Yesterday you showed the same resistors, but the results today are different and more reliable. What's changed since yesterday?
I don't know what changes break the measurement value,
now i use "k" firmware with freq counter fix (not use yesterday firmware) then i fix 5V LDO now my LDO shows 5.04V as you said i have to replace better LDO.
however, the resistor measurement after 1 hour I repeat the measurement without calibration then, the result will change. every low resistor measurement I have to calibrate, or I have to do a 3pin jumper probe but when there is a request for selftest mode... ? I looked at it and saw the results of the short probe 0.00 0.00 so I did a resistor check and the results were appropriate. but if I do not calibrate and see the 3 pin probe jumper then the result is 0.07 0.00 this will damage the current value of the measurement. can you find out where this error lies?
the change in the value of short check pin 123 occurs after a few minutes or after a few hours, so I have to calibrate again to get the result 0.00 - 0.00, as if eeprom didn't save the last calibration result.