« on: September 15, 2011, 01:36:17 pm »
I finally got the Agilent U1252B. And I followed the standard procedure here of 'Don't turn it on, take it apart'.



If you want more info, or something covered more in depth, feel free to ask. This is my first time doing a teardown and a review, so I know I missed stuff. Man, I need more enthusiasm.
Brandon Falk, Systems Software Engineer
http://gamozo.org/ - http://gamozolabs.com/
 Catalog your components - http://eesdb.org/
 

Offline shadowless

  • Regular Contributor
  • *
  • Posts: 126
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #1 on: September 15, 2011, 05:47:59 pm »
Great review.

Quick question. Blown the 11a fuse in the u1252b, can i use fluke fuse in place of the SIBA fuses.Fluke fuses seems to be cheaper.
 

Offline gamozoTopic starter

  • Regular Contributor
  • *
  • Posts: 86
  • Country: us
    • Personal Website
Re: Agilent U1252B Teardown/Review
« Reply #2 on: September 15, 2011, 05:59:12 pm »
I'm not quite sure on that. You may have to resort to directly contacting Agilent (although they may be biased just to get your money). Unless you can find the exact fuse specs somewhere online. I'm assuming there would be no issue, as they both are 11A HRC fuses, but maybe there are tiny things that could change compatibility. (Perhaps the size of the fuse in the first place, I wonder if it would even fit)
Brandon Falk, Systems Software Engineer
http://gamozo.org/ - http://gamozolabs.com/
 Catalog your components - http://eesdb.org/
 

Offline shadowless

  • Regular Contributor
  • *
  • Posts: 126
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #3 on: September 15, 2011, 06:10:52 pm »
Seems to be the same size base on the same amp rating. Everyone sells fluke fuses and they seems cheaper. Probably get the fluke ones.
 

alm

  • Guest
Re: Agilent U1252B Teardown/Review
« Reply #4 on: September 15, 2011, 06:23:33 pm »
Agilent uses fuses with a higher interrupt rating in some meters. I wouldn't use fuses with worse specs, all safety tests were performed with these specs in mind.
 

Offline Fox

  • Regular Contributor
  • *
  • Posts: 54
  • Country: de
  • Electronics Engineer
Re: Agilent U1252B Teardown/Review
« Reply #5 on: September 15, 2011, 07:00:13 pm »
Good Review gamozo
I like the case of Agilent Multimeters too, i have a U1241B and it has basically the same case as yours.

@shadowless
i think you can use fuses from SIBA, they have dedicated fuses for usage in a Multimeter see the attached Pdf.
Fluke uses fuses from Bussmann DMM-B Series if you looken for an other supplier than Fluke directly.
The sizees of the fuses are standardised 10x38mm for the 11A and 10x35mm for the 440mA one.
A closed Switch should have zero Ohms or less!
 

alm

  • Guest
Re: Agilent U1252B Teardown/Review
« Reply #6 on: September 15, 2011, 07:10:43 pm »
The Agilent U1252B manual states that the fuses have an interrupt current of at least 30kA. Fluke fuses were 20kA last time I checked. Can't find the spec for Siba. My assumption is that since Agilent uses better fuses, they probably changed some other parameters so the expected fault current is higher. Why else would they go for more expensive fuses if the inputs are rated for the same CAT IV 600V / CAT III 1000V?
 

Offline Fox

  • Regular Contributor
  • *
  • Posts: 54
  • Country: de
  • Electronics Engineer
Re: Agilent U1252B Teardown/Review
« Reply #7 on: September 15, 2011, 07:21:40 pm »
You can find the specs on the SIBA Site, thou they are not that easy to find. Search for the URZ Type Fuses.
I attached the only specs for the 10x38mm Fuse, cause the Pdf for all of the URZ TYpe Fuse is just too big for the forum limits.
A closed Switch should have zero Ohms or less!
 

alm

  • Guest
Re: Agilent U1252B Teardown/Review
« Reply #8 on: September 15, 2011, 07:38:36 pm »
Thanks for the datasheet, a cursory search on the Siba website only yielded product catalogs. The specs you posted appear to match the Agilent specs, so I agree that the fuse should be fine. Funny how the 11A version is missing from many listings, though. You would expect it to be more popular than the 12A version.
 

Offline saturation

  • Super Contributor
  • ***
  • Posts: 4787
  • Country: us
  • Doveryai, no proveryai
    • NIST
Re: Agilent U1252B Teardown/Review
« Reply #9 on: September 15, 2011, 08:20:52 pm »
Thanks for the review!  You can find other reviews on eevblog on the 1252a, is twin brother that's just a bit blue, just to compare notes.
Best Wishes,

 Saturation
 

Offline Wartex

  • Frequent Contributor
  • **
  • Posts: 411
  • Country: ca
    • http://headsplosive.com
Re: Agilent U1252B Teardown/Review
« Reply #10 on: September 15, 2011, 11:52:26 pm »
Did you film this with a toilet plunger?
 

Offline shadowless

  • Regular Contributor
  • *
  • Posts: 126
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #11 on: September 16, 2011, 03:26:29 am »
Yeah the SIBA fuses are 30ka and they are hard to find. Trying to ask for the agent in US. What is the difference between the 30ka and 20ka rating?

I thought i will never blow them till yesterday.  They are really fast blow, accidental touch of a battery terminals, one spark and they are gone. 

 

Offline Kiriakos-GR

  • Super Contributor
  • ***
  • !
  • Posts: 3525
  • Country: gr
  • User is banned.
    • Honda AX-1 rebuild
Re: Agilent U1252B Teardown/Review
« Reply #12 on: September 16, 2011, 04:27:24 am »
If you take a CAT III 1000 V meter with the test leads in the amps jacks, you will have a series
resistance of approximately 0.1 ohms (0.01 for the shunt, 0.04 for the test leads and 0.05 for the
fuse and circuit board conductors) between the leads.
Now when you accidentally place the leads across a 1,000 volt source, by
Ohms Law you will generate a current of 10,000 amps (E/R=I, 1,000/0.1 = 10,000).
You want a fuse that will break that current and do it quickly.

For the full story read this..  :)
http://assets.fluke.com/Appnotes/2041429_w.pdf
 

Offline Amarbir[Lynx-India]

  • Frequent Contributor
  • **
  • Posts: 431
  • Country: in
  • Indian Dealer
    • Lynx-India - Visit Us For Not So Boring Electronic Instruments
Re: Agilent U1252B Teardown/Review
« Reply #13 on: September 16, 2012, 06:47:10 pm »
I finally got the Agilent U1252B. And I followed the standard procedure here of 'Don't turn it on, take it apart'.



If you want more info, or something covered more in depth, feel free to ask. This is my first time doing a teardown and a review, so I know I missed stuff. Man, I need more enthusiasm.

Hi,
 I would like to give you intense appreciation for this video teardown .Have USD 200 to spend and cannot think of anything better .any suggestions guys other then buying this
Regards

Amarbir Singh Dhillon [ Lynx-India ] , Chandigarh [ India ] - > www.lynxdealerstore.com , www.lynx-india.com
Indian Distributor For  [ Autoelectric , Sofitech , IDEOfy ,Peak Electronic Design [UK ] , Anatek And Creatronica ]
My Electronics Blog - > www.lynxchandigarh.com
 

Offline Baliszoft

  • Frequent Contributor
  • **
  • Posts: 277
  • Country: hu
Re: Agilent U1252B Teardown/Review
« Reply #14 on: September 16, 2012, 07:16:18 pm »
Usd200 wont buy you a 1252. I  prefer it over on the 87V, if that helps.
 

Offline nanofrog

  • Super Contributor
  • ***
  • Posts: 5446
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #15 on: September 16, 2012, 08:16:22 pm »
If you're after an Agilent, Baliszoft is correct. You'd actually need to double your budget for that meter in the US, and probably a bit more since you're in Europe. The U1242B would be much closer to your budget I suspect (~$216 - 226USD in the US).

BTW, for anyone that is interested, the cheapest source for fuses I found (US) is directly from Agilent. Shipping was free, so I only paid ~$6USD for the 11A unit (pricing was less for the 440mA fuse).
 

Offline Bored@Work

  • Super Contributor
  • ***
  • Posts: 3932
  • Country: 00
Re: Agilent U1252B Teardown/Review
« Reply #16 on: September 16, 2012, 09:43:21 pm »
and probably a bit more since you're in Europe.

Sure, India is now in Europe ...
I delete PMs unread. If you have something to say, say it in public.
For all else: Profile->[Modify Profile]Buddies/Ignore List->Edit Ignore List
 

alm

  • Guest
Re: Agilent U1252B Teardown/Review
« Reply #17 on: September 16, 2012, 11:02:16 pm »
India and Europe are now two different countries? When did that happen?
 

Offline nanofrog

  • Super Contributor
  • ***
  • Posts: 5446
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #18 on: September 17, 2012, 01:52:54 am »
andSure, India is now in Europe ...
Accidentally saw Baliszoft's country flag rather than Amarbir's during my reply in regard to location. It was an honest mistake.  ::)
« Last Edit: September 17, 2012, 06:36:35 am by nanofrog »
 

Offline Amarbir[Lynx-India]

  • Frequent Contributor
  • **
  • Posts: 431
  • Country: in
  • Indian Dealer
    • Lynx-India - Visit Us For Not So Boring Electronic Instruments
Re: Agilent U1252B Teardown/Review
« Reply #19 on: September 17, 2012, 04:19:29 am »
Regards

Amarbir Singh Dhillon [ Lynx-India ] , Chandigarh [ India ] - > www.lynxdealerstore.com , www.lynx-india.com
Indian Distributor For  [ Autoelectric , Sofitech , IDEOfy ,Peak Electronic Design [UK ] , Anatek And Creatronica ]
My Electronics Blog - > www.lynxchandigarh.com
 

Offline Achilles

  • Regular Contributor
  • *
  • Posts: 229
  • Country: de
Re: Agilent U1252B Teardown/Review
« Reply #20 on: September 21, 2012, 12:17:02 pm »
About the Fuse question again:
Try to check out on the Agilent site or write your service center. I got my spare fuses there and they were much cheaper than in the online shops.
The 11A Siba fuse was 7,54€ incl. VAT. A fluke 11A fuse is around 8,5€ incl. shipping.
So it may be worth checking. You can check the prices only on the Agilent site if you go to the U1252B meter, on the right side you can click on a "spare parts" and see the fuses there. The prices there are excl. VAT and may change (mine were a bit cheaper).

Well, they do have a minimum order of 35Euro here, so I bought a bunch of them and paid no shipping.


...just in case if you need spare fuses ;)
 

Offline LaurenceW

  • Frequent Contributor
  • **
  • Posts: 258
  • Country: gb
    • It's Time, Jim, but not as we know it
Re: Agilent U1252B Teardown/Review
« Reply #21 on: September 22, 2012, 09:28:52 pm »
Wow - do you REALLY have to strip the meter down that far, just to change a fuse? Hmmm >:(
If you don't measure, you don't get.
 

Offline T4P

  • Super Contributor
  • ***
  • Posts: 3697
  • Country: sg
    • T4P
Re: Agilent U1252B Teardown/Review
« Reply #22 on: September 23, 2012, 07:11:52 am »
Wow - do you REALLY have to strip the meter down that far, just to change a fuse? Hmmm >:(
It's a good meter but that's just nuts, having to get to the other side of the board just to change the fuses  >:(
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Agilent U1252B Teardown/Review
« Reply #23 on: September 23, 2012, 09:43:11 am »
cos you are not supposed to pop the fuses :P
 

Offline nanofrog

  • Super Contributor
  • ***
  • Posts: 5446
  • Country: us
Re: Agilent U1252B Teardown/Review
« Reply #24 on: September 23, 2012, 12:14:30 pm »
Wow - do you REALLY have to strip the meter down that far, just to change a fuse? Hmmm >:(

It's a good meter but that's just nuts, having to get to the other side of the board just to change the fuses  >:(
Not the simplest fuse changes, but assuming the user isn't blowing one very often, I see it as an acceptable compromise for what the meter can do/offers in value, particularly for bench use. And it doesn't take that much time (could see it as a real PITA for field work though, particularly in regard to finding dropped/lost screws).
 