Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #225 on: August 22, 2023, 09:38:55 pm »
Maybe there are sporadic problems with the autodetection of the probe.
Or the probe is faulty.
Have you checked by using other probes on the same channel ?
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #226 on: August 22, 2023, 09:45:11 pm »
Hi Rob,

No, because the freeze thing was a one-time thing - before and after that I didn't have such a crash.
The thing with the 1x/10x, I had however more often, mostly if the Bodeplot mode was terminated, then one channel remained on 10x, the other on 1x - although both were before on 10x.
Maybe this will be resolved when the firmware update is released at some point.
In principle, you have an "old" scope with a firmware version from almost a year ago.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #227 on: August 24, 2023, 08:02:26 pm »
Here is a current example regarding the error of not being able to remember the thresholds.
"Yesterday" the scope was set to 1.70V, today directly after power on the thresholds are at 7.1V and 4.1V.
Which makes no sense at all, at logic levels.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline teddychn

  • Regular Contributor
  • *
  • Posts: 51
  • Country: 00
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #228 on: September 08, 2023, 07:17:54 am »
"Yesterday" the scope was set to 1.70V, today directly after power on the thresholds are at 7.1V and 4.1V.
Which makes no sense at all, at logic levels.

This should be corrected. No one would want to readjust the logic level every day before using the logic analyzer.
 

Offline JPortici

  • Super Contributor
  • ***
  • Posts: 3515
  • Country: it
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #229 on: September 08, 2023, 07:54:25 am »
Here is a current example regarding the error of not being able to remember the thresholds.
"Yesterday" the scope was set to 1.70V, today directly after power on the thresholds are at 7.1V and 4.1V.
Which makes no sense at all, at logic levels.

That is probably the same problem that is present in the SDS5000X.
Disconnect the probe, see if defaults to 1:1 (or to anything else than what the probe is)
If this is the case, there's your problem.
Something bizarre happens during initialization: I don't know how the thresholds are saved internally, but the data gets mangled when the probe type changes during boot.

The scope will start with a preset probe ratio for the channel (and all the values that are tied to that channel will be applied with that scale). Then the autodetect will detect the probe and adapt all the values to the new scale.

scope saves thresholds in FORMAT B (because probe is connected) then on boot loads FORMAT B into FORMAT A because channel is in its default state then autodetect and FORMAT A goes into FORMAT B

The problem here is that autodetect overrides the probe format, but doesn't store it
 
The following users thanked this post: 2N3055

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4132
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #230 on: September 08, 2023, 07:59:01 am »
Here is a current example regarding the error of not being able to remember the thresholds.
"Yesterday" the scope was set to 1.70V, today directly after power on the thresholds are at 7.1V and 4.1V.
Which makes no sense at all, at logic levels.

I can not repeat this. (SDS2504X HD   FW: 1.2.1.1)

1st image. Setup as it looks like in your image.

Shutdown. 

Power on again.

2nd image.

I have repeated this many times, always result is same.




 1.


 2.



ETA:  When I write this msg  I have not read @JPortici.  His observations and suspicions may be significant also here.


« Last Edit: September 08, 2023, 08:21:09 am by rf-loop »
BEV of course. Cars with smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the (strong)wises gone?
 

Offline teddychn

  • Regular Contributor
  • *
  • Posts: 51
  • Country: 00
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #231 on: September 08, 2023, 08:30:52 am »
Here are my wishes for the SDS2000X HD:

1. Add an option for a comments field when printing or taking screenshots.

2. Channel labels can become difficult to distinguish when they overlap with each other. It would be great to either stagger the overlapped labels or display them elsewhere, perhaps in a side pop-up box or anywhere else on the screen.

3. Include bandwidth information alongside the ERES message.

4. Enable customization of XY chart colors. The dark blue is hard to read in screen captures.
 
The following users thanked this post: rf-loop

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 7117
  • Country: hr
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #232 on: September 08, 2023, 09:04:49 am »
Here are my wishes for the SDS2000X HD:

4. Enable customization of XY chart colors. The dark blue is hard to read in screen captures.

Can you explain this, please?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4132
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #233 on: September 08, 2023, 10:58:37 am »

picture link from:
https://github.com/microsoft/WSL/issues/930

I believe that everyone who has, for example, designed important cockpits / control rooms in industry, knows the basics of visual ergonomics (and other human ergonomics and how body and how all the senses, sight, hearing, touch, muscle control and position sense and fine motor skills work as well as one have to know a little about behavior as well. It would be good for measuring device manufacturers to know these things too...including Siglent also,  least the basics of these things.
But who would teach these things if they themselves don't feel like studying. In the short term, stock prices on the stock exchange are more important...

RGB  B alone on Black background is problematic in many situations..


« Last Edit: September 08, 2023, 11:02:05 am by rf-loop »
BEV of course. Cars with smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the (strong)wises gone?
 

Offline teddychn

  • Regular Contributor
  • *
  • Posts: 51
  • Country: 00
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #234 on: September 08, 2023, 11:02:40 am »
Here are my wishes for the SDS2000X HD:

4. Enable customization of XY chart colors. The dark blue is hard to read in screen captures.

Can you explain this, please?

This is a screenshot from the SDS2104X Plus, but I believe the HD should be similar. I find it challenging to me to distinguish the vertical blue line in the picture. The SDS2000X HD allows you to assign trace colors, and I wish the XY chart could be customized in the same way.

By the way, Siglent should include bookmarks in their PDF manuals; it would make them easier to use.
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4132
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #235 on: September 08, 2023, 11:10:22 am »
Also same for digital (LA) input signals.
BEV of course. Cars with smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the (strong)wises gone?
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #236 on: September 08, 2023, 11:17:39 am »
Also same for digital (LA) input signals.
Should chance as I proposed.  ;)
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 7117
  • Country: hr
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #237 on: September 08, 2023, 12:56:08 pm »
Here are my wishes for the SDS2000X HD:

4. Enable customization of XY chart colors. The dark blue is hard to read in screen captures.

Can you explain this, please?



This is a screenshot from the SDS2104X Plus, but I believe the HD should be similar. I find it challenging to me to distinguish the vertical blue line in the picture. The SDS2000X HD allows you to assign trace colors, and I wish the XY chart could be customized in the same way.

By the way, Siglent should include bookmarks in their PDF manuals; it would make them easier to use.

Here is a snapshot from my scope. On screen blue trace is much brighter, lighter blue than on this image. That is why I was confused, I have no trouble seeing it.
But I think your request is good one, it should be user configurable..

As a side note, to make it easier to read (if you have problem with blue) enable Color grading in Display menu... Maybe that helps a bit..
 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #238 on: September 08, 2023, 05:39:30 pm »
Hi,
Here a clip about the threshold thing:
https://drive.google.com/file/d/1srETzAoYwZZwdh1fOC-6uWFfBZE9bN9M/view?usp=sharing

(Meh, why is the video quality so bad...(EDIT: Because it´s 360p,can be changed to 1080))

Thresholds changing also when you re-boot the scope.
Threshold-Level is different from channel to channel, for example Ch 1 is falling back to 8.2V, Ch2 something between 800...833mV, Ch3 14.2V, Ch4 I didn´t test so far.

Quote
I can not repeat this. (SDS2504X HD   FW: 1.2.1.1)

Got the same FW...
I will test it again tomorrow, but then without the probes - Possibly the autosense function messes the whole thing up.
(For example, I currently have all probes on the scope and after each start it "tells" me that it has detected a probe on channel 4, although I had not disconnected and reconnected them).
« Last Edit: September 08, 2023, 08:21:44 pm by Martin72 »
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline JPortici

  • Super Contributor
  • ***
  • Posts: 3515
  • Country: it
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #239 on: September 08, 2023, 08:55:20 pm »
Hi,
Here a clip about the threshold thing:
https://drive.google.com/file/d/1srETzAoYwZZwdh1fOC-6uWFfBZE9bN9M/view?usp=sharing

(Meh, why is the video quality so bad...(EDIT: Because it´s 360p,can be changed to 1080))

Thresholds changing also when you re-boot the scope.
Threshold-Level is different from channel to channel, for example Ch 1 is falling back to 8.2V, Ch2 something between 800...833mV, Ch3 14.2V, Ch4 I didn´t test so far.

Quote
I can not repeat this. (SDS2504X HD   FW: 1.2.1.1)

Got the same FW...
I will test it again tomorrow, but then without the probes - Possibly the autosense function messes the whole thing up.
(For example, I currently have all probes on the scope and after each start it "tells" me that it has detected a probe on channel 4, although I had not disconnected and reconnected them).

Disconnect probes. set channels mode to 1:10, then retry ;) you are describing exactly the issue i have with the SDS5000X ;)
 
The following users thanked this post: Martin72

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #240 on: September 08, 2023, 08:59:31 pm »
Makes me so curious, I´ll test it in the next minutes.. ;)
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #241 on: September 08, 2023, 09:17:59 pm »
Bingo... ;)

Disconnecting the probes, setting all the channels manually to 1:10, then adjusting the decoder threshold-levels.
Re-booting the scope, now the threshold values remain.
Siglent, we have a bug in the autosense circuit (programm?)... 8)
And this seems to affect not only the level setting for the decoding, because I had already told early that when fall back from the bodeplot mode a channel is set back to 1:1, although a 1:10 autosense probe was connected and recognized as such before starting the bodeplot.
Or how the scope froze for me for the first time a few weeks ago, with the message that it had detected a probe on channel 3, although the probe was already connected.
So it seems to influence many things, a rather serious bug.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline teddychn

  • Regular Contributor
  • *
  • Posts: 51
  • Country: 00
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #242 on: September 15, 2023, 08:24:42 am »
Here is a snapshot from my scope. On screen blue trace is much brighter, lighter blue than on this image. That is why I was confused, I have no trouble seeing it.
But I think your request is good one, it should be user configurable..

As a side note, to make it easier to read (if you have problem with blue) enable Color grading in Display menu... Maybe that helps a bit..

Thanks 2N3055,

I tried your idea, and it seems a bit better.

Actually, I can see the dark blue details on the oscilloscope screen, but they appeared too dark to be seen on my laptop.

Here are the new captures.
 

Offline Ulrich.G

  • Regular Contributor
  • *
  • Posts: 64
  • Country: de
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #243 on: October 12, 2023, 09:43:41 am »
Firmware update V1.2.2.5 is available!

https://int.siglent.com/upload_file/zip/firmware/Oscilloscope/SDS2000X_HD_V1.2.2.5_EN.zip

Its not just a software update of the OS, but also reprogramming the FPGA to version 2023-05-30  :o
« Last Edit: October 12, 2023, 11:11:48 am by Ulrich.G »
 
The following users thanked this post: 2N3055, Martin72

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #244 on: October 12, 2023, 09:52:36 am »
YES  :-+
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: 2N3055

Offline Ulrich.G

  • Regular Contributor
  • *
  • Posts: 64
  • Country: de
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #245 on: October 12, 2023, 01:32:20 pm »
With the new firmware version 1.2.2.5, we not only got 4 math channels, but also 4 memory channels. The possibilities seem to be almost endless. Import data from other channels, files, math functions...

I'm a little overwhelmed with it, but great feature.

Found the decription in the SDS6000A manual - page 260-262
 
The following users thanked this post: 2N3055, Martin72, JimKnopf

Offline kladit

  • Regular Contributor
  • *
  • Posts: 51
  • Country: de
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #246 on: October 14, 2023, 06:16:26 pm »
The description of what?

I  looked at the "SDS6000A User Manual" and at page 260 found a description of "Mask Setup".

Neither "memory channel" nor "memory trace" can be found in this pdf.

So please give link to the pdf you are talking of.


 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #247 on: October 14, 2023, 06:31:18 pm »
The description of what?

I  looked at the "SDS6000A User Manual" and at page 260 found a description of "Mask Setup".

Neither "memory channel" nor "memory trace" can be found in this pdf.

So please give link to the pdf you are talking of.
Try this manual version:
https://int.siglent.com/u_file/download/22_10_24/SDS6000A_UserManual_EN01B.pdf

Chapter 21 Memory P260
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: kladit, Ulrich.G

Online Martin72Topic starter

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #248 on: October 14, 2023, 07:06:03 pm »
To the beta testers active here:
Do any of you have a 2kHD at home ?
Then please send me a PM. ;)

As for the new firmware:
The datum thing works.. :-+
Window font size I also found and set to "small", I haven't noticed a direct difference yet.
And a bugfix not listed in the revision I also discovered:
The save message now disappears by itself again.... :-+
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: 2N3055, KungFuJosh

Offline kripton2035

  • Super Contributor
  • ***
  • Posts: 2670
  • Country: fr
    • kripton2035 schematics repository
Re: Siglent SDS2000X HD, missing Features and Bugs
« Reply #249 on: October 14, 2023, 07:56:12 pm »
I've posted it in the other 2000HD thread, but this one on the bug is a better place
with last firmware, if you press default, the 10x probe is still not detected. you must unplug and plug it to be detected as 10x
 