There are several others at least on eevblog, I am one. My 2c to slburris' input.
I haven't calibrated my Fluke 85-I since 1989. I recently compared it against a calibrated voltage source. I used an HP 3456A lab DVM to test it across all its Vdc ranges and its accurate to 0.04% Vdc, at worst. At best it was 0.01%. I also did these measurements across a continuous 2 hour timespan, to check for short term drift, and it was rock steady. I also tracked an c1981 DMM I made as a kit that has a user adjustable calibration, rated at 0.1%, and found that it really was more 0.5% and only 0.1% at the 1V scale. I was testing my Mastech power supply, and comparing its built in DVM the manual claims is 0.5%, well it ranged from 0.3%-4.0%. Also, the value would drift 1 digit while the others were rock steady.
This stability isn't something you'll read on the advertising literature when comparing DMM. Fluke also understate specs, Fluke 85-I is rated at 0.5%+1 digit across all Vdc ranges, actual is more ~10x better.
Unlike excellent competitors from Agilent or Gossen Metrawatt, some Fluke models such as the 80s series, are still made, almost the same, continuously, for 20+ years. So there is a good chance you'll get a Fluke with a similar real long term stability, not just a lab tested 'accelerated aging' test.
Agilent has been in-out of the
handheld DMM business, so there is inconsistency. Gossen is closer to Fluke as competitor, quality wise, in hand held DMM.
I have two Fluke 6.5 digit bench meters from the 1980's and they
agree with each other within a few counts of the last digit generally
and also agree with my Fluke 87-V to within the limits of that meter,
so I have to say I'm quite impressed with how well these meters age.
I heard stories of Flukes years since their last calibration that are still
well within their specifications, but I have no way to confirm that.
Scott