I've got a small selection of hand held DMMs but always wanted a bench one.  I've now decided to go for it and having looked at the excellent spreadsheet on here I did some research on the internet, ebay etc and there are a few possibles.  New I was looking at possibly the Rigol DM3058E but then I looked at some older ones and spotted the Philips PM2525....

My question is, the Philips is £210 whilst the Rigol is £435.  I can build an interface for IEEE-488 to USB for a few pounds and the 2525 includes the IEEE-488 interface so effectively the 2525 is half the price so what advantage is there in getting a new one?  Warranty is 6 months for the 2nd hand one so that's not as good and the calibration is not guaranteed.

Is there another DMM I should look at?  My budget is around £500 but I could wait another month or so if there was a brilliant one at around £600

EDIT:  Oh I should add what I'm doing with it!  Well digital circuit design, microprocessors, power supplies, vintage computers (inc modding), repairs to older and some new stuff..... 
« Last Edit: January 12, 2018, 08:25:30 pm by cowasaki »
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1551
  • Country: be
Re: New bench DMM or older one
« Reply #1 on: January 12, 2018, 08:50:19 pm »
210£ for a Philips PM2525 seems very high, I would expect these to go for below 100£.
 
The following users thanked this post: cowasaki

Online tautech

  • Super Contributor
  • ***
  • Posts: 29096
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: New bench DMM or older one
« Reply #2 on: January 12, 2018, 08:58:51 pm »
Welcome to the forum.

A couple from the SDM3000 range of Siglent bench meters fall into your budget.
https://www.siglentamerica.com/digital-multimeters/
http://siglenteu.com/pd.aspx?tid=37&T=1

They're 3 year warranted and come with Cal certs.

There's more about them here:
https://www.eevblog.com/forum/testgear/siglent-new-bench-dmm-sdm3055/

The 5 1/2 digit model was the first released and they had some problems but all the new ones with later firmware work good.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: cowasaki

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 6040
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: New bench DMM or older one
« Reply #3 on: January 12, 2018, 09:21:27 pm »
210£ for a Philips PM2525 seems very high, I would expect these to go for below 100£.
I agree with this. IIRC some folks reported the Philips/Fluke was not as robust as other brands of their time, but I have never had one. Just FYI, some users reported important things to look for on a used PM2525.

As tautech mentioned, Siglent also has some interesting models if you are looking for something brand new. Also, a resident expert here (nctnico) swears by the low cost VC8145 meter.

Your usage looks quite similar to mine. I have some older bench meters (Keithley 191, Racal Dana 5001, HP 3478A) that I bought used but very cheaply. I don't necessarily need the full 5-1/2 digit precision, but I like the fact they can be left on all day long without problems with batteries or auto-off. I also love the 4-wire measurements.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 
The following users thanked this post: cowasaki

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: New bench DMM or older one
« Reply #4 on: January 12, 2018, 09:51:05 pm »
I generally prefer older bench DMMs, some of them are quite good, even ancient instruments that were very expensive originally. My oldest is a Keithly nixie display DMM, it's only 3.5 digits but it has some very low ranges for measuring tiny currents and voltages. Checked against a calibrator it was dead on despite the last calibration being something like 30 years ago.
 
The following users thanked this post: cowasaki

Offline tggzzz

  • Super Contributor
  • ***
  • Posts: 20228
  • Country: gb
  • Numbers, not adjectives
    • Having fun doing more, with less
Re: New bench DMM or older one
« Reply #5 on: January 12, 2018, 10:03:32 pm »
Thanks for adding your expected usage :)

The answer to your question depends on either what is inadequate about your existing meters, and/or what you hope to gain with new meters.
There are lies, damned lies, statistics - and ADC/DAC specs.
Glider pilot's aphorism: "there is no substitute for span". Retort: "There is a substitute: skill+imagination. But you can buy span".
Having fun doing more, with less
 
The following users thanked this post: cowasaki

Offline Performa01

  • Super Contributor
  • ***
  • Posts: 1701
  • Country: at
Re: New bench DMM or older one
« Reply #6 on: January 12, 2018, 10:53:43 pm »
I’ve just looked around in my lab and counted a total of 16 bench multimeters – and there are some more in the cabinet. This is a collection that has built up over time, as bothering with handhelds in a lab has never been a tempting idea for me (of course I do have a couple of handhelds for field use though).

Regarding my bench DMM collection, none of these instruments is a Philips. I remember that for some reason they have never appeared that attractive and competitive – at least not for my requirements.

That said, you have clearly stated your requirements and having had a look at the datasheet, I cannot see a reason why a Philips 2525 would not be for you. Its accuracy seems rather low for 5.5 digits, but then again, high accuracy isn’t important for your applications anyway. On the other hand it has some great features like the 1µA and 10µA current ranges and the low burden voltage in general. But as stated before, I have absolutely no experience with Philips DMMs, and given the age of these instruments, they might be unreliable and in need of repair (with some obsolete and hard to source parts) by now.

For your requirements, a 4.5 digit meter would be all you need. 20000 counts has been the standard for lab meters (without special requirements) for many decades. In my collection, there are the Fluke 45, HP3466A and Tektronix DM2510 that belong in this category. All these are very nice meters, but…

The Fluke 45 is fairly modern and reliable, but a bit expensive on the used market for what it is, at least in Europe.
The HP3466A is much cheaper, but it’s rather old and most of them got unreliable by now (mainly because of the mechanical switches), so I cannot recommend it anymore.
The Tektronix DM2510 is fairly modern and reliable just like the Fluke 45, but it is very hard to find and because of this, prices are sometimes a little on the high side.

So even though all three meters would fit your budget, they are either hard to find (at least at a reasonable price) or unreliable, so I would not recommend them unless you can actually find a great deal on either the Fluke or the Tektronix.

As I said before, the Philips looks nice if the price is right, but I cannot comment on the reliability.

The problem with older 5.5 or even 6.5 digit DMMs is that they usually lack decent current ranges – and this would be something really useful (if not even essential) for your tasks. If it weren’t for that, and your requirements would be very different and just call for exceptional accuracy and stability, then I might have recommended a Fluke 8840A or even a Keithley 2000 or 2015 THD. But all these meters won’t fit the bill when it comes to current ranges. One has to be particularly cautious when purchasing a used Fluke 8840A, because True RMS AC measurements are an option and advertisements for used instruments often don’t clearly state which options are fitted. IEEE488 is also an option by the way.

I don’t have experience with the Rigol DM3058E and from the datasheet it would fit the bill, but from the comments I’ve read here in the forum, it doesn’t appear to be that great. At one point I made my department at work purchase a couple Rigol DM3068, as I figured they would be okay for our needs – and they certainly are, but then wasn’t overly impressed when I personally evaluated them. After that, I abandoned the idea to get one of these meters for my personal lab – because my requirements are quite obviously different from yours and also from the requirements at work. But that’s only a side note, as this meter is not within your budget anyway.

From the datasheet, the Siglent SDM3045X looks like a good fit for your needs and is very similar to the Rigol DM3058 - except for the resolution, which is 60000 Siglent vs. 200000 Rigol). In any case it’s a bit cheaper and has the fancier display for sure. But then again, I have no personal experience with Siglent DMMs either.
 
The following users thanked this post: cowasaki

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1551
  • Country: be
Re: New bench DMM or older one
« Reply #7 on: January 13, 2018, 07:06:45 am »
Regarding my bench DMM collection, none of these instruments is a Philips. I remember that for some reason they have never appeared that attractive and competitive – at least not for my requirements.

The Philips ones were also sold by fluke (fluke pm2525), so I think they are quite ok (but for sure not worth 210£) I have 2 of these, but I almost never use them. In contrast with what was reported here, I prefer to use handhelds (BM869s) instead of bench multimeters, because I can place them in the "line of sight" when I am probing.

The only use for a bench multimeter for me would be high accuracy, but then we are not talking about a 5 1/2 digits one off course.
I would gladly trade my 2 PM2525s for 1 additional BM869s, this is in my opinion a much more useful tool on the bench!
 
The following users thanked this post: cowasaki

Offline Tom45

  • Frequent Contributor
  • **
  • Posts: 556
  • Country: us
Re: New bench DMM or older one
« Reply #8 on: January 13, 2018, 03:52:54 pm »
The Rigol has trend graphs and histograms, features that  I think are missing from the Phillips. I use the trend and histogram features of my Keysight 34465a far more often than I need the full accuracy.
 
The following users thanked this post: cowasaki

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1551
  • Country: be
Re: New bench DMM or older one
« Reply #9 on: January 13, 2018, 09:12:18 pm »
The Rigol has trend graphs and histograms, features that  I think are missing from the Phillips. I use the trend and histogram features of my Keysight 34465a far more often than I need the full accuracy.
Yes, I can imagine that indeed a useful feature. When I would buy a bench meter, I would for sure go for a recent one, and the rigol or siglent look like nice units. Off course if the budget stretches until a 34465a, I would not say no also...
 
The following users thanked this post: cowasaki

Offline KrudyZ

  • Frequent Contributor
  • **
  • Posts: 289
  • Country: us
Re: New bench DMM or older one
« Reply #10 on: January 13, 2018, 09:35:22 pm »
I've got a small selection of hand held DMMs but always wanted a bench one.  I've now decided to go for it and having looked at the excellent spreadsheet on here I did some research on the internet, ebay etc and there are a few possibles.  New I was looking at possibly the Rigol DM3058E but then I looked at some older ones and spotted the Philips PM2525....

My question is, the Philips is £210 whilst the Rigol is £435.  I can build an interface for IEEE-488 to USB for a few pounds and the 2525 includes the IEEE-488 interface so effectively the 2525 is half the price so what advantage is there in getting a new one?  Warranty is 6 months for the 2nd hand one so that's not as good and the calibration is not guaranteed.

Is there another DMM I should look at?  My budget is around £500 but I could wait another month or so if there was a brilliant one at around £600

EDIT:  Oh I should add what I'm doing with it!  Well digital circuit design, microprocessors, power supplies, vintage computers (inc modding), repairs to older and some new stuff.....

Considering your use case, I see no good reason to get a bench meter. I would spend my money on time domain instruments instead, i.e. a scope, logic analyzer, protocol analyzer or an MSO with all of that in one instrument.
If you really want a bench DMM, you could easily get a used 34401A within your budget and it would hold its (remaining) value for a lot longer than any of the lower resolution meters.
 
The following users thanked this post: cowasaki

Offline Macbeth

  • Super Contributor
  • ***
  • Posts: 2571
  • Country: gb
Re: New bench DMM or older one
« Reply #11 on: January 13, 2018, 10:09:07 pm »
The Rigol has trend graphs and histograms, features that  I think are missing from the Phillips. I use the trend and histogram features of my Keysight 34465a far more often than I need the full accuracy.
Yes, I can imagine that indeed a useful feature. When I would buy a bench meter, I would for sure go for a recent one, and the rigol or siglent look like nice units. Off course if the budget stretches until a 34465a, I would not say no also...

Rigol DM3068 has trend graph and histograms (that update in real time).

The DM3058/E only has histogram, and it does not update on it's tiny LCD screen without user intervention to keep mashing the buttons. Utterly pathetic. Cheaper Siglent models appear much better than this >10 year old dinosaur that has NO SUPPORT WHATSOEVER from RIGOL.  :horse:
 
The following users thanked this post: cowasaki

Offline GreyWoolfe

  • Supporter
  • ****
  • Posts: 3652
  • Country: us
  • NW0LF
Re: New bench DMM or older one
« Reply #12 on: January 13, 2018, 10:19:34 pm »
Tautech made a comment on the HP 3466 and switch unreliability.  I have 2 of them bought for about $100 USD total.  The LED indicators didn't match up to the switches.  Some switch cleaner and much vigorous button pushing cleaned everything up and the indicators match the selection switch.  They both are right there with my GW Instek GDM 8251A 5.5 digit meter.  I really like the red LED display, just wish the meters used shrouded banana jacks but that's not a mod I feel like doing.  Consider it as it would leave you money for more test equipment.  One can never have enough, right? >:D
"Heaven has been described as the place that once you get there all the dogs you ever loved run up to greet you."
 
The following users thanked this post: cowasaki

Online tautech

  • Super Contributor
  • ***
  • Posts: 29096
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: New bench DMM or older one
« Reply #13 on: January 13, 2018, 10:36:22 pm »
Tautech made a comment on the HP 3466 and switch unreliability.
I couldn't find it with a search but it would likely have been a generic comment based on the reliability of mechanical switches in older test equipment. Meters, scopes etc all suffer from switch degradation and cleaning and exercise is another maintenance task needed for older equipment.
For the hobbyist that's aware of this, it's not so big deal.
« Last Edit: January 14, 2018, 01:45:43 am by tautech »
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 6040
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: New bench DMM or older one
« Reply #14 on: January 13, 2018, 11:41:36 pm »
Yes, my Keithley 191 had this issue as well. After some continuous repetition pressing all keys and they are like new again. In my case I suspect it was a meter that stayed on the same scale for most of its previous life.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline GreyWoolfe

  • Supporter
  • ****
  • Posts: 3652
  • Country: us
  • NW0LF
Re: New bench DMM or older one
« Reply #15 on: January 14, 2018, 01:54:30 am »
Tautech made a comment on the HP 3466 and switch unreliability.
I couldn't find it with a search but it would likely have been a generic comment based on the reliability of mechanical switches in older test equipment. Meters, scopes etc all suffer from switch degradation and cleaning and exercise is another maintenance task needed for older equipment.
For the hobbyist that's aware of this, it's not so big deal.

My bad, it was Performa01 not you.  Sorry, Sir.  My train of thought obviously derailed. |O
"Heaven has been described as the place that once you get there all the dogs you ever loved run up to greet you."
 
The following users thanked this post: tautech

Offline Jester

  • Frequent Contributor
  • **
  • Posts: 883
  • Country: ca
Re: New bench DMM or older one
« Reply #16 on: January 14, 2018, 03:32:15 am »
Re PM2525.....I eventually sold every LCD based, bench DMM meter that I had acquired over the years, simply because in comparison to VFD or even LED they are just not pleasant to look at.

What about a Instek GDM-8251A, these are selling for about $100-200 on ebay. Large bright dual display and 10A range. It's the least accurate of all the bench DMM's I have, however it was really inexpensive and basically works fine for basic measurements. IMO, great value for the price.
« Last Edit: January 14, 2018, 11:45:40 am by Jester »
 
The following users thanked this post: cowasaki

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1551
  • Country: be
Re: New bench DMM or older one
« Reply #17 on: January 14, 2018, 07:55:35 am »
Rigol DM3068 has trend graph and histograms (that update in real time).

The DM3058/E only has histogram, and it does not update on it's tiny LCD screen without user intervention to keep mashing the buttons. Utterly pathetic. Cheaper Siglent models appear much better than this >10 year old dinosaur that has NO SUPPORT WHATSOEVER from RIGOL.  :horse:

Stating it like this, I would indeed also go for the Siglent then...  :wtf:

Have not used the Siglent or the Rigol multimeter in the past, but have used product from both vendors, and liked them. They sometimes have some quirks, but price/performance is excellent, even when compared to older A-brand EBay gear.

For me personally, I purchase based on the following guideline:
-   When the equipment is cheap/affordable to buy new from a good supplier, I buy new.
-   When the type of equipment that has benefit a lot from the digital revolution (much more functions, big screen, smaller footprint, remote connectivity…), but it is too expensive to buy from an A-brand supplier, a still try to buy new, but also consider B and C brand suppliers
-   For all other gear, I go the Ebay route. Sometimes I risk to buy broken, sometimes I buy working. Typically I do not buy HPAK gear, because they go for way too high on Ebay. Advantest, R&S, Yokogawa, EG&G, SRS… can be had for much better prices. For a repair project, HPAC is recommended because of excellent documentation and info on the net.
 
The following users thanked this post: cowasaki

Offline HalFET

  • Frequent Contributor
  • **
  • Posts: 512
  • Country: 00
Re: New bench DMM or older one
« Reply #18 on: January 14, 2018, 03:22:19 pm »
Regarding my bench DMM collection, none of these instruments is a Philips. I remember that for some reason they have never appeared that attractive and competitive – at least not for my requirements.

The Philips ones were also sold by fluke (fluke pm2525), so I think they are quite ok (but for sure not worth 210£) I have 2 of these, but I almost never use them. In contrast with what was reported here, I prefer to use handhelds (BM869s) instead of bench multimeters, because I can place them in the "line of sight" when I am probing.

The only use for a bench multimeter for me would be high accuracy, but then we are not talking about a 5 1/2 digits one off course.
I would gladly trade my 2 PM2525s for 1 additional BM869s, this is in my opinion a much more useful tool on the bench!

The Philips PM25x5 meters weren't that great, you can pretty much knock one off the digit rating. The PM2535 claims to be 6.5, but it's really only 5.5. That being said, these things do seem to keep their cal pretty well, mine is still nearly spot on even though it's last cal was in the early 90s. The 2525 is definitely not worth 210 GBP, but it's not that bad of a meter either for all its flaws. (See C. Arnoux for really atrocious quality from the same era.) The build quality on them seems pretty good, just be sure to read the manual before you ever turn one on.

And if anyone is interested, I am planning to dispose of that PM2535 in my stash rather soon. :)
 
The following users thanked this post: cowasaki

Offline cowasakiTopic starter

  • Frequent Contributor
  • **
  • Posts: 605
  • Country: gb
Re: New bench DMM or older one
« Reply #19 on: January 14, 2018, 09:57:38 pm »
Thanks for adding your expected usage :)

The answer to your question depends on either what is inadequate about your existing meters, and/or what you hope to gain with new meters.

Absolutely nothing is inadequate about my main meter which is a UNI-T 60E.  I am hoping to move away from micro controllers and start experimenting with other things.  I have a UNI-T UTD4202C scope and liked that so I just bought the UNI-T 60E when my current DMM broke.  I've always wanted a bench DMM so I can stick it on the shelf with the scope etc and make use of it there rather than having the hand held one getting in the way.  It also means I can have it back lit with no battery issues. I have a larger budget but there are other things I would like to buy.  I quite liked the Philips/fluke stuff having used it in my first proper job around 1991 when I spent quite a lot of work's money on a 100MHz storage scope.

If I can find a PM2525 for under £100 I would buy it even if I got another DMM later.  I know that the PM2525 will be better than my hand held device!
 

Offline cowasakiTopic starter

  • Frequent Contributor
  • **
  • Posts: 605
  • Country: gb
Re: New bench DMM or older one
« Reply #20 on: January 14, 2018, 10:03:58 pm »
Regarding my bench DMM collection, none of these instruments is a Philips. I remember that for some reason they have never appeared that attractive and competitive – at least not for my requirements.

The Philips ones were also sold by fluke (fluke pm2525), so I think they are quite ok (but for sure not worth 210£) I have 2 of these, but I almost never use them. In contrast with what was reported here, I prefer to use handhelds (BM869s) instead of bench multimeters, because I can place them in the "line of sight" when I am probing.

The only use for a bench multimeter for me would be high accuracy, but then we are not talking about a 5 1/2 digits one off course.
I would gladly trade my 2 PM2525s for 1 additional BM869s, this is in my opinion a much more useful tool on the bench!

I've just looked at your recommendation of a BM869s which really does look like a decent DMM.  I do have a DMM which does what I need at the moment but obviously you can't have too many.

On another note if you feel like selling a PM2525 let me know, especially if it has an IEEE833 interface like the one I've seen.  I actually thought the PM2525 I saw was a bit expensive.
 

Offline cowasakiTopic starter

  • Frequent Contributor
  • **
  • Posts: 605
  • Country: gb
Re: New bench DMM or older one
« Reply #21 on: January 14, 2018, 10:13:33 pm »
I've got a small selection of hand held DMMs but always wanted a bench one.  I've now decided to go for it and having looked at the excellent spreadsheet on here I did some research on the internet, ebay etc and there are a few possibles.  New I was looking at possibly the Rigol DM3058E but then I looked at some older ones and spotted the Philips PM2525....

My question is, the Philips is £210 whilst the Rigol is £435.  I can build an interface for IEEE-488 to USB for a few pounds and the 2525 includes the IEEE-488 interface so effectively the 2525 is half the price so what advantage is there in getting a new one?  Warranty is 6 months for the 2nd hand one so that's not as good and the calibration is not guaranteed.

Is there another DMM I should look at?  My budget is around £500 but I could wait another month or so if there was a brilliant one at around £600

EDIT:  Oh I should add what I'm doing with it!  Well digital circuit design, microprocessors, power supplies, vintage computers (inc modding), repairs to older and some new stuff.....

Considering your use case, I see no good reason to get a bench meter. I would spend my money on time domain instruments instead, i.e. a scope, logic analyzer, protocol analyzer or an MSO with all of that in one instrument.
If you really want a bench DMM, you could easily get a used 34401A within your budget and it would hold its (remaining) value for a lot longer than any of the lower resolution meters.

Thanks, I have a UNI-T UTD4202C 200MHz digital storage scope with logic analyser now and to be honest I don't NEED a bench DMM it's rather that I WANT one :)

I've found a 34401A in Israel for nearly £300 inc postage but it's a bit scabby.......  The better condition ones seem to be from the US at maybe £420 BUT if customs stop it I'm looking at another £84 VAT and £25 handling charge so £529
« Last Edit: January 14, 2018, 10:43:42 pm by cowasaki »
 

Offline cowasakiTopic starter

  • Frequent Contributor
  • **
  • Posts: 605
  • Country: gb
Re: New bench DMM or older one
« Reply #22 on: January 14, 2018, 10:17:52 pm »
Tautech made a comment on the HP 3466 and switch unreliability.  I have 2 of them bought for about $100 USD total.  The LED indicators didn't match up to the switches.  Some switch cleaner and much vigorous button pushing cleaned everything up and the indicators match the selection switch.  They both are right there with my GW Instek GDM 8251A 5.5 digit meter.  I really like the red LED display, just wish the meters used shrouded banana jacks but that's not a mod I feel like doing.  Consider it as it would leave you money for more test equipment.  One can never have enough, right? >:D

Used equipment is massively cheaper in the USA than it is in Europe/UK.  I would like more stuff so buying a cheaper bench DMM that still does everything would be better.  It's brilliant to get a new box with a new machine in it and a decent warranty but the extra cost....... (especially with 20% VAT)
 

Offline cowasakiTopic starter

  • Frequent Contributor
  • **
  • Posts: 605
  • Country: gb
Re: New bench DMM or older one
« Reply #23 on: January 14, 2018, 10:23:49 pm »
Regarding my bench DMM collection, none of these instruments is a Philips. I remember that for some reason they have never appeared that attractive and competitive – at least not for my requirements.

The Philips ones were also sold by fluke (fluke pm2525), so I think they are quite ok (but for sure not worth 210£) I have 2 of these, but I almost never use them. In contrast with what was reported here, I prefer to use handhelds (BM869s) instead of bench multimeters, because I can place them in the "line of sight" when I am probing.

The only use for a bench multimeter for me would be high accuracy, but then we are not talking about a 5 1/2 digits one off course.
I would gladly trade my 2 PM2525s for 1 additional BM869s, this is in my opinion a much more useful tool on the bench!

The Philips PM25x5 meters weren't that great, you can pretty much knock one off the digit rating. The PM2535 claims to be 6.5, but it's really only 5.5. That being said, these things do seem to keep their cal pretty well, mine is still nearly spot on even though it's last cal was in the early 90s. The 2525 is definitely not worth 210 GBP, but it's not that bad of a meter either for all its flaws. (See C. Arnoux for really atrocious quality from the same era.) The build quality on them seems pretty good, just be sure to read the manual before you ever turn one on.

And if anyone is interested, I am planning to dispose of that PM2535 in my stash rather soon. :)
[/quote

I might be interested if the price is right.....  Even if I get a really decent bench DMM a second DMM is always a good idea if I want to measure two values at the same time.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 6365
  • Country: ca
Re: New bench DMM or older one
« Reply #24 on: January 14, 2018, 10:50:20 pm »
@cowaski    well in some case as you said ... generally used equipment is cheaper in usa, but buying 2 gossen 28s meters months ago, and their usb-rs232 interfaces has costed me a lot, (mostly the shipping fees), a german EEVblog friend helped me to get them from germany.

You have to be sure of what you need. I needed them period  loll 

Had an  mastech 22000 count meter and a brymen 859s but needed more resolution, i dont care about the speed, as i posted on other threads, i hate the mv settings, i only need 1 range of ac or dc measurements. Totally love the auto select input, ac dc or ohms detection ...

Luckily in the lasts months, i was able to score an Gossen 29s on ebay Usa :)  and an old 50k count B&K Precision 5490 bench meter, i was very lucky to obtain the calibration software from Chauvin Arnoux, since it is a clone of mx554 - 556 models.

But the first thing i've checked is for the nearest calibration facility in my country, if there is one ... before i've bouht the Gossen's, since they use specialized software for them.

In Canada you have Transcat who do ton's of brand - models, in usa i think there is an Dranetz facility ...
 