I'm missing the possibility to trigger on the waveform generator. Also, is the waveform generator clock derived from the external reference clock input? And what is the frequency setting accuracy of the waveform generator?
We plan to support triggering on the built-in waveform generator with an upcoming update. The datasheet will be updated with new features as soon as they are implemented and tested.
The sampling as well as the waveform generator are based on the same reference clock. You can choose between an internal or external reference ref-clock source, which then applies to both the sampling and the generator.
The frequency setting accuracy of the waveform generator is 1 µHz.
Edit: another thing that doesn't make sense to me is the deskew resolution. It is limited to the sample time interval which is 800ps minimum. That makes no sense to me; deskew should be adjustable in picoseconds given the 100ps channel-to-channel skew, minimum of 200ps/div and the trigger jitter.
Yes, the deskew function is limited to the full ADC sampling rate. Maybe later we will be able to relate it to the interpolated sampling rate, but this is not yet on our roadmap for the next few month.
Will it be possible to change the sampling rate manually or is just set automatically according to the number of channels used and the horizontal time scale?
The AD converters operate continuously at the highest sample rate (1 GSa/s or 1.6 GSa/s). If the set (or automatically selected) memory depth is not sufficient to save all samples over the selected time scale, the sample rate is automatically reduced before saving. A manual setting of the sample rate is currently not provided.
I'm wondering why there is no picture from the backside. I would like to see which connectors are available and how the optional AWG will be mounted.
I attach two quick photos of one of the prototypes. One picture shows the back of the Magnova with the cover removed before installing the module, one shows the back after installation.
The production units will look better than this prototype.
![Wink ;)](https://www.eevblog.com/forum/Smileys/default/xwink.gif.pagespeed.ic.cldandycH0.png)
The nameplates and the final paintwork are missing, but it should be enough to show the assembly of the generator.
I don't understand the thing with the 320 Mpts and the 12 divs. When do you have 80 Mpts with 4 channels?
With 4 channels and 5 ms/div (=50 ms/screen), you have 60 mpts on each channel. With a slightly higher time scale of 6.666 ms/div (~80 ms/screen), you have 80 mpts on each channel. Of course, this time scale is somewhat unusual and is not in the 1-2-5 steps.
Another possibility would be to set the memory from "Auto" to
60 80 Mpts at e.g. 5 ms/div. Then the Magnova will sample and store 16 divs (12 on the screen and 3 to the left and right of it).
But even if you don't want that, 320 mpts helps for example to increase the number of saved waveforms in the history (by 50% compared to a memory limit of 240 mpts, which will fit more to 12 divs/screen).