Author Topic: Function generators with 50ohm termination question?  (Read 3179 times)

0 Members and 1 Guest are viewing this topic.

Offline BlueBillTopic starter

  • Regular Contributor
  • *
  • Posts: 169
  • Country: ca
Function generators with 50ohm termination question?
« on: February 13, 2015, 03:00:44 am »
I have a Rigol DS1054z-S with waveform generator, it's got a 50ohm and hi-z option.

Do all function generators have 50ohm series resistance? Does enabling the 50ohm mode make it like the left image and hi-z the right image?

 

Offline w2aew

  • Super Contributor
  • ***
  • Posts: 1780
  • Country: us
  • I usTa cuDnt speL enjinere, noW I aR wuN
    • My YouTube Channel
Re: Function generators with 50ohm termination question?
« Reply #1 on: February 13, 2015, 03:07:40 am »
I have a Rigol DS1054z-S with waveform generator, it's got a 50ohm and hi-z option.

Do all function generators have 50ohm series resistance? Does enabling the 50ohm mode make it like the left image and hi-z the right image?



Many of them do.  Some of them allow you to specify what the load impedance is (for when it is not 50, and not very high Z). The 50 output impedance is fixed.  Specifying the load allows the generator to mathematically compute the voltage at the load and adjust the source amplitude as needed to achieve the user specified output amplitude at the load.

My video on the topic

YouTube channel: https://www.youtube.com/w2aew
FAE for Tektronix
Technical Coordinator for the ARRL Northern NJ Section
 

Offline BlueBillTopic starter

  • Regular Contributor
  • *
  • Posts: 169
  • Country: ca
Re: Function generators with 50ohm termination question?
« Reply #2 on: February 13, 2015, 03:59:09 pm »
Wow thanks, and a video too! I was subscribed to your YouTube channel too, just haven't had a chance to watch them all.
« Last Edit: February 13, 2015, 04:00:42 pm by BlueBill »
 