How much noise is normal with a TDS7104
« on: August 22, 2021, 09:10:44 pm »
Hello experts,
without a probe, the oscilloscope will show this noise.


Is that normal?
Regards
 Its Me
« Last Edit: August 22, 2021, 09:47:41 pm by Its me »
 

Offline Johnny10

  • Frequent Contributor
  • **
  • Posts: 900
  • Country: us
Re: How much noise is normal with a TDS7104
« Reply #1 on: August 22, 2021, 09:57:12 pm »
Mine shows 1.1-2.2 mV peak to peak.
Just after start-up

Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline Specmaster

  • Super Contributor
  • ***
  • Posts: 14483
  • Country: gb
Re: How much noise is normal with a TDS7104
« Reply #2 on: August 22, 2021, 10:14:50 pm »
Depending on what the settings are on the inputs, yes it is normal.
Who let Murphy in?

Brymen-Fluke-HP-Thurlby-Thander-Tek-Extech-Black Star-GW-Avo-Kyoritsu-Amprobe-ITT-Robin-TTi
 

Offline RolandK

  • Regular Contributor
  • *
  • Posts: 109
  • Country: de
Re: How much noise is normal with a TDS7104
« Reply #3 on: August 23, 2021, 05:15:39 am »
Your scope is broken - give it to me  8)


To see the noise of the scope you should prevent any pickup with an open input, short the input or put a 50 Ohm Termination on it. And no wire loop...
Why do old shaffner filters blow? - because there are rifas inside.
Why do rifas blow? Only time shows if the best new thing is really best. Here it is not.
 

Offline Its meTopic starter

  • Contributor
  • Posts: 10
  • Country: de
Re: How much noise is normal with a TDS7104
« Reply #4 on: August 23, 2021, 10:28:04 pm »
Your scope is broken - give it to me  8)
Das könnte Dir so passen  :)

Even with short-circuited inputs, the noise does not change.
When I connect the probe to ground nothing changes in the noise.
A connection to Probe Compensation provides this result.

I suspected that the capacitors were failing on my TDS7104.
My Tektronix TDS 684A oscilloscope that was not used for 9 years
also showed this noise. Shortly afterwards it was broken.
As it turned out, many capacitors on the TDS 684A have leaked.

If the noise on the TDS7104 is normal, I don't need to worry.