Author Topic: Siglent SDS2000X Plus  (Read 938499 times)

0 Members and 10 Guests are viewing this topic.

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27659
  • Country: nl
    • NCT Developments
Re: Siglent SDS2000X Plus
« Reply #2575 on: March 23, 2021, 06:25:10 pm »
A lot of fuss about $5 probes. Many people probably toss the probes in the bin and buy (or already have) proper ones anyway. Probes that come with scopes are a courtesy item -nothing more nothing less-.

That is not a normal comment from you. They paid it, they are entitled to it even if costed $0.01.

In no place in the invoice it is said that it's a "GIFT"!
But the invoice doesn't state a price either... Anyway, it is rather wasteful to go through a lot of trouble for an item which many people want to replace anyway. If this is about special and/or high quality probes I'd agree that you should get what is ordered / promised. But making so much fuss about low end probes is just not worth it. Think about the extra amount of e-waste, time & energy spend with sending out extra low quality probes. IMHO it would be much better to sell scopes without probes so the buyers can choose what kind of probes they want. Added to that is the fact that it is questionable whether you can actually get a good measurement for signals over 100MHz using High-Z probes anyway due to the large capacitive loading of the probe. All it boils down is demanding 'better' probes from a principle while getting very little practical / real added value.

Besides that it seems to me Siglent has changed its mind and is now shipping 100MHz probes with 100MHz scopes which seems logical. Likely the marketing material has not been updated yet.
« Last Edit: March 23, 2021, 07:35:43 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline Self Bias

  • Contributor
  • Posts: 20
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #2576 on: March 23, 2021, 07:56:45 pm »
New owner of a SDS2104X+ here. I also received 4x PP510 probes, going to e-mail the seller now. What might be helpful at least for people in Europe:

https://www.siglenteu.com/wp-content/uploads/dlm_uploads/2017/10/Probe_List_12302020.pdf

This is the list of standard probes supplied for each device. If I read the timestamp correctly, it's from the 30th of December 2020, so fairly recent.

I also had a weird malfunction with one of the probes: In 1:10 mode it seemed to only have capacitive coupling. The problem was reproducible and followed when I changed channels with the probe. 1:1 mode worked fine.
I tried it again a few hours later and the issue was gone, never reappeared since. Screenshot attached with another probe and this one connected to the CAL output.
« Last Edit: March 23, 2021, 08:03:02 pm by Self Bias »
 

Offline oz2cpu

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: dk
    • webx.dk private hobby and diy stuff
Re: Siglent SDS2000X Plus
« Reply #2577 on: March 23, 2021, 08:09:30 pm »
to self bias : it is the plastic sleeve of the probe, it is loose, that affect how the 10:1 switch is, and how much it can move,
you need to press the outher sleeve back, now the switch works,
happens to me all the time, need to figure out how to glue this
« Last Edit: March 23, 2021, 08:13:54 pm by oz2cpu »
Radioamateur OZ2CPU, Senior EE at Prevas
EMC RF SMPS SI PCB LAYOUT and all that stuff.
youtube : oz2cpu teardown
 
The following users thanked this post: Self Bias

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2578 on: March 23, 2021, 08:11:31 pm »
I don't think its a lot of fuss when you spend $1400 dollars on a scope and get probes with half the bandwidth than you were supposed to get. I've been saving up for the majority of my internship to try and build a decent lab. Although I do plan on buying other probes at some point, I would like to receive everything I have been quoted
My thoughts exactly.

We do know from frequency sweep tests done by Performa01 that PP510 and PP215 do perform very similar to out past their respective -3dB rated BW to 300 MHz.
Those results are here:
https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-in-depth-review/msg1434665/#msg1434665


But rather than the 200 MHz PP215 it is my view the 100 and 200 MHz models should be shipped with the $59 300 MHz 10x SP2030A Auto-Sense probe that with these DSO's and their 500uV/div sensitivity still permits selection of 5mV/div max sensitivity which is more than adequate for the vast majority of work.


OTOH I have written to high levels at Siglent to investigate why units are being supplied with PP510 probes instead of the datasheet specified PP215 probes.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: TK, KungFuJosh, Martin72, profanum429, Jeff C

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27659
  • Country: nl
    • NCT Developments
Re: Siglent SDS2000X Plus
« Reply #2579 on: March 23, 2021, 09:00:52 pm »
I don't think its a lot of fuss when you spend $1400 dollars on a scope and get probes with half the bandwidth than you were supposed to get. I've been saving up for the majority of my internship to try and build a decent lab. Although I do plan on buying other probes at some point, I would like to receive everything I have been quoted
My thoughts exactly.

We do know from frequency sweep tests done by Performa01 that PP510 and PP215 do perform very similar to out past their respective -3dB rated BW to 300 MHz.
Those results are here:
https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-in-depth-review/msg1434665/#msg1434665
Unfortunately such tests say absolutely nothing about probing in real circuits  :palm: Especially for frequencies over 100MHz. It has been explained before and any experienced engineer can tell you that. Or put differently: I never use a High-Z probe to probe signals over 100MHz because I know what I get on screen from a High-Z probe is a fantasy signal.
« Last Edit: March 23, 2021, 09:02:51 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6620
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #2580 on: March 23, 2021, 10:06:08 pm »
I think, it´s a general (and interesting)thing and should be forwarded in an explicit thread, doesn´t have to do with the SDS2K+ only.

"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline KungFuJosh

  • Super Contributor
  • ***
  • Posts: 2444
  • Country: us
  • TEAS is real.
Re: Siglent SDS2000X Plus
« Reply #2581 on: March 23, 2021, 11:35:19 pm »
All it boils down is demanding 'better' probes from a principle while getting very little practical / real added value.

Besides that it seems to me Siglent has changed its mind and is now shipping 100MHz probes with 100MHz scopes which seems logical. Likely the marketing material has not been updated yet.

I'm with you in that I don't use the stock probes (I have some nice ProbeMaster probes). But if a product is advertised with specific accessories, and they send something of lesser quality, that's false advertising, and it's generally illegal.
"Right now I’m having amnesia and déjà vu at the same time. I think I’ve forgotten this before." - Steven Wright
 

Online mawyatt

  • Super Contributor
  • ***
  • Posts: 3735
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #2582 on: March 23, 2021, 11:55:07 pm »
If you order a new 200MPH Porsche 911 Twin Turbo you expect the tires to be better rated than 100MPH!! The factory supplies tires that are rated for 200MPH, if you want to change them you certainly can, but they are rated for 200MPH, and you paid for those 200MPH tires ;)

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 
The following users thanked this post: oz2cpu

Offline oz2cpu

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: dk
    • webx.dk private hobby and diy stuff
Re: Siglent SDS2000X Plus
« Reply #2583 on: March 24, 2021, 08:51:37 am »
well we "chip-tuned" our scopes to run double as fast, surely we cant expect the stock probes to follow our new "demand"
Radioamateur OZ2CPU, Senior EE at Prevas
EMC RF SMPS SI PCB LAYOUT and all that stuff.
youtube : oz2cpu teardown
 

Offline jemangedeslolos

  • Frequent Contributor
  • **
  • Posts: 386
  • Country: fr
Re: Siglent SDS2000X Plus
« Reply #2584 on: March 24, 2021, 09:07:36 am »
If you order a new 200MPH Porsche 911 Twin Turbo you expect the tires to be better rated than 100MPH!! The factory supplies tires that are rated for 200MPH, if you want to change them you certainly can, but they are rated for 200MPH, and you paid for those 200MPH tires ;)

Best,

You chose the wrong example.
If you want to talk about precision on numbers, you must avoid talking about a brand that belongs to the Volkwagen group  :-DD
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 7111
  • Country: hr
Re: Siglent SDS2000X Plus
« Reply #2585 on: March 24, 2021, 10:08:27 am »
If you order a new 200MPH Porsche 911 Twin Turbo you expect the tires to be better rated than 100MPH!! The factory supplies tires that are rated for 200MPH, if you want to change them you certainly can, but they are rated for 200MPH, and you paid for those 200MPH tires ;)

Best,

You chose the wrong example.
If you want to talk about precision on numbers, you must avoid talking about a brand that belongs to the Volkwagen group  :-DD

911 is pretty much purebred Porsche.. Other models not so much, but 911 is...

To be honest, I personally consider 911 only real Porsche... >:D

OT I think that if SDS2000X+ says in datasheet that it comes with 200Mhz model probes, it should come with those.. For correctness of data, if nothing else.

OTOH, like oz2cpu says, if you buy 200MHz scope and put a turbo on it, you need to buy better tires yourself... Or something like that  :-DD
 

Online mawyatt

  • Super Contributor
  • ***
  • Posts: 3735
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #2586 on: March 24, 2021, 12:59:08 pm »
If you order a new 200MPH Porsche 911 Twin Turbo you expect the tires to be better rated than 100MPH!! The factory supplies tires that are rated for 200MPH, if you want to change them you certainly can, but they are rated for 200MPH, and you paid for those 200MPH tires ;)

Best,

You chose the wrong example.
If you want to talk about precision on numbers, you must avoid talking about a brand that belongs to the Volkwagen group  :-DD

Not the wrong example, Porsche just belongs to a larger group but doesn't convey they are performing or even held to the Volkswagen metrics which I do consider questionable, same goes for Audi. The Porsche 911 is recognized worldwide as the benchmark for all sports cars, why we have two (BTW the red 911 is a 993 model which is the last of the air cooled 911s, and worth more today than when purchased new in 1996!!) ;D

LeCroy belongs to Teledyne, but they still perform to their own established metrics!! I mean what does Teledyne know about oscilloscopes??

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 

Online mawyatt

  • Super Contributor
  • ***
  • Posts: 3735
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #2587 on: March 24, 2021, 01:16:20 pm »
If you order a new 200MPH Porsche 911 Twin Turbo you expect the tires to be better rated than 100MPH!! The factory supplies tires that are rated for 200MPH, if you want to change them you certainly can, but they are rated for 200MPH, and you paid for those 200MPH tires ;)

Best,

You chose the wrong example.
If you want to talk about precision on numbers, you must avoid talking about a brand that belongs to the Volkwagen group  :-DD

911 is pretty much purebred Porsche.. Other models not so much, but 911 is...

To be honest, I personally consider 911 only real Porsche... >:D

OT I think that if SDS2000X+ says in datasheet that it comes with 200Mhz model probes, it should come with those.. For correctness of data, if nothing else.

OTOH, like oz2cpu says, if you buy 200MHz scope and put a turbo on it, you need to buy better tires yourself... Or something like that  :-DD

Agree, the 911 is Porsche!! The 911 just won the 12 hours of Sebring in the GTLM and GTD classes, this is Porsche 100th total class wins at Sebring since 1st racing in ~1954!!

Cars are like scopes, the faster the better, but you need proper rated tires and probes to reach the full potential of either ;D

Best,
Curiosity killed the cat, also depleted my wallet!
~Wyatt Labs by Mike~
 
The following users thanked this post: 2N3055, oz2cpu

Offline mubes

  • Regular Contributor
  • *
  • Posts: 238
  • Country: gb
  • Do Not Boil
Re: Siglent SDS2000X Plus
« Reply #2588 on: March 25, 2021, 10:06:13 am »
I don't suppose anyone knows a way to list the installed options on this scope via SCPI, do they? The 'normal' *OPT? doesn't seem to deliver the goods, and there's nothing in the programming manual suggesting a route.

DAVE
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2589 on: March 25, 2021, 10:32:38 am »
I don't suppose anyone knows a way to list the installed options on this scope via SCPI, do they? The 'normal' *OPT? doesn't seem to deliver the goods, and there's nothing in the programming manual suggesting a route.

DAVE
Why ?  :-//

If you go to the Options page in the Utility menu and hit Print you can get that screenshot sent to a USB stick if you need some record of what's installed.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27659
  • Country: nl
    • NCT Developments
Re: Siglent SDS2000X Plus
« Reply #2590 on: March 25, 2021, 10:53:28 am »
I don't suppose anyone knows a way to list the installed options on this scope via SCPI, do they? The 'normal' *OPT? doesn't seem to deliver the goods, and there's nothing in the programming manual suggesting a route.

DAVE
Why ?  :-//
Simple: say you want to write a piece of software which depends on a certain option; you will want to detect whether that option is installed and if that is not the case present a clear warning to the user instead of failing in an undefined way. Probably you can achieve the same using checks to access optional functionality but in general that is not the easiest way.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto

Offline mubes

  • Regular Contributor
  • *
  • Posts: 238
  • Country: gb
  • Do Not Boil
Re: Siglent SDS2000X Plus
« Reply #2591 on: March 25, 2021, 11:00:18 am »
Simple: say you want to write a piece of software which depends on a certain option; you will want to detect whether that option is installed and if that is not the case present a clear warning to the user instead of failing in an undefined way. Probably you can achieve the same using checks to access optional functionality but in general that is not the easiest way.

This. Porting all the functionality of scopehal is going to be a bit tricky without such a capability....
 

Offline Steamdonkey

  • Newbie
  • Posts: 4
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #2592 on: March 25, 2021, 01:13:37 pm »
What probes are included from the factory with this machine?
  • SDS2104X+: 4x PP215 - Correct?
  • SDS2204X+: 4x PP430 ?
  • SDS2354X+: 4x PP430 ?

My SDS2104X+ came with PP510, the 100MHz Siglent probes.

Same here. Ordered an SDS2104X+ from tequipment.net, which arrived yesterday, and it came with 4x PP510 100MHz probes.  Not sure what brand they are. 
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2593 on: March 25, 2021, 01:35:16 pm »
What probes are included from the factory with this machine?
  • SDS2104X+: 4x PP215 - Correct?
  • SDS2204X+: 4x PP430 ?
  • SDS2354X+: 4x PP430 ?

My SDS2104X+ came with PP510, the 100MHz Siglent probes.

Same here. Ordered an SDS2104X+ from tequipment.net, which arrived yesterday, and it came with 4x PP510 100MHz probes. 
Ask them to replace them with PP215's as listed in the datasheet.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Re: Siglent SDS2000X Plus
« Reply #2594 on: March 25, 2021, 09:44:34 pm »
What probes are included from the factory with this machine?
  • SDS2104X+: 4x PP215 - Correct?
  • SDS2204X+: 4x PP430 ?
  • SDS2354X+: 4x PP430 ?

My SDS2104X+ came with PP510, the 100MHz Siglent probes.

Same here. Ordered an SDS2104X+ from tequipment.net, which arrived yesterday, and it came with 4x PP510 100MHz probes. 
Ask them to replace them with PP215's as listed in the datasheet.
Received my unit today from tequipment... calibration date is February 2021, received 4 x PP510 probes.  It is clear by now that someone at the factory decided to cut some cost
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2595 on: March 25, 2021, 10:01:23 pm »
Received my unit today from tequipment... calibration date is February 2021, received 4 x PP510 probes.  It is clear by now that someone at the factory decided to cut some cost
Actually not but I won't go into this until I have official confirmation.

Please see you claim for the correct PP215 probe as replacements.
« Last Edit: March 25, 2021, 10:11:08 pm by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Steamdonkey

  • Newbie
  • Posts: 4
  • Country: us
Re: Siglent SDS2000X Plus
« Reply #2596 on: March 25, 2021, 10:07:44 pm »


My SDS2104X+ came with PP510, the 100MHz Siglent probes.

Same here. Ordered an SDS2104X+ from tequipment.net, which arrived yesterday, and it came with 4x PP510 100MHz probes. 
Ask them to replace them with PP215's as listed in the datasheet.
I can't seem to find any current documentation that lists the probes that come with the SDS2104X+.  Maybe with the current limited time offer (free software - I2S, FlexRay, Mil-1553B, CANFD, Power Analysis and Built-in 50MHz AWG) that runs through the end of the month they're shipping it with cheaper probes. 
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2597 on: March 25, 2021, 10:10:29 pm »


My SDS2104X+ came with PP510, the 100MHz Siglent probes.

Same here. Ordered an SDS2104X+ from tequipment.net, which arrived yesterday, and it came with 4x PP510 100MHz probes. 
Ask them to replace them with PP215's as listed in the datasheet.
I can't seem to find any current documentation that lists the probes that come with the SDS2104X+.  Maybe with the current limited time offer (free software - I2S, FlexRay, Mil-1553B, CANFD, Power Analysis and Built-in 50MHz AWG) that runs through the end of the month they're shipping it with cheaper probes.
P3 of the datasheet.
See here:
https://www.eevblog.com/forum/testgear/siglent-sds2000x-plus-coming/msg3525558/#msg3525558

If you received PP510 probes they are wrong and claim for the correct 200 MHz PP215 probes.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6620
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #2598 on: March 25, 2021, 11:45:29 pm »
Regardless of wrong or right cheapest probes delivered, it´s a strange policy from siglent, giving the cheapest probes they can get to this not-so-cheap-anymore-scope.
I would expect that for their very low price models, not for this one.
When rigol could offer 350Mhz probes even with their 70Mhz model of the 5000 series, which costs nearly 500 bucks less...
When Lecroy doesn´t have a problem to pack their 350Mhz scope with 4 500Mhz probes, where each of them will costs extremely when buying separately..
...It tells me, that probes itselfs will cost not much for rigol or lecroy.
But for siglent ?
What will happen, when I´m upgrading the bandwith official ?
Do I get new probes with the license key too ?
I think not..
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 29185
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #2599 on: March 26, 2021, 12:21:47 am »
Regardless of wrong or right cheapest probes delivered, it´s a strange policy from siglent,
It's not policy but nonadherence to the official datasheet.

Let me say this again....anyone that receives PP510 probes with a new SDS2104X Plus instead of the specified PP215 probes contact your supplier to have them replaced.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf