5000 MHz / 128 ?
Oh yes, of course!
« Last Edit: December 25, 2023, 03:51:13 pm by Performa01 »
 

Online gf

  • Super Contributor
  • ***
  • Posts: 1314
  • Country: de
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #351 on: December 25, 2023, 07:57:47 pm »
5000 MHz / 128 ?
Oh yes, of course!

What still irritates me is the low sample rate of only 39M that the ADC cores would have and the large number of interleaved cores. For comparison, the HMCAD1511 which is used in several low-cost 1GSa/s scope models has 8 interleaved ADC cores, with 125MSa/s each. And for faster ADCs, I'd rather expect an even higher sample rate per core. But who knows...
 

Online hpw

  • Frequent Contributor
  • **
  • Posts: 409
  • Country: 00
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #352 on: December 26, 2023, 08:46:14 am »
5000 MHz / 128 ?
Oh yes, of course!

What still irritates me is the low sample rate of only 39M that the ADC cores would have and the large number of interleaved cores. For comparison, the HMCAD1511 which is used in several low-cost 1GSa/s scope models has 8 interleaved ADC cores, with 125MSa/s each. And for faster ADCs, I'd rather expect an even higher sample rate per core. But who knows...

This looks as a clear 5000 MHz / 128  = 39.0625Mhz SQUARE Wave or other combination(s)... guessing as exact 128 diff by the reference frequency (synthesizer) or oscillation on the VRef power.
The other one as about -100dB are for me the spurs...

Hopefully the upcoming SDS3000 12 bit, SDS7000 12 bit are any better or just send them back to the development group.

Hp

 
 
The following users thanked this post: egonotto

Online mawyatt

  • Super Contributor
  • ***
  • Posts: 3750
  • Country: us
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #353 on: December 26, 2023, 03:43:36 pm »
Some interesting high speed ADCs utilizing multiple individual channels, recall from ~2008 when we saw the early Fuji 56GPSP ADC later expanded to 64GSPS, this was developed in 65nm CMOS with 2 ADCs per chip.

Later some other interesting ADCs utilizing multiple individual channels emerged, like the 2014 IBM 90GSPS in 32nm SOI, with 5.2 ENOB at 20GHz which utilized 64 ADC channels @ 1.4GSPS per ADC, and the 2017 Broadcom 64GSPS in 20nm with 128 channels per ADC (has 4 ADCs per chip).

https://indico.cern.ch/event/121657/attachments/68435/98170/ADC_DAC_CERN.pdf

https://www.electronicspecifier.com/products/mixed-signal-analog/fujitsu-launches-ultrafast-56gsa-s-8-bit-adc-technology-for-100g-coherent-receivers-and-high-performance-test-equipment

https://www.fujitsu.com/uk/Images/c45.pdf

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 
The following users thanked this post: Performa01, egonotto, 2N3055

Online Performa01

  • Super Contributor
  • ***
  • Posts: 1706
  • Country: at
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #354 on: January 01, 2024, 10:29:00 am »
Trigger Jitter

The datasheet specifies the trigger jitter as follows:

•   <9ps RMS (typical) for ≥300MHz sine and ≥6 divisions peak to peak amplitude for vertical gain settings from 2.5mV/div to 10V/div.
•   <5ps RMS (typical) for ≥500MHz sine and ≥6 divisions peak to peak amplitude for vertical gain settings from 2.5mV/div to 10V/div.

The EXT trigger jitter is not exactly stellar at <200 ps RMS – but then again this is a classical analog (comparator) trigger and not the fully digital trigger system that we get when using one of the input channels as trigger source.

Now let’s verify this with a 1 GHz sine signal from an OCXO-driven AWG (SDG7102A), fed into channels 3 and 4 of an SDS6204 H12 Pro (which has an individual ADC for each channel) via a 12.4 GHz resistive power splitter. This way we can observe the jitter in the trigger channel as well as a not triggered channel, where both are using different ADCs, hence are completely unrelated.

The high quality 1 GHz sine signal has been chosen for its fast edges and low inherent jitter – after all we want to characterize the DSO and not the signal source. See the attached screenshot which has been taken after more than 15 minutes running with infinite persistence:


SDS6204 Pro H12_Trigger_Jitter_1GHz_15m

At a timebase of 100 ps/div, we can see the peak to peak jitter in the triggered as well as the non-triggered channel after more than 15 minutes at infinite persistence.

The jitter measurements are as follows:
Triggered channel: 6 ps pk-pk, 698 fs rms;
Un-triggered channel: 11.3 ps pk-pk, 1.564 ps rms;
Skew Ch.3-Ch.4: 12.06 ps pk-pk, 1.594 ps rms;
 
The following users thanked this post: Martin72

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #355 on: January 31, 2024, 09:59:28 pm »
Hi,

For a new project we want to look at the signals of a 1000 Base T connection.
I have roughly thought about it and would consider a 1Ghz scope for this - is that enough or would we have to go even higher?
Either way, only the 6000A would come into question, it is one of the cheapest on the market despite its performance.
The only question is, 1Ghz or 2Ghz.... ;)
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7117
  • Country: hr
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #356 on: January 31, 2024, 11:24:06 pm »
Hi,

For a new project we want to look at the signals of a 1000 Base T connection.
I have roughly thought about it and would consider a 1Ghz scope for this - is that enough or would we have to go even higher?
Either way, only the 6000A would come into question, it is one of the cheapest on the market despite its performance.
The only question is, 1Ghz or 2Ghz.... ;)

Short version, 2GHz version would be better, but 1GHz should be usable..
Some lit... Holler if you need more..

DS7000A has compliance package  >:D.... Just saying... :-DD
« Last Edit: January 31, 2024, 11:27:14 pm by 2N3055 »
 
The following users thanked this post: Performa01, egonotto, KungFuJosh, Martin72

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #357 on: January 31, 2024, 11:57:03 pm »
Hi my friend,

DS7000A is too big for this project in financial terms.. ;)

Incidentally, this is an issue that affects both models.
Both want to be Pro models, and the data sheet says they are - but it's astonishing how few options come with these scopes.
The protocol and analysis options for our waverunner fill a whole page, unfortunately the 500Mhz are not enough.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 27673
  • Country: nl
    • NCT Developments
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #358 on: February 01, 2024, 01:09:31 am »
Look at the Tektronix appnote. Is says 8ns bit times. Just like 100Mbit ethernet, 1Gbit ethernet has a maximum frequency content of around 125MHz. CAT5E wiring just ain't going to pass higher frequencies. I'm not sure what the goal is here, but a 500MHz oscilloscope is going to do just fine to look at the signals. Compliance testing is of little use as you likely can't change anything on the transmitter / receiver chip (unless you are designing these kind of chips). What is left is the transmission system (wiring, filters, transformers) and for that you can use a network analyser to check the impedances of the transmission lines / system. All in all: what is the goal of the measurements?
« Last Edit: February 01, 2024, 01:26:09 am by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7117
  • Country: hr
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #359 on: February 01, 2024, 06:34:43 am »
Look at the Tektronix appnote. Is says 8ns bit times. Just like 100Mbit ethernet, 1Gbit ethernet has a maximum frequency content of around 125MHz. CAT5E wiring just ain't going to pass higher frequencies. I'm not sure what the goal is here, but a 500MHz oscilloscope is going to do just fine to look at the signals. Compliance testing is of little use as you likely can't change anything on the transmitter / receiver chip (unless you are designing these kind of chips). What is left is the transmission system (wiring, filters, transformers) and for that you can use a network analyser to check the impedances of the transmission lines / system. All in all: what is the goal of the measurements?

Says minimum 400ps risetime, BW 10x of 125Mhz, sample rate of 5GS/S. That fits to roughly 1 GHz scope. That is minimum
125Mhz is not the problem here but risetime, as always. 125Mhz is guaranteed for long cable runs. For short cables it will be more..
You will see that they performed actual testing with 4 GHz scope with 10GS/s...

Naturally, if you want to measure what characteristics are is different to just checking if it passes basic requirements.

I don't know what requirements are. From what little was said, it is not cable testing... And for that, a LAN analyser with compliance testing would be better match anyways.
« Last Edit: February 01, 2024, 07:02:01 am by 2N3055 »
 
The following users thanked this post: egonotto

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #360 on: February 01, 2024, 06:45:06 am »
I only got the flippant announcement "to snoop around in it"...
But I'll ask again.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 27673
  • Country: nl
    • NCT Developments
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #361 on: February 01, 2024, 10:31:56 am »
Look at the Tektronix appnote. Is says 8ns bit times. Just like 100Mbit ethernet, 1Gbit ethernet has a maximum frequency content of around 125MHz. CAT5E wiring just ain't going to pass higher frequencies. I'm not sure what the goal is here, but a 500MHz oscilloscope is going to do just fine to look at the signals. Compliance testing is of little use as you likely can't change anything on the transmitter / receiver chip (unless you are designing these kind of chips). What is left is the transmission system (wiring, filters, transformers) and for that you can use a network analyser to check the impedances of the transmission lines / system. All in all: what is the goal of the measurements?

Says minimum 400ps risetime, BW 10x of 125Mhz, sample rate of 5GS/S. That fits to roughly 1 GHz scope. That is minimum
No, that is Tektronix wanting to sell you an oscilloscope which is extreme overkill. Probably their lower bandwidth oscilloscope lack the analysis capabilities. If you look through the appnote you'll see that some measurements end at 100MHz and most of the accesoires (like filters and splitter) are specced up to a couple of hundred MHz.
« Last Edit: February 01, 2024, 12:17:30 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7117
  • Country: hr
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #362 on: February 01, 2024, 12:06:15 pm »
Look at the Tektronix appnote. Is says 8ns bit times. Just like 100Mbit ethernet, 1Gbit ethernet has a maximum frequency content of around 125MHz. CAT5E wiring just ain't going to pass higher frequencies. I'm not sure what the goal is here, but a 500MHz oscilloscope is going to do just fine to look at the signals. Compliance testing is of little use as you likely can't change anything on the transmitter / receiver chip (unless you are designing these kind of chips). What is left is the transmission system (wiring, filters, transformers) and for that you can use a network analyser to check the impedances of the transmission lines / system. All in all: what is the goal of the measurements?

Says minimum 400ps risetime, BW 10x of 125Mhz, sample rate of 5GS/S. That fits to roughly 1 GHz scope. That is minimum
No, that is Tektronix wanting to sell you an oscilloscope which is extreme overkill. Probably their lower bandwidth oscilloscope lack the analysis capabilities. If you look through the appnote you'll see that some measurements end at 100MHz.

Not Tektronix, the actual working group...
 
The following users thanked this post: egonotto

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #363 on: February 01, 2024, 09:08:16 pm »
In the meantime, the wishes have become more concrete, so the Siglent models are out.
Because protocol analyzes should also be possible and, as I already mentioned, the available options for the large models are rather limited.
For our Waverunner model, however, there are dozens of options available, but the hope that its bandwidth (500Mhz, smallest model) would be sufficient has died.
Lecroy recommends 1Ghz for T100 and 2.5Ghz for T1000.
But this doesn't belong here, maybe I'll open a new thread.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: egonotto

Online nctnico

  • Super Contributor
  • ***
  • Posts: 27673
  • Country: nl
    • NCT Developments
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #364 on: February 01, 2024, 09:29:15 pm »
In the meantime, the wishes have become more concrete, so the Siglent models are out.
Because protocol analyzes should also be possible and, as I already mentioned, the available options for the large models are rather limited.
For our Waverunner model, however, there are dozens of options available, but the hope that its bandwidth (500Mhz, smallest model) would be sufficient has died.
Lecroy recommends 1Ghz for T100 and 2.5Ghz for T1000.
But this doesn't belong here, maybe I'll open a new thread.
Opening a new thread sounds like a good plan. Having quite a bit of hands-on experience with various types of electrical and optical ethernet and other types of high speed serial interfaces, I strongly doubt an oscilloscope or protocol analyser is going to help solve a problem you may have. Getting the problem definition clear is likely way more useful as it will point into the direction of the solution.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto, 2N3055, Martin72

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #365 on: May 01, 2024, 11:10:01 am »
New firmware for SDS6000A models.

Version: V1.4.9.3
131MB
https://int.siglent.com/upload_file/zip/firmware/Oscilloscope/SDS6000_V1.4.9.3_EN.zip

Release notes
1. Supported new probes: Siglent’s SCP5030, SCP5030A, SCP5150, SCP5500 and LeCroy’s CP150, HVD3106A
2. Option to hide Memory and Math traces
3. Webserver: Screenshot can be directly copied and pasted
4. Fixed several bugs
a) Decode Threshold bug
b) NO persistence available with frames captured by SEQUENCE
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Ulrich.G

  • Regular Contributor
  • *
  • Posts: 64
  • Country: de
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #366 on: August 07, 2024, 08:29:59 am »
Any signs to see a pro model, at least 10bit, for the western market?
 
The following users thanked this post: tautech

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #367 on: August 07, 2024, 08:57:02 am »
Any signs to see a pro model, at least 10bit, for the western market?
I'm just hoping future FW will give us 12bit capability as a gift....which would cover the existing 12bit BW hole between SDS3104X HD and SDS7304A.
 :popcorn:
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Dennis Frie

  • Contributor
  • Posts: 25
  • Country: dk
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #368 on: August 07, 2024, 09:08:33 am »
While that would be great, I guess they will focus on expanding their HD-series rather than updating a fairly old scope (everything considered) ;)
I just purchased a SDS6000A for high speed verification at work, but sure would love to get 12 bit. There's a fairly big "hole" between the SDS3000X HD and SDS7000A
 
The following users thanked this post: tautech

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #369 on: August 07, 2024, 09:19:35 am »
While that would be great, I guess they will focus on expanding their HD-series rather than updating a fairly old scope (everything considered) ;)
I just purchased a SDS6000A for high speed verification at work, but sure would love to get 12 bit. There's a fairly big "hole" between the SDS3000X HD and SDS7000A
Coming are SDS7000A 6 and 8 GHz models.....
« Last Edit: August 07, 2024, 09:25:24 am by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #370 on: August 07, 2024, 09:41:20 am »
Any signs to see a pro model, at least 10bit, for the western market?

If you don't necessarily need 2Ghz, the 3000X HD is the gap filler.
Although of course I would also prefer a larger screen. ;)
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline Ulrich.G

  • Regular Contributor
  • *
  • Posts: 64
  • Country: de
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #371 on: August 07, 2024, 09:59:23 am »
If you don't necessarily need 2Ghz, the 3000X HD is the gap filler.
Although of course I would also prefer a larger screen. ;)

Its not necessarily the bandwidth, but the four separate ADCs at 5GSa/sec, the bigger screen, the eye diagram & jitter analysis,  ...

« Last Edit: August 07, 2024, 10:03:12 am by Ulrich.G »
 

Offline Ulrich.G

  • Regular Contributor
  • *
  • Posts: 64
  • Country: de
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #372 on: August 07, 2024, 10:02:40 am »
I'm just hoping future FW will give us 12bit capability as a gift....which would cover the existing 12bit BW hole between SDS3104X HD and SDS7304A.

Do you have informations, that at least some of the pro models (lower bandwidth & 5GSa/sec) sharing the same hardware with the SDS6000A?
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #373 on: August 07, 2024, 10:43:26 am »
The 6000A has the same ADCs as the 6000 Pro H10 , H12.
It is the lower cost Pro model with the same features as the much more expensive H10 and H12.
And you can guess why the 3000 X HD has up to 16bit ERES instead of 12 + 3 bits.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: Ulrich.G

Online hpw

  • Frequent Contributor
  • **
  • Posts: 409
  • Country: 00
Re: Siglent SDS6000A DSO's 500MHz-2GHz
« Reply #374 on: August 07, 2024, 04:02:48 pm »
While that would be great, I guess they will focus on expanding their HD-series rather than updating a fairly old scope (everything considered) ;)
I just purchased a SDS6000A for high speed verification at work, but sure would love to get 12 bit. There's a fairly big "hole" between the SDS3000X HD and SDS7000A
Coming are SDS7000A 6 and 8 GHz models.....

hmm, this means any faster ADC and better ref. oszillator required. So the prices as >> 20k$.

Let's see what Keysight will show on September as now 14 bits. If really 14 bit for 1GHz, high ENOB 12 bits, this will be the next game changer for 22k$.