution :)
I played around with the (still excellent!) python script by tomud and also made a small script myself for reading and activating options (got tired of typing those license keys).
Then I noticed in the Siglent SSA3000X Programming Guide that valid options are Meas | EMI | CAT | TG. No mention of 3032, but that works.
So AMK is Meas and Refl is CAT. I added Meas and CAT to otheropt in tomud's script:
Code: [Select]
otheropt = ('AMK', 'EMI', 'TG', 'Refl', 'DMA', 'AMA', 'WDMA', 'DTF', 'VNA', 'RT40', '3021', '3032', '3050', '3075', 'ALL', 'Meas', 'CAT')And now I have all options unlocked and full 3.2GHz bandwidth 8)

Huge thanks to tomud, tv84 and all the others who contributed to this!
 
The following users thanked this post: yo3ggx

Offline DL4JY

  • Regular Contributor
  • *
  • Posts: 68
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2137 on: January 06, 2023, 09:25:00 pm »
as in reply #2007?
 

Offline Thomas

  • Regular Contributor
  • *
  • Posts: 118
  • Country: no
SSA3000X Solution
« Reply #2138 on: January 07, 2023, 05:08:24 am »
as in reply #2007?

Oh, someone did answer my question, sorry I missed that :-[
And replies #2006-2009 summarize the same as I found. I missed that too :-[

Summary:
3000X (non-Plus) can get all options and full bandwidth with python on reply #1968.
5 options work: Meas, EMI, CAT, TG and 3032.
ALL does not work, options must be activated separately.
 

Offline cubitus

  • Regular Contributor
  • *
  • Posts: 62
  • Country: fr
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2139 on: January 09, 2023, 03:47:27 pm »
Hello
This afternoon we turned on the analyzer the curve was displayed but everything remained frozen... I turned it off and restarted it and since then it remains on a black screen with the logo Siglent to display  :palm:
I still have access to the ssa3000 via ssh
What do you advise me to do?
SSA 3021 plus transformed into SVA1032 with all enabled options and the latest firmware.

Code: [Select]
/usr/bin/siglent # ps
PID   USER     TIME   COMMAND
    1 root       0:00 {linuxrc} init
    2 root       0:00 [kthreadd]
    3 root       0:00 [ksoftirqd/0]
    5 root       0:00 [kworker/0:0H]
    6 root       0:00 [kworker/u4:0]
    7 root       0:00 [rcu_preempt]
    8 root       0:00 [rcu_sched]
    9 root       0:00 [rcu_bh]
   10 root       0:00 [migration/0]
   11 root       0:00 [migration/1]
   12 root       0:00 [ksoftirqd/1]
   13 root       0:00 [kworker/1:0]
   14 root       0:00 [kworker/1:0H]
   15 root       0:00 [khelper]
   16 root       0:00 [kdevtmpfs]
   19 root       0:00 [perf]
   56 root       0:01 [kworker/1:1]
  202 root       0:00 [writeback]
  205 root       0:00 [bioset]
  207 root       0:00 [kblockd]
  238 root       0:00 [edac-poller]
  326 root       0:01 [kworker/0:1]
  327 root       0:00 [rpciod]
  341 root       0:00 [kswapd0]
  342 root       0:00 [fsnotify_mark]
  343 root       0:00 [nfsiod]
  441 root       0:00 [spi0]
  455 root       0:00 [spi1]
  564 root       0:00 [irq/23-f8007100]
  571 root       0:00 [deferwq]
  574 root       0:00 [kworker/0:1H]
  575 root       0:00 [kworker/1:1H]
  582 root       0:00 [ubi_bgt1d]
  586 root       0:00 [ubi_bgt2d]
  590 root       0:00 [ubi_bgt3d]
  594 root       0:00 [ubi_bgt0d]
  597 root       0:00 [ubifs_bgt1_0]
  600 root       0:00 [ubifs_bgt3_0]
  602 root       0:00 [ubifs_bgt0_0]
  648 root       0:00 [goodix_wq]
  692 root       0:00 telnetd
  716 root       0:00 [ci_otg]
  730 root       0:00 [kworker/0:2]
  739 root       0:00 /usr/bin/siglent/vsftpd
  742 1          0:00 portmap
  774 root       0:00 [kworker/u4:3]
  882 root       0:00 /usr/bin/siglent/vncserver
  887 root       0:00 /usr/sbin/lighttpd -f /usr/bin/siglent/lighttpd.conf
  888 root       0:00 /usr/bin/php-cgi -c /usr/bin/siglent/php.ini
  890 root       0:00 /sbin/getty -L ttyPS0 115200 vt100
  894 root       0:00 /usr/bin/php-cgi -c /usr/bin/siglent/php.ini
  895 root       0:00 -sh
  897 root       0:00 ps
RIGOL DM3068, DG4062, DP832A MSO4014
FLIR E4 SSA3021X Plus
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1745
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2140 on: January 09, 2023, 06:06:19 pm »
Post the result of a df -k
My guess is that the flash is full.

Offline cubitus

  • Regular Contributor
  • *
  • Posts: 62
  • Country: fr
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2141 on: January 09, 2023, 06:09:01 pm »
Code: [Select]
/usr/bin/siglent # df -k
Filesystem           1K-blocks      Used Available Use% Mounted on
/dev/root                63152     63152         0 100% /
devtmpfs                 69852         0     69852   0% /dev
none                     78140         4     78136   0% /tmp
ubi1_0                   30388     20864      9524  69% /usr/bin/siglent
ubi2_0                    5848       192      5656   3% /usr/bin/siglent/firmdata0
ubi3_0                    5848        24      5824   0% /usr/bin/siglent/log
ubi0_0                   84752      7784     76968   9% /usr/bin/siglent/usr
/usr/bin/siglent #
RIGOL DM3068, DG4062, DP832A MSO4014
FLIR E4 SSA3021X Plus
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1745
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2142 on: January 09, 2023, 09:46:10 pm »
Ask Tautech for a magic file.
It's not the flash being full.

Offline DeDe

  • Contributor
  • Posts: 18
  • Country: de
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2143 on: January 09, 2023, 10:14:38 pm »
Is the process list complete? I don't see the main application process /usr/bin/siglent/Aladdin?
Please check also the log files in /usr/bin/siglent/usr/logs (debug.log for instance)

You may try to start main application manually (see how it is done in usr/bin/siglent/startup_app.sh):
/usr/bin/siglent/Aladdin &

Additionally please post output of command: dmesg

Hello
This afternoon we turned on the analyzer the curve was displayed but everything remained frozen... I turned it off and restarted it and since then it remains on a black screen with the logo Siglent to display  :palm:
I still have access to the ssa3000 via ssh
What do you advise me to do?
SSA 3021 plus transformed into SVA1032 with all enabled options and the latest firmware.

Code: [Select]
/usr/bin/siglent # ps
PID   USER     TIME   COMMAND
    1 root       0:00 {linuxrc} init
    2 root       0:00 [kthreadd]
    3 root       0:00 [ksoftirqd/0]
    5 root       0:00 [kworker/0:0H]
    6 root       0:00 [kworker/u4:0]
    7 root       0:00 [rcu_preempt]
    8 root       0:00 [rcu_sched]
    9 root       0:00 [rcu_bh]
   10 root       0:00 [migration/0]
   11 root       0:00 [migration/1]
   12 root       0:00 [ksoftirqd/1]
   13 root       0:00 [kworker/1:0]
   14 root       0:00 [kworker/1:0H]
   15 root       0:00 [khelper]
   16 root       0:00 [kdevtmpfs]
   19 root       0:00 [perf]
   56 root       0:01 [kworker/1:1]
  202 root       0:00 [writeback]
  205 root       0:00 [bioset]
  207 root       0:00 [kblockd]
  238 root       0:00 [edac-poller]
  326 root       0:01 [kworker/0:1]
  327 root       0:00 [rpciod]
  341 root       0:00 [kswapd0]
  342 root       0:00 [fsnotify_mark]
  343 root       0:00 [nfsiod]
  441 root       0:00 [spi0]
  455 root       0:00 [spi1]
  564 root       0:00 [irq/23-f8007100]
  571 root       0:00 [deferwq]
  574 root       0:00 [kworker/0:1H]
  575 root       0:00 [kworker/1:1H]
  582 root       0:00 [ubi_bgt1d]
  586 root       0:00 [ubi_bgt2d]
  590 root       0:00 [ubi_bgt3d]
  594 root       0:00 [ubi_bgt0d]
  597 root       0:00 [ubifs_bgt1_0]
  600 root       0:00 [ubifs_bgt3_0]
  602 root       0:00 [ubifs_bgt0_0]
  648 root       0:00 [goodix_wq]
  692 root       0:00 telnetd
  716 root       0:00 [ci_otg]
  730 root       0:00 [kworker/0:2]
  739 root       0:00 /usr/bin/siglent/vsftpd
  742 1          0:00 portmap
  774 root       0:00 [kworker/u4:3]
  882 root       0:00 /usr/bin/siglent/vncserver
  887 root       0:00 /usr/sbin/lighttpd -f /usr/bin/siglent/lighttpd.conf
  888 root       0:00 /usr/bin/php-cgi -c /usr/bin/siglent/php.ini
  890 root       0:00 /sbin/getty -L ttyPS0 115200 vt100
  894 root       0:00 /usr/bin/php-cgi -c /usr/bin/siglent/php.ini
  895 root       0:00 -sh
  897 root       0:00 ps
« Last Edit: January 09, 2023, 10:24:36 pm by DeDe »
 

Offline cubitus

  • Regular Contributor
  • *
  • Posts: 62
  • Country: fr
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2144 on: January 10, 2023, 08:33:03 am »
Hello and thank you for your help  :-+

In the logs, there are almost only warnings, s I did not find anything relevant...
By launching aladdin by hand, the screen appears but this immediately blocks with the error message below.



I also put you the command dmesg


Code: [Select]
/usr/bin/siglent # dmesg
[    0.336218] TCP: reno registered
[    0.336234] UDP hash table entries: 256 (order: 1, 8192 bytes)
[    0.336264] UDP-Lite hash table entries: 256 (order: 1, 8192 bytes)
[    0.336460] NET: Registered protocol family 1
[    0.336805] RPC: Registered named UNIX socket transport module.
[    0.336818] RPC: Registered udp transport module.
[    0.336827] RPC: Registered tcp transport module.
[    0.336836] RPC: Registered tcp NFSv4.1 backchannel transport module.
[    0.336867] PCI: CLS 0 bytes, default 64
[    0.337196] hw perfevents: enabled with armv7_cortex_a9 PMU driver, 7 counters available
[    0.338540] futex hash table entries: 512 (order: 3, 32768 bytes)
[    0.340126] jffs2: version 2.2. (NAND) © 2001-2006 Red Hat, Inc.
[    0.341154] io scheduler noop registered
[    0.341174] io scheduler deadline registered
[    0.341231] io scheduler cfq registered (default)
[    0.343272] dma-pl330 f8003000.dmac: Loaded driver for PL330 DMAC-241330
[    0.343360] dma-pl330 f8003000.dmac:         DBUFF-128x8bytes Num_Chans-8 Num_Peri-4 Num_Events-16
[    0.343828] e0001000.serial: ttyPS0 at MMIO 0xe0001000 (irq = 145, base_baud = 2500000) is a xuartps
[    0.917296] console [ttyPS0] enabled
[    0.921446] xdevcfg f8007000.devcfg: ioremap 0xf8007000 to 4a878000
[    0.928185] [drm] Initialized drm 1.1.0 20060810
[    0.940673] brd: module loaded
[    0.947643] loop: module loaded
[    0.956927] libphy: MACB_mii_bus: probed
[    1.043419] macb e000b000.ethernet eth0: Cadence GEM rev 0x00020118 at 0xe000b000 irq 149 (00:0a:35:00:01:23)
[    1.053307] macb e000b000.ethernet eth0: attached PHY driver [Generic PHY] (mii_bus:phy_addr=e000b000.etherne:1e, irq=-1)
[    1.064867] i2c /dev entries driver
[    1.070148] rtc-ds1307 0-0068: rtc core: registered ds1340 as rtc0
[    1.076593] cdns-i2c e0004000.i2c: 400 kHz mmio e0004000 irq 143
[    1.083835] zynq-edac f8006000.memory-controller: ecc not enabled
[    1.090051] Xilinx Zynq CpuIdle Driver started
[    1.095137] ledtrig-cpu: registered to indicate activity on CPUs
[    1.102140] nand: device found, Manufacturer ID: 0x2c, Chip ID: 0xda
[    1.108452] nand: Micron MT29F2G08ABAEAWP
[    1.112430] nand: 256 MiB, SLC, erase size: 128 KiB, page size: 2048, OOB size: 64
[    1.120023] nand: WARNING: pl353-nand: the ECC used on your system is too weak compared to the one required by the NAND chip
[    1.131531] Bad block table found at page 131008, version 0x01
[    1.137807] Bad block table found at page 130944, version 0x01
[    1.143935] 12 ofpart partitions found on MTD device pl353-nand
[    1.149785] Creating 12 MTD partitions on "pl353-nand":
[    1.155041] 0x000000000000-0x000000780000 : "fsbl"
[    1.160857] 0x000000780000-0x000000b80000 : "kerneldata"
[    1.167142] 0x000000b80000-0x000000c00000 : "device-tree"
[    1.173491] 0x000000c00000-0x000001100000 : "Manufacturedata"
[    1.180156] 0x000001100000-0x000001600000 : "reserved1"
[    1.186367] 0x000001600000-0x000003e00000 : "rootfs"
[    1.192260] 0x000003e00000-0x000004800000 : "firmdata0"
[    1.198384] 0x000004800000-0x000006c00000 : "siglent"
[    1.204412] 0x000006c00000-0x00000d000000 : "datafs"
[    1.210426] 0x00000d000000-0x00000da00000 : "log"
[    1.216087] 0x00000da00000-0x00000f800000 : "upgrade_cramdisk"
[    1.222900] 0x00000f800000-0x000010000000 : "reserved2"
[    1.230968] TCP: cubic registered
[    1.234458] NET: Registered protocol family 17
[    1.239084] Registering SWP/SWPB emulation handler
[    1.245704] rtc-ds1307 0-0068: setting system clock to 2023-01-10 08:53:07 UTC (1673340787)
[    1.259101] cramfs_fill_nand blocks is 320-----------------------
[    1.259101]
[    1.259101]
[    1.259101]
[    1.272121] VFS: Mounted root (cramfs filesystem) readonly on device 31:5.
[    1.279013] devtmpfs: mounted
[    1.282156] Freeing unused kernel memory: 196K (40553000 - 40584000)
[    1.554972] UBI-1: ubi_attach_mtd_dev:attaching mtd7 to ubi1
[    1.703267] UBI-1: scan_all:scanning is finished
[    1.714262] UBI-1 warning: print_rsvd_warning: cannot reserve enough PEBs for bad PEB handling, reserved 3, need 40
[    1.725400] UBI-1: ubi_attach_mtd_dev:attached mtd7 (name "siglent", size 36 MiB)
[    1.732818] UBI-1: ubi_attach_mtd_dev:PEB size: 131072 bytes (128 KiB), LEB size: 126976 bytes
[    1.741476] UBI-1: ubi_attach_mtd_dev:min./max. I/O unit sizes: 2048/2048, sub-page size 2048
[    1.749975] UBI-1: ubi_attach_mtd_dev:VID header offset: 2048 (aligned 2048), data offset: 4096
[    1.758672] UBI-1: ubi_attach_mtd_dev:good PEBs: 288, bad PEBs: 0, corrupted PEBs: 0
[    1.766427] UBI-1: ubi_attach_mtd_dev:user volume: 1, internal volumes: 1, max. volumes count: 128
[    1.775376] UBI-1: ubi_attach_mtd_dev:max/mean erase counter: 27/10, WL threshold: 4096, image sequence number: 1996219743
[    1.786425] UBI-1: ubi_attach_mtd_dev:available PEBs: 0, total reserved PEBs: 288, PEBs reserved for bad PEB handling: 3
[    1.797314] UBI-1: ubi_thread:background thread "ubi_bgt1d" started, PID 582
[    1.799673] UBI-2: ubi_attach_mtd_dev:attaching mtd6 to ubi2
[    1.843974] UBI-2: scan_all:scanning is finished
[    1.854764] UBI-2 warning: print_rsvd_warning: cannot reserve enough PEBs for bad PEB handling, reserved 9, need 40
[    1.865756] UBI-2: ubi_attach_mtd_dev:attached mtd6 (name "firmdata0", size 10 MiB)
[    1.873376] UBI-2: ubi_attach_mtd_dev:PEB size: 131072 bytes (128 KiB), LEB size: 126976 bytes
[    1.881961] UBI-2: ubi_attach_mtd_dev:min./max. I/O unit sizes: 2048/2048, sub-page size 2048
[    1.890504] UBI-2: ubi_attach_mtd_dev:VID header offset: 2048 (aligned 2048), data offset: 4096
[    1.899203] UBI-2: ubi_attach_mtd_dev:good PEBs: 80, bad PEBs: 0, corrupted PEBs: 0
[    1.906857] UBI-2: ubi_attach_mtd_dev:user volume: 1, internal volumes: 1, max. volumes count: 128
[    1.915819] UBI-2: ubi_attach_mtd_dev:max/mean erase counter: 158/101, WL threshold: 4096, image sequence number: 1168290863
[    1.927041] UBI-2: ubi_attach_mtd_dev:available PEBs: 0, total reserved PEBs: 80, PEBs reserved for bad PEB handling: 9
[    1.937849] UBI-2: ubi_thread:background thread "ubi_bgt2d" started, PID 586
[    1.944942] UBI-3: ubi_attach_mtd_dev:attaching mtd9 to ubi3
[    1.990185] UBI-3: scan_all:scanning is finished
[    2.000959] UBI-3 warning: print_rsvd_warning: cannot reserve enough PEBs for bad PEB handling, reserved 9, need 40
[    2.011967] UBI-3: ubi_attach_mtd_dev:attached mtd9 (name "log", size 10 MiB)
[    2.019099] UBI-3: ubi_attach_mtd_dev:PEB size: 131072 bytes (128 KiB), LEB size: 126976 bytes
[    2.027668] UBI-3: ubi_attach_mtd_dev:min./max. I/O unit sizes: 2048/2048, sub-page size 2048
[    2.036189] UBI-3: ubi_attach_mtd_dev:VID header offset: 2048 (aligned 2048), data offset: 4096
[    2.044899] UBI-3: ubi_attach_mtd_dev:good PEBs: 80, bad PEBs: 0, corrupted PEBs: 0
[    2.052534] UBI-3: ubi_attach_mtd_dev:user volume: 1, internal volumes: 1, max. volumes count: 128
[    2.061508] UBI-3: ubi_attach_mtd_dev:max/mean erase counter: 16/9, WL threshold: 4096, image sequence number: 479844541
[    2.072388] UBI-3: ubi_attach_mtd_dev:available PEBs: 0, total reserved PEBs: 80, PEBs reserved for bad PEB handling: 9
[    2.083185] UBI-3: ubi_thread:background thread "ubi_bgt3d" started, PID 590
[    2.085348] UBI-0: ubi_attach_mtd_dev:attaching mtd8 to ubi0
[    2.485725] UBI-0: scan_all:scanning is finished
[    2.497905] UBI-0: ubi_attach_mtd_dev:attached mtd8 (name "datafs", size 100 MiB)
[    2.505374] UBI-0: ubi_attach_mtd_dev:PEB size: 131072 bytes (128 KiB), LEB size: 126976 bytes
[    2.513957] UBI-0: ubi_attach_mtd_dev:min./max. I/O unit sizes: 2048/2048, sub-page size 2048
[    2.522464] UBI-0: ubi_attach_mtd_dev:VID header offset: 2048 (aligned 2048), data offset: 4096
[    2.531179] UBI-0: ubi_attach_mtd_dev:good PEBs: 800, bad PEBs: 0, corrupted PEBs: 0
[    2.538933] UBI-0: ubi_attach_mtd_dev:user volume: 1, internal volumes: 1, max. volumes count: 128
[    2.547886] UBI-0: ubi_attach_mtd_dev:max/mean erase counter: 19/12, WL threshold: 4096, image sequence number: 1984375540
[    2.558932] UBI-0: ubi_attach_mtd_dev:available PEBs: 0, total reserved PEBs: 800, PEBs reserved for bad PEB handling: 40
[    2.569907] UBI-0: ubi_thread:background thread "ubi_bgt0d" started, PID 594
[    2.583511] UBIFS: background thread "ubifs_bgt1_0" started, PID 597
[    2.615063] UBIFS: recovery needed
[    2.686073] UBIFS: recovery completed
[    2.689742] UBIFS: mounted UBI device 1, volume 0, name "siglent"
[    2.695795] UBIFS: LEB size: 126976 bytes (124 KiB), min./max. I/O unit sizes: 2048 bytes/2048 bytes
[    2.704917] UBIFS: FS size: 34410496 bytes (32 MiB, 271 LEBs), journal size 4952064 bytes (4 MiB, 39 LEBs)
[    2.714577] UBIFS: reserved for root: 0 bytes (0 KiB)
[    2.719610] UBIFS: media format: w4/r0 (latest is w4/r0), UUID DFEF06B5-93BD-405D-955B-0D192442F33A, small LPT model
[    2.785946] UBIFS: mounted UBI device 2, volume 0, name "firm0", R/O mode
[    2.792670] UBIFS: LEB size: 126976 bytes (124 KiB), min./max. I/O unit sizes: 2048 bytes/2048 bytes
[    2.801829] UBIFS: FS size: 7237632 bytes (6 MiB, 57 LEBs), journal size 1650688 bytes (1 MiB, 13 LEBs)
[    2.811213] UBIFS: reserved for root: 0 bytes (0 KiB)
[    2.816263] UBIFS: media format: w4/r0 (latest is w4/r0), UUID 89BA13B0-5EF6-4E0F-BE2E-B95726632087, small LPT model
[    2.830634] UBIFS: background thread "ubifs_bgt3_0" started, PID 600
[    2.862228] UBIFS: recovery needed
[    2.884958] UBIFS: recovery completed
[    2.888634] UBIFS: mounted UBI device 3, volume 0, name "log"
[    2.894332] UBIFS: LEB size: 126976 bytes (124 KiB), min./max. I/O unit sizes: 2048 bytes/2048 bytes
[    2.903463] UBIFS: FS size: 7237632 bytes (6 MiB, 57 LEBs), journal size 1650688 bytes (1 MiB, 13 LEBs)
[    2.912839] UBIFS: reserved for root: 0 bytes (0 KiB)
[    2.917903] UBIFS: media format: w4/r0 (latest is w4/r0), UUID DBC2ADDB-2E45-46BE-8553-F5AEAB2F89BD, small LPT model
[    2.931544] UBIFS: background thread "ubifs_bgt0_0" started, PID 602
[    2.963172] UBIFS: recovery needed
[    3.044372] UBIFS: recovery completed
[    3.048042] UBIFS: mounted UBI device 0, volume 0, name "rootfs"
[    3.054006] UBIFS: LEB size: 126976 bytes (124 KiB), min./max. I/O unit sizes: 2048 bytes/2048 bytes
[    3.063119] UBIFS: FS size: 94597120 bytes (90 MiB, 745 LEBs), journal size 9023488 bytes (8 MiB, 72 LEBs)
[    3.072786] UBIFS: reserved for root: 0 bytes (0 KiB)
[    3.077836] UBIFS: media format: w4/r0 (latest is w4/r0), UUID 27A2060B-95C4-4CE0-ACBD-280A13E9C15D, small LPT model
[    5.516861] Unhandled fault: external abort on non-linefetch (0x818) at 0x36f2e380
[    5.542077] irq = 170
[    5.553152] <<-GTP-INFO->> GTP driver installing...
[    5.558186] <<-GTP-INFO->> GTP Driver Version: V2.4<2014/11/28>
[    5.564081] <<-GTP-INFO->> GTP Driver Built@22:06:06, Apr 27 2018
[    5.570124] <<-GTP-INFO->> GTP I2C Address: 0x14
[    5.574775] 0-0014 supply vdd_ana not found, using dummy regulator
[    5.580961] 0-0014 supply vcc_i2c not found, using dummy regulator
[    5.587265] <<-GTP-INFO->> Guitar reset
[    5.713798] <<-GTP-INFO->> IC Version: 928_1060
[    5.724470] <<-GTP-INFO->> X_MAX: 1024, Y_MAX: 600, TRIGGER: 0x00
[    5.743323] <<-GTP-INFO->> create proc entry gt9xx_config success
[    5.749597] input: goodix-ts as /devices/virtual/input/input0
[    5.755587] <<-GTP-INFO->> GTP works in interrupt mode.
[    5.771618] gpib_usb_init
[    5.774381] usbcore: registered new interface driver usb-gpib
[    7.043496] macb e000b000.ethernet eth0: link up (100/Full)
[   10.602623] ehci_hcd: USB 2.0 'Enhanced' Host Controller (EHCI) Driver
[   10.621906] ehci-pci: EHCI PCI platform driver
[   10.680048] e0002000.usb supply vbus not found, using dummy regulator
[   10.687358] ci_hdrc ci_hdrc.0: ChipIdea HDRC found, lpm: 0; cap: 4abfc100 op: 4abfc140
[   10.687380] ci_hdrc ci_hdrc.0: It is OTG capable controller
[   10.687400] ci_hdrc ci_hdrc.0: EHCI Host Controller
[   10.692233] ci_hdrc ci_hdrc.0: new USB bus registered, assigned bus number 1
[   10.723343] ci_hdrc ci_hdrc.0: USB 2.0 started, EHCI 1.00
[   10.728863] usb usb1: New USB device found, idVendor=1d6b, idProduct=0002
[   10.735636] usb usb1: New USB device strings: Mfr=3, Product=2, SerialNumber=1
[   10.742817] usb usb1: Product: EHCI Host Controller
[   10.747704] usb usb1: Manufacturer: Linux 3.19.2-omg-xilinx-svn35686 ehci_hcd
[   10.754839] usb usb1: SerialNumber: ci_hdrc.0
[   10.760001] hub 1-0:1.0: USB hub found
[   10.763823] hub 1-0:1.0: 1 port detected
[   10.768426] e0003000.usb supply vbus not found, using dummy regulator
[   10.775198] ci_hdrc ci_hdrc.1: ChipIdea HDRC found, lpm: 0; cap: 4abfe100 op: 4abfe140
[   10.775220] ci_hdrc ci_hdrc.1: It is OTG capable controller
[   10.831060] usbcore: registered new interface driver usb-storage
[   10.859701] usbcore: registered new interface driver usbhid
[   10.868287] usbhid: USB HID core driver
[   10.882227] mousedev: PS/2 mouse device common for all mice
[   12.540768] random: Aladdin urandom read with 21 bits of entropy available
[   12.892722] spidev spi1.1: setup: unsupported mode bits 60
[   13.036710] UBIFS: background thread "ubifs_bgt2_0" started, PID 756
[   13.080274] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.121244] UBIFS: background thread "ubifs_bgt2_0" started, PID 761
[   13.159043] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.177131] UBIFS: background thread "ubifs_bgt2_0" started, PID 778
[   13.221521] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.240604] UBIFS: background thread "ubifs_bgt2_0" started, PID 788
[   13.250671] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.265515] UBIFS: background thread "ubifs_bgt2_0" started, PID 794
[   13.293002] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.309112] UBIFS: background thread "ubifs_bgt2_0" started, PID 808
[   13.327784] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.351298] UBIFS: background thread "ubifs_bgt2_0" started, PID 813
[   13.366858] UBIFS: background thread "ubifs_bgt2_0" stops
[   13.822426] <<-GTP-INFO->> System resume.
[   13.822437] gpio-963 (GTP INT IRQ): _gpiod_direction_output_raw: tried to set a GPIO tied to an IRQ as output
[   13.833489] <<-GTP-INFO->> GTP wakeup sleep.
[   13.833498] gpio-963 (GTP INT IRQ): _gpiod_direction_output_raw: tried to set a GPIO tied to an IRQ as output
[   13.962236] UBIFS: background thread "ubifs_bgt2_0" started, PID 825
[   13.999042] UBIFS: background thread "ubifs_bgt2_0" stops
[   14.033277] UBIFS: background thread "ubifs_bgt2_0" started, PID 830
[   14.048697] UBIFS: background thread "ubifs_bgt2_0" stops
[   14.337720] UBIFS: background thread "ubifs_bgt2_0" started, PID 837
[   14.369317] UBIFS: background thread "ubifs_bgt2_0" stops
[   16.548160] UBIFS: background thread "ubifs_bgt2_0" started, PID 844
[   16.576941] UBIFS: background thread "ubifs_bgt2_0" stops
[   22.610170] input: PolyVision Touch Screen as /devices/virtual/input/input1
[  137.330684] random: nonblocking pool is initialized
RIGOL DM3068, DG4062, DP832A MSO4014
FLIR E4 SSA3021X Plus
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1745
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2145 on: January 10, 2023, 08:48:56 am »
Send a message to Tauchtech asking for help. I already told you to do that...

Offline cubitus

  • Regular Contributor
  • *
  • Posts: 62
  • Country: fr
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2146 on: January 10, 2023, 09:03:04 am »
It's done  ;)
RIGOL DM3068, DG4062, DP832A MSO4014
FLIR E4 SSA3021X Plus
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28915
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2147 on: January 10, 2023, 09:17:18 am »
It's done  ;)
Sorry don’t have the correct tools.
Contact mailto:info-eu@siglent.com
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline cubitus

  • Regular Contributor
  • *
  • Posts: 62
  • Country: fr
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2148 on: January 10, 2023, 10:17:51 am »
How am I going to shoot this?
Hello Siglent, I bought an SSA3021+, I tinkered with it in SVA1032, it is down what can you do for me? :-BROKE
Won't there be something to try before? Do a full reset?
RIGOL DM3068, DG4062, DP832A MSO4014
FLIR E4 SSA3021X Plus
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3263
  • Country: pt
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #2149 on: January 10, 2023, 10:51:23 am »
Try to reinstall the latest SVA FW again. Can you do it from the boot sequence?
 