The voltages in the following table where checked. These voltages are generate using offsets n x 40 mV and wave voltages from -20 mV to 20 mV.
All voltages below 660 mV had +/- 1 digit accuracy, thus maximal 0.1 mV off. I did not enter the true MM values, because that would take a lot of time, and going better than 1 digit off would not be worth the efford.
Voltages above 660 mV where measured in a different range, but I'm convinced that they'll be as accurate as well, because it seems that the wave and offset errors are independent of each other. So the error correction will continue in the same manner. (I could measure this with an channel 2 offset as well, but an accuracy that is in par with my Fluke is OK to me)
the EXP is again the voltage that "should" be outputted by the AWG using the requested parameters. Although the standard accuracy seems OK most of the time, the worse deviations are improved by more than 10x.
Voltage TARGET: 0,0 mV; EXP: -0,3 mV; MM: 0,0 mV
Voltage TARGET: 5,0 mV; EXP: 4,6 mV; MM: 5,0 mV
Voltage TARGET: 10,0 mV; EXP: 9,6 mV; MM: 10,0 mV
Voltage TARGET: 15,0 mV; EXP: 14,6 mV; MM: 15,0 mV
Voltage TARGET: 20,0 mV; EXP: 20,5 mV; MM: 20,0 mV
Voltage TARGET: 25,0 mV; EXP: 25,5 mV; MM: 25,0 mV
Voltage TARGET: 30,0 mV; EXP: 30,4 mV; MM: 30,0 mV
Voltage TARGET: 35,0 mV; EXP: 35,4 mV; MM: 35,0 mV
Voltage TARGET: 40,0 mV; EXP: 40,4 mV; MM: 40,0 mV
Voltage TARGET: 45,0 mV; EXP: 45,3 mV; MM: 45,0 mV
Voltage TARGET: 50,0 mV; EXP: 50,3 mV; MM: 50,0 mV
Voltage TARGET: 55,0 mV; EXP: 55,3 mV; MM: 55,0 mV
Voltage TARGET: 60,0 mV; EXP: 60,0 mV; MM: 60,0 mV
Voltage TARGET: 65,0 mV; EXP: 65,0 mV; MM: 65,0 mV
Voltage TARGET: 70,0 mV; EXP: 69,9 mV; MM: 70,0 mV
Voltage TARGET: 75,0 mV; EXP: 74,9 mV; MM: 75,0 mV
Voltage TARGET: 80,0 mV; EXP: 79,9 mV; MM: 80,0 mV
Voltage TARGET: 85,0 mV; EXP: 84,8 mV; MM: 85,0 mV
Voltage TARGET: 90,0 mV; EXP: 89,8 mV; MM: 90,0 mV
Voltage TARGET: 95,0 mV; EXP: 94,8 mV; MM: 95,0 mV
Voltage TARGET: 100,0 mV; EXP: 100,5 mV; MM: 100,0 mV
Voltage TARGET: 105,0 mV; EXP: 105,5 mV; MM: 105,0 mV
Voltage TARGET: 110,0 mV; EXP: 110,4 mV; MM: 110,0 mV
Voltage TARGET: 115,0 mV; EXP: 115,4 mV; MM: 115,0 mV
Voltage TARGET: 120,0 mV; EXP: 120,4 mV; MM: 120,0 mV
Voltage TARGET: 125,0 mV; EXP: 125,3 mV; MM: 125,0 mV
Voltage TARGET: 130,0 mV; EXP: 130,3 mV; MM: 130,0 mV
Voltage TARGET: 135,0 mV; EXP: 135,3 mV; MM: 135,0 mV
Voltage TARGET: 140,0 mV; EXP: 140,5 mV; MM: 140,0 mV
Voltage TARGET: 145,0 mV; EXP: 145,5 mV; MM: 145,0 mV
Voltage TARGET: 150,0 mV; EXP: 150,4 mV; MM: 150,0 mV
Voltage TARGET: 155,0 mV; EXP: 155,4 mV; MM: 155,0 mV
Voltage TARGET: 160,0 mV; EXP: 160,4 mV; MM: 160,0 mV
Voltage TARGET: 165,0 mV; EXP: 165,3 mV; MM: 165,0 mV
Voltage TARGET: 170,0 mV; EXP: 170,3 mV; MM: 170,0 mV
Voltage TARGET: 175,0 mV; EXP: 175,3 mV; MM: 175,0 mV
Voltage TARGET: 180,0 mV; EXP: 179,9 mV; MM: 180,0 mV
Voltage TARGET: 185,0 mV; EXP: 184,9 mV; MM: 185,0 mV
Voltage TARGET: 190,0 mV; EXP: 189,8 mV; MM: 190,0 mV
Voltage TARGET: 195,0 mV; EXP: 194,8 mV; MM: 195,0 mV
Voltage TARGET: 200,0 mV; EXP: 199,8 mV; MM: 200,0 mV
Voltage TARGET: 205,0 mV; EXP: 204,7 mV; MM: 205,0 mV
Voltage TARGET: 210,0 mV; EXP: 209,7 mV; MM: 210,0 mV
Voltage TARGET: 215,0 mV; EXP: 214,7 mV; MM: 215,0 mV
Voltage TARGET: 220,0 mV; EXP: 219,8 mV; MM: 220,0 mV
Voltage TARGET: 225,0 mV; EXP: 224,8 mV; MM: 225,0 mV
Voltage TARGET: 230,0 mV; EXP: 229,7 mV; MM: 230,0 mV
Voltage TARGET: 235,0 mV; EXP: 234,7 mV; MM: 235,0 mV
Voltage TARGET: 240,0 mV; EXP: 239,7 mV; MM: 240,0 mV
Voltage TARGET: 245,0 mV; EXP: 244,6 mV; MM: 245,0 mV
Voltage TARGET: 250,0 mV; EXP: 249,6 mV; MM: 250,0 mV
Voltage TARGET: 255,0 mV; EXP: 254,6 mV; MM: 255,0 mV
Voltage TARGET: 260,0 mV; EXP: 260,1 mV; MM: 260,0 mV
Voltage TARGET: 265,0 mV; EXP: 265,1 mV; MM: 265,0 mV
Voltage TARGET: 270,0 mV; EXP: 270,0 mV; MM: 270,0 mV
Voltage TARGET: 275,0 mV; EXP: 275,0 mV; MM: 275,0 mV
Voltage TARGET: 280,0 mV; EXP: 280,0 mV; MM: 280,0 mV
Voltage TARGET: 285,0 mV; EXP: 284,9 mV; MM: 285,0 mV
Voltage TARGET: 290,0 mV; EXP: 289,9 mV; MM: 290,0 mV
Voltage TARGET: 295,0 mV; EXP: 294,9 mV; MM: 295,0 mV
Voltage TARGET: 300,0 mV; EXP: 300,3 mV; MM: 300,0 mV
Voltage TARGET: 305,0 mV; EXP: 305,3 mV; MM: 305,0 mV
Voltage TARGET: 310,0 mV; EXP: 310,2 mV; MM: 310,0 mV
Voltage TARGET: 315,0 mV; EXP: 315,2 mV; MM: 315,0 mV
Voltage TARGET: 320,0 mV; EXP: 320,2 mV; MM: 320,0 mV
Voltage TARGET: 325,0 mV; EXP: 325,1 mV; MM: 325,0 mV
Voltage TARGET: 330,0 mV; EXP: 330,1 mV; MM: 330,0 mV
Voltage TARGET: 335,0 mV; EXP: 335,1 mV; MM: 335,0 mV
Voltage TARGET: 340,0 mV; EXP: 340,5 mV; MM: 340,0 mV
Voltage TARGET: 345,0 mV; EXP: 345,5 mV; MM: 345,0 mV
Voltage TARGET: 350,0 mV; EXP: 350,4 mV; MM: 350,0 mV
Voltage TARGET: 355,0 mV; EXP: 355,4 mV; MM: 355,0 mV
Voltage TARGET: 360,0 mV; EXP: 360,4 mV; MM: 360,0 mV
Voltage TARGET: 365,0 mV; EXP: 365,3 mV; MM: 365,0 mV
Voltage TARGET: 370,0 mV; EXP: 370,3 mV; MM: 370,0 mV
Voltage TARGET: 375,0 mV; EXP: 375,3 mV; MM: 375,0 mV
Voltage TARGET: 380,0 mV; EXP: 380,6 mV; MM: 380,0 mV
Voltage TARGET: 385,0 mV; EXP: 385,6 mV; MM: 385,0 mV
Voltage TARGET: 390,0 mV; EXP: 390,5 mV; MM: 390,0 mV
Voltage TARGET: 395,0 mV; EXP: 395,5 mV; MM: 395,0 mV
Voltage TARGET: 400,0 mV; EXP: 400,5 mV; MM: 400,0 mV
Voltage TARGET: 405,0 mV; EXP: 405,4 mV; MM: 405,0 mV
Voltage TARGET: 410,0 mV; EXP: 410,4 mV; MM: 410,0 mV
Voltage TARGET: 415,0 mV; EXP: 415,4 mV; MM: 415,0 mV
Voltage TARGET: 420,0 mV; EXP: 420,9 mV; MM: 420,0 mV
Voltage TARGET: 425,0 mV; EXP: 425,9 mV; MM: 425,0 mV
Voltage TARGET: 430,0 mV; EXP: 430,8 mV; MM: 430,0 mV
Voltage TARGET: 435,0 mV; EXP: 435,8 mV; MM: 435,0 mV
Voltage TARGET: 440,0 mV; EXP: 440,8 mV; MM: 440,0 mV
Voltage TARGET: 445,0 mV; EXP: 445,7 mV; MM: 445,0 mV
Voltage TARGET: 450,0 mV; EXP: 450,7 mV; MM: 450,0 mV
Voltage TARGET: 455,0 mV; EXP: 455,7 mV; MM: 455,0 mV
Voltage TARGET: 460,0 mV; EXP: 461,1 mV; MM: 460,0 mV
Voltage TARGET: 465,0 mV; EXP: 466,1 mV; MM: 465,0 mV
Voltage TARGET: 470,0 mV; EXP: 471,0 mV; MM: 470,0 mV
Voltage TARGET: 475,0 mV; EXP: 476,0 mV; MM: 475,0 mV
Voltage TARGET: 480,0 mV; EXP: 481,0 mV; MM: 480,0 mV
Voltage TARGET: 485,0 mV; EXP: 485,9 mV; MM: 485,0 mV
Voltage TARGET: 490,0 mV; EXP: 490,9 mV; MM: 490,0 mV
Voltage TARGET: 495,0 mV; EXP: 495,9 mV; MM: 495,0 mV
Voltage TARGET: 500,0 mV; EXP: 501,3 mV; MM: 500,0 mV
Voltage TARGET: 505,0 mV; EXP: 506,3 mV; MM: 505,0 mV
Voltage TARGET: 510,0 mV; EXP: 511,2 mV; MM: 510,0 mV
Voltage TARGET: 515,0 mV; EXP: 516,2 mV; MM: 515,0 mV
Voltage TARGET: 520,0 mV; EXP: 521,2 mV; MM: 520,0 mV
Voltage TARGET: 525,0 mV; EXP: 526,1 mV; MM: 525,0 mV
Voltage TARGET: 530,0 mV; EXP: 531,1 mV; MM: 530,0 mV
Voltage TARGET: 535,0 mV; EXP: 536,1 mV; MM: 535,0 mV
Voltage TARGET: 540,0 mV; EXP: 541,9 mV; MM: 540,0 mV
Voltage TARGET: 545,0 mV; EXP: 546,9 mV; MM: 545,0 mV
Voltage TARGET: 550,0 mV; EXP: 551,8 mV; MM: 550,0 mV
Voltage TARGET: 555,0 mV; EXP: 556,8 mV; MM: 555,0 mV
Voltage TARGET: 560,0 mV; EXP: 561,8 mV; MM: 560,0 mV
Voltage TARGET: 565,0 mV; EXP: 566,7 mV; MM: 565,0 mV
Voltage TARGET: 570,0 mV; EXP: 571,7 mV; MM: 570,0 mV
Voltage TARGET: 575,0 mV; EXP: 576,7 mV; MM: 575,0 mV
Voltage TARGET: 580,0 mV; EXP: 582,2 mV; MM: 580,0 mV
Voltage TARGET: 585,0 mV; EXP: 587,2 mV; MM: 585,0 mV
Voltage TARGET: 590,0 mV; EXP: 592,1 mV; MM: 590,0 mV
Voltage TARGET: 595,0 mV; EXP: 597,1 mV; MM: 595,0 mV
Voltage TARGET: 600,0 mV; EXP: 602,1 mV; MM: 600,0 mV
Voltage TARGET: 605,0 mV; EXP: 607,0 mV; MM: 605,0 mV
Voltage TARGET: 610,0 mV; EXP: 612,0 mV; MM: 610,0 mV
Voltage TARGET: 615,0 mV; EXP: 617,0 mV; MM: 615,0 mV
Voltage TARGET: 620,0 mV; EXP: 622,0 mV; MM: 620,0 mV
Voltage TARGET: 625,0 mV; EXP: 627,0 mV; MM: 625,0 mV
Voltage TARGET: 630,0 mV; EXP: 631,9 mV; MM: 630,0 mV
Voltage TARGET: 635,0 mV; EXP: 636,9 mV; MM: 635,0 mV
Voltage TARGET: 640,0 mV; EXP: 641,9 mV; MM: 640,0 mV
Voltage TARGET: 645,0 mV; EXP: 646,8 mV; MM: 645,0 mV
Voltage TARGET: 650,0 mV; EXP: 651,8 mV; MM: 650,0 mV
Voltage TARGET: 655,0 mV; EXP: 656,8 mV; MM: 655,0 mV
Voltage TARGET: 660,0 mV; EXP: 662,0 mV; MM: 660,0 mV
Voltage TARGET: 665,0 mV; EXP: 667,0 mV; MM: 665,0 mV
Voltage TARGET: 670,0 mV; EXP: 671,9 mV; MM: 670,0 mV
Voltage TARGET: 675,0 mV; EXP: 676,9 mV; MM: 675,0 mV
Voltage TARGET: 680,0 mV; EXP: 681,9 mV; MM: 680,0 mV
Voltage TARGET: 685,0 mV; EXP: 686,8 mV; MM: 685,0 mV
Voltage TARGET: 690,0 mV; EXP: 691,8 mV; MM: 690,0 mV
Voltage TARGET: 695,0 mV; EXP: 696,8 mV; MM: 695,0 mV
Voltage TARGET: 700,0 mV; EXP: 702,4 mV; MM: 700,0 mV
Voltage TARGET: 705,0 mV; EXP: 707,4 mV; MM: 705,0 mV
Voltage TARGET: 710,0 mV; EXP: 712,3 mV; MM: 710,0 mV
Voltage TARGET: 715,0 mV; EXP: 717,3 mV; MM: 715,0 mV
Voltage TARGET: 720,0 mV; EXP: 722,3 mV; MM: 720,0 mV
Voltage TARGET: 725,0 mV; EXP: 727,2 mV; MM: 725,0 mV
Voltage TARGET: 730,0 mV; EXP: 732,2 mV; MM: 730,0 mV
Voltage TARGET: 735,0 mV; EXP: 737,2 mV; MM: 735,0 mV
Voltage TARGET: 740,0 mV; EXP: 742,0 mV; MM: 740,0 mV
Voltage TARGET: 745,0 mV; EXP: 747,0 mV; MM: 745,0 mV
Voltage TARGET: 750,0 mV; EXP: 751,9 mV; MM: 750,0 mV
Voltage TARGET: 755,0 mV; EXP: 756,9 mV; MM: 755,0 mV
Voltage TARGET: 760,0 mV; EXP: 761,9 mV; MM: 760,0 mV
Voltage TARGET: 765,0 mV; EXP: 766,8 mV; MM: 765,0 mV
Voltage TARGET: 770,0 mV; EXP: 771,8 mV; MM: 770,0 mV
Voltage TARGET: 775,0 mV; EXP: 776,8 mV; MM: 775,0 mV
Voltage TARGET: 780,0 mV; EXP: 782,0 mV; MM: 780,0 mV
Voltage TARGET: 785,0 mV; EXP: 787,0 mV; MM: 785,0 mV
Voltage TARGET: 790,0 mV; EXP: 791,9 mV; MM: 790,0 mV
Voltage TARGET: 795,0 mV; EXP: 796,9 mV; MM: 795,0 mV
Voltage TARGET: 800,0 mV; EXP: 801,9 mV; MM: 800,0 mV
Voltage TARGET: 805,0 mV; EXP: 806,8 mV; MM: 805,0 mV
Voltage TARGET: 810,0 mV; EXP: 811,8 mV; MM: 810,0 mV
Voltage TARGET: 815,0 mV; EXP: 816,8 mV; MM: 815,0 mV
Voltage TARGET: 820,0 mV; EXP: 822,2 mV; MM: 820,0 mV
Voltage TARGET: 825,0 mV; EXP: 827,2 mV; MM: 825,0 mV
Voltage TARGET: 830,0 mV; EXP: 832,1 mV; MM: 830,0 mV
Voltage TARGET: 835,0 mV; EXP: 837,1 mV; MM: 835,0 mV
Voltage TARGET: 840,0 mV; EXP: 842,1 mV; MM: 840,0 mV
Voltage TARGET: 845,0 mV; EXP: 847,0 mV; MM: 845,0 mV
Voltage TARGET: 850,0 mV; EXP: 852,0 mV; MM: 850,0 mV
Voltage TARGET: 855,0 mV; EXP: 857,0 mV; MM: 855,0 mV
Voltage TARGET: 860,0 mV; EXP: 861,8 mV; MM: 860,0 mV
Voltage TARGET: 865,0 mV; EXP: 866,8 mV; MM: 865,0 mV
Voltage TARGET: 870,0 mV; EXP: 871,7 mV; MM: 870,0 mV
Voltage TARGET: 875,0 mV; EXP: 876,7 mV; MM: 875,0 mV
Voltage TARGET: 880,0 mV; EXP: 881,7 mV; MM: 880,0 mV
Voltage TARGET: 885,0 mV; EXP: 886,6 mV; MM: 885,0 mV
Voltage TARGET: 890,0 mV; EXP: 891,6 mV; MM: 890,0 mV
Voltage TARGET: 895,0 mV; EXP: 896,6 mV; MM: 895,0 mV
Voltage TARGET: 900,0 mV; EXP: 901,5 mV; MM: 900,0 mV
Voltage TARGET: 905,0 mV; EXP: 906,5 mV; MM: 905,0 mV
Voltage TARGET: 910,0 mV; EXP: 911,4 mV; MM: 910,0 mV
Voltage TARGET: 915,0 mV; EXP: 916,4 mV; MM: 915,0 mV
Voltage TARGET: 920,0 mV; EXP: 921,4 mV; MM: 920,0 mV
Voltage TARGET: 925,0 mV; EXP: 926,3 mV; MM: 925,0 mV
Voltage TARGET: 930,0 mV; EXP: 931,3 mV; MM: 930,0 mV
Voltage TARGET: 935,0 mV; EXP: 936,3 mV; MM: 935,0 mV
Voltage TARGET: 940,0 mV; EXP: 941,5 mV; MM: 940,0 mV
Voltage TARGET: 945,0 mV; EXP: 946,5 mV; MM: 945,0 mV
Voltage TARGET: 950,0 mV; EXP: 951,4 mV; MM: 950,0 mV
Voltage TARGET: 955,0 mV; EXP: 956,4 mV; MM: 955,0 mV
Voltage TARGET: 960,0 mV; EXP: 961,4 mV; MM: 960,0 mV
Voltage TARGET: 965,0 mV; EXP: 966,3 mV; MM: 965,0 mV
Voltage TARGET: 970,0 mV; EXP: 971,3 mV; MM: 970,0 mV
Voltage TARGET: 975,0 mV; EXP: 976,3 mV; MM: 975,0 mV
Voltage TARGET: 980,0 mV; EXP: 982,4 mV; MM: 980,0 mV
Voltage TARGET: 985,0 mV; EXP: 987,4 mV; MM: 985,0 mV
Voltage TARGET: 990,0 mV; EXP: 992,3 mV; MM: 990,0 mV
Voltage TARGET: 995,0 mV; EXP: 997,3 mV; MM: 995,0 mV
Voltage TARGET: 1000,0 mV; EXP: 1002,3 mV; MM: 1000,0 mV