Possible bad calibration, bad touch on ch1 during calibration.

v0.020p checking the limits of calibration values. If the calibration is bad, it will say calibration failed
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: facekim

Offline Evi

  • Regular Contributor
  • *
  • Posts: 115
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1776 on: January 14, 2024, 01:30:18 pm »
I got every time Calibration fail until I switched to stock FW and made there Baseline Calirbation.
Returnig back to new FW Calibration ended successfuly.
 
The following users thanked this post: facekim

Offline Evi

  • Regular Contributor
  • *
  • Posts: 115
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1777 on: January 14, 2024, 01:36:13 pm »
I guess it'll be comfortably to have ability to make partial calibration:
e.g. only baseline or to stop on 3, 6 or 7.5V.
You do not allways have all available voltages on site.
« Last Edit: January 14, 2024, 01:44:19 pm by Evi »
 
The following users thanked this post: facekim

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1778 on: January 14, 2024, 01:53:05 pm »
Which is the stock firmware?  The calibration of the input dividers needs to be performed only once.
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 

Offline Evi

  • Regular Contributor
  • *
  • Posts: 115
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1779 on: January 14, 2024, 02:03:30 pm »
I made a fast switching between SW(described some pages ago). What is version I do not remeber it is one of the Fnirsi's releases.
I do not know does it needed be done once or twiсe, but the fact is fact.
« Last Edit: January 14, 2024, 02:05:02 pm by Evi »
 

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1780 on: January 14, 2024, 03:16:34 pm »
Evi upload firmware 0.021, go to the main menu. Push the usb connection. On the blue screen, tap on the PC screen. Take a photo of the displayed values and post it on the forum.
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 

Offline Evi

  • Regular Contributor
  • *
  • Posts: 115
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1781 on: January 14, 2024, 03:23:41 pm »
Can I ask about purpose?
 

Offline boojum

  • Contributor
  • Posts: 13
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1782 on: January 14, 2024, 03:33:15 pm »
Quite strange that the header is not found to be correct, because it is based on the guide lines given here.

Maybe this will help fix the problem:
I look Pillow code. It check header size and raise error because header size is 56 bytes.

After add value 56 to check header size condition Pillow successfully opens a 768070 bytes bmp.
Maybe you can try just increasing the BMP header to 64 bytes?
As I understand it, 56 is a non-standard header size, which causes a problem in many cases.

I found easy working solution for fix BMP format :)
Just change only one byte - 56 to 40 here (i use binary hack for .bmp, change one byte at 0x0e offset )
After that Firefox, Telegram and Python successufllly recognized screenshots.
 

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1783 on: January 14, 2024, 03:45:48 pm »
won't I be missing some data in the header?

v0.021a  here I changed the //Size of DIB header
   40, 0, 0, 0, //56,0,0,0

Can I ask about purpose?
I want to know what the tolerance is on the input part that did not pass the calibration.

I hope that you did the first part of the calibration without USB and with the probe disconnected, and then set the voltage on the test and confirm OK
« Last Edit: January 14, 2024, 04:19:46 pm by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: facekim, boojum

Offline pcprogrammer

  • Super Contributor
  • ***
  • Posts: 4248
  • Country: nl
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1784 on: January 14, 2024, 03:52:53 pm »
won't I be missing some data in the header?

Who knows.

Nice those semi standards. According to the wikipedia page about bitmap file format it is a specific header format to use when additional information needs to be specified. In this case for the correct color mapping.

I tried the modification of the single byte on a file I have on my system and Firefox failed on it before, but opened it just fine after the change. The default image viewer on Linux Mint has no problem with either.

Offline boojum

  • Contributor
  • Posts: 13
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1785 on: January 14, 2024, 04:01:09 pm »
won't I be missing some data in the header?
v0.021a
Position of pixel array still correct - it not changed.
The last 16 bytes of the DIB header just won't be used, I think.
 

Offline pcprogrammer

  • Super Contributor
  • ***
  • Posts: 4248
  • Country: nl
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1786 on: January 14, 2024, 05:22:54 pm »
The last 16 bytes of the DIB header just won't be used, I think.

And yet these bytes tell the layout of the color information within the 16 bits per pixel. That is why I choose this header format.

Ah well, if it works it works.  :)

Online bffargo

  • Regular Contributor
  • *
  • Posts: 50
  • Country: us
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1787 on: January 14, 2024, 05:56:39 pm »
fnirsi_1013d.bin v0.020j.bin
Something wrong with CH1 below 500mV after calibration

I had similar results yesterday pulling 20J.; though mine was that it didn't work correctly in higher V/div settings only. I recalibrated 3 times (which is a huge pain as I had to hack some interesting BNC splitters and bench DMM probes to detect exact voltages at the inputs and tweak with analog power supply dials). Manually fiddling with volt range on each channels got signals showing correctly.

But auto set is completely broken it doesn't detect anything right no matter channels being on or voltages or waves applied.
 

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1788 on: January 14, 2024, 05:57:54 pm »
Fargo, an image with data as I wrote above. (auto settings not work in DC mode?)

I will take one BNC T-piece, insert it into ch1, connect it to ch2, and from the BNC T-piece I will stretch the cable with alligator clips to the measuring device and the power source.

Tokar: I know about your request for DC mode, I'll come back to it later.

 I am now working on scrolling the record in memory. 

And somehow there is quite often a request for roll mode or long recording. 

I would appreciate some help with auto DC mods if anyone has time.
« Last Edit: January 14, 2024, 06:04:09 pm by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 

Online bffargo

  • Regular Contributor
  • *
  • Posts: 50
  • Country: us
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1789 on: January 14, 2024, 06:19:26 pm »
Fargo, an image with data as I wrote above. (auto settings not work in DC mode?)

I will take one BNC T-piece, insert it into ch1, connect it to ch2, and from the BNC T-piece I will stretch the cable with alligator clips to the measuring device and the power source.

Yeah that's pretty much how I did it. DMM'd one side of the T and through various interconnects tied CH1 to CH2 over two T's. I just don't have exact parts to make it line up perfectly, but once I got it all it was stable and same power going into both sides. But the numbers seemingly below show that it was inconsistent for some reason? (just reading into the variances and seem to be correlated to having problems on one end for CH1 and other end for CH2)
 
The following users thanked this post: facekim

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1790 on: January 14, 2024, 06:27:42 pm »
Ch1 5V and ch2 100mV are out of calibration.  For calibration, use a properly filtered voltage, possibly a 7815 and a potentiometer.  You have to remember that the oscilloscope takes a sample and when there is a peak, it calibrates according to it.

50mV is just 100mV * 2

I will modify the software tomorrow.  It would be a problem to calibrate this way.  I'm not sure that this software will let you calibrate. 

Now the only option is to record the 00h.bin mentioned about 2 pages ago into the oscilloscope, it will erase everything.  And you will recalibrate... 

That's why I wanted the loader to be supplemented with a button that confuses the configuration data.
« Last Edit: January 14, 2024, 06:42:30 pm by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: facekim

Online bffargo

  • Regular Contributor
  • *
  • Posts: 50
  • Country: us
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1791 on: January 14, 2024, 06:38:26 pm »
Am using a SMPS bench power supply with Owon bench DMM connected metering the voltage right at the BNC. I can switch to a Linear one if you think the frequency of the SMPS is causing an issue
« Last Edit: January 14, 2024, 06:40:22 pm by bffargo »
 
The following users thanked this post: facekim

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1792 on: January 14, 2024, 06:44:27 pm »
I wanted to use the rd6006, but when I saw what it had on the output, I used an analog source.  See text up.
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: facekim

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1793 on: January 15, 2024, 10:19:55 am »
 :scared: V0. 021b mistake   :-//

 :popcorn: 0.021c ok  :phew:

Added possibility to reset calibration values. (if a bad calibration occurs) Added notification of a faulty calibration if the offset cannot be set (only if there is some hardware problem, bad ad converter, missing ad converter, problem with the offset voltage setting)

Input divider calibration is enabled when the oscilloscope is turned on for the first time and baseline calibration is used.

Calling up the calibration of the input divider is possible only through the menu by resetting the calibration values. (or by deleting the data on the sd card where the configuration data is located. NEED SDCART READER No It can be done with a card stored in the oscilloscope. )

After successful calibration of the input divider and switching off the oscilloscope, the data is written to the SD card. (Unless you change the burnt components on the input circuits, it will no longer be needed - that is, if you do not erase, do not replace the sd card)

If the calibration is started again, only the base level calibration will be performed. i.e. 0 is found. (What sometimes needs to be done)

By uploading the firmware, the calibration of the input divider DOES NOT CHANGE.

« Last Edit: January 16, 2024, 06:48:51 am by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: wolfy007, tokar, facekim, boojum

Offline facekim

  • Newbie
  • Posts: 7
  • Country: es
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1794 on: January 15, 2024, 01:26:24 pm »

With the version updated to v0.021c, resetting the calibration in the USB menu does not work.
At least in my case.

Thank you
 

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1795 on: January 15, 2024, 01:27:34 pm »
Does the button respond to touch?  Make a video.
« Last Edit: January 15, 2024, 01:30:32 pm by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: facekim

Offline facekim

  • Newbie
  • Posts: 7
  • Country: es
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1796 on: January 15, 2024, 01:29:14 pm »
After returning to the factory firmware and putting the modified one again if it comes out.

Sorry
Thank you
 

Offline tokar

  • Regular Contributor
  • *
  • Posts: 75
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1797 on: January 15, 2024, 09:13:45 pm »

With the version updated to v0.021c, resetting the calibration in the USB menu does not work. At least in my case.
I confirm.
 Procedure: v0021a, “Baseline calibration”, calibration of 0.3-15V dividers was successful.  I closed the menu, turned it off and turned on the device.  I click the menu - USB connection - PC window.  We get a table of constants. 
I'm flashing it via loader v0021c.  Notification enable, Baseline calibration, I get "Calibration successful".  Click the menu - USB connection - PC window.  No reaction.  I fill sector 700 ""00h". I repeat the above steps - there is no expected result.
I fill sectors from 16 to 2047 "00h". Turn it off - turn on the oscilloscope, the Fnirsi screen saver. Calibrate. Flash v0021c. I repeat the above steps - there is no expected result.
 

Offline Atlan

  • Frequent Contributor
  • **
  • Posts: 392
  • Country: sk
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1798 on: January 16, 2024, 06:45:50 am »
It is no longer necessary to enable the notification.  Resetting the values ​​does not work, because they are written again during a correct shutdown.(use external sdcard reader) The option is to click on the screen starting in usb mode.  You have to click right in the middle of the pc screen.  Very precisely in the center of the pc screen.
« Last Edit: January 16, 2024, 06:49:52 am by Atlan »
FNIRSI 1013D Always provide a picture or video with the problem where the parameters of the oscilloscope are visible, and a picture of the diagnostic screen with the values.
 
The following users thanked this post: tokar, facekim

Offline tokar

  • Regular Contributor
  • *
  • Posts: 75
  • Country: ru
Re: FNIRSI-1013D "100MHz" tablet oscilloscope
« Reply #1799 on: January 16, 2024, 07:05:11 am »
Yes, it works. Thank you.
 
The following users thanked this post: facekim