Author Topic: why Fluke use bandpass filter in their DMMs?  (Read 2006 times)

0 Members and 2 Guests are viewing this topic.

Offline emm386Topic starter

  • Newbie
  • Posts: 8
  • Country: us
why Fluke use bandpass filter in their DMMs?
« on: October 04, 2023, 05:46:54 am »
Hi all,

While I was skimming over the service manual of Fluke's 87-V and 189-II (figures attached), I realized that they have a bandpass-looking filter in their custom IC.
According to the Fluke 87-V service manual, it is referred as a low-pass filter, but its feedback structure makes it looking like an active bandpass filter (reference: https://cheever.domains.swarthmore.edu/Ref/FilterBkgrnd/Filters.html).
Can anyone please explain me the purpose of this filter and why this is configured as a bandpass filter ?

Plus, where would the 'FAMP_O' pin in the first attachment be connected?




« Last Edit: October 04, 2023, 05:52:14 am by emm386 »
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14638
  • Country: de
Re: why Fluke use bandpass filter in their DMMs?
« Reply #1 on: October 04, 2023, 07:28:13 am »
The filter as 2nd order low pass filters. It is a configuration often found in DMMs, but not so much in the standard books.
In the first schematics C27+C36+R78 and the Filter op amp form a kind of "super (2nd order)" capacitor.
With a cross over at some 10 Hz (crude estimate) this filter is likely there to suppress mains hum (as part of the signal) before the gain stage. 
 

Offline emm386Topic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: why Fluke use bandpass filter in their DMMs?
« Reply #2 on: October 04, 2023, 07:41:00 am »
Could you explain me how this forms a second order filter? Even in my LTSpice simulation this shows BP response...
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14638
  • Country: de
Re: why Fluke use bandpass filter in their DMMs?
« Reply #3 on: October 04, 2023, 08:20:57 am »
The low pass ouput is not from the output of the OP-amp, but from the right of R2.
 
The following users thanked this post: robert.rozee, emm386

Offline MegaVolt

  • Frequent Contributor
  • **
  • Posts: 930
  • Country: by
Re: why Fluke use bandpass filter in their DMMs?
« Reply #4 on: October 04, 2023, 12:18:46 pm »
Perhaps these images will help you understand these filters.
 
The following users thanked this post: Exosia, robert.rozee, tridac, emm386

Offline emm386Topic starter

  • Newbie
  • Posts: 8
  • Country: us
Re: why Fluke use bandpass filter in their DMMs?
« Reply #5 on: October 04, 2023, 05:39:56 pm »
MegaVolt and Kleinstein, thanks for the clarification and answering my question! this is very helpful
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17050
  • Country: us
  • DavidH
Re: why Fluke use bandpass filter in their DMMs?
« Reply #6 on: October 04, 2023, 09:11:55 pm »
The advantage of this filter configuration is that it is DC accurate; the DC offset voltage of the amplifier does not contribute to the output.  There is a similar configuration for a DC accurate precision rectifier which is AC coupled on its input and output.

« Last Edit: October 04, 2023, 09:25:43 pm by David Hess »
 
The following users thanked this post: robert.rozee, emm386

Online tszaboo

  • Super Contributor
  • ***
  • Posts: 7841
  • Country: nl
  • Current job: ATEX product design
Re: why Fluke use bandpass filter in their DMMs?
« Reply #7 on: October 04, 2023, 10:36:32 pm »
Looking at the second picture, my question is rather:
Why do they have so many lightbulbs in a multimeter? 
:-DD
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17050
  • Country: us
  • DavidH
Re: why Fluke use bandpass filter in their DMMs?
« Reply #8 on: October 05, 2023, 07:08:56 am »
Why do they have so many lightbulbs in a multimeter?

As shown below, what light bulbs in multimeters might look like.
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14638
  • Country: de
Re: why Fluke use bandpass filter in their DMMs?
« Reply #9 on: October 05, 2023, 07:48:25 am »
In a few cases light bulbs are used as a low current fuse.
In the plan from the start the circle with x symbol is not for light bulbs but for CMOS switches.  So no light bulbs but chip internal CMOS switches for the signal.

P.s. : A light bulb is not a good fuse and may fail with an arc flash. It is not uncommon for a failing mains bulb to take out the breaker from that.
« Last Edit: October 05, 2023, 07:50:26 am by Kleinstein »
 

Offline MegaVolt

  • Frequent Contributor
  • **
  • Posts: 930
  • Country: by
Re: why Fluke use bandpass filter in their DMMs?
« Reply #10 on: October 05, 2023, 10:19:40 am »
There is a similar configuration for a DC accurate precision rectifier which is AC coupled on its input and output.
Do you have an example circuit diagram?
 

Offline tridac

  • Regular Contributor
  • *
  • Posts: 125
  • Country: gb
Re: why Fluke use bandpass filter in their DMMs?
« Reply #11 on: October 05, 2023, 12:15:09 pm »
Been trying to find more info on this filter design. Asked in another thread and pointed towards a Burr Brown third gen op amp techniques book, since found on ABE books. Still interested in a more thorough treatment, so which book or source did you find the above 5th order example ?.

Thanks, still collecting data...
« Last Edit: October 05, 2023, 12:20:12 pm by tridac »
Test gear restoration, hardware and software projects...
 

Offline MegaVolt

  • Frequent Contributor
  • **
  • Posts: 930
  • Country: by
Re: why Fluke use bandpass filter in their DMMs?
« Reply #12 on: October 05, 2023, 12:26:45 pm »
Still interested in a more thorough treatment, so which book or source did you find the above 5th order example ?
 
The following users thanked this post: tridac, emm386

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17050
  • Country: us
  • DavidH
Re: why Fluke use bandpass filter in their DMMs?
« Reply #13 on: October 05, 2023, 12:40:06 pm »
There is a similar configuration for a DC accurate precision rectifier which is AC coupled on its input and output.
Do you have an example circuit diagram?

The Tektronix DM501 and DM502 use it with a 318 type of operational amplifier which otherwise would contribute large DC errors.
 
The following users thanked this post: MegaVolt, tridac

Online tszaboo

  • Super Contributor
  • ***
  • Posts: 7841
  • Country: nl
  • Current job: ATEX product design
Re: why Fluke use bandpass filter in their DMMs?
« Reply #14 on: October 05, 2023, 03:40:47 pm »
In a few cases light bulbs are used as a low current fuse.
In the plan from the start the circle with x symbol is not for light bulbs but for CMOS switches.  So no light bulbs but chip internal CMOS switches for the signal.

P.s. : A light bulb is not a good fuse and may fail with an arc flash. It is not uncommon for a failing mains bulb to take out the breaker from that.

Why do they have so many lightbulbs in a multimeter?

As shown below, what light bulbs in multimeters might look like.


Come on, take a look at the picture, there are at least 20 light bulbs in that multimeter according to the schematic.
Or someone used that symbol as the selector switch's switching part as a lightbulb. If you look at it, 29-30-31-32 clearly sets a gain of an amplifier.
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11726
  • Country: us
Re: why Fluke use bandpass filter in their DMMs?
« Reply #15 on: October 05, 2023, 04:04:38 pm »
internal switches, not lamps.

check here for example:
http://www.cyrustek.com.tw/en/wp-content/uploads/ES51997.pdf
« Last Edit: October 05, 2023, 04:08:15 pm by joeqsmith »
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17050
  • Country: us
  • DavidH
Re: why Fluke use bandpass filter in their DMMs?
« Reply #16 on: October 05, 2023, 05:05:17 pm »
Come on, take a look at the picture, there are at least 20 light bulbs in that multimeter according to the schematic.
Or someone used that symbol as the selector switch's switching part as a lightbulb. If you look at it, 29-30-31-32 clearly sets a gain of an amplifier.

Any light bulb is a switch, if driven hard enough, and the reverse.
 

Offline TimFox

  • Super Contributor
  • ***
  • Posts: 8168
  • Country: us
  • Retired, now restoring antique test equipment
Re: why Fluke use bandpass filter in their DMMs?
« Reply #17 on: October 05, 2023, 05:18:51 pm »
Symbolism can be difficult.
 

Offline tridac

  • Regular Contributor
  • *
  • Posts: 125
  • Country: gb
Re: why Fluke use bandpass filter in their DMMs?
« Reply #18 on: October 06, 2023, 03:45:50 pm »
There are references in that article, one (9), being Analog Filter Design by Val Valkenburg. Looked that up and lloks like it may have material on the filter topology of interest. Found a copy on ABE Books for less than $10, soon to be on it's way to me...
Test gear restoration, hardware and software projects...
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf