Thanks. Added to the collection. Also made a specifix FW changer package for it.
However I smell more Hantek "engeneereng"...
Looks like Hantek's testing department is non-existent, but they seem to respond reasonably well at bug reports.
So, what about making a single post and keep it with all bugs being found, ex. like this example?
I'd have no problem keeping it updated.
» Current software version «
» 2022.02.10
» Vertical view «
» Horizontal view «
» Serious artifacts caused by undersampling, ex. a 10MHz signal shows garbage below 20us/div, you might not notice there's a signal there.
» Confirmed in FWs: 3102
» Zoom window can't be moved to the left past the trigger position, didn't happen before 2021.02.23
» Confirmed in FWs: 3102
» Math «
» Trigger «
» Wave generator «
» Amplitude can be set too high and cause distortion if offset was modified first. Backwards, it works correctly.
» Confirmed in FWs: 3102
» Duty moves in 1% steps, but after rotating fast and reaching 99%, it'll stick to 10% steps until manually reaching 0%.
» Confirmed in FWs: 3102
» Decoding «
» Impossible to decode SPI no matter the trigger options? Displays a "0" or "?" on each clock cycle instead every 8/16/32.
» Confirmed in FWs: 3102
» Display «
» Measure «
» Statistics don't reset after disabling/enabling, you need to go into Type and deselect/select all.
» Confirmed in FWs: 3102
» Poor accuracy in voltage measurements
» Confirmed in FWs: 3102
» Cursor «
» Acquire «
» Other issues «