At 200us/div and max 2Mpts FFT a sweep speed of ~1/s is obtained.
100us/div = ~2/s
50us/div = ~4/s

Rigol is three or four times faster than that.

These results are only applicable to the scope settings in use. FFT 2Mpts remained unchanged although it's auto managed by the scope. Memory depth is just what it was, didn't check.
Higher FFT sweep speeds are likely possible if one takes the trouble to optimize them.
Pick your poison, speed or accuracy.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7123
  • Country: hr
Re: SDS800x HD vs DHO800x FFT
« Reply #26 on: July 16, 2024, 08:03:15 am »
To be honest, only scope in less than 7000-8000 € range that has something resembling realtime SA is GW-Instek MDO-2000A series.
And that is because it literally has realtime SA mode implemented. It works well but it's ADC are only 8 bit so limited dynamic range. Also scopes are in 1500€ ranges, a bit old fashioned by now etc.....

All other scopes in that price range will have only FFT as a math function, with implementation variations.
Short version for the two scopes mentioned:

- if you want just something bouncing on the screen to see if there is something happening at some frequency, DHO800 will have faster refresh at some settings.
- if you want control of what are you measuring, accurate windows, accurate level measurements, Peak Hold and Averaging, various markers and peek markers and control of them, then SDS800xHD is the one.

If you want SA display on your scope that resembles R&S MXO4 or MXO5 or Tektronix MSO 6 digital down conversion tech, you will have to pay for those...
 
The following users thanked this post: Martin72

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29214
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDS800x HD vs DHO800x FFT
« Reply #27 on: July 16, 2024, 08:10:08 am »
To be honest, only scope in less than 7000-8000 € range that has something resembling realtime SA is GW-Instek MDO-2000A series.
And that is because it literally has realtime SA mode implemented. It works well but it's ADC are only 8 bit so limited dynamic range. Also scopes are in 1500€ ranges, a bit old fashioned by now etc.....

..........
Which buys a SSA3021X Plus with dedicated EMI mode.

See Reply #8
« Last Edit: July 16, 2024, 08:22:57 am by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17067
  • Country: 00
Re: SDS800x HD vs DHO800x FFT
« Reply #28 on: July 16, 2024, 09:31:05 am »
As I have mentioned before, FFT sweep speed is related to the settings you use, just as they are for an analyzer.
Yes, but one can still be slower than another.
Which one wise guy ?

That's what we're trying to establish here, not just to sell Siglents.
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17067
  • Country: 00
Re: SDS800x HD vs DHO800x FFT
« Reply #29 on: July 16, 2024, 09:32:21 am »
various markers and peek markers and control of them, then SDS800xHD is the one.

The Rigol shows peaks, etc.
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7123
  • Country: hr
Re: SDS800x HD vs DHO800x FFT
« Reply #30 on: July 16, 2024, 12:12:31 pm »
various markers and peek markers and control of them, then SDS800xHD is the one.

The Rigol shows peaks, etc.

Don't start with this. You know well how little capabilities DHO800 has in that regard, compared to SDS800XHD.
Rigol only shows peaks. And that is it. No manual control of markers, harmonic markers, delta, etc. etc..
It does not even have Peak hold mode or average mode.
All it does is show jumpy waveform and jumpy peak markers on waveform and ever changing list in a table. That is it.
It is useful only for static signals, which is irony, because OP wants it for fast changing signals.
Fact is that using SD800xHD in peak detect mode will let you detect (and measure) where some fast frequency peaks, while Rigol will blink something on screen for a moment and if it was too fast to read you will lose it...


 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17067
  • Country: 00
Re: SDS800x HD vs DHO800x FFT
« Reply #31 on: July 16, 2024, 12:33:12 pm »
various markers and peek markers and control of them, then SDS800xHD is the one.

The Rigol shows peaks, etc.

Don't start with this.

I'm not starting anything, just correcting a factual error.
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7123
  • Country: hr
Re: SDS800x HD vs DHO800x FFT
« Reply #32 on: July 16, 2024, 01:53:14 pm »
various markers and peek markers and control of them, then SDS800xHD is the one.

The Rigol shows peaks, etc.

Don't start with this.

I'm not starting anything, just correcting a factual error.

No you are not. You are INSINUATING capabilities Rigol does not have.

I didn't say Rigol has no Peak markers at all. I said SDS800xHD has order of magnitude better capabilities in that regard..
I also explained additionally factually what it has.
Rigol shows realtime peaks. Full stop. There is no etc.
There is nothing else there as you would insinuate...

So yes, please don't start with this.
 

Offline rpro

  • Regular Contributor
  • *
  • Posts: 54
  • Country: us
Re: SDS800x HD vs DHO800x FFT
« Reply #33 on: July 16, 2024, 02:26:08 pm »
Secondly, the Rigol, like all Rigol models, lacks useful additional functions such as average and peak detect.
At least average should be "mandatory".
This would also be easy to solve if rigol were to attach importance to it.
I agree, I would guess it would probably take their SA software people a few hours (not even a day) to implement and test standard avg. and peak-hold in their DHO and MSO series.
On the positive side I have found their color heat map scheme useful to see/measure FFT peaks, etc., e.g., see attached. 
 
The following users thanked this post: 2N3055, Martin72

Offline cyo3fffTopic starter

  • Contributor
  • Posts: 11
  • Country: ro
Re: SDS800x HD vs DHO800x FFT
« Reply #34 on: July 16, 2024, 04:08:26 pm »
SDS814X HD FFT
A few assorted screenshots, all showing a 400 MHz span representative of you might use for EMI work.
Signal from SDG6022X, dual tone with the principle @100 MHz.

These results are only applicable to the scope settings in use. FFT 2Mpts remained unchanged although it's auto managed by the scope. Memory depth is just what it was, didn't check.
Higher FFT sweep speeds are likely possible if one takes the trouble to optimize them.

At 200us/div and max 2Mpts FFT a sweep speed of ~1/s is obtained.
100us/div = ~2/s
50us/div = ~4/s
As we increase only timebase the sweep gets faster where at 10us/div its speed is indiscernible and at 500ns/div detail is starting to be lost without changing other settings.

Max hold and Average screenshots thrown in for good measure.

All other settings should be plain to see in screenshots.


Dear Tautech,
Thank you very much to be the first to broke the ice with real facts!
This is the information I was expected from my post.

I will not answer to all of your messages but will try to summarize theme up to now. Please correct me if I'm wrong or add more if I miss something!
===========
Rigol DHO800x vs Siglent SDS800x HD

- FFT speed: Rigol is 4 times (average) faster than Siglent
- FFT accuracy: Siglent is better but slower than Rigol
- FFT features (Peak, Avg, Max Hold, customize parameters of the FFT such RBW, manual markers, Log scale, etc.): Siglent is better than Rigol
- FFT color heat map:Rigolis the only one with this feature (useful too)
- Balance between speed and accuracy: NONE of the scopes have such thing; you can chose only one of this features like speed=Rigol OR accuracy (and FFT better features)=Siglent (4 times slower)

Abot smell, Rigol have "annoying" smell, not Siglent.
=============

I hope this summary will be helpful also for the future reader in need to find an answer to those questions.
I will not conclude which one is better because it would be a false answer. Best is to say NONE of theme are good for this task or "it depends".

To be honest, only scope in less than 7000-8000 € range that has something resembling realtime SA is GW-Instek MDO-2000A series.
And that is because it literally has realtime SA mode implemented. It works well but it's ADC are only 8 bit so limited dynamic range. Also scopes are in 1500€ ranges, a bit old fashioned by now etc.....

All other scopes in that price range will have only FFT as a math function, with implementation variations.
Short version for the two scopes mentioned:

- if you want just something bouncing on the screen to see if there is something happening at some frequency, DHO800 will have faster refresh at some settings.
- if you want control of what are you measuring, accurate windows, accurate level measurements, Peak Hold and Averaging, various markers and peek markers and control of them, then SDS800xHD is the one.

If you want SA display on your scope that resembles R&S MXO4 or MXO5 or Tektronix MSO 6 digital down conversion tech, you will have to pay for those...


2N3055 (I'd liked that transistor anyways :)) thank you for this answer. I have no experience with GW-Instek scopes therefore I cannot judge your statement.
On the other hand (I'm an EMC engineer) I have experience with Keysight 20GB/s scopes used with FFT for EMC troubeshooting and they are good...in their price range.
One advantage of a scope vs RTSA is the dual more (time/frequency domain) and very wide real time bandwidth on the scope side (professional scopes I mean).
Not to mention the trigger functions that can correlate aggressor with observed emission, things that cannot be achievable with a simple SA.

A lats word about Rigol DHO804. The reason I'm stick to its FFT speed performance is that I've used it for the EMC tasks and it was enough for basic investigation but yes, in terms of markers and SA management is not very useful. The reason I've stopped to SDS824x HD is its better implementation of the FFT Math function and much closer look to a SA. Unfortunately its speed is not that good and I have no idea why they failed on that. I guess is due to the processing power available on the HW. Who knows.

Thank you very much to all of your for help.

Cristian
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7123
  • Country: hr
Re: SDS800x HD vs DHO800x FFT
« Reply #35 on: July 16, 2024, 05:47:18 pm »
About "heatmap", Siglent also (and many other) support color grading.. I have no idea why people think it is something "only Rigol has".

As for speed, Siglent didn't fail on it. Rigol is faster because it does a fraction of things, with much less points. Basically Rigol optimized for speed at the expense of everything else. Siglent chose to make full implementation.

I did "have a feeling" that you are talking about FFT (or spectrum mode) on higher end devices as a reference.
On Siglent SDS7000A FFT is also very fast.  >:D

Anyways, I don't know if Rigol have fixed wrong windowing, and if they did and if all other shortcomings are not important to you maybe for you Rigol will do..
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6634
  • Country: de
  • Testfield Technician
Re: SDS800x HD vs DHO800x FFT
« Reply #36 on: July 16, 2024, 09:34:44 pm »
To be honest, only scope in less than 7000-8000 € range that has something resembling realtime SA is GW-Instek MDO-2000A series.
And that is because it literally has realtime SA mode implemented. It works well but it's ADC are only 8 bit so limited dynamic range. Also scopes are in 1500€ ranges, a bit old fashioned by now etc.....

I had a look at it earlier, the built-in SA has a bandwidth of DC-500Mhz.
I find that interesting, other SAs only start at 9kHz (why actually, because of possible interference on the mains voltage?), much to my regret.
I find it “stupid” to add a (software) analyzer for the LF range.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29214
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDS800x HD vs DHO800x FFT
« Reply #37 on: July 16, 2024, 10:43:07 pm »
cyo3fff

As I have said numerous times and you as a EMC engineer should know only too well, sweep speed is directly related to settings in use.
Speed vs accuracy, pick your poison.

Before I need setup gear for other tasks, some closing screenshots.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: Martin72


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf