My experience is that EXACTLY with the likes of Fluke, it is NOT safe to read specification any other way than corporate lawyers thought when they gave final edit on the specification documents.
Flukes safety information is very clear on the 116:
Do not apply more than the rated voltage, as marked on Meter, between terminals or between any terminal and earth ground.
And both the meter and the specifications says 600V
"Caution" statement identifies conditions and actions that could damage the Meter or the equipment under test.
To avoid possible electric shock or personal injury, follow these guidelines:
•
Use the Meter only as specified in this manual or the protection provided by the Meter might
be impaired.
• Do not use the Meter or test leads if they appear damaged, or if the Meter is not operating
properly.
•
Always use proper terminals, switch position, and range for measurements.• Verify the Meter's operation by measuring a known voltage. If in doubt, have the Meter
serviced.
• Do not apply more than the rated voltage, as marked on Meter, between terminals or
between any terminal and earth ground. So reading that, I can only see that you
shouldn't apply
more than 600V when measuring
VOLTS.
So, if you connect 600V source with no current limit to the meter
contrary to the
proper measurement procedure in manual to
a wrong setting of uA, then
all bets are off..
All they guarantee is safety according to :
Complies with ANSI/ISA 82.02.01 (61010-1) 2004,
CAN/CSA-C22.2 No 61010-1-04, UL 6101B (2003) and
IEC/EN 61010-1 2nd Edition for measurement Category III,
600 V, Pollution Degree 2, EMC EN61326-1
Which mostly means it won't kill you but they don't even mention if it gets nonfunctional...
Don't get me wrong, it might as well be that it would routinely survive 600V on uA range, it's just Fluke carefully avoided making any promises it would.. Which is suspicious because that would be great bragging point...