But it does look weird...

It looks typical of most FFT images on the web.

eg. https://www.gaussianwaves.com/2020/09/window-function-figure-of-merits/

As you say, it's probably related to number of bins/number of sample points in the FFT.
 

Offline Alex-lab

  • Contributor
  • Posts: 13
  • Country: gb
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #801 on: February 12, 2023, 12:14:14 pm »
Hi, Anybody knows, if that HiRes issue was fixed in available FW updates? 02.04? Thanks.
 

Offline switchabl

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: de
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #802 on: February 12, 2023, 01:09:45 pm »
The "flattop" screenshot definitely looks weird. At first glance it might look like "a" windowed FFT. But look closer:
- It is nothing like a flattop window. A flattop produces a wide main-lobe and the side-lobes shouldn't have a smooth envelope.
- Why do we see wide side-lobes at all? Has the number of points suddenly dropped? The RBW display hasn't changed but unfortunately that is bugged anyway (it seems to be off by a constant factor of 1000 though; that would suggest we are still at 1 Mpts, the maximum according to the datasheet).
- Look at the main-lobe again. It actually hasn't changed much at all, it is still very sharp. And so are the harmonics. So we must still be at full resolution.

Maybe there is just a typo in the formula for the window function.  :-// It is definitely possible to construct a window that behaves like this (for example take a "normal" window function and superimpose it with a low-amplitude, much narrower window), but you wouldn't really do that on purpose.
« Last Edit: February 12, 2023, 01:29:34 pm by switchabl »
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7108
  • Country: hr
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #803 on: February 12, 2023, 01:27:13 pm »
But it does look weird...

It looks typical of most FFT images on the web.

eg. https://www.gaussianwaves.com/2020/09/window-function-figure-of-merits/

As you say, it's probably related to number of bins/number of sample points in the FFT.

No it does not.

If you look at images on internet (wikipedia for instance), they demonstrate windowing with 11 or 13 sample points..
That is why you see something similar... With 1 million points FFT you don't see it.

And if I had to choose what it looks like it looks more like Parzen window with 11 sample points......

Because of all that i presume there is some "magic" happening with Rigol's FFT choice of how many bins it uses automatically...

Also one thing to note: They state RBW : 4.999m  ???????? What is that ???

FFT is definitely buggy and to me not finished. Few critical control variables user cannot control, and are not reported. 
« Last Edit: February 12, 2023, 01:28:58 pm by 2N3055 »
 

Online Njk

  • Frequent Contributor
  • **
  • Posts: 288
  • Country: ru
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #804 on: February 12, 2023, 01:28:41 pm »
Looks like FFT is still a toy function in any scope. It's not for measurements, it's just for reference. The values for SNR or dynamic range are not provided. The only related parameter in the spec is channel isolation, which is of 1:100. That's 40 dB, so anything on the FFT chart that is below that level (-60 dB, -100 dB, whatever) is just FYI. Accuracy is not officially specified. The most fierce debates are usually about toys  :)
« Last Edit: February 12, 2023, 01:36:13 pm by Njk »
 
The following users thanked this post: dreamcat4

Offline markone

  • Frequent Contributor
  • **
  • Posts: 730
  • Country: it
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #805 on: February 12, 2023, 02:14:34 pm »
Looks like FFT is still a toy function in any scope. It's not for measurements, it's just for reference. The values for SNR or dynamic range are not provided. The only related parameter in the spec is channel isolation, which is of 1:100. That's 40 dB, so anything on the FFT chart that is below that level (-60 dB, -100 dB, whatever) is just FYI. Accuracy is not officially specified. The most fierce debates are usually about toys  :)

That's the point, if you need reliable spectrum analysis you buy a full specified spectrum analyzer, a DSO is designed to be used mainly as YT instrument, for sure the FFT implementation is not the key point when you look for a new instrument.

What make me really laugh is that in this forum we can see dozens of DSO's FFT screen with 100dB (or even more) vertical dynamic range and infinite debates spent to demonstrate that they are actual  ::)

Anyhow, let's all keep in mind that the HDO1074 is available in EU for a little less than 1200 euro VAT included, it offers most of HDO4K features with a respectable 200MHz BW, 100Mpts , 2GSa/s.

This is the real bargain from this new line and there is nothing on the market that come close to it, i.e. the cheapest SDS2000X-HD cost 3 times more.
 

Offline switchabl

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: de
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #806 on: February 12, 2023, 02:59:25 pm »
It looks to me like the FFT mode doesn't do real decimation after all (at least in the screenshots). The noise graphs published earlier in this thread seem to indicate that the Rigol has a bit less noise in the kHz region anyway. And it looks like it is running 1 Mpts FFT vs 512 kpts, so that is another 3 dB of processing gain. If there was real 500 MSa/s -> 5 MSa/s decimation (and not just subsampling) that should give another 20 dB and I'm not seeing that.

So I think it is still not quite clear what is happening with respect to resampling. Does the noise floor decrease if you switch acquisition mode to Hi-Res?

Btw I disagree that spectrum analysis does not matter much on an oscilloscope. The main point of having 12 bits is that you can analyze small details on large signals that you often can't see in a time-domain display. Spectrum analysis is a major part of the toolbox. I do agree that specifications lack a lot of detail compared to high-resolution options from Lecroy or R&S. But those are exactly the gaps that people are trying to fill here.
« Last Edit: February 12, 2023, 03:35:48 pm by switchabl »
 

Offline gf

  • Super Contributor
  • ***
  • Posts: 1312
  • Country: de
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #807 on: February 12, 2023, 03:20:50 pm »
And it looks like it is running 1 Mpts FFT

Indeed, basically looks like 1MPts (~5Hz bin spacing), OTOH the displayed RBW=5mHz @5MSa/sa would imply 1Gpts, which can't be true.

However, the "flattop" trace rather suggests a window size of only ~50k points (~100Hz bin spacing), zero-padded to a larger FFT size, leading to interpolation in the frequency domain then (which makes the shape of the side lobes visible). And as you already said, the main lobe is way to narrow for a flattop window. Any common window function has a main lobe width >= 2x the width of side lobes. Something is weird here.
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7108
  • Country: hr
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #808 on: February 12, 2023, 03:23:42 pm »
Looks like FFT is still a toy function in any scope. It's not for measurements, it's just for reference. The values for SNR or dynamic range are not provided. The only related parameter in the spec is channel isolation, which is of 1:100. That's 40 dB, so anything on the FFT chart that is below that level (-60 dB, -100 dB, whatever) is just FYI. Accuracy is not officially specified. The most fierce debates are usually about toys  :)

I think that is a bit simplified look at things..

SDS2000X HD: CH to CH
Isolation (@50Ω)
> 60 dBc, < 500 MHz
> 70 dBc, < 350 MHz

Also CH2CH isolation is relevant only if you use several channels at the same time.. With single channel not so much..
At that time only SFDR is relevant......

So it is very useful but need to be realistic what it is and what is it not...
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7108
  • Country: hr
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #809 on: February 12, 2023, 03:42:40 pm »

What make me really laugh is that in this forum we can see dozens of DSO's FFT screen with 100dB (or even more) vertical dynamic range and infinite debates spent to demonstrate that they are actual  ::)


A 12 bit scope cannot have 120 dB dynamic range.. But it can have -120dBm noise floor, and it can have larger than ADC dynamic range in FFT because of software gain.
Sometimes large dynamic range is not needed, good sensitivity is enough...

GPS signals are in the noise of preamp. It is literally not visible in time domain...
Using clever software techniques they extract useful signal from the noise...

Math don't lie. But it can be misapplied and misunderstood.
And I agree that it is better to be cautious and have a healthy dose of scepticism than believe in fairy tales..
But sometimes we can squeeze a bit more performance with a bit of cleverness. 
 
The following users thanked this post: rf-loop, egonotto, Bad_Driver

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #810 on: February 12, 2023, 04:59:34 pm »
Strange things gentlemen...
Playing with the FFT, same Setup as before.
Setting Memory to 100 Mpoints, hanning window.
Playing with the timebase..
From say 500ms to 50ms the memory remains the same.
Then switching to 20ms - the memory drops down to 10Mpoints.
Switching back to 50ms, the memory remains 10 and the FFT is freaking out....
You can go in the menu and set the memory to 100 again, FFT becomes normal again.
Will upload a video in a few minutes.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: egonotto

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: thm_w, Detlev

Offline switchabl

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: de
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #812 on: February 12, 2023, 05:50:49 pm »
Looks like the FFT input is being truncated (without proper windowing). Hard to tell what exactly is going on. Some buffer size mismatch maybe?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4132
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #813 on: February 12, 2023, 05:54:38 pm »
Here is small example (dynamic). (SDS2504X HD)

And note, For this particular purpose these FFT etc settings are not at all best optimal ones.

Signals: (note that these said levels are not accurate, just entered set values using SDG1032X (but I know these levels do not have high error)

If everything is ideal and accurate these are levels.

(marker 1) 5000Hz, +22dBm
(marker 2) 5100Hz -48dBm (carrier level, AM modulation with 20Hz, and modulation depth 20%). 70dB below marker 1
(marker 3) 5120Hz -68dBm (AM USB -20dBc because 20% mod depth) 90dB below marker 1

FFT mode is just Normal and acquisition just normal, no any averaging or other things. Just normal acquisition.
FFT window Flattop.

As can see displayed levels are not bad when also think what is signal source and what are all specifications. 

« Last Edit: February 12, 2023, 05:57:48 pm by rf-loop »
BEV of course. Cars with smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the (strong)wises gone?
 
The following users thanked this post: Martin72

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #814 on: February 12, 2023, 05:59:16 pm »
Hi,

the flattop thing from yesterday...
Here a video, look at the FFT on various timebases, especially how the form of the fundamental wave looks like.

https://youtu.be/83K0Y8Bov2k

Weird: With the flattop window memorypoints won´t change...
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: egonotto

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #815 on: February 12, 2023, 07:08:07 pm »
The rest of the FFT test...
Playing from 500ms timebase till 5ms, watch the info on the window (RBW, samplerate), I´ve changed nothing else except the timebase.
I can not guarantee the correct order of the pics as the forum software somehow shuffled them since a couple of months. ::)
More FFT test I don´t make, imho it´s too buggy.
Next play is connecting the bodnar pulser. ;)
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: egonotto

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #816 on: February 12, 2023, 08:31:29 pm »
Last for today, the bodnar pulser.
I´ve been looking for switching the display mode (dot/vectors), to my surprise you can´t change the mode, it´s always vector.
(BTW, to the owners: I haven´t found the possibility to save all the screen content to the stick, when doing a screenshot, all open menus won´t be captured, so I must use a camera)
Then connecting the pulser, measure the rise time: 1.6ns sounds plausible to me, while the signal looks a little bit "filtered".
Following an idea, I´ve activated the second channel and to my surprise, the pulse on ch1 is getting "faster" and overshoots came up.
(1.2ns when 2 channels are active).
When all channels are active, the overshoot of the pulse gets "blurried" while the risetimes decreases to 1.5ns .
Hm...

« Last Edit: February 12, 2023, 08:33:02 pm by Martin72 »
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: egonotto

Online tautech

  • Super Contributor
  • ***
  • Posts: 29173
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #817 on: February 12, 2023, 08:46:12 pm »
Last for today, the bodnar pulser.
I´ve been looking for switching the display mode (dot/vectors), to my surprise you can´t change the mode, it´s always vector.
Purposely done to foil rf-loop's investigations of false dots.  ;)
Quote
(BTW, to the owners: I haven´t found the possibility to save all the screen content to the stick, when doing a screenshot, all open menus won´t be captured, so I must use a camera)
Inspect the Save/Recall menu for the switch to show or hide menus in screenshots.
Do you forget your Siglent has that ?

Quote
...............
Hm...
Indeed.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #818 on: February 12, 2023, 08:56:12 pm »
Quote
Inspect the Save/Recall menu for the switch to show or hide menus in screenshots.
Do you forget your Siglent has that ?

No, why should I...This was the first section I´ve looked for. ;)

Edit:
BTW, the  \$\Omega\$ symbol in the channelbox stands for 50 Ohm input... :P
« Last Edit: February 12, 2023, 09:00:47 pm by Martin72 »
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #819 on: February 12, 2023, 09:09:52 pm »
Quote from: tautech
Purposely done to foil rf-loop's investigations of false dots.

Snip of the manual:

Quote
19.1 Display Type
This series oscilloscope provides the "Vector" display mode in which the sample
points are connected by lines and displayed. In most cases, this mode can provide the
most vivid waveform for you to view the steep edge of the waveform (such as square
waveform).

LOL....

And the manual telling me another thing, about the screenshot...
There is no possibility to save the whole screen content.... :palm:
« Last Edit: February 12, 2023, 09:16:11 pm by Martin72 »
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline skander36

  • Frequent Contributor
  • **
  • Posts: 827
  • Country: ro
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #820 on: February 12, 2023, 09:17:15 pm »

(BTW, to the owners: I haven´t found the possibility to save all the screen content to the stick, when doing a screenshot, all open menus won´t be captured, so I must use a camera)


On the web interface you have menu "Print screen" . There you have both options Screenshot and Screen Capture.
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #821 on: February 12, 2023, 09:21:13 pm »
Thank you but it should be also possible without web control..Something for the wishlist.
Not for me,but for the current and future owners.


"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline gf

  • Super Contributor
  • ***
  • Posts: 1312
  • Country: de
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #822 on: February 12, 2023, 09:26:32 pm »
Following an idea, I´ve activated the second channel and to my surprise, the pulse on ch1 is getting "faster" and overshoots came up.
(1.2ns when 2 channels are active).
When all channels are active, the overshoot of the pulse gets "blurried" while the risetimes decreases to 1.5ns .
Hm...

Sinc interpolation is supposed to introduce pre-shoot and overshoot, when applied to an undersampled edge. I guess the frontend is too fast to prevent aliasing at 1GSa/s and 2GSa/s. Is it possible to turn off sinc interpolation?
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6616
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #823 on: February 12, 2023, 09:45:49 pm »
Just powered on again*, there is no possibility to do it.

*(When plug in the scope to mains, it will switch on for a second (fan starts running, display lit)before it reaches the standby condition)
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline JeremyC

  • Regular Contributor
  • *
  • Posts: 148
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #824 on: February 12, 2023, 10:07:07 pm »

Sinc interpolation is supposed to introduce pre-shoot and overshoot, when applied to an undersampled edge. I guess the frontend is too fast to prevent aliasing at 1GSa/s and 2GSa/s. Is it possible to turn off sinc interpolation?

No, you can't turn off sin(x)/x interpolation in these scopes, it has been confirmed by Rigol.
Check reply #180 https://www.eevblog.com/forum/testgear/rigol-hdo1000-and-hdo4000-12bit-oscilloscopes-launched-in-china/msg4426576/#msg4426576
 
The following users thanked this post: gf, Martin72