Author Topic: Tektronix TDS5054B: Record Length Option: Solved!  (Read 18937 times)

0 Members and 2 Guests are viewing this topic.

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #25 on: August 28, 2023, 06:05:58 am »
Hi again,

For a TDS6154C with SN B010620 and Options 4M, ST, SM, RTE, JA3, PTD it is:

PS C:\Users\????\????\python\tek-ssc-master> C:\Python27\python .\gen.py TDS6154B B010620 08 44 00 00 00 00 80 00 28 00 00 00 00 00 00 00 00 00 (thus string ends in this case with 18 Zero's).

X3URE-768T6-LSDKG-DZ2KE-L2GUZ-6K6J4-FVZ4L-NQ6G4-923A

In general can the number of Zero's or F's be reduced to make it work. This is a Scope dependent Trial and Error Procedure.

Succes,

Calibrationfixture
« Last Edit: September 18, 2023, 06:30:48 pm by calibrationfixture »
 

Online tv84

  • Super Contributor
  • ***
  • Posts: 3271
  • Country: pt
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #26 on: September 17, 2023, 05:07:36 pm »
Hi again,

For a TDS6154C with SN B010620 and Options 4M, ST, SM, RTE, JA3, PTD it is:

PS C:\Users\????\????\python\tek-ssc-master> C:\Python27\python gen.py TDS6154B B010620 08 44 00 00 00 00 80 00 28 00 00 00 00 00 00 00 00 00 (thus string ends in this case with 18 Zero's).
X3URE-768T6-LSDKG-DZ2KE-L2GUZ-6K6J4-FVZ4L-NQ6G4-923A

In general can the number of Zero's or F's be reduced to make it work. This is a Scope dependent Trial and Error Procedure.

Succes,

Calibrationfixture

This format should work and is much simpler:

DZSMA-M6GKW-46MHA-MKSSS-CADDB-YDSE
 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #27 on: September 18, 2023, 06:44:29 pm »
Hi tv84,

Or:

NHA5C-M5FLG-T2BTA-TEAAJ-AAAAA-EASC

There is a slight discrepancy in the UID.

It is namely a real TDS6154C on eBay.com. It is a pity that there is no Follow-up from Ko. I didn't have a TDS6000C Series on hand, thus evaluations are appreciated..

Calibrationfixture
« Last Edit: September 19, 2023, 07:17:32 am by calibrationfixture »
 

Online tv84

  • Super Contributor
  • ***
  • Posts: 3271
  • Country: pt
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #28 on: September 22, 2023, 09:34:08 pm »
There is a slight discrepancy in the UID.

It is namely a real TDS6154C on eBay.com. It is a pity that there is no Follow-up from Ko. I didn't have a TDS6000C Series on hand, thus evaluations are appreciated..

Just for reference:

Your license would work in a 6154C but it's not "officially" correct as it's oversized.

I attach a picture that graphically shows what I'm saying. The last license is the correct/official license for the UID/options combo that you were creating.

Pay attention to the blue color bytes...  ;)
 
The following users thanked this post: calibrationfixture

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #29 on: September 23, 2023, 05:44:27 am »
Hi tv84,

Thank you for your reply. I know that the Option Mask String I provided was too long but Ko asked for:

I know the relations.
However the generated code is not accepted with the scope.
I think the cause is for the incorrect length of option mask or similar problem which I have experienced for TDS5034B.

Do you have any examples of option code for TDS6254C?

In the past I created working Option Keys for a real DPO7254, DPO72004 and TDS7404B. To enable all possible Options I experienced then that the string for a TDS7404B can be shorter than those for the DPO7254 and DPO72004 and observed slight UID differences between DSA and DPO Models. I don't know how to create "officially" correct Keys. You are invited :)

Calibrationfixture
« Last Edit: September 23, 2023, 06:04:04 am by calibrationfixture »
 

Online tv84

  • Super Contributor
  • ***
  • Posts: 3271
  • Country: pt
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #30 on: September 23, 2023, 08:55:55 am »
I don't know how to create "officially" correct Keys. You are invited :)

It could be nitpicking but I've seen occasions where it had some repercussions (long time...). From memory: the public keygen code has some shortcomings and one of them is not always generating the shortest license code possible (assuming 0x1C as the minimum qualifying size for the bitmask). Sometimes, I suspect this can lead to buffer overruns when loading the license, because the scope app is not expecting further bytes in its buffer.

But there are plenty of examples where even Tek's official "STD" licenses have a huge bitmask  (with lots of bits enabled) although this type of option is a 0-bit bitmask option, so the license should have a 0x1C length bitmask with no bit activated. It kind of shows that even Tek has different keygen implementations (for different models/factories  :-//), some of them more loosely implemented.

Also if UID are different (even if just in 1 bit) then we should be talking different devices. Usually applies for the B and C models...
 

Offline Joebeazelman

  • Contributor
  • Posts: 43
  • Country: yt
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #31 on: November 19, 2023, 07:44:56 am »
Sorry for my newbie question, but what are the differences between the 5000 and 5000B series. I was told there's a difference between the two series. I've compared the datasheets and I could only find minor differences in the scope part. Most of the differences I found were in the CPU board or PC motherboard.

Tektronix did offer an upgrade kit for the TDS, but they didn't include the tds5000 only the tds5000b series. My guess is that the tds5000 was past it's service life. Are there significant differences that would make it impossible to upgrade my damaged motherboard to the Advantech one mentioned in the upgrade kit?
 

Offline colorburst

  • Contributor
  • Posts: 38
  • Country: us
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #32 on: November 19, 2023, 09:29:01 am »
Specification-wise they're virtually identical. Most notably, the B series has double the record length and comes with a newer computer (still running W2k though). There may be extra Java applications and trigger options supported as well, I'm not too familiar with those. It's basically the same thing.

Internally it's an update of the older design and is technically a different product. Although the improvements are incremental, parts and software from one generation are not interchangeable with the other.

Because the instrument was built as an integrated system, altering the qualified configuration is risky. You may trip up on hacks and bugs that laid dormant or were deemed not worth fixing, like the PCI error or some more insidious crash that happens 0.001% of the time. Proceed at your own risk.
 
The following users thanked this post: Joebeazelman

Offline Ko

  • Contributor
  • Posts: 11
  • Country: jp
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #33 on: January 06, 2024, 03:42:53 am »
Sorry for delaying my response.

I tried to make code, but I haven't succeeded the correct code yet.

My serial number is B010492 which is also displayed on the option installation dialog.

The generated option keys of geb.py which I inputed into my osc are;

-------------------------------------
(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 084400000000800028000000000000000000 (18 digits as your explanation)
X3UNE-6N4T2-QQHRF-Q988Y-Q9H99-49QN7-E8966-8L5HZ-Z8RA

(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 0844000000008000280000000000000000 (1 byte digit reduced)
PTC2E-NPMBU-NN27P-4MV43-9N9MR-Z5B7Q-6LRT2-T5J9D-HD

(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 08440000000080002800000000000000 (2 byte digit reduced)
FTUJE-NNLBS-QJ46P-4MV43-9N9MR-Z5B7Q-6LRT2-T5J9D

(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 084400000000800028000000000000 (3 byte digit reduced)
78A7A-47V34-6YSNP-Q998Z-99999-59R77-68976-955D

(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 0844000000008000280000000000 (4 byte digit reduced)
X8SPA-46U32-8UUPP-Q998Z-99999-59R77-68976-9M

(py27) C:\tektronix\tek_options\tool>python gen.py TDS6154B B010492 08440000000080002800000000 (5 byte digit reduced)
PYA3A-L7DKU-6SP9F-4MU42-QNHMR-Y5ANQ-ELRS2-A
-------------------------------------
Any code couldn't work on my TDS6254C.

My TDS is STD which doesn't have any option. The result of validation.py is following;

--------------------------------------------------
(py27) C:\tektronix\tek_options\tool>python validate.py 37C6E-8K5V4-E6CRH-M545W
B9 0B 4E 7C DA 53 13 2E DE 59 5B 6F 0A
1C FC 28 00 03 0A 18 09 3E 00 00 00 E0 70 10 10 00 00 00 00 00 00 00 00 00 00 00
This key is for UID 180A030028FC (S/N 10492, model TDS/DSA/DPO6154B):
CRC: 3E09
Key is valid, active options:
00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00
--------------------------------------------------

Any ideas?
 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #34 on: January 07, 2024, 06:44:06 am »
Hi Ko,

Too many Zero's. Try this:

PS C:\Users\????\Documents\Python\tek-ssc-master\tek-ssc-master> C:\Python27\python .\gen.py TDS6154B B010492 0844FFFFFFFFFFFFFF (14 F's)
PZA2A-LPDKU-6WZ4N-VHKZA-2HZWK-LETN

Succes,

Calibrationfixture


« Last Edit: January 07, 2024, 07:17:15 am by calibrationfixture »
 

Offline Ko

  • Contributor
  • Posts: 11
  • Country: jp
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #35 on: January 08, 2024, 04:09:57 am »
I tried it. However, key didn't work.

I'm not sure why it can't work correctly. Can you help me?
 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #36 on: January 08, 2024, 07:47:39 am »
Hi,

Intriging. I assume you also replaced the MB Battery. The Scope is still out of date: 26 April ....)

You didn't get the Message: The Option Key isn't valid. Instead the Scope didn't succeed in enabling the New Options and Applications. Wow. It seems a Hardware issue. I assume you reinstalled Current Oscilloscope Application (V5.1.7 is the latest for TDS6000C's).

Calibrationfixture
« Last Edit: January 08, 2024, 08:15:52 am by calibrationfixture »
 

Offline Ko

  • Contributor
  • Posts: 11
  • Country: jp
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #37 on: January 08, 2024, 01:48:09 pm »
Latest firmware is v518.

Please see https://www.tek.com/en/support/software/firmware/tds6000c-series-firmware-upgrade-v518.

I  will recheck your key after I change date into current.
 
The following users thanked this post: calibrationfixture

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #38 on: January 08, 2024, 06:19:45 pm »
Hi,

Your link gives a 404 Error (the.)? But these:

https://www.tek.com/en/support/software/firmware/tds6000c-series-firmware-upgrade

https://www.tek.com/en/support/software/firmware/tds6000c-series-firmware-upgrade-v518

are working. There are two "latest" Firmware Upgrades. I am still learning :)

Calibrationfixture
« Last Edit: January 08, 2024, 06:37:09 pm by calibrationfixture »
 

Offline Ko

  • Contributor
  • Posts: 11
  • Country: jp
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #39 on: January 13, 2024, 08:16:51 am »
Hello
Today I change the date and time to current.
However my TDS could not accept the key.
Umm, I have not found the solution yet….
 

Offline Jaak

  • Newbie
  • Posts: 9
  • Country: ca
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #40 on: March 13, 2024, 05:13:50 am »
Hi Ko,

I think these are the Scope Model Numbers to use for the TDS5000B Series:

TDS5104B = TDS7216
TDS5054B = TDS7166
TDS5034B = TDS7146

I didn't have a TDS5000B Series on hand thus evaluations are appreciated.

Greetings,

Calibrationfixture

That totally did it on my 5034B; thank you very much!

Thanks for this!  Fully optioned my TDS5054B.  :)
 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #41 on: April 17, 2024, 07:00:50 am »
Hi alexandraquirosh,

I cannot PM you, because (Zeroposter?) your Mailbox is full. Please empty it, or post your request in this Thread.

Calibrationfixture
 

Offline Circuitauger

  • Newbie
  • Posts: 8
  • Country: us
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #42 on: May 14, 2024, 05:22:10 pm »
So I also happen to be having the exact same issue as KO, using a TDS6154C.

I get the message "The new options and applications were not successfully installed", I will note however that typing in gibberish as a key yields the same error, furthermore if I type in a known good key I get the message that the option was successfully enabled, so I do not believe this is an issue with installing the key rather the key is indeed invalid.

What I'm currently looking at now is how a known good key (like the one found on eBay) will validate and accurately be able to decode the model number and serial # however attempts to recreate this known good key will fail, in other-words taking the mask and other info from the validation and trying to apply it doesn't generate the same key.... Not sure why this would be as the ability to validate would mean you accurately know all the sbins? lol been a while since my crypto class.

Additionally I've noticed at least the first 5 chars of the lic will not change for a given mask, so once you've picked a mask you can put any model number and serial (even all zeros) and it will remain the same
so idk if this is a hint towards determining the proper mask for recreating the ebay key who's first 5 chars are "NHA5C"



 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #43 on: May 15, 2024, 06:19:41 am »
Hi Circuitauger,

Allright, thus it is not a HW issue. Because using alternative Scopenumbers did work for the TDS5000B Series I hoped it also works for the TDS6000B/C series. Because I don't have these Models at hand, you were all invited. The same for DPO7000C and DSA/DPO/MSO70000(B/C) Series.

Calibrationfixture
 

Offline Circuitauger

  • Newbie
  • Posts: 8
  • Country: us
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #44 on: May 15, 2024, 02:44:19 pm »
Well ultimately I do still think there's a good chance for this to work given that the validate.py is able to spit out valid info, and it sounds like from other posts that this works for both TDS5000 and TDS7000 scopes, so hopefully there's nothing weird about TDS6000 scopes.
Probably looking at redoing w/e work shuvodutta did to discover the code mishandling the B variant. Sounds like they aren't contactable though.

There's the post where you mentioned
Quote
I think these are the Scope Model Numbers to use for the TDS5000B Series:
TDS5104B = TDS7216
TDS5054B = TDS7166
TDS5034B = TDS7146
How did you determine this was the correct nomenclature to use?

Also in this post https://www.eevblog.com/forum/testgear/tek-csa7404-repair-project/msg4617574/#msg4617574
Quote
I used for my TDS7404B Option Masks: 10 44 FF FF FF FF FF FF FF
How did you generate that option mask? Looking in the option.py file, there's a description of masking but I don't see how your mask would be constructed.

Thanks,
-Circuitauger
 

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #45 on: May 16, 2024, 06:38:51 am »
Hi Circuitauger,

Using validate.py on a lot of existing TDS5000B License Keys leads me to the alternative and working Scope Numbers. The same with the TDS6124C and TDS6154C Models. Using validate.py gives TDS/DSA/DPO6124B and TDS/DSA/DPO6154B as Scope Numbers in case of existing TDS6000C License Keys, but these don't work as we know now. Maybe using DSA or DPO instead of TDS is the solution.

Using F's is the most time/cost effective way to enable as much as possible Options. A suggestion somewhere in the:

https://www.eevblog.com/forum/testgear/tek-csa7404-repair-project/600/

Thread.

Calibrationfixture
« Last Edit: May 16, 2024, 07:14:09 am by calibrationfixture »
 

Offline Circuitauger

  • Newbie
  • Posts: 8
  • Country: us
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #46 on: May 16, 2024, 05:47:33 pm »
Alright actually got it, I suspect this is similar to the mod made to support B models

The clue was in trying to validate the known good keys for TDS6154C variants you’ll end up getting a UUID error:
Quote
UID mismatch !
UID in key:     7C 29 00 03 0A 18
Calculated UID: 7C 29 00 02 0A 18
Which was suspiciously 1 off from what it should be... like C is to B

Basically the model number is getting expressed as a wrapped (maybe slightly scrambled) hex representation in the UUI
e.g. TDS6155C will yield a Calculated UID of: 7C 29 00 02 0B 18

So for anyone trying to repair their C variant scope in “key.py” you need to

change line 63 to
Code: [Select]
mdl_int = int(model[3:].strip('C')And add after line 68 lines:
Code: [Select]
if model[-1]=='C':
sn_int |= 0x03000000

You should now be able to type in your model number normally as TDS6154C into gen.py

Of course for a long term fix which allows for B & C variants the proper way would be to check for a char as the ending value and apply the appropriate strip and hex value set (the above change to line 63 will break the code if you try to type in a B model number as it now doesn't know what to do with the 'B')
« Last Edit: May 16, 2024, 06:14:23 pm by Circuitauger »
 
The following users thanked this post: coromonadalix, calibrationfixture

Offline calibrationfixture

  • Regular Contributor
  • *
  • Posts: 129
  • Country: nl
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #47 on: May 18, 2024, 07:14:37 am »
Hi Circuitauger,

I think TDS6000C's are the only exception. With TDS5000B, DPO7000(C) and DPO/DSA/MSO70000(B/C) Models you always get only Digits as Scope Type.  By the way contrary to TDS5000B's the CRC mismatches on existing License Keys are seldom when using validate.py in splitting out the Option Masks.

Calibrationfixture
 

Offline nicnac117

  • Contributor
  • Posts: 31
  • Country: gb
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #48 on: June 09, 2024, 08:08:35 pm »
Just a quick question ? Does any of this knowledge apply to a TDS5104 ?Your comments would be appreciated .
 

Offline colorburst

  • Contributor
  • Posts: 38
  • Country: us
Re: Tektronix TDS5054B: Record Length Option: Solved!
« Reply #49 on: June 09, 2024, 08:56:45 pm »
Yes, everything discussed here is applicable to TDS5104.


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf