Author Topic: Multimeter spreadsheet (80 manufacturers and over 400 meters listed)  (Read 869440 times)

0 Members and 5 Guests are viewing this topic.

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #300 on: January 07, 2016, 12:32:52 pm »

I will reply others later.
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #301 on: January 08, 2016, 01:50:41 am »
LoZ
  TY720 has it. Other TY710, TY530, TY520 don't.
Again, no LoZ on TY720. Read the manual.

This is the manual.

 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #302 on: January 08, 2016, 01:52:24 am »
This is for HIOKI.

Accuracy
  DT4252 and DT4253 -> 30/-/5
  DT4256 -> 30/-/3
4252 600mV 20/-/5
4253 600mV 50/-/5

This is the images for accuracy.



 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #303 on: January 08, 2016, 07:44:27 am »
LoZ
  TY720 has it. Other TY710, TY530, TY520 don't.
Again, no LoZ on TY720. Read the manual.

This is the manual.


That's not a loZ function, as used to cancel false readings due to ghost voltage presence (capacitive coupling)
Read up on ghost voltage if you don't know what it is. (No 3 on the first post of this thread)
« Last Edit: January 08, 2016, 07:46:05 am by Wytnucls »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #304 on: January 08, 2016, 07:50:32 am »
UL listing for 530 and 520 is now correct (2016)
The document I referred to was from Dec 2011.
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #305 on: January 08, 2016, 10:05:57 am »
This is for HIOKI.

Accuracy
  DT4252 and DT4253 -> 30/-/5
  DT4256 -> 30/-/3
4252 600mV 20/-/5
4253 600mV 50/-/5

This is the images for accuracy.




The Hioki 4252 has a special high accuracy 600mV DC range, 0.2%+5.
« Last Edit: January 08, 2016, 10:07:36 am by Wytnucls »
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #306 on: January 09, 2016, 09:35:43 am »
You are correct, the TY710 doesn't come with VFD. The other meters do. All the meters have port blanking. The spreadsheet will be corrected later.
« Last Edit: January 09, 2016, 09:41:55 am by Wytnucls »
 

Offline shempe

  • Contributor
  • Posts: 29
  • Country: de
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #307 on: January 16, 2016, 06:12:05 pm »
One of my favourite Bench DMM Meters .

maybe you can include it.

Texio DL-2060G  http://www.texio.co.jp/en/03prod_01_04_08.html

 

Offline PedroDaGr8

  • Super Contributor
  • ***
  • Posts: 1283
  • Country: us
  • A sociable geek chemist
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #308 on: January 19, 2016, 12:04:39 am »
One of my favourite Bench DMM Meters .

maybe you can include it.

Texio DL-2060G  http://www.texio.co.jp/en/03prod_01_04_08.html

Pretty certain that is just a rebadge of the Picotest M3500A, just with a different color scheme
The very existence of flamethrowers proves that some time, somewhere, someone said to themselves, "You know, I want to set those people over there on fire, but I'm just not close enough to get the job done." -George Carlin
 

Offline shempe

  • Contributor
  • Posts: 29
  • Country: de
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #309 on: January 19, 2016, 12:19:59 am »
Quote
Pretty certain that is just a rebadge of the Picotest M3500A, just with a different color scheme

Yes indeed, it is the same.

Im sorry, I was not aware of the Picotest M3500A.

And the Picotest M3500A look´s like the Keithley 2100

Curious question, are the Firmware Update´s  from the other Unit´s usable with my Unit ?
« Last Edit: January 19, 2016, 12:33:42 am by shempe »
 

Offline PedroDaGr8

  • Super Contributor
  • ***
  • Posts: 1283
  • Country: us
  • A sociable geek chemist
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #310 on: January 21, 2016, 01:28:35 am »
Quote
Pretty certain that is just a rebadge of the Picotest M3500A, just with a different color scheme

Yes indeed, it is the same.

Im sorry, I was not aware of the Picotest M3500A.

And the Picotest M3500A look´s like the Keithley 2100

Curious question, are the Firmware Update´s  from the other Unit´s usable with my Unit ?

The Keithley 2100 is also a M3500A rebadge, as you noticed.

I have no clue on the firmware to be honest. It would be a crap shoot and you risk bricking your meter.
The very existence of flamethrowers proves that some time, somewhere, someone said to themselves, "You know, I want to set those people over there on fire, but I'm just not close enough to get the job done." -George Carlin
 

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 6071
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #311 on: February 03, 2016, 11:57:06 am »
Wytnucls, one detail regarding battery life on Fluke's 175/177/179: in both manuals I have (an original from my 179 and a copy I downloaded from their website), I see it listed as being 400 hours. Perhaps you have a different source?
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #312 on: February 03, 2016, 12:49:57 pm »
I can't remember where I got that figure from. I see Element 14 has it as 200hrs also. I will change it to 400hrs at the next revision. Thanks for your input.
http://in.element14.com/fluke/fluke-177/multimeter-digital-hand-held-6000/dp/3558680?mckv=svf5CnTY5|pcrid|86562217480&CMP=KNC-GOO-SKU-G30-FLU
« Last Edit: February 03, 2016, 01:15:14 pm by Wytnucls »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #313 on: February 10, 2016, 07:33:51 am »
I can't remember where I got that figure from. I see Element 14 has it as 200hrs also. I will change it to 400hrs at the next revision.
Wrt to the 200 and 400 hour confusion ....

https://www.eevblog.com/forum/testgear/fluke-77iv-and-179-differences-with-teardown-photos/msg804693/#msg804693
 

Offline hammy

  • Supporter
  • ****
  • Posts: 465
  • Country: 00
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #314 on: February 11, 2016, 10:36:56 pm »
Thank you!  :-+
 

Offline willocks1718

  • Newbie
  • Posts: 3
  • Country: us
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #315 on: February 15, 2016, 01:14:37 am »
I am narrowing down your multimeter list to purchase a new one soon and was looking at the Brymen BM857.  In column H (uA, best DC current resolution) it says 10nA but instead either green, white, yellow, or red for high, average, low or danger, it is a tan color that looks like it was maybe missing feature.  However the tan color isn't the same as with your legend so figured it meant something else.  What does that color mean in your table?
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #316 on: February 15, 2016, 05:07:34 pm »
Not tan, but light green, to differentiate from meters with a 1nA resolution (full green).
Use MS Excel viewer to pick up the right colors.
 

Offline broz

  • Regular Contributor
  • *
  • Posts: 50
  • Country: ca
  • Pronounced like the "bros." in "Super Mario bros."
    • My Wordpress Page
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #317 on: February 16, 2016, 05:31:37 pm »
Thank you for this! I'm in the process of researching the purchase of my first DMM and was only aware of a few brands, and a few models for each of those brands. This spreadsheet blew my mind as well as opened my eyes!
Slowly but surely making my way through EE school
 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #318 on: February 18, 2016, 12:51:41 pm »
Family picture of all TRMS 6000+ count multimeters (new and old) :
 

Offline fredboivin

  • Regular Contributor
  • *
  • Posts: 52
  • Country: ca
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #319 on: February 21, 2016, 12:54:06 am »
I can't open the new file, it says it's been encrypted with an encoding that's not available in my country. Am I the only one? I was able to open the previous ones.

EDIT: Just open it in Google Spreadsheet.
« Last Edit: February 21, 2016, 12:57:48 am by fredboivin »
 

Offline fanOfeeDIY

  • Supporter
  • ****
  • Posts: 412
  • Country: jp
    • YouTube Channel
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #320 on: February 22, 2016, 09:54:28 am »
This is about Yokogawa.

All TY720, TY710, TY530, TY520 have:

Jack
  Physical port shutter.

Hi Wytnucls,

Hope this helps.

http://tmi.yokogawa.com/files/uploaded/TY700_User_Manual.pdf
http://tmi.yokogawa.com/files/uploaded/IMTY530_E_030.pdf







« Last Edit: February 22, 2016, 10:19:40 am by fanOfeeDIY »
 

Offline willocks1718

  • Newbie
  • Posts: 3
  • Country: us
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #321 on: February 27, 2016, 02:01:54 am »
Hey, thanks for this resource, the spreadsheet is great.  I saw the how to choose a DMM link at the top but I personally would find a simpler more concise table easier and more helpful.  I am looking for a multimeter for some beginner electronics work and some basic home mains wiring.  So if there was somewhere linked with your master spreadsheet that said for electronics you would recommend a meter this accurate or with at least this DC current resolution. And home wiring you should look for DMMs with these values. 

Can you recommend some values from your table you think would be the most important to have for basic electronics and some home mains (120 V) work?

Sent from my SM-G900V using Tapatalk

 

Offline WytnuclsTopic starter

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #322 on: February 27, 2016, 11:10:19 am »
If you don't know what to buy, get the EEVBlog Brymen 235. Safe, useful and priced right.
 

Offline hammy

  • Supporter
  • ****
  • Posts: 465
  • Country: 00
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #323 on: February 27, 2016, 03:35:02 pm »
If you don't know what to buy, get the EEVBlog Brymen 235. Safe, useful and priced right.

+1  :-+
 

Offline mos6502

  • Frequent Contributor
  • **
  • Posts: 537
  • Country: aq
Re: Multimeter spreadsheet (38 brands and 210 meters listed)
« Reply #324 on: February 28, 2016, 04:28:42 am »
Here's a correction: I just found out that the backlight on the Fluke 87V does NOT turn off automatically! How awesome is that? I get to decide when the backlight turns off! I didn't even know such a meter existed!
for(;;);
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf