I have followed Reply #1624 to upgrade my SSA3021 PLUS.
After step I) Initial backup of SSA3021X Plus
There's no back files in my U disk.
What's wrong happened?
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1726 on: October 12, 2021, 03:20:28 pm »
I have followed Reply #1624 to upgrade my SSA3021 PLUS.
After step I) Initial backup of SSA3021X Plus
There's no back files in my U disk.
What's wrong happened?

Make sure your flash drive is FAT32 formatted and empty (except for the ADS backup file).
Try reformatting it on a windows PC.

Then try again. It took me two tries as well, not sure why ...

rudi
 

Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1727 on: October 13, 2021, 05:18:21 am »
I have tried several U disks.
All the U disks were formatted as FAT32 on Windows 10 PC.
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1728 on: October 13, 2021, 06:09:30 am »
I have tried several U disks.
All the U disks were formatted as FAT32 on Windows 10 PC.

Did you unzip SSA3000X+_backup_NAND+mem+firmdata0_11411.zip  ?
Only SSA3000X+_backup_11411.ADS should be on the flash drive.

rudi
 

Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1729 on: October 13, 2021, 09:06:23 am »
yes, I have unzipped the file.
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1730 on: October 13, 2021, 09:31:30 am »
yes, I have unzipped the file.

Make sure your SSA3021X has the latest firmware, BEFORE you do any of the steps
below. It should have SSA FW 3.2.2.4.0 !


ok, so you now have a USB Flash drive, FAT32 formatted, larger than 2GB, the only
file on the flash drive is "SSA3000X+_backup_11411.ADS"

Plug this in to your SSA3021X, and power it on.
Once it is booted, go to "System->System Info->Firmware Upgrade"
Navigate to the U-disk0, select the ads file, and then click "Open".
That should run the backup script ...

What happens now ?

rudi
 

Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1731 on: October 13, 2021, 11:04:06 am »
I have followed your advice. After excuting the .ads file,the SSA3000 showing “upgrading …”,and reboot automatically.
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1732 on: October 13, 2021, 11:30:54 am »
I have followed your advice. After excuting the .ads file,the SSA3000 showing “upgrading …”,and reboot automatically.

Thats good news. Now shutdown the SSA, remove the flash drive and copy all files to a secure location for backup.
Check that the total size of all files is around ~483MB

And  then proceed with the upgrade instructions !

rudi
 

Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1733 on: October 13, 2021, 12:16:39 pm »
There's no other files in the USB disk except ssa3000x+_backup_11411.ads after rebooting.

« Last Edit: October 13, 2021, 12:23:18 pm by zkf0100007 »
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1734 on: October 14, 2021, 10:29:57 am »
There's no other files in the USB disk except ssa3000x+_backup_11411.ads after rebooting.

Strange, I really don't know why you are not getting a backup.

Keep on trying different USB Flash drives.

If you want you can open your unit and attach cables to UART port, that should tell you what is going on and why no backup ...

rudi
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1762
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1735 on: October 14, 2021, 10:43:14 am »
If the Product ID of the device does not match the Product ID in the ADS file, then the upgrade will not be executed.

The current Product ID depends on the currently installed FW and, of course, if it has been tampered during the hack.

That might be one reason why the backup is not working.

Regards,
Vitor


Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1736 on: October 15, 2021, 05:34:39 am »
Actually,I want to backup my SSA.
I followed the instructions,but the backup seemed to be failed
 

Offline zkf0100007

  • Contributor
  • Posts: 18
  • Country: cn
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1737 on: October 15, 2021, 07:32:24 am »
How to check the product id in the ads file?
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1738 on: October 15, 2021, 07:55:43 am »
How to check the product id in the ads file?

In some of them I included the ProdID in the file's name.

ssa3000x+_backup_11411.ads  -> means it ONLY works in a machine with ProdID=11411.

If you read the thread you'll see that Siglent has been changing the ProdIDs of some of these equipments and that's why some FWs are not retro-compatible.

See here.
« Last Edit: October 15, 2021, 08:00:45 am by tv84 »
 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1739 on: October 15, 2021, 08:30:29 am »
How to check the product id in the ads file?

Telnet in to your SSA.

Edit   /usr/bin/siglent/firmdata0/NSP_trends_config_info.xml

make sure the pid line looks like this:

      <pid>0x11411</pid>


rudi

 

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1740 on: October 15, 2021, 06:18:05 pm »
Guys,

I recently upgraded my SSA to SVA.

Should I copy the calibration files from my SSA back to my new SVA ?

Thanks,
rudi
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1762
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1741 on: October 15, 2021, 07:00:27 pm »
These files are not touched by the upgrade/crossgrade.

Offline SharpEars

  • Supporter
  • ****
  • Posts: 634
  • Country: us
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1742 on: October 15, 2021, 07:31:41 pm »
Damn! I made a bigger mistake!!!    :palm:

Now the SSA3032X-R became a SSA3075X-R.   


There goes the neighborhood...

I presume that this is a joke, since there don't appear to be any instructions on all of eevblog on how to get apply this sort of hack to an SSA3000X-R series model. (only models without the -R)
 

Offline Bicurico

  • Super Contributor
  • ***
  • Posts: 1762
  • Country: pt
    • VMA's Satellite Blog
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1743 on: October 15, 2021, 07:46:37 pm »
Damn! I made a bigger mistake!!!    :palm:

Now the SSA3032X-R became a SSA3075X-R.   


There goes the neighborhood...

I presume that this is a joke, since there don't appear to be any instructions on all of eevblog on how to get apply this sort of hack to an SSA3000X-R series model. (only models without the -R)

It is not a joke, but the hack is not public.

Offline SharpEars

  • Supporter
  • ****
  • Posts: 634
  • Country: us
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1744 on: October 16, 2021, 12:43:47 am »
Damn! I made a bigger mistake!!!    :palm:

Now the SSA3032X-R became a SSA3075X-R.   


There goes the neighborhood...

I presume that this is a joke, since there don't appear to be any instructions on all of eevblog on how to get apply this sort of hack to an SSA3000X-R series model. (only models without the -R)

It is not a joke, but the hack is not public.

So what is stopping this hack from going public? All the other hacks have full guides on how to perform them here. It's hard to believe that someone has actually figured out a way and is not sharing it with the rest of us, unless this is just a big fat lie (fake screenshots or not).
 

Offline SharpEars

  • Supporter
  • ****
  • Posts: 634
  • Country: us
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1745 on: October 16, 2021, 12:46:07 am »
Damn! I made a bigger mistake!!!    :palm:

Now the SSA3032X-R became a SSA3075X-R.   


There goes the neighborhood...

How do we know this hack is legit if no other person has duplicated your steps so far and a almost year has gone by. These could be just images from a higher end model superimposed (i.e., Photoshoped) over the screen of the SSA3032X-R for all we know!
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1746 on: October 16, 2021, 07:00:20 am »
So what is stopping this hack from going public?

It's called "freedom of choice".

How do we know this hack is legit if no other person has duplicated your steps so far and a almost year has gone by. These could be just images from a higher end model superimposed (i.e., Photoshoped) over the screen of the SSA3032X-R for all we know!

I use Paintbrush.

PS: Those type of comments/demands are usually the best reason for not disclosing some of these things publicly. Think about it.
 
The following users thanked this post: Performa01, citizenrich, electrophil

Offline luudee

  • Frequent Contributor
  • **
  • Posts: 283
  • Country: th
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1747 on: October 16, 2021, 07:26:43 am »
So what is stopping this hack from going public?

It's called "freedom of choice".

How do we know this hack is legit if no other person has duplicated your steps so far and a almost year has gone by. These could be just images from a higher end model superimposed (i.e., Photoshoped) over the screen of the SSA3032X-R for all we know!

I use Paintbrush.

PS: Those type of comments/demands are usually the best reason for not disclosing some of these things publicly. Think about it.

Hi TV84,

I understand and respect you wanting to keep your work to yourself.

But, I am still an Engineer! ;)

Can you at least tell us if it is a SW/FW hack only, or if HW needed to
be "adjusted" as well?

Thank you!
rudi
 

Offline tv84

  • Super Contributor
  • ***
  • Posts: 3288
  • Country: pt
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1748 on: October 16, 2021, 07:38:41 am »
Can you at least tell us if it is a SW/FW hack only, or if HW needed to be "adjusted" as well?

Never touched the HW and that's why it is shown like that.
 
The following users thanked this post: luudee

Offline dibro

  • Contributor
  • Posts: 26
  • Country: nl
Re: Siglent SSA3000X and SSA3000X-Plus Spectrum Analyzers
« Reply #1749 on: October 16, 2021, 08:32:30 am »
With the information I found in this forum and experience with my SSA3021X I've got my SSA3032X-R now in factory mode.
Didn't find a way to keep the serial number thou, but this works for me.