Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #175 on: September 20, 2022, 08:42:32 pm »
Unboxing

Thank you for this first impression.
At last a rigol with low noise input... ;)
Optically I don´t like it, but the rest seems to be nice.
And the price is very hot, no question.

Quote
They must have changed it on the HD model then, because that wasn't the case for previous Siglents I've looked at

The 6000A and the 2000HD got it from the beginning, on the SDS5000 it´s "new" with firmware update.

"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline balnazzar

  • Frequent Contributor
  • **
  • Posts: 417
  • Country: it
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #176 on: September 20, 2022, 09:06:41 pm »
Why can't they just copycat the UI from R&S or Keysight, I don't know..

It seems they greatly value the somewhat videogame-ish appearance.

 
The following users thanked this post: Martin72

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6407
  • Country: de
  • Testfield Technician
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #177 on: September 20, 2022, 09:20:06 pm »
When you´re growing up with a (android) smartphone in your hand, this could be the result when you have to design the UI.
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: balnazzar

Online Someone

  • Super Contributor
  • ***
  • Posts: 4696
  • Country: au
    • send complaints here
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #178 on: September 20, 2022, 10:58:10 pm »
Why can't they just copycat the UI from R&S or Keysight, I don't know..
Some people dont like those UI's! Scopes are products that still have not settled on the "right" way to do a touchscreen.
 
The following users thanked this post: balnazzar

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #179 on: September 20, 2022, 11:41:19 pm »
Rigol have confirmed that you cannot turn off the Sin x/x interpolation.
 
The following users thanked this post: egonotto, thm_w, dreamcat4

Offline egonotto

  • Frequent Contributor
  • **
  • Posts: 863
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #180 on: September 20, 2022, 11:57:14 pm »
Hello,

Maybe we can make a noise comparison between Rigol and Siglent 12 bit scope. For this purpose, 8 files (binary) per device would be useful.

Each with 50 Ohm and 1 MOhm

1 mV/div with 200MHz and 20 MHz with the largest possible memory and highest sampling rate.

And

1 V/div with 200MHz and 20 MHz with the largest possible memory and highest sampling rate.

The files would probably be quite large. But there are file hosters with free access and hopefully enough storage space.

Best regards
egonotto




 
The following users thanked this post: EEVblog

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #181 on: September 21, 2022, 12:19:55 am »
5:50 in the video I totally missed that bias is actually different to position offset. You can actually DC bias the input to any level required which is pretty cool.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #182 on: September 21, 2022, 12:27:58 am »
Maybe we can make a noise comparison between Rigol and Siglent 12 bit scope. For this purpose, 8 files (binary) per device would be useful.
Each with 50 Ohm and 1 MOhm

1 mV/div with 200MHz and 20 MHz with the largest possible memory and highest sampling rate.

And

1 V/div with 200MHz and 20 MHz with the largest possible memory and highest sampling rate.

The files would probably be quite large. But there are file hosters with free access and hopefully enough storage space.

Getting those now.
Maybe 1M memory is better as a standard?
 
The following users thanked this post: egonotto

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #183 on: September 21, 2022, 12:42:29 am »
Noise data for 1M memory depth, 1mV/1V, 20M/200M, and 50ohm/1Mopen all at 1ms/DIV, all at 50MS/s
https://eevblog.com/files/RigolHDO4000data.zip
« Last Edit: September 21, 2022, 04:31:14 am by EEVblog »
 
The following users thanked this post: egonotto, jord4231, adonishong

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #184 on: September 21, 2022, 02:03:59 am »
Noise data for 1M memory depth, 1mV/1V, 20M/200M, and 50ohm/1Mopen all at 1ms/DIV
https://eevblog.com/files/RigolHDO4000data.zip

And 250M memory data for 50ohm input at 4GS/s 1ms timebase
https://eevblog.org/files/RigolHDO4000-250M-data.zip
« Last Edit: September 21, 2022, 04:31:35 am by EEVblog »
 
The following users thanked this post: egonotto, adonishong

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #185 on: September 21, 2022, 03:17:58 am »
Thanks! Can you confirm the LSB scaling? For RigolDS1-50ohm1mV20M.bin I am getting STD=19.8uV assuming full scale of 10 mV in 2^16 bits.

Also need to know sampling rate to calculate PSD.
« Last Edit: September 21, 2022, 03:24:24 am by maxwell3e10 »
 
The following users thanked this post: egonotto

Offline egonotto

  • Frequent Contributor
  • **
  • Posts: 863
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #186 on: September 21, 2022, 03:52:50 am »

....

Getting those now.
Maybe 1M memory is better as a standard?

With max memory an max sample rate than with Rigol one get 62.5 ms, With Siglent one get 100 ms.
This allows you to observe the noise over a large frequency range.
And you can try hires algorithms.

Best regards
egonotto


 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #187 on: September 21, 2022, 04:14:02 am »
Thanks! Can you confirm the LSB scaling? For RigolDS1-50ohm1mV20M.bin I am getting STD=19.8uV assuming full scale of 10 mV in 2^16 bits.
Also need to know sampling rate to calculate PSD.

4GS/s for all files.
For actual LSB I don't know if there is a way to actually get this exact figure?
As I showed in my video there is some extra dynamic range outside of the displayed window. So it's at least 8 divisions + whatever extra there is. Maybe 10V full scale?
EDIT: Hang on, I think I know how to get this.
 
The following users thanked this post: egonotto

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #188 on: September 21, 2022, 04:21:32 am »
For 1 M Sample files it can't be 4GS/sec and 1 msec/div.
Actually checking your video, it seems sometime the horizontal scale is 10 divisions and sometimes it is 20 divisions.

But the spectra are looking nice... Just want to confirm scaling.
 
The following users thanked this post: egonotto

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #189 on: September 21, 2022, 04:27:20 am »
Thanks! Can you confirm the LSB scaling? For RigolDS1-50ohm1mV20M.bin I am getting STD=19.8uV assuming full scale of 10 mV in 2^16 bits.
Also need to know sampling rate to calculate PSD.

4GS/s for all files.
For actual LSB I don't know if there is a way to actually get this exact figure?
As I showed in my video there is some extra dynamic range outside of the displayed window. So it's at least 8 divisions + whatever extra there is. Maybe 10V full scale?
EDIT: Hang on, I think I know how to get this.

I have confirmed on the 100mV range that the dynamic signal acquisition range is about 960mVpp (assuming my Keysight 3000 sig gen is accurate.
I did this by feeding in a triange wave, capturing, and seeing where it clips. 970mVpp definitely clipped.
I assume the same on the 1V and 1mV ranges, so call it 9.60Vpp/4096 = 2.34mV/bit
 
The following users thanked this post: egonotto

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #190 on: September 21, 2022, 04:30:13 am »
For 1 M Sample files it can't be 4GS/sec and 1 msec/div.

Oops, it's 50MS/s on the 1M memory captures. 4GS/s on the 250M samples.
 
The following users thanked this post: egonotto

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #191 on: September 21, 2022, 04:39:04 am »
I have confirmed on the 100mV range that the dynamic signal acquisition range is about 960mVpp (assuming my Keysight 3000 sig gen is accurate.
I did this by feeding in a triange wave, capturing, and seeing where it clips. 970mVpp definitely clipped.
I assume the same on the 1V and 1mV ranges, so call it 9.60Vpp/4096 = 2.34mV/bit
It could be clipping before reaching full ADC range. Also the binary numbers are encoded in 16 bit unsigned integer. There should be some meta file saved  along with binary data that would specify it. But 10 divisions full scale is probably fairly accurate.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #192 on: September 21, 2022, 04:43:36 am »
I have confirmed on the 100mV range that the dynamic signal acquisition range is about 960mVpp (assuming my Keysight 3000 sig gen is accurate.
I did this by feeding in a triange wave, capturing, and seeing where it clips. 970mVpp definitely clipped.
I assume the same on the 1V and 1mV ranges, so call it 9.60Vpp/4096 = 2.34mV/bit
It could be clipping before reaching full ADC range. Also the binary numbers are encoded in 16 bit unsigned integer. There should be some meta file saved  along with binary data that would specify it. But 10 divisions full scale is probably fairly accurate.

There is no extra file saved.
Yes it could be clipping before the ADC, but there is no easy way to know this?
I confirmed the 200mV range is the same FS ratio.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #193 on: September 21, 2022, 04:51:41 am »
I have confirmed on the 100mV range that the dynamic signal acquisition range is about 960mVpp (assuming my Keysight 3000 sig gen is accurate.
I did this by feeding in a triange wave, capturing, and seeing where it clips. 970mVpp definitely clipped.
I assume the same on the 1V and 1mV ranges, so call it 9.60Vpp/4096 = 2.34mV/bit
It could be clipping before reaching full ADC range. Also the binary numbers are encoded in 16 bit unsigned integer. There should be some meta file saved  along with binary data that would specify it. But 10 divisions full scale is probably fairly accurate.

The 1V range is different, 8.75Vpp. At 8.8V I can see the clipping.
 

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #194 on: September 21, 2022, 05:26:16 am »
Here is the noise spectrum using 250Msample files and assuming 10 div full scale. Looks very clean up to 200 MHz!
On 1 V/div (10V full scale) the white noise level is 240 nV/Hz^(1/2), I think that corresponds to almost 10 ENOB. If I actually use proper equations instead of just "thinking", it comes out to 8.3 ENOB.
« Last Edit: September 21, 2022, 06:23:49 pm by maxwell3e10 »
 
The following users thanked this post: EEVblog, egonotto, thm_w

Offline egonotto

  • Frequent Contributor
  • **
  • Posts: 863
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #195 on: September 21, 2022, 05:50:10 am »
Hello maxwell3e10,

How do you make your graphics?

Best regards
egonotto

 

Offline hhappy1

  • Contributor
  • Posts: 42
  • Country: kr
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #196 on: September 21, 2022, 06:19:29 am »
The noise looks very low.
1mv, 50ohm, 14bit mode.
1M poit fft, 0 ~ 2Mhz span, 100Msa/s.
If you show me the fft noise screen, I appreciate it.

There's no fft average view, right?
« Last Edit: September 21, 2022, 06:22:51 am by hhappy1 »
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #197 on: September 21, 2022, 06:25:00 am »
Quick video on getting the dynamic range:
 
The following users thanked this post: egonotto

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38123
  • Country: au
    • EEVblog
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #198 on: September 21, 2022, 06:53:23 am »
It looks like it has a different amplifier path for 50 Ohm, giving lower noise at low frequency.
The ffts and plots are made with Matlab.
EDIT:
Actually, to be sure, was the scope input grounded externally when on 1 MOhm setting, Dave? Because it gives std=48 uV, instead of 19 uV. I think you mentioned in the video it does not make a difference when using on-screen rms measurements.

1M tests used an open input. If I used an external 50ohm termination then that's effectively the same as the 50 ohm mode, unless there is actually different paths for the 50ohm and 1M input which is usually not the case for scopes.
So that graph you just posted in correct. I thought you wanted an open input?
 

Offline maxwell3e10

  • Frequent Contributor
  • **
  • Posts: 870
  • Country: us
Re: Rigol HDO1000 and HDO4000 12bit oscilloscopes launched in China
« Reply #199 on: September 21, 2022, 07:07:55 am »
Some scopes do have a different amplifier path, for example RTM3k and other higher-end scopes. One can tell if there is any difference in spectrum between 50 Ohm external short and internal 50 Ohm. It looks like there is not in this case, at least based on RMS measurements.

I am not sure how interesting is the spectrum with 1 MOhm open input. It might get better if one puts a shielding BNC cup (with no central pin) over the input.
 