Author Topic: Hioki 7020 - looking for service manual  (Read 945 times)

0 Members and 1 Guest are viewing this topic.

Offline DavidKoTopic starter

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: cz
Hioki 7020 - looking for service manual
« on: October 25, 2021, 08:50:47 pm »
Hello,
I'm searching for service manual for Hioki 7020. Hioki official statement is that it is long time unsupported device. Some frequencies are out of the range. There is lot of trimpots and test points inside, but the marking does not give a hint.

At least I have dumped the FW 1.20 if someone can have a use for it
 

Offline DavidKoTopic starter

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: cz
Re: Hioki 7020 - looking for service manual
« Reply #1 on: November 17, 2021, 11:05:27 pm »
To my surprise the device uses universal transformer. After some measurements I have changed the "jumper" to from 100V to 240V one (one cable is not necessary, but I do not waste him, so I have only doubled the connection that is already there).

I have also replaced the input filter before doing this. I have bad experience with X2 caps in old devices
 

Offline DavidKoTopic starter

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: cz
Re: Hioki 7020 - looking for service manual
« Reply #2 on: November 17, 2021, 11:17:54 pm »
Some pictures

The picture of the first board is missing. During the first repair I have dumped the FW, but I forgot to make a picture  :(. It cannot be removed without removing the front panel.
 

Offline DavidKoTopic starter

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: cz
Re: Hioki 7020 - looking for service manual
« Reply #3 on: November 18, 2021, 10:14:16 am »
It looks that service manual is unobtainium. I have tried to ask for it, but the reply is that they will not calibrate it and it is unsupported device.

At least the user manual is still available online on their site https://www.hioki.com/download/29153
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf