Hi everybody,

I'm playing with the fisrt time with the counter of my DG1022. I dont know if it's me but it's really hard to use.

For exemple. I setup a simple square waveform from ch1 and enter it in the counter with a coax. 1 Mhz. 2v peak. No offset.
I check the output signal on scope.... very clean 0 to 2v 50% duty square wave.

To be able to read anything I must set the trigger sense to 95%, level to 0 and coupling is DC. The gate time change only the accuracy.

If I change trigger sens to 85%... result jump all over the place
If I rise trigger lever to only 100mv... nothing
If I lower the frequency to 100khz or 1Khz same result.

If I probe with my dm3068 (On Freq as well), it is showing me the right frequency.

And this is just one exemple. Is it me ? Or it's very hard use. It's pratically useless.

I looked on other thread but nobody is talking about it. It's why I think it is maybe defective ?

Thank you

Yannick






« Last Edit: November 14, 2023, 02:26:57 pm by Yannick99 »
 

Offline AVGresponding

  • Super Contributor
  • ***
  • Posts: 4742
  • Country: england
  • Exploring Rabbit Holes Since The 1970s
Re: Can my Rigol dg1022 counter be defective ?
« Reply #1 on: November 13, 2023, 05:30:27 pm »
Have you tried changing it to a sine wave? Square waves have a lot of harmonics, your counter might be sensitive enough to be confused by them.
nuqDaq yuch Dapol?
Addiction count: Agilent-AVO-BlackStar-Brymen-Chauvin Arnoux-Fluke-GenRad-Hameg-HP-Keithley-IsoTech-Mastech-Megger-Metrix-Micronta-Racal-RFL-Siglent-Solartron-Tektronix-Thurlby-Time Electronics-TTi-UniT
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #2 on: November 13, 2023, 07:55:44 pm »
Yes of course, sine wave 1vpp gave me the same thing. If I rise to 2Vpp I can lover the  trigger sens to 75%.

As well with a sinewave I chossing ac coupling instead DC but anyway I have the same problem in both.

It's like my signal is detected very dim or fadeor too low ?

Yannick
 

Online xrunner

  • Super Contributor
  • ***
  • Posts: 7573
  • Country: us
  • hp>Agilent>Keysight>???
Re: Can my Rigol dg1022 counter be defective ?
« Reply #3 on: November 13, 2023, 08:07:05 pm »
Hi everybody,

I'm playing with the fisrt time with the counter of my DG1022. I dont know if it's me but it's really hard to use.

Yannick

I had a DG1022 back in 2016. I had problems with the counter as well. The thread is linked below. Mostly it wasn't accurate even with an external GPSDO reference. I reported it to Rigol and they worked a deal for me to return it and get a DG1032Z plus a few $$. I just don't think the counter in those DG1022s are any good at all.

https://www.eevblog.com/forum/testgear/rigol-dg1022a-function-gen-frequency-counter-function/msg926368/#msg926368
I told my friends I could teach them to be funny, but they all just laughed at me.
 
The following users thanked this post: AVGresponding

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #4 on: November 13, 2023, 08:58:40 pm »
Yes I read your thread. I think I have something similar. I have the new model but I do not know if the internal hardware is updated or better. If I look my reading, I don't think so.
 

 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23045
  • Country: gb
Re: Can my Rigol dg1022 counter be defective ?
« Reply #5 on: November 13, 2023, 09:01:42 pm »
Ex DG1022 owner. The counter is shit. Ex DG800 owner. The counters are shit too. Ex Siglent SDG1000 owner. The counter is terrible too.

Advice: buy a separate box. 1980s HP junker off ebay is good.
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #6 on: November 13, 2023, 09:20:16 pm »
I know what you mean. I have my old HP. Same signal, same cable.

I thought with time, thing should be better :) no...

Yannick
« Last Edit: November 13, 2023, 09:25:41 pm by Yannick99 »
 
The following users thanked this post: bd139

Offline Circlotron

  • Super Contributor
  • ***
  • Posts: 3234
  • Country: au
Re: Can my Rigol dg1022 counter be defective ?
« Reply #7 on: November 13, 2023, 11:19:34 pm »
I'll have to check mine when I get home.
You'd think the sig gen and the counter would run off the same reference so there would always be zero "error" even if the reference drifted.

Edit -> I misread the title. I thought it was talking about the later DG1022Z.
« Last Edit: November 14, 2023, 12:50:22 am by Circlotron »
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #8 on: November 14, 2023, 12:07:26 am »
Exactly, I run it under external gpsdo but, the reference is the same. I sould be reading bang on. Even if I were using internal oscillator. But the problem isn't just that. It is very hard to bring the counter counting. Let me know please. Thank you, appreciated.

Yannick
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #9 on: November 14, 2023, 02:27:41 pm »


Edit -> I misread the title. I thought it was talking about the later DG1022Z.

Yes sorry it's dg1022Z
 

Offline Njk

  • Frequent Contributor
  • **
  • Posts: 250
  • Country: ru
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #10 on: November 14, 2023, 05:33:24 pm »
Anyway for DG800 I witnessed it's working somehow. Perhaps there can be variations from one sample to the other.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23045
  • Country: gb
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #11 on: November 14, 2023, 09:11:18 pm »
It'll work for a bit untily you accidentally exceed the miniscule voltage it'll handle before dying  :-DD
 

Offline Njk

  • Frequent Contributor
  • **
  • Posts: 250
  • Country: ru
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #12 on: November 14, 2023, 10:25:50 pm »
Well, there is some protection in the form of clamping diodes. Although the input range is quite limited indeed and generally it's not a unique instrument in that regard. As usual, RTFM before use.
 

Offline Circlotron

  • Super Contributor
  • ***
  • Posts: 3234
  • Country: au
Re: Can my Rigol dg1022 counter be defective ?
« Reply #13 on: November 14, 2023, 10:40:34 pm »


Edit -> I misread the title. I thought it was talking about the later DG1022Z.

Yes sorry it's dg1022Z
The running title changes from post #10 onward, oddly enough….

Okay, tried out mine last night and and the counter seems to be very level sensitive. Started off with 10V p/p sine wave and the counter sat at zero. At 2V it was a bit jumpy and wouldn’t go above 1.3MHz. At 1V p/p it worked great all the way to 25MHz. Input clamping diodes were mentioned. I’ll put the scope on it and see if that is happening.
« Last Edit: November 14, 2023, 10:43:12 pm by Circlotron »
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022 counter be defective ?
« Reply #14 on: November 15, 2023, 02:00:29 pm »
Quote
The running title changes from post #10 onward, oddly enough….

Yes I corrected the title :p

Quote
At 1V p/p it worked great all the way to 25MHz

Do you remember the trigger sense and trigger level it was set ?

Quote
Input clamping diodes were mentioned. I’ll put the scope on it and see if that is happening

Ok thanks, let me know!

Yannick
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #15 on: February 01, 2024, 12:32:19 am »
Finally, I fixed it. It was indeed defective.

The input resistor 26.1 Ohm was open. I do not really know how this happen but anyway. It's fixed and the counter is working well.



Yannick
 
The following users thanked this post: xrunner, Circlotron

Offline Circlotron

  • Super Contributor
  • ***
  • Posts: 3234
  • Country: au
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #16 on: February 01, 2024, 03:50:34 am »
If you measure DC resistance from counter input to ground, what do you find? If other people have the same problem they can check this without opening the unit up.
 

Offline Yannick99Topic starter

  • Contributor
  • Posts: 17
  • Country: ca
    • My instructables
Re: Can my Rigol dg1022Z counter be defective ?
« Reply #17 on: February 01, 2024, 01:46:23 pm »
Yes sure it's a good idea, could be helpful.

The input resistance on the counter bnc connector (center to GND) is 2.361 MegOhm at Off state and 1.2 Meg when powered On.
Before the replacement, it was just Open, OL.

The resistor replaced was a 26.1 Ohm in a 1206 Package named Counter-In