Author Topic: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair  (Read 15672 times)

0 Members and 1 Guest are viewing this topic.

Offline hafrse

  • Regular Contributor
  • *
  • Posts: 117
  • Country: se
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #25 on: February 26, 2017, 09:49:24 pm »
I'm starting this topic because I have bought 2 Advantest R3273 Analyzers on ebay.
Both with different defects.
So far I have been able to replace the step attenuator and got the signal path below 3GHz working.
There was a defect bias circuit for the first mixer which put DC voltage in the first stage and degraded the signal greatly.

Since I also had to replace the mainboard all the cal data is incorrect.
This is not a problem for the range <3.5GHz because the inbuilt cal puts everything back in order.
However for frequncies above 3.5Ghz they go through a different frontend (YTF -> YIG tuned filter) and this is not calibrated with the built in cal function but has to be calibrated thorugh service menu.
Sadly I'm missing the cal manual and the password for the service menu.
I have extracted the firmware / OS which is on a CF card but so far haven't been able to find the password anywhere.
The image can be found here:
good_spectrum.img

Also I'm not sure what architecture the CPU is.
I've taken pictures.
http://imgur.com/a/gGEWs
Maybe someone on the forum is able to point me in the right direction.

Regards

Hello,

I think the  image file you posted is removed from the link you have, can you please check ? I am curious to check the files there.
Thanks

Regards
 

Online PA0PBZ

  • Super Contributor
  • ***
  • Posts: 5169
  • Country: nl
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #26 on: February 26, 2017, 10:00:35 pm »
I think the  image file you posted is removed from the link you have, can you please check ? I am curious to check the files there.

Uploaded here: http://www.filedropper.com/goodspectrum
Keyboard error: Press F1 to continue.
 

Offline hafrse

  • Regular Contributor
  • *
  • Posts: 117
  • Country: se
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #27 on: February 27, 2017, 04:41:36 pm »
I think the  image file you posted is removed from the link you have, can you please check ? I am curious to check the files there.

Uploaded here: http://www.filedropper.com/goodspectrum
Thanks
 

Offline dschiedschTopic starter

  • Contributor
  • Posts: 44
  • Country: at
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #28 on: February 27, 2017, 10:10:33 pm »
Hi.
No I haven't found someone with a signal source yet.
I'm talking to someone at the university which looks good but I don't have that much time at the moment so it will take another month or so to do it.
Since someone else reuploaded the image I didn't reupload my copy. I hope thats ok. Otherwise I can upload mine as well
 

Offline hafrse

  • Regular Contributor
  • *
  • Posts: 117
  • Country: se
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #29 on: March 02, 2017, 09:46:27 pm »
You are right, no way to save calibration values , write protection is on and can not be changed....
 

Offline lowtension

  • Contributor
  • Posts: 24
  • Country: de
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #30 on: April 10, 2017, 03:32:54 pm »
Not as interesting as you might expect.
If you enter a wrong password for the service menu the analyzer just does nothing.
It doesn't display a message like key invalid.
I guess this string is used for the option licensing maybe?

Pressing SHIFT + TRANSIENT will also ask for a password. 882323 will not work. any clues what this password is?

Thanks
 

Offline mvanocht

  • Contributor
  • Posts: 10
  • Country: us
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #31 on: June 07, 2017, 07:24:08 pm »
I also have this model analyzer.  Documentation says they offered an "Option 5" Audio Demodulation.  I wondered if it was just enabled through software or if it included some hardware.  I had the front panel off a few days ago, and found the internal speaker is there, along with the headphone jack, and everything is wired up and therefore I think it is just a software option.
 

Offline 3g57

  • Regular Contributor
  • *
  • Posts: 72
  • Country: ua
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #32 on: November 22, 2020, 08:54:54 am »
Hello! hafrse
look at write protection here!
Protection is enabled by default!  ;)





Best Regards, 3G57
« Last Edit: November 22, 2020, 09:01:05 am by 3g57 »
 
The following users thanked this post: jjoonathan

Offline 3g57

  • Regular Contributor
  • *
  • Posts: 72
  • Country: ua
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #33 on: November 22, 2020, 12:08:35 pm »
Hi there!
I uploaded some datasheets to IC of the Advantest R3273 PC board




1   mPC4072 J-FET INPUT LOW-NOISE DUAL OPERATIONAL AMPLIFIER
    https://datasheet.octopart.com/UPC4072C-NEC-datasheet-101880.pdf

2   DM74LS245 3-STATE Octal Bus Transceiver
    https://pdf1.alldatasheet.com/datasheet-pdf/view/51057/FAIRCHILD/74LS245.html

3   74VHC245   Octal Bidirectional Transceiver with 3-STATE Outputs
    https://pdf1.alldatasheet.com/datasheet-pdf/view/50653/FAIRCHILD/74VHC245.html

4   AD7538KR LC2 MOS Microprocessor-Compatible 14-Bit DAC
    https://eu.mouser.com/datasheet/2/609/AD7538-877338.pdf

5   LT1027/E5   LT1027 is a precision reference with extra-low drift
    https://www.analog.com/media/en/technical-documentation/data-sheets/1027fcs.pdf

6   LV245A SN74LV245AT   OCTAL BUS TRANSCEIVER WITH 3-STATE OUTPUTS
    https://pdf1.alldatasheet.com/datasheet-pdf/view/192063/TI/LV245A.html

7   AC125  74AC125  Quad Buffer with 3-STATE Outputs
    https://pdf1.alldatasheet.com/datasheet-pdf/view/50119/FAIRCHILD/74AC125.html

8   HD153110  Color Palette with Triple 8-bit DA Converter
    http://www.datasheetq.com/datasheet-download/883579/1/Hitachi/HD153110

9   M38802M2-220HP  Microcomputer for key-board control(Mask ROM:ROM-8K,RAM-384K)
    https://www.datasheets360.com/pdf/-6417315541921908420

10  M514260C-60J  262,144-Word x 16-Bit DYNAMIC RAM
    https://datasheet4u.com/datasheet-parts/MSM514260C-datasheet.php?id=505572

11  HM538253BTT-7  262,144-word x 8-bit Multiport CMOS Video RAM
    https://www.datasheets360.com/pdf/3616523693885072771

12   SIM-65802GN105-1  Specialized microcircuit

13  SH-2 HD6417604F28 (Hitaсhi) Renesas 32-Bit RISC Microcomputer SuperHTM RISC engine Family
    https://www.renesas.com/us/en/doc/products/mpumcu/001/rej09b0171_superh.pdf
    http://www.farnell.com/datasheets/44795.pdf

14  RTC-62423   4-bit REAL TIME CLOCK MODULE
    https://pdf1.alldatasheet.com/datasheet-pdf/view/272654/EPSONTOYOCOM/RTC-62423.html

15  7673  ICL7673 Automatic Battery Back-Up Switch
    https://www.renesas.com/jp/ja/www/doc/datasheet/icl7673.pdf

16  TC551001CF-70L  8-BIT STATIC RAM   
    https://pdf1.alldatasheet.com/datasheet-pdf/view/31713/TOSHIBA/TC551001CF-70L.html

17  74VHC595  8-Bit Shift Register with Output Latches
    https://www.mouser.com/datasheet/2/149/74VHC595-101446.pdf

18  74AC157  Quad 2-Input Multiplexer
    https://pdf1.alldatasheet.com/datasheet-pdf/view/50125/FAIRCHILD/74AC157.html

19  74VHC04  HEX INVERTER
    https://pdf1.alldatasheet.com/datasheet-pdf/view/21830/STMICROELECTRONICS/74VHC04.html

20  74LV245A  Octal bus transceiver; 3-state
    https://assets.nexperia.com/documents/data-sheet/74LV245A.pdf

21  HM5165165FTT6  64M EDO DRAM (4-Mword x 16-bit) 8k refresh/4k refresh
    https://datasheetspdf.com/pdf-file/526944/HitachiSemiconductor/HM5165165F/1

22  TMS99C14AFNL   GPIB TO SERIAL CONVERTER
    https://accelconf.web.cern.ch/ica05/proceedings/pdf/P3_047.pdf

23  LSI1032E   High-Density Programmable Logic
    https://datasheetspdf.com/pdf-file/684967/LatticeSemiconductor/LSI1032E/1

24  PC97338V   ACPI 1.0 and PC98/99 Compliant SuperI/O
    https://pdf1.alldatasheet.com/datasheet-pdf/view/9318/NSC/PC97338VJG.html

25  MC68340PV25E    High-performance 32-bit integrated processor with direct memory access (DMA)
    https://www.datasheets360.com/pdf/-1055304228673071662

26  29F400BC-90PF1N   4M (512K x 8/256K x 16) BIT FLASH MEMORY
    https://datasheetspdf.com/pdf-file/1114535/Fujitsu/MBM29F400BC-90/1

27  SIM-190G12S (90G12F0009)  Specialized microcircuit


28  SIM-65801GD082-1    Specialized microcircuit


29  74VHC595   8-Bit Shift Register with Output Latches
    https://www.mouser.com/datasheet/2/149/74VHC595-101446.pdf
   
30  MB88346B   R-2R TYPE 8-BIT D/A CONVERTER WITH OPERATIONAL AMPLIFIER OUTPUT BUFFERS
    http://pdf.datasheetcatalog.com/datasheet/FujitsuMicroelectronics/mXrtvyw.pdf

31  MAX4581  Low-Voltage, CMOS Analog Multiplexers/Switches
    https://datasheets.maximintegrated.com/en/ds/MAX4581-MAX4583.pdf

32  QUALCOMM CD90-24093-1   Specialized microcircuit


33  HI5735KCB   12-Bit, 80 MSPS, High Speed Video D/A Converter
    https://www.renesas.com/cn/en/www/doc/datasheet/hi5735.pdf

34  LM6172   Dual High Speed, Low Power, LowDistortion,Voltage Feedback Amplifiers
    https://pdf1.alldatasheet.com/datasheet-pdf/view/9004/NSC/LM6172.html

35  SN75160B   Octal General-Purpose Interface Bus Transceiver (Rev. B)
    https://www.ti.com/lit/ds/symlink/sn75160b.pdf

36  SN75162B  OCTAL GENERAL-PURPOSE INTERFACE BUS TRANSCEIVERS
    http://pdf.datasheetcatalog.com/datasheet/texasinstruments/sn75162b.pdf

37  LM2991S   Negative Low Dropout Adjustable Regulator
    https://pdf1.alldatasheet.com/datasheet-pdf/view/545554/TI/LM2991S.html

38  LM2990S (-5.0V)  Negative Low-Dropout Regulator   
    https://www.promelec.ru/fs/sources/e6/fd/69/62/d4f1d60f4e4c09941d668328.pdf

39  UPC1663  DC to VHF WIDEBAND DIFFERENTIAL INPUT AND OUTPUT AMPLIFIER IC
    https://pdf1.alldatasheet.com/datasheet-pdf/view/6675/NEC/UPC1663.html

40  961EUA (MAX961EUA-T)   Single/Dual/Quad, Ultra-High-Speed, +3V/+5V, Beyond-the-Rails Comparators
    https://pdf1.alldatasheet.es/datasheet-pdf/view/434192/MAXIM/MAX961EUA-T.html

41  74VHC14  Hex Schmitt Inverter
    https://pdf1.alldatasheet.com/datasheet-pdf/view/223002/FAIRCHILD/74VHC14.html

42  MC145583V  3.3 VOLT ONLY DRIVER / RECEIVER WITH AN INTEGRATED STANDBY MODE
    https://pdf1.alldatasheet.com/datasheet-pdf/view/158429/MOTOROLA/MC145583VF.html

43  K6T1008C2E-GB70   128Kx8 bit Low Power CMOS Static RAM
    https://pdf1.alldatasheet.com/datasheet-pdf/view/37113/SAMSUNG/K6T1008C2E-GB70.html

44  M51953B   Voltage Detecting, System Resetting IC Series
   https://pdf1.alldatasheetru.com/datasheet-pdf/view/196325/RENESAS/M51953B.html
   

   


Please pay attention to item
#15  ICL7673 Automatic Battery Back-Up Switch
How does it work:
This item give You possibility for Hot-replace the battery. You should made this very careful.




Best regards,  3G57
« Last Edit: November 23, 2020, 11:34:06 am by 3g57 »
 
The following users thanked this post: jjoonathan, hua

Offline romal76

  • Contributor
  • Posts: 23
  • Country: ru
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #34 on: April 11, 2022, 11:57:35 am »
Anyone please upload Advantest R3273 CF card image file again, old link's is not working.
 

Offline chick0n

  • Regular Contributor
  • *
  • Posts: 92
  • Country: de
Re: Advantest R3273 100Hz-26.5GHz Spectrum Analyzer repair
« Reply #35 on: October 26, 2022, 10:33:57 am »

There was a defect bias circuit for the first mixer which put DC voltage in the first stage and degraded the signal greatly.


I think i have the same problem. The DC is everywhere in the circuit. This cant be right.

Could you explain the repair a bit more?
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf