Author Topic: Making accurate low frequency measurements with your scope  (Read 9989 times)

0 Members and 3 Guests are viewing this topic.

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #25 on: December 15, 2015, 12:51:49 am »
 :--   Need to see if it really is the INL or something else.  If it is, I plan to model it with a polynomial. 

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #26 on: December 15, 2015, 04:47:56 am »
Looks repeatable.  I ran 3 sweeps on each channel.   Shown is INL in counts for channel 0 in counts (again, after oversample).

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #27 on: December 15, 2015, 04:52:21 am »
Notice how that graph looked very similar to the single sweep I had made the other day....

Showing INL for channel 1.   

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #28 on: December 16, 2015, 12:35:10 am »
Imagine trying to fit that stupid INL curve...  Table lookup.   Green,  least squares.  White is raw (after 256X oversample). Red is compensated.   :-+  Some hope ....  almost.....

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #29 on: December 16, 2015, 12:39:17 am »
Shown is INL before (green) and after (red) compensation.  Raw INL was filtered with IIR (0-phase) using Butterworth to create the lookup table.   Starts out really good, then all hell breaks loose.   Suspect better filter could remove some of it but really, there are bigger problems with it...

Offline joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 12033
  • Country: us
Re: Making accurate low frequency measurements with your scope
« Reply #30 on: December 16, 2015, 12:45:05 am »
And here is the money shot.  It would be very interesting to know more about the design of this DSO.   That's not the input signal doing that.  Why DNL goes in the crapper at a volt, no idea.  Good question for LeCroy designers.

So, in the end, there are gains to be had.  This will help out in my power measurement program but if you want good data, don't use a scope... :-DD

Hope it was helpful.