Ok, I go the java worked out.
Ran the test, here are the results:
Should these values be this far off? The display of the power supply (DP832) did show a much closer number when the steps were selected to the meter and not what the steps are...
---------------------------- START OF TEST ---------------------------
Please disconnect all cables from the DMM
Enter 1 to begin automated voltage readback tests
1
Testing remote voltage measurement
meas val: 0.0015v
meas val: 0.0015v
meas val: 0.0015v
meas val: 0.0015v
meas val: 0.0015v
Testing local voltage measurement
Please check that instrument is in DC volts, 10npcl, local mode
Enter 1 to begin automated current readback tests
1
Testing remote 1A to 10A current measurement
meas val: -0.0001A
meas val: -0.0001A
meas val: -0.0001A
meas val: -0.0001A
meas val: -0.0001A
Testing local current measurement
Please check that instrument is in DC current, 10npcl, local mode
Enter y to test reading output of PSU cal points or n to skip
y
---------------------- SELECT READBACK TEST TYPE ----------------------
Note 1: If you're having trouble with voltage or current calibrations
performing a full voltage + current calibration will fix this
Note 2: Automated current calibration requires DMM with >= 3A range
Enter value of desired test type for ch1
0: skip ch1 test
1: voltage + manual current test
2: voltage + automated current test
3: voltage test (automated entry)
4: current test (manual entry)
5: current test (automated entry)
2
-------------------- VOLTAGE READBACK ROUTINE --------------------
Please connect DMM voltage input to ch1 output of PSU
Enter 1 to begin automated voltage readback
1
ch1 DAC-V readback
step 0, cal point: 0.2v, meas val: 0.0119v
step 1, cal point: 0.5v, meas val: 0.3104v
step 2, cal point: 1.2v, meas val: 1.0007v
step 3, cal point: 2v, meas val: 1.7894v
step 4, cal point: 3.2v, meas val: 2.9779v
step 5, cal point: 4.1v, meas val: 3.8682v
step 6, cal point: 5.2v, meas val: 4.9582v
step 7, cal point: 6.9v, meas val: 6.6497v
step 8, cal point: 7.5v, meas val: 7.2426v
step 9, cal point: 8.7v, meas val: 8.4378v
step 10, cal point: 10.1v, meas val: 9.8239v
step 11, cal point: 11.8v, meas val: 11.5075v
step 12, cal point: 12.6v, meas val: 12.3022v
step 13, cal point: 13.5v, meas val: 13.1974v
step 14, cal point: 15v, meas val: 14.6870v
step 15, cal point: 15.8v, meas val: 15.4879v
step 16, cal point: 16.5v, meas val: 16.1856v
step 17, cal point: 17.3v, meas val: 16.9773v
step 18, cal point: 18.5v, meas val: 18.1661v
step 19, cal point: 19.1v, meas val: 18.7651v
step 20, cal point: 19.9v, meas val: 19.5579v
step 21, cal point: 20.2v, meas val: 19.8532v
step 22, cal point: 20.8v, meas val: 20.4471v
step 23, cal point: 21.8v, meas val: 21.4467v
step 24, cal point: 22.4v, meas val: 22.0422v
step 25, cal point: 22.7v, meas val: 22.3420v
step 26, cal point: 23.9v, meas val: 23.5388v
step 27, cal point: 24.3v, meas val: 23.9386v
step 28, cal point: 25.7v, meas val: 25.3328v
step 29, cal point: 26.9v, meas val: 26.5311v
step 30, cal point: 27.9v, meas val: 27.5246v
step 31, cal point: 28.5v, meas val: 28.1186v
step 32, cal point: 28.9v, meas val: 28.5138v
step 33, cal point: 29.8v, meas val: 29.4050v
step 34, cal point: 30.2v, meas val: 29.8022v
step 35, cal point: 32v, meas val: 31.5926v
ch1 ADC-V readback
step 0, cal point: 0v, meas val: -0.1861v
step 1, cal point: 0.05v, meas val: -0.1374v
step 2, cal point: 0.1v, meas val: -0.0878v
step 3, cal point: 0.5v, meas val: 0.3101v
step 4, cal point: 1v, meas val: 0.8039v
step 5, cal point: 5v, meas val: 4.7599v
step 6, cal point: 10v, meas val: 9.7236v
step 7, cal point: 12.8v, meas val: 12.5013v
step 8, cal point: 20v, meas val: 19.6557v
step 9, cal point: 30v, meas val: 29.6043v
step 10, cal point: 32v, meas val: 31.5924v
-------------------- CURRENT READBACK ROUTINE --------------------
Connect ch1 output to DMM 10A current input
Enter 1 to begin automated current readback
1
ch1 DAC-I readback
step 0, cal point: 0.1A, meas val: 0.0862v
step 1, cal point: 0.25A, meas val: 0.2358v
step 2, cal point: 0.5A, meas val: 0.4855v
step 3, cal point: 0.8A, meas val: 0.7837v
step 4, cal point: 1A, meas val: 0.9828v
step 5, cal point: 1.25A, meas val: 1.2307v
step 6, cal point: 1.5A, meas val: 1.4785v
step 7, cal point: 1.75A, meas val: 1.7257v
step 8, cal point: 1.9A, meas val: 1.8756v
step 9, cal point: 2.15A, meas val: 2.1228v
step 10, cal point: 2.35A, meas val: 2.3225v
step 11, cal point: 2.5A, meas val: 2.4712v
step 12, cal point: 2.75A, meas val: 2.7187v
step 13, cal point: 3A, meas val: 2.9663v
step 14, cal point: 3.2A, meas val: 3.1632v
ch1 ADC-I readback
step 0, cal point: 0A, meas val: -0.0129v
step 1, cal point: 0.01A, meas val: -0.0028v
step 2, cal point: 0.1A, meas val: 0.0861v
step 3, cal point: 1A, meas val: 0.9827v
step 4, cal point: 2A, meas val: 1.9741v
step 5, cal point: 3A, meas val: 2.9662v
step 6, cal point: 3.2A, meas val: 3.1631v
---------------------------- END OF TEST ----------------------------