Author Topic: DC Block response  (Read 4089 times)

0 Members and 1 Guest are viewing this topic.

Offline JesterTopic starter

  • Frequent Contributor
  • **
  • Posts: 887
  • Country: ca
DC Block response
« on: September 30, 2016, 10:35:41 pm »
I picked up one of these on eBay  http://www.ebay.com/itm/Attenuator-DC-Block-N-male-to-Female-DC-6-0GHZ-50ohms-RF-coaxial-Power-2W-200V-/361279079925?hash=item541de945f5:g:XAQAAOSwPgxVOetN

Initial test was with a signal generator and scope, it blocks DC as advertised

Next test I swept on the SA and it looks like this, what am I doing wrong?




 
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27895
  • Country: nl
    • NCT Developments
Re: DC Block response
« Reply #1 on: September 30, 2016, 10:42:23 pm »
Your screendump doesn't make any sense. The centre frequency is supposed to be 15MHz but your marker shows 89kHz. Perhaps you should reset the spectrum analyser and test again.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline hendorog

  • Super Contributor
  • ***
  • Posts: 1627
  • Country: nz
Re: DC Block response
« Reply #2 on: September 30, 2016, 10:48:34 pm »
Your screendump doesn't make any sense. The centre frequency is supposed to be 15MHz but your marker shows 89kHz. Perhaps you should reset the spectrum analyser and test again.

It's a log frequency scale - the 'centre' value is 1.5 divisions in from the Right edge of the screen!
 
The following users thanked this post: tautech

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 27895
  • Country: nl
    • NCT Developments
Re: DC Block response
« Reply #3 on: September 30, 2016, 10:54:19 pm »
Your screendump doesn't make any sense. The centre frequency is supposed to be 15MHz but your marker shows 89kHz. Perhaps you should reset the spectrum analyser and test again.
It's a log frequency scale - the 'centre' value is 1.5 divisions in from the Right edge of the screen!
I missed that! But appearantly so did to OP because it would mean the left part of the screen is in the Hz range which is where (AFAIK) all spectrum analysers show a zero level.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline JesterTopic starter

  • Frequent Contributor
  • **
  • Posts: 887
  • Country: ca
Re: DC Block response
« Reply #4 on: September 30, 2016, 10:56:56 pm »
Ooops and not to mention the TG minimum frequency is 100kHz
 

Offline hendorog

  • Super Contributor
  • ***
  • Posts: 1627
  • Country: nz
Re: DC Block response
« Reply #5 on: September 30, 2016, 11:07:13 pm »
Your screendump doesn't make any sense. The centre frequency is supposed to be 15MHz but your marker shows 89kHz. Perhaps you should reset the spectrum analyser and test again.
It's a log frequency scale - the 'centre' value is 1.5 divisions in from the Right edge of the screen!
I missed that! But appearantly so did to OP because it would mean the left part of the screen is in the Hz range which is where (AFAIK) all spectrum analysers show a zero level.

Yep, you passed the test - you are human :)

It's still within spec for the SA at 90kHz at the marker (spec says 9kHz) so I'm thinking the LO feedthrough shouldn't take over at that point.

However google says that that a Siglent TG starts at 100kHz, so just out of spec for that.
https://mediacdn.eu/m/media/wysiwyg/siglent/Downloads/DataSheet/SSA3000X_DataSheet.pdf

The 'stair-step' looks a bit odd too. Like it is operating at the limit of the ADC, perhaps the atten/gain settings are not sensible for this measurement? Hit preset and try again?
 

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: DC Block response
« Reply #6 on: September 30, 2016, 11:26:29 pm »
I love that the listing says it is good from DC-6 GHz  - yet it blocks DC.
VE7FM
 
The following users thanked this post: rx8pilot

Offline Jay_Diddy_B

  • Super Contributor
  • ***
  • Posts: 2764
  • Country: ca
Re: DC Block response
« Reply #7 on: September 30, 2016, 11:52:30 pm »
Hi,

The frequency response is given by:

F = 1/2 x PI x RC

If R= 50 Ohms (Spectrum analyzer input)

And C = 100pF

The F (-3dB point) = 31.8 MHz

This is consistent with your measurements.

If you use the same DC block on a scope with a 1M Ohm input

The -3dB frequency is 1.59kHz

I assume that in this context DC-6GHz means low frequencies to 6Ghz.

But for some applications, like EMC measurements, a higher value capacitor is needed.

22nF will give you a -3dB point at 150kHz with a 50 Ohm SA.

Regards,

Jay_Diddy_B

 
The following users thanked this post: Jester

Offline JesterTopic starter

  • Frequent Contributor
  • **
  • Posts: 887
  • Country: ca
Re: DC Block response
« Reply #8 on: October 01, 2016, 01:41:38 am »
Hi,

The frequency response is given by:

F = 1/2 x PI x RC

If R= 50 Ohms (Spectrum analyzer input)

And C = 100pF

The F (-3dB point) = 31.8 MHz

This is consistent with your measurements.

If you use the same DC block on a scope with a 1M Ohm input

The -3dB frequency is 1.59kHz

I assume that in this context DC-6GHz means low frequencies to 6Ghz.

But for some applications, like EMC measurements, a higher value capacitor is needed.

22nF will give you a -3dB point at 150kHz with a 50 Ohm SA.

Regards,

Jay_Diddy_B

J-D,

Your analysis is correct, looks like I need to find a DC block with a larger capacitor for conducted emissions testing.

Updated graph with 50 \$\Omega\$ scope termination as well as TG/SA from 100k to 100MHz

 

Offline Performa01

  • Super Contributor
  • ***
  • Posts: 1712
  • Country: at
Re: DC Block response
« Reply #9 on: October 01, 2016, 10:58:24 pm »
... looks like I need to find a DC block with a larger capacitor for conducted emissions testing.

I use this one from Mini Circuits and can highly recommend it:

https://www.minicircuits.com/pdfs/BLK-89+.pdf


Even though it’s rated from 100kHz to 8GHz, for most practical purposes it actually starts much earlier:



This means -0.5dB error at 15kHz.

It has SMA connectors though.
Mini Circuits also have an N-type DC-Block, but that also is 10MHz to 6GHz, hence very similar to the one you have.
 

Offline JesterTopic starter

  • Frequent Contributor
  • **
  • Posts: 887
  • Country: ca
Re: DC Block response
« Reply #10 on: October 02, 2016, 11:45:38 pm »
... looks like I need to find a DC block with a larger capacitor for conducted emissions testing.

I use this one from Mini Circuits and can highly recommend it:

https://www.minicircuits.com/pdfs/BLK-89+.pdf


Even though it’s rated from 100kHz to 8GHz, for most practical purposes it actually starts much earlier:



This means -0.5dB error at 15kHz.

It has SMA connectors though.
Mini Circuits also have an N-type DC-Block, but that also is 10MHz to 6GHz, hence very similar to the one you have.

Thanks for that, I found that exact one when I searched after J-D's response.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf