Author Topic: Handheld meter robustness testing  (Read 1262108 times)

0 Members and 47 Guests are viewing this topic.

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1911
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2325 on: January 20, 2018, 08:13:07 pm »
Yikes! Now that's a porn shot! You must have taken all night to render too (not to mention shoot it). Nobody does it like Joe - period.
 
The following users thanked this post: joeqsmith

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7115
  • Country: hr
Re: Handheld meter electrical robustness testing.
« Reply #2326 on: January 20, 2018, 09:07:29 pm »
Price of BM839 in Welectron (Germany) is 150€ (no Vat)... It is actually more expensive than BM867, that is 133€ (no Vat)... I don't see it as a better than BM867... Funny...
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17056
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2327 on: January 20, 2018, 11:23:35 pm »
Does this mean we get to add another 100,000 cycles to the Brymen and Fluke 17B+?

Duel to the death!

PS: Did the Fluke's switch wipers go under the Microscope? It's not in the original video.
« Last Edit: January 20, 2018, 11:41:16 pm by Fungus »
 

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1911
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2328 on: January 20, 2018, 11:36:01 pm »
No need. Just have blind faith, Fluke's contacts are certified to hold up during nuclear shock waves!  :-DD
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17056
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2329 on: January 20, 2018, 11:39:36 pm »
No need. Just have blind faith, Fluke's contacts are certified to hold up during nuclear shock waves!  :-DD

The Fluke will win easily if the contact resistance graphs are any indication.  :popcorn:
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2330 on: January 21, 2018, 01:05:16 am »
If there is a contest, the Keysight meter would win some sort of award for their design.   The mark of true quality right here.   

Offline kalel

  • Frequent Contributor
  • **
  • Posts: 880
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2331 on: January 21, 2018, 10:46:53 am »
The UT210E's added bandwidth was enough to get the video flagged for copyright infringement.  The sound quality has improved beyond what the UT90A was capable of.   :-DD   
If they start to place adds on it, I will most likely remove it.

Can you use some music that doesn't trigger those things? There is "open" music (public domain or creative commons). That is, if you're interested in making such tests in the future, otherwise it really doesn't matter.
 

Offline Hydrawerk

  • Super Contributor
  • ***
  • Posts: 2607
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2332 on: January 21, 2018, 05:48:52 pm »
I am happy that Brymen introduced a DMM powered by AA cells and with easy removable fuses. I noticed that the internal PCB is a bit unpopulated. Are there new improved meters coming?
On the other hand, the Brymen BM839 is not much feature wise.
« Last Edit: January 21, 2018, 09:07:48 pm by Hydrawerk »
Amazing machines. https://www.youtube.com/user/denha (It is not me...)
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17056
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2333 on: January 21, 2018, 05:51:01 pm »
Can you use some music that doesn't trigger those things? There is "open" music (public domain or creative commons). That is, if you're interested in making such tests in the future, otherwise it really doesn't matter.

There's actually some "Youtube friendly" music supplied by Google that you use. I assume there's no way that can be flagged.
 

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1911
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2334 on: January 21, 2018, 06:39:57 pm »
If he has to re-post this 90min video again, he'll have to re-shoot the segment and render. What a pain! That was fair use of fast fingers!
(for keyboards, try this for fast http://bit.ly/2F2Hm3E8)
 

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1911
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2335 on: January 21, 2018, 11:29:21 pm »
Wow.. Nothing can top that. I bet YT would never be able to identify that second guy - frigging awesome!
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Handheld meter electrical robustness testing.
« Reply #2336 on: January 22, 2018, 05:37:49 am »
I like the big bands and am a bit of a Buddy Rich fan myself.  Here with Gene Krupa and Sammy Davis Jr. 


Excellent! Amazing the good stuff you find on this forum. ;D
TEA is the way. | TEA Time channel
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2337 on: January 23, 2018, 05:23:14 am »
Having a wimpy meter may be alright for some people but not for me.   While testing has shown that my favorite meter can take a pretty big hit before it is damaged,  it just does not seem right that the cheep little ZT102 / AN8002 can out perform it. 

If you like wimpy meters, this next video is not for you but if you want to see my favorite meter step up it's game, stay tuned.   It's 14KV or BUST!!

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1911
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2338 on: January 23, 2018, 12:51:48 pm »
Having a wimpy meter may be alright for some people but not for me.   While testing has shown that my favorite meter can take a pretty big hit before it is damaged,  it just does not seem right that the cheep little ZT102 / AN8002 can out perform it. 

If you like wimpy meters, this next video is not for you but if you want to see my favorite meter step up it's game, stay tuned.   It's 14KV or BUST!!
Let me guess.. You could call it the yellow wonder. A totally potted Harbor Freight blue-light special?
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2339 on: January 23, 2018, 12:57:36 pm »
It's had it's rotary switch cycled 50,000 times,  it's swam with the fishes and it has been exposed to transient levels that damaged it's input.   Can it survive a 14,000 volt transient with a $1.00 budget and no PCB changes?   Watch and find out. 

https://youtu.be/u9P4N5HPwf4

Offline kcbrown

  • Frequent Contributor
  • **
  • Posts: 896
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2340 on: January 24, 2018, 02:57:30 am »
It's had it's rotary switch cycled 50,000 times,  it's swam with the fishes and it has been exposed to transient levels that damaged it's input.   Can it survive a 14,000 volt transient with a $1.00 budget and no PCB changes?   Watch and find out. 

https://youtu.be/u9P4N5HPwf4

$1 in parts and $10 million in engineering later ... :-DD


Sent from my iPhone using Tapatalk
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2341 on: January 24, 2018, 02:33:17 pm »
$1 in parts and $10 million in engineering later ... :-DD
:-DD  :-+

Offline Scottjd

  • Frequent Contributor
  • **
  • Posts: 436
  • Country: us
    • YouTube Gadget Reveiws
Re: Handheld meter electrical robustness testing.
« Reply #2342 on: January 24, 2018, 05:26:03 pm »
$1 in parts and $10 million in engineering later ... :-DD
Just finished watching the video.  :popcorn:
Well those transistors are not cheap, maybe a little more then $1 for parts.
But the time spent in testing and engineering, the custom build generators, home made high voltage probe on the oscilloscope.
Hats off to you sir,  :clap: Truly a great accomplishment, and great work.

So what’s the lesson learned from this. Do they need to put better parts that are equal to or of higher value in front of the PTC’s and what they are rated for? So if the PTC is rated for X then why put transistors that can only handle C in front of them?
This kind of sounds like common sense now when I ask it like that. I’m not trying to over simplify it, but it might sound like that and I couldn’t figure another way to ask the questions.

That’s one impressive meter for sure, especially with the “Joe Touch” mod. Maybe you should start award levels now? The highest awarded would be to meet or exceed your modified meter testing results, but being stock with no mods or updated parts.

Questions, maybe a dumb one but I’ll ask it anyway.
First, how much can your high voltage scope probe handle, will you need to make a new one if meters start to handle more power then 15KV?

I recently tore down a dead USB QC capable power supply for a Lanova Yoga laptop and noticed spark gaps between the common mode choke on the AC input side. So I was wondering, what if they had still used the cheaper transitors but added some spark gaps on the PCB before them with proper insulation? Would or could this result in the same affect as the higher end transitors? Or am I thinking crazy thoughts?

Side Note: Great reference about USB switch mode power supplies. I found this page full of good information. Since I never repaired a USB power supply (or any power supply) I went searching and found this page to be full of good info:http://lygte-info.dk/info/SMPS%20workings%20UK.html

Thanks again for all the hard work, great testing and time spent editing and doing these videos.
Maybe now it’s time to take a break.  :phew:
Scott
Please be sure to check out my YouTube channel and subscribe if you like the videos. https://www.youtube.com/c/GadgetReviewVideos

By people subscribing and giving thumbs up I know what I am doing is still wanted and adding value, then will continue to release new videos. Thank you for your support.
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 17056
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2343 on: January 24, 2018, 05:39:20 pm »
So what’s the lesson learned from this. Do they need to put better parts...

The CAT rating for this meter goes up to 8000 volt transients, the parts currently in the meter meet those requirements (presumably - I know Joe's tests aren't the same as a real CAT rating test) so there's no requirement to replace them.

Joe's just showing off by making it into a 12kV+ meter.  :popcorn:

(and learning...)
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2344 on: January 24, 2018, 05:44:39 pm »
Joe's just showing off by making it into a 12kV+ meter.  :popcorn:

(and learning...)
Always learning.... 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2345 on: January 24, 2018, 06:15:36 pm »
$1 in parts and $10 million in engineering later ... :-DD
Just finished watching the video.  :popcorn:
Well those transistors are not cheap, maybe a little more then $1 for parts.

So what’s the lesson learned from this. Do they need to put better parts that are equal to or of higher value in front of the PTC’s and what they are rated for? So if the PTC is rated for X then why put transistors that can only handle C in front of them?

That’s one impressive meter for sure, especially with the “Joe Touch” mod. Maybe you should start award levels now? The highest awarded would be to meet or exceed your modified meter testing results, but being stock with no mods or updated parts.

Questions, maybe a dumb one but I’ll ask it anyway.
First, how much can your high voltage scope probe handle, will you need to make a new one if meters start to handle more power then 15KV?

....
Thanks again for all the hard work, great testing and time spent editing and doing these videos.
Maybe now it’s time to take a break.  :phew:

As stated, my cost was under $1 w/ tax.

PTCs are before the transistors, not after.   Meters will be different.  If we are talking about this one specifically, you have the PTC in series with the two resistors.  Again, the PTC is not going to do much of anything beyond act as a resistor in the HV transient tests.  If in resistance mode for example, we have the drop of the clamp.  Basically nothing.  Then you need to consider the transient.  It's not a matter of just looking at the voltages present.  I know I get a fair number of MOT and DC comments from people but the reality is this is not at all what we are looking at. 

I doubt when a company like Brymen designs a meter they are thinking we need to meet the JQS transient tests and survive.  They will design around what ever standards they want to meet and use what ever internal best practices they have come up with.  From what I have seen, companies all follow the same basic concepts.  It's all fairly well understood.  I can't understand why a brand new meter like the 121GW for example would not be very robust.  If you have watched some of my videos, it's not a lot of added cost. 

I have a playlist for the home made HV probes, linked below.  I called it 40KV but I've ran tests far above that to make sure it was not going to be a problem.  You mention power rather than voltage.  I assume a mistake in terms but If you watch the videos, I do talk about the life of the parts based on how hard you run them.   It may be shorter than you think.   

Break??  lol.  I make the videos for the fun of it.  I don't ask for handouts and rarely accept meters to run.  If I need a break, that means I have lost interest and it's no longer fun.  When that time comes, I will just stop making them.  Growing the channel and making money was never part of it.   

https://www.youtube.com/playlist?list=PLZSS2ajxhiQCAQ6gIp6s-WoKiIEb1gHPD

Offline 3db

  • Frequent Contributor
  • **
  • Posts: 331
  • Country: gb
Re: Handheld meter electrical robustness testing.
« Reply #2346 on: January 27, 2018, 10:18:44 pm »
Nice one Joe.
Thanks for the time and effort.
3DB  ;D
 
The following users thanked this post: joeqsmith

Offline rsjsouza

  • Super Contributor
  • ***
  • Posts: 6045
  • Country: us
  • Eternally curious
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #2347 on: February 03, 2018, 12:29:53 pm »
Oh, no! You know now that you are mandated by TEA law to amend your warning statements, right?  "Warning: multimeters AND probes were harmed in the making of this video. TEA viewer discretion is advised."

I remember very well the HV probe discussion as I was interested in making one as well and even got the parts, etc. One thing lead to another and i still didn't find the time to finish this project...  :palm:

Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2348 on: February 04, 2018, 12:15:37 am »
Oh, no! You know now that you are mandated by TEA law to amend your warning statements, right?  "Warning: multimeters AND probes were harmed in the making of this video. TEA viewer discretion is advised."

I remember very well the HV probe discussion as I was interested in making one as well and even got the parts, etc. One thing lead to another and i still didn't find the time to finish this project...  :palm:

As a hobbyist, I enjoy designing and building projects like this and have respect for anyone who attempts them.   Making a DC probe is one thing but add over 10KV to the mix and top it off with AC responses in the MHz, it turns into several hours of play time. 

We are back in business.
https://youtu.be/kWOe803atvg

 

Online joeqsmithTopic starter

  • Super Contributor
  • ***
  • Posts: 11756
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2349 on: February 16, 2018, 12:19:00 am »
With the HV probe repaired and checked out, it's time to look at another meter.    I offered to let people pick the meter at 2000 subscribers and everyone wanted to see the 121GW.  Sadly the meter is still not available.   But I can offer to run another Fluke 87V which was peoples third pick after the Gossen.

We know the 87V I looked at could not come close to the performance of any of the the Chinese made Flukes I have looked.   I was asked about the revision of the meter I last ran and it was not the most recent.    What  I am thinking is to buy a brand new one, directly from Fluke and repeat all of the transient tests.  Maybe run a temperature sweep on it. 

I am not so sure where the 87V is made anymore.  The US laws on what it means to be made in the USA are pretty relaxed.  I would have to see the assembly line, here in the USA with this meter being made to know what it means anymore.    Danaher always seems to be listed in the top companies for unrepatriated profits.  It is what it is.   If China is doing the majority of the work now, it may prove to be a more robust meter.    Even that its a later design that the last one I looked at, maybe they improved it. 

The Fluke 17B+ is still the best performer from my life cycle testing.  I have been wanting to run another Fluke and it seems only fitting that the king pin of Fluke, the meter that all others are ranked against, gets a shot at that as well.   Not a 50,000 half stroke test , but the full rotation for 50K cycles that all of the other meters have been tested to.     

It may seem I have an axe to grind with Fluke.  I don't.  I am more than willing to give a brand new one a fair shake and let the data once again stand on it's own.   

So stay tuned.... 


Share me