I didn't expect that of a Fluke meter, even if it is a low end model.
Expect what of a fluke meter? We have three low cost devices in a dubious test situation. What is to say any of them are correct?
The error is less than 1% the quoted accuracy of the 17B is DC Volt : 400mV / 4V / 40V / 400V / 1000V , +/-1.0% + 10
I see some people just invent figures and quote them to sound like they know what they are talking about:
The actual quoted accuracy spec of the 17B in the DC 4.000/40.00/400.0/1000 volts range is +/-(0.5% +3 counts LSD).
Only the DC 400.0 Millivolts range is quoted at +/-(1.0% +10 counts LSD).
Even so, the meter (it's my meter, by the way) was within spec.
5.0000 Volts with +/-(0.5% +3 counts LSD) means any reading between 4.95 and 5.05 Volts is within spec on that (40.00V) range.
So it made it ...just.
But I think it shows somewhat of a lack of care and attention to detail to sell a meter that far off accuracy.
Yes, if one has a "within spec is good enough" attitude, then it's okay. But with that sort of attitude, one doesn't maintain a company reputation for accuracy and care for detail.
But, it is a relatively low cost meter. So whatever. I've adjusted it now and it reads correctly.
If I was a Digital multimeter manufacturer, I would not be satisfied selling someone a meter that read 4.96V from a 5.0000 V source, even if it was "within spec".
If I was a Fluke employee sitting on the manufacturing line responsible for calibrating the meters as they go out, with a high precision calibrator in front of me, and a device came through reading just a tad within spec, but that far off accuracy, I would not be content to let it go through without correcting it. 5 Volts is 5 Volts, regardless of how many zeros you put after the decimal point.
The TI chip accuracy is quoted at 1%. I doubt very much the millispec part is used, I can see the rest of the parts are no better than 1%. What is the accuracy repeatability of the rest of the device?
Sure it's checked against a laboratory device. That is before shipping storage, operator manhandling etc. Were the temperature, humidity etc identical to those When was the calibration pot was last tweaked (knocked)?
More off the top of your head spouting...
The TI 5050 standard grade chip is quoted at 0.1% inaccuracy MAX. But that isn't really the main concern in this instance.
The device is aged for >250 hours then it is carefully trimmed using a precision 10 turn trim pot to exactly 5.0000 using a calibrated 8.5 digit HP 3458A DMM that has an accuracy of 0.0008%. (Tektronix Certificate of Calibration # 4787923 valid through January 19, 2012).
The temperature was recorded at 21 degrees C at the time of calibration of the voltage reference.
So more important than the accuracy of the TI chip, is the stability, or level of drift, which is stated at 8ppm/degree C MAX temperature drift, and 5ppm/1000 hours long term stability. But much of this possible time related drift happens in the first 250 hours (as proven under testing and shown in the data sheet), which is why Doug - who makes the voltage reference device - ages the chip for 250+ hours prior to trimming.
The rest of the components are high quality parts but don't affect the voltage reference, if you know how the chip works.
Also, I don't know how you "can see the rest of the parts are no better than 1%", when if you actually looked at the device, and knew what you were talking about, there are several resistors that are clearly marked to be of 0.01% accuracy (either in writing on the brown resistors, or with a violet coloured 5th band).
The date of calibration of the voltage reference I was using was exactly one week prior to the date the photos were taken.
The temperature at the time the photos were taken was also 21 degrees C (+/-1 degree), the same as that at the time of calibration.
As for transit, the device has had a touch of enamel put on the trim pot adjuster after calibration to prevent change, and it is still in place un-altered.
It was carefully packaged and shipped.
There was no "operator manhandling" thank you.
Yes the temperature was within a degree of when it was calibrated 1 week prior, and no, it hasn't been "knocked".
The DMMCHECK device is guaranteed to be accurate to within 0.01% (plus or minus 500uV) for a minimum of 6 months, but as it has just been fine tuned, it is much more accurate than even that at present.
Seriously if a couple of sub $100 meters and a mail order reference board are that close in a knock about workshop test what is the problem?
The Fluke could be spot on, there could be error each way or thing could be as pictured.
The Fluke was not spot on, but it is now that I've adjusted it. Now it matches the readings of my 87 V.
The reason a difference of .04 V can be of concern to me, is when charging li-ion cells.
The difference between charging a li-ion cell to 4.20 V and 4.24 V can have a big impact on the long term life of the cell. They really are that sensitive.
Safety of the charge is also impacted if a reading is out by even such a small amount.
The same is true with a .04 V difference with lead acid batteries, The difference in voltage between nearly 0% SOC and nearly 100% SOC is only about 0.6 V in some cases, so a .04 V difference can be important, especially if working with individual cells.
Anything requiring better accuracy shouldn't be using this class of device.
Quite right, and that is why I bought a Fluke 87 V as well.
A hobbyist can perceivably have need for good accuracy, but not have the funds to spend on something like a Fluke 87 V or something else in that class.
There is no need for elitist snobbery, it simply makes sense when on a very truncated budget to attempt to get the best item one can afford.
If all one can afford is a <$100 meter, and one wants to get the best possible DMM one can within those constraints,
in the past, one could be forgiven for thinking that by getting a Fluke there was a certain promise of care and accuracy in manufacturing and setting them in the factory before shipping.
Clearly, with these cheaper, Chinese manufactured Flukes, that isn't the case.
Having said that, the build quality of the 17B seems really good, it has good components, and is nicely soldered and well put together and designed.
It was just let down by the lack of a tiny turn of a couple of trim pots. A lack of care and attention to detail.
Yes it was within spec, but surely they should make an attempt to have the units leave the factory "ACCURATE" not just "WITHIN SPEC" It's the matter of a slight turn of the trim pot. and makes a huge difference to the end user, who may not have access to the high accuracy calibration devices that they have at the manufacturing location.
If then, after a few years it's reading slightly out, well one might expect that, but this device was only a couple of Months old, and should have been accurate, in my opinion.
/rant.