Author Topic: Repair of a Fluke 87 lll  (Read 1052 times)

0 Members and 1 Guest are viewing this topic.

Offline QmavamTopic starter

  • Regular Contributor
  • *
  • Posts: 190
  • Country: us
Repair of a Fluke 87 lll
« on: June 05, 2023, 12:12:57 pm »
Hi,
  I have a Fluke 87 lll that had over voltage AC applied.
I checked all the MOVs and they seem fine.
 Any experience fixing an 87 lll that had too much voltage applied?
 What should I look at?
                   Thanks, Mikek
 

Offline BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1316
  • Country: us
Re: Repair of a Fluke 87 lll
« Reply #1 on: June 05, 2023, 02:01:11 pm »
What does it do now? It doesn't power up? Overvotage AC applied how high was the voltage and in which mode the meter was in?
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17045
  • Country: 00
Re: Repair of a Fluke 87 lll
« Reply #2 on: June 05, 2023, 07:33:03 pm »
Any experience fixing an 87 lll that had too much voltage applied?

How much is "too much"?

What should I look at?

If it powers on and responds then try removing the two clamping transistors to see what happens.
 

Offline BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1316
  • Country: us
Re: Repair of a Fluke 87 lll
« Reply #3 on: June 05, 2023, 08:10:20 pm »
The company I work for had a number of 87III and they tend to have problem with the display. None had any other problems.
 

Offline Ugur

  • Regular Contributor
  • *
  • Posts: 87
  • Country: tr
Re: Repair of a Fluke 87 lll
« Reply #4 on: June 06, 2023, 05:58:01 am »
Turn it into Ohms and do you see OL on the screen? If not, check the varistors.
 

Online EEVblog

  • Administrator
  • *****
  • Posts: 38411
  • Country: au
    • EEVblog
Re: Repair of a Fluke 87 lll
« Reply #5 on: June 06, 2023, 07:25:28 am »
Does the display still work but it just reads zero?
Check the series input resistor. I think it's a fusible type.
 
The following users thanked this post: paulbt

Offline QmavamTopic starter

  • Regular Contributor
  • *
  • Posts: 190
  • Country: us
Re: Repair of a Fluke 87 lll
« Reply #6 on: June 19, 2023, 03:21:52 pm »
It turned out to be one MOV that was defective.
            Thanks all, Qmavam
 