I'm looking for some kind of multi-channel multimeter, controllable via USB, with voltage (and ideally current) measurement.

For voltages, they can all have a common reference. For currents, it would need to have a floating connection. But if such a beast is hard to find or doesn't exist, I can do with just voltage measurement - I'll add some current-to-voltage circuitry.

- 4 channels min. for voltage measurement, ~0.1% max error, range from a few tens of µV to about 5V.
- 1 channel min. for current measurement, ~0.5% max error, range from a few tens of nA to about 1mA. As said, this part can be done without if hard to find in a single device - then at least 5 channels of voltage measurement would be required.

Ideally, the USB interface would come with a simple API to program the device on potentially Linux, macOS and Windows.

Some could say that "USB oscllioscopes" might do here (if just for the voltage measurements), but I don't know of any (at least reasonably priced) that would have max. 0.1% error in amplitude, and support low input ranges down to a few tens of µV. Hence why I called that more "multi-channel multimeter".

Some others could say, just buy several multimeters with an USB interface. I would prefer a single device.

Regarding price, $1000 to $1500 would be good.

Note: I know and have used the LabJack T7 in the past: https://labjack.com/products/labjack-t7
it was good for production testing of a simple product, but I'm looking for something a bit more polished and more accurate.
 

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Multi-channel multimeter with USB interface?
« Reply #1 on: July 13, 2023, 10:02:10 pm »
Hi,

What about a multimeter with scannercard ?

Siglent for example offer a scannercard for their 5.5 digit meter, with 12 multipurpose channels and 4 current channels.

"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6982
  • Country: ca
  • Non-expert
Re: Multi-channel multimeter with USB interface?
« Reply #2 on: July 13, 2023, 10:09:37 pm »
Did you see the T8? https://labjack.com/products/t8
8 isolated channels.

Then there is the DMM6500 with DIY/used scanner card, but that is above your budget.
Then the older Agilent DAQ stuff but it doesn't have USB (34970A).

The siglent card Martin mentioned seems quite reasonably priced
https://siglentna.com/product/sdm3055-sc/ ($745)
https://siglentna.com/digital-multimeters/sdm3065x-6-%C2%BD-digits-dual-display-digital-multimeters/ ($995)
https://siglentna.com/product/scan-card-option/
Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 

Offline SiliconWizardTopic starter

  • Super Contributor
  • ***
  • Posts: 15172
  • Country: fr
Re: Multi-channel multimeter with USB interface?
« Reply #3 on: July 13, 2023, 10:26:50 pm »
Thanks.

- I'm going to have a look at the Siglent stuff.
- I did see the T8. Might be a candidate. Plus is: reasonably priced, I already know their products. Minus: not quite as sturdy as regular lab equipment.
- I'm going to have alook at the DMM6500 anyway as well, in case I can get a higher budget for this.
 

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Multi-channel multimeter with USB interface?
« Reply #4 on: July 13, 2023, 10:37:06 pm »
Another thing:

We (work) have two of the hioki datalogger, super small size, 10 isolated channels, via software you can control/monitoring up to 50 channels.
Good accuracy, but only voltage or temperature input, so for current measures a shunt would be needed.

https://www.farnell.com/datasheets/2847482.pdf
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multi-channel multimeter with USB interface?
« Reply #5 on: July 13, 2023, 10:47:41 pm »
I'm going to have a look at the Siglent stuff.
SC models must be purchased as such as the SC option cannot be added later.
https://siglentna.com/wp-content/uploads/2017/12/SC1016_Datasheet_DS60030-E02A.pdf

For USB or LAN use, SDM models can be used with free EasyDMM SW or with SCPI scripts.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline SiliconWizardTopic starter

  • Super Contributor
  • ***
  • Posts: 15172
  • Country: fr
Re: Multi-channel multimeter with USB interface?
« Reply #6 on: July 13, 2023, 10:54:55 pm »
Saw that. Thanks.
The scan card seems out of stock though: https://siglentna.com/product/scan-card-option/
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 7997
  • Country: us
Re: Multi-channel multimeter with USB interface?
« Reply #7 on: July 13, 2023, 11:12:23 pm »
The scan card seems out of stock though: https://siglentna.com/product/scan-card-option/

The -SC will come with a scan card.  You'd only need extras if you wanted to prewire them and exchange them for different setups.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: tautech, SiliconWizard

Offline switchabl

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: de
Re: Multi-channel multimeter with USB interface?
« Reply #8 on: July 13, 2023, 11:23:46 pm »
NI USB-6341 or cDAQ-V1100 maybe? They have multiple ranges (+-10V down to +-200mV full scale) and 8 or 16 differential inputs, unfortunately no current inputs. You could build a cDAQ system with an additional current module but that would be way over your budget.
https://www.ni.com/docs/en-US/bundle/usb-6341-specs/page/specs.html
https://www.ni.com/docs/en-US/bundle/ni-9205-specs/page/specs.html

Software/API is typical NI, somewhat bloated but works well and has been very stable for me. Documentation is decent. DAQmx API is available on Windows and Linux for C and .NET. Python bindings and Matlab toolbox is also available. (And LabVIEW of course  :-X)
https://www.ni.com/docs/en-US/bundle/ni-daqmx/page/daqhelp/daqhelp.html
https://www.ni.com/docs/en-US/bundle/ni-daqmx-c-api-ref/page/cdaqmx/help_file_title.html
« Last Edit: July 14, 2023, 12:04:59 am by switchabl »
 

Offline thm_w

  • Super Contributor
  • ***
  • Posts: 6982
  • Country: ca
  • Non-expert
Re: Multi-channel multimeter with USB interface?
« Reply #9 on: July 13, 2023, 11:24:58 pm »
Semi off topic:

From Daves teardown they seem to not have the scanner card connection installed in his 3055 unit at all. So thats presumably why its separate, saves a few dollars of parts.
Here is a teardown of one with the scan card connector board (looks to be the same connector as used on the DMM6500/keithley units):

Profile -> Modify profile -> Look and Layout ->  Don't show users' signatures
 
The following users thanked this post: SiliconWizard

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multi-channel multimeter with USB interface?
« Reply #10 on: July 13, 2023, 11:45:47 pm »
Saw that. Thanks.
The scan card seems out of stock though: https://siglentna.com/product/scan-card-option/
You can get additional scan cards if/as required.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multi-channel multimeter with USB interface?
« Reply #11 on: July 13, 2023, 11:50:38 pm »
From Daves teardown they seem to not have the scanner card connection installed in his 3055 unit at all. So thats presumably why its separate, saves a few dollars of parts.
Yep, this.

There's quite a few extra bits and bobs required that includes the mounting tray and fastener HW for the SC.
Did some considerable research on this some years back.

It quickly became obvious if you want the SC capability purchase that model.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: thm_w

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29200
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Multi-channel multimeter with USB interface?
« Reply #12 on: July 13, 2023, 11:54:30 pm »
With some imagination, creativity and a 3D printer you can do this:


https://www.eevblog.com/forum/testgear/siglent-new-bench-dmm-sdm3055/msg3070270/#msg3070270
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: ch_scr, Martin72, bdbell

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6630
  • Country: de
  • Testfield Technician
Re: Multi-channel multimeter with USB interface?
« Reply #13 on: July 13, 2023, 11:56:17 pm »
WTF ???!!!???  ;D
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online Someone

  • Super Contributor
  • ***
  • Posts: 4856
  • Country: au
    • send complaints here
Re: Multi-channel multimeter with USB interface?
« Reply #14 on: July 14, 2023, 12:12:56 am »
- I'm going to have alook at the DMM6500 anyway as well, in case I can get a higher budget for this.
We're having troubles controlling the DMM6500 over USB. No way to reset the system state or clear buffers, requiring physical power cycle to recover. Support did not engage or consider the provided test cases, and basically told us to use it the way they imagined it would be (ie, never have any errors or work between reboots/disconnects of the host).
So if automation/remote over USB is the primary goal probably best to look elsewhere.
 
The following users thanked this post: SiliconWizard

Offline SiliconWizardTopic starter

  • Super Contributor
  • ***
  • Posts: 15172
  • Country: fr
Re: Multi-channel multimeter with USB interface?
« Reply #15 on: July 14, 2023, 01:24:22 am »
I've also seen the Pico ADC-24. Downside is that its inputs are limlited to +/-2.5V.
 

Online Someone

  • Super Contributor
  • ***
  • Posts: 4856
  • Country: au
    • send complaints here
Re: Multi-channel multimeter with USB interface?
« Reply #16 on: July 14, 2023, 03:11:51 am »
Software/API is typical NI, somewhat bloated but works well and has been very stable for me. Documentation is decent. DAQmx API is available on Windows and Linux for C and .NET. Python bindings and Matlab toolbox is also available. (And LabVIEW of course  :-X)
There are also the "cheap" USB-600x models from NI. Was sure there was a tiny little modular USB chassis that wasn't far off in price but can't find it and NI are aggressively pushing all links to their shop which doesn't show any content in my region  :-//
 

Offline switchabl

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: de
Re: Multi-channel multimeter with USB interface?
« Reply #17 on: July 14, 2023, 11:45:02 am »
The USB-600x don't have the low voltage range which the OP wants. I completely agree about the NI website, it's absolutely impossible to find anything there. I swear it became even worse when they switched to the new design (green logo) a couple of years ago.

I also realize now that Linux support doesn't include the USB DAQs. So if this is a priority, you need to look elsewhere (ideally for something with USBTMC or CDC that doesn't need proprietary driver).
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11757
  • Country: us
Re: Multi-channel multimeter with USB interface?
« Reply #18 on: July 14, 2023, 11:53:28 am »
I have a cheap LabJack as well that's very old.  Have you looked recently to see if they offer  more accurate products that your current one?   

Offline SiliconWizardTopic starter

  • Super Contributor
  • ***
  • Posts: 15172
  • Country: fr
Re: Multi-channel multimeter with USB interface?
« Reply #19 on: July 14, 2023, 08:18:13 pm »
I have a cheap LabJack as well that's very old.  Have you looked recently to see if they offer  more accurate products that your current one?

thm_w mentioned the T8. The T7-Pro was already not too bad, but not isolated. The T8 is, which is a nice plus. The specs (if they are met) look ok for my needs. Had a deeper look though, and they say the T8 is in some "pre-release" state, with some features unfinished software-wise as far as I got it. Might be a bit too early to go with that.
 