Hi everyone, finally I had to register myself to clarify something, but was reading the forum for a
while, so not a huge loss...
I am starting a new one for the BM789 as it does not seems to have an own topic yet.
So I got the bm789 with FW. version 08. (The Brymen multimeter, not the other bm789 equipment one can find on youtube,of which I was hoping to see an in depth review, done by the lady presenting it, after accidentally clicking on it, and watching the whole presentation.)
So I found two problems:
1.If I use the mV Scale and feed in a squarewave with 50% duty cycle without(or very minimal like 100-200mV) zero crossing (though it behaves almost the same way with less or more than 50%) something above 1 V p-p (so above the specs for the mV scale) it reads around
650mv on the AC scale, and it stays so even if I raise the voltage to around 30 V .(Can't get further)
DC or AC+DC overranges properly.
If I change the duty % it will be a bit less than 650mV, around 500mV AC. So it peaks at 50%
The strange thing is, that if I chane the polarity of the leads, it behaves correctly and overranges around 650 mV AC which it should. So with one way the leads connected it behaves as expected, but not the other way around. I assume it depends on which polarity the squarevawe has.
Do you thing this is normal or acceptable? Before anyone mentions it, I know that I should use the higher ranges for signals like this, but when someone measures a lower level signal, and it raises up
above the limit, there is no way of knowing it, you just stop seeing anything above this level, while the voltagemight be skyhigh. I can not test it with higher voltages unfortunately, but would be nice if someone could do that.
So I am not vorrying about accuracy above the specified limit, but about the fact that it does not show overrange on the AC scale on the mV scale.
How do other meters behaves in this scenario?
2. Thats a minor issue, but still annoying: When measuring milli or micro amps, if on auto range, the measured currentis above the displayed range, so the meter has to switch range, on AC+DC mode first it displays "OL" for a second or more and only if you keep the leads connected it displays the proper values.
I thought, it is just a question of a timeout value in the FW, but Brymen says it is not possible to change it.
Should I keep it or not?
Ps:
Thanks to Dave for the inspiring Videos!
And also thanks to Joe Smith for your extensive work, wish we could see this level of testing with some more consumer goods as well, it would mean much less garbage be produced!
I would be especially interested in your oppinion on this, and whether it is a firmware bug or maybe hardware related.
I contacted Brymen through the seller, but they say it is all good and fine with the meter.