Any recommendation for a high impedance / low capacitance scope probe for testing 5V open-drain outputs with high pull-up resistance, e.g. 100k, such that its effect on the observed signal is minimal, as opposed to existing 10M/15pF probes supplied with scope?

TIA.
« Last Edit: April 09, 2019, 02:44:30 pm by euzer »
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16854
  • Country: 00
For 5V you could build a little board with an op-amp (voltage follower) and look at the ouput.
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27323
  • Country: nl
    • NCT Developments
Less than 15pf of input capacitance will be hard to beat. Perhaps look for a probe with a low capacitance at the tip. 10M Ohm isn't a problem. With an impedance at the load the error is in the 1% ball park which is better than your oscilloscope anyway.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline schmitt trigger

  • Super Contributor
  • ***
  • Posts: 2272
  • Country: mx
« Last Edit: April 03, 2019, 09:51:02 pm by schmitt trigger »
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28887
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Less than 15pf of input capacitance will be hard to beat. Perhaps look for a probe with a low capacitance at the tip. 10M Ohm isn't a problem. With an impedance at the load the error is in the 1% ball park which is better than your oscilloscope anyway.
Actually not at all hard if your scope has sufficient input sensitivity.
Use a 100:1 probe and at least halve 10:1 probe tip capacitance and they cost just a few 10's of $ as opposed to an active probe $$$ !

http://www.pintek.com.tw/files/pintek/CP3308R.pdf
Avid Rabid Hobbyist.
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: euzer

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16889
  • Country: us
  • DavidH
An active probe will be required to get a low enough capacitance.  Below is an example from Bob Pease but you may require something faster.
 
The following users thanked this post: schmitt trigger

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4130
  • Country: fi
  • Born in Finland with DLL21 in hand
I have used this. Not bad at all. But it is  1.5GHz RF probe, not DC coupled.  With its price, $15 shipped not bad.

It is just finished and perhaps somehow improved version of Elektor's article "Poor Mans 1 GHz" by David Jewsbury (pdf)

It may also give some idea how to make quite cheap DC coupled low capacitance high impedance  "poor mans" probe for predefined usage situation.

So or so but active probe is one good solution.

If cheap is not needed then there is available more in 1pF class:
https://www.picotech.com/accessories/high-bandwidth-active-probes/tetris-1-ghz-active-probe

Of course there is also  cheap (active)differential probes, but these kind probes typically are not at all low capacitance. (example EEEVBlog HVP70  5.5pF)
« Last Edit: April 04, 2019, 05:21:49 am by rf-loop »
BEV of course. Cars with smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the (strong)wises gone?
 

Offline Berni

  • Super Contributor
  • ***
  • Posts: 5002
  • Country: si
Active probes have to be selected carefully because they often only have 1M or 100K DC resistance. The advantage of these probes comes above about 10MHz where the 20pF capacitance of a passive scope probe starts to load down the circuit significantly more than the actual probe resistance.

But given that he is looking at 5V signals i think a DIY 100:1 scope probe will do just fine to get you to 100M input impedance. That is if your frequencies of interest are reasonably low in the few MHz range.

If 100M still is not enough then the way to go is a good opamp in non inverting configuration. This gets you in to the GOhm range, with very careful opamp selections even in the TOhm range. Obviously the opamp will have limitations in terms of bandwidth, slewrate and voltage swing.

Tho if you are interested in just static operation the tool for the job is a bench multimeter with a HighZ mode.
« Last Edit: April 04, 2019, 05:28:29 am by Berni »
 

Offline euzerTopic starter

  • Contributor
  • Posts: 45
Less than 15pf of input capacitance will be hard to beat. Perhaps look for a probe with a low capacitance at the tip. 10M Ohm isn't a problem. With an impedance at the load the error is in the 1% ball park which is better than your oscilloscope anyway.
Actually not at all hard if your scope has sufficient input sensitivity.
Use a 100:1 probe and at least halve 10:1 probe tip capacitance and they cost just a few 10's of $ as opposed to an active probe $$$ !

http://www.pintek.com.tw/files/pintek/CP3308R.pdf
Thanks. It's showing as discontinued https://www.rapidonline.com/pintek-cp-3308r-3000mhz-oscilloscope-probe-85-2235
« Last Edit: April 17, 2019, 04:13:51 pm by euzer »
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4130
  • Country: fi
  • Born in Finland with DLL21 in hand
Less than 15pf of input capacitance will be hard to beat. Perhaps look for a probe with a low capacitance at the tip. 10M Ohm isn't a problem. With an impedance at the load the error is in the 1% ball park which is better than your oscilloscope anyway.
Actually not at all hard if your scope has sufficient input sensitivity.
Use a 100:1 probe and at least halve 10:1 probe tip capacitance and they cost just a few 10's of $ as opposed to an active probe $$$ !


http://www.pintek.com.tw/files/pintek/CP3308R.pdf
Thanks. It's showing as https://www.rapidonline.com/pintek-cp-3308r-3000mhz-oscilloscope-probe-85-2235

This "rapidonline" information is junk. It really is not 3000MHz probe.

In this pdf (pintek) it ios right and it is of course 300MHz probe.

But now.
It is good to think...

Quote
low capacitance scope probe for testing 5V open-drain outputs with high pull-up resistance, e.g. 100k, such that its effect on the observed signal is minimal, as opposed to existing 10M/15pF probes supplied with scope?

Ok this Pintek have then 100Mega ohm DC resistance... so itvreally do not affect.

But and then BIG but.

Pintek 3308R input advertised "approximated" (perhaps they forget how to measure  ;) )  capacitance is 5.5pF. For your circuit it is not at all low capacitance.
Think this 100k pull up as source impedance and look what 5.5pF load do...
in your hand is now "300MHz" probe what freq response (-3dB) in this case is 300kHz
risetime roughly 1.2µs!

(there is many RC filter calculators with bodeplot etc:
 http://sim.okawa-denshi.jp/en/CRtool.php )