Measuring parasitics of multimeters is a very healthy exercise!
![ThumbsUp :-+](https://www.eevblog.com/forum/Smileys/default/icon_smile_thumbsup.gif.pagespeed.ce._JElyJQqdB.gif)
The capacitance would also depend on the voltage range setting, when the input divider is connected the capacitance from the "+" input terminal might drop significantly. My measurements with a hand held capacitance meter (1kHz) for the DC voltage mode only:
Keithley 2015 Positive input to mains ground - 10V range and below ~100pF, 100V and 1000V ~25pF
Negative input to mains ground ~ 180pF
Between input terminals ~300pF for 10V range and below, ~20pF for 100V and 1000V ranges
HP3456APositive input to mains ground - 10V range and below ~350pF, 100V and 1000V ~85pF
Negative input to mains ground ~ 1nF with the guard open, ~2.2nF with the guard to negative
Between input terminals ~600pF for 10V range and below, ~60pF for 100V and 1000V ranges
Hioki DM7275-03 Positive input to mains ground - 10V range and below <30pF, 100V and 1000V <10pF
Negative input to mains ground <30pF
Between input terminals ~1500pF for 10V range and below*, <10pF for 100V and 1000V ranges.
Cheers
Alex
* - possibly something to do with the contact checking feature, even if it is OFF. The cure is simple (if the input C is a problem) - to put a 10K resistor in series with the positive lead and use Hi-Z, on 10V and below the meter has ~20pA input current or ~0.2uV added error with the resistor.